Report Generator[edit]
Keywords[edit]
Keep in mind that:
- In a 'test-specific' template the Main Keywords are not used, (only the Second Keywords). 'Test-specific' templates are a new feature since version 5.2. For more information on this topic, please read the paragraph 'Test-specific templates' (under 'Special features') further below.
Unrecognized Keywords[edit]
Generating a report[edit]
Making a template file[edit]
(Be advised, even though the Microsoft Excel report generator editor has been present in earlier versions of RadiMation®, most of these functionalities are only included in RadiMation® version 5.3 and newer.)
Specific Features[edit]
- If the 'Client Contact Person' data has been entered in the EUT file, the code above will be replaced (with this data) and the text (including the data) will be visible in the report.
- If this data has not been entered in the EUT file, the code will not be replaced and the text (as well as the 'unrecognized' code) will not be visible in the report.
Test-specific templates[edit]
Introduction
||EUT|ALL TESTS||
Conclusion
- SIGNAL POWER GRAPH
- CURRENT GRAPH
- VOLTAGE GRAPH
- FORWARD POWER GRAPH
- REFLECTED POWER GRAPH
- NET POWER GRAPH
- MID POWER GRAPH
- OUT POWER GRAPH
- IMPEDANCE GRAPH
- ERROR GRAPH
- CALCULATED FIELD GRAPH
- AVERAGE FIELD GRAPH
- CALCULATED VOLTAGE GRAPH
- CALCULATED CURRENT GRAPH
- AD GRAPH 1
- AD GRAPH 2
- AD GRAPH 3
- AD GRAPH 4
- AD GRAPH 5
- AD GRAPH 6
- AD GRAPH 7
- AD GRAPH 8
- FIELD GRAPH 1
- FIELD GRAPH 2
- FIELD GRAPH 3
- FIELD GRAPH 4
- FIELD GRAPH 5
- FIELD GRAPH 6
- FIELD GRAPH 7
- FIELD GRAPH 8
- CALIBRATED POWER GRAPH
- EUT ATTENUATION GRAPH
- EUT DYNAMIC RANGE GRAPH
- Frequency
- Frequency
- Frequency
- FREQUENCY
- SIGNAL POWER
- FORWARD POWER
- REFLECTED POWER
- NET POWER
- CURRENT
- VOLTAGE
- CALCULATED FIELD
- AVERAGE FIELD
- FIELD 1
- FIELD 2
- FIELD 3
- FIELD 4
- FIELD 5
- FIELD 6
- FIELD 7
- FIELD 8
- AD CHANNEL 1
- AD CHANNEL 2
- AD CHANNEL 3
- AD CHANNEL 4
- AD CHANNEL 5
- AD CHANNEL 6
- AD CHANNEL 7
- AD CHANNEL 8
- CALIBRATION FILE
- TEST ENGINEER
- TEST DATE
- TEST TIME
- START TIME
- STOP TIME
- TEMPERATURE
- HUMIDITY
- PRESSURE
- START FREQUENCY
- STOP FREQUENCY
- STEP FREQUENCY
- LOGARITHMIC STEP
- LOGARITHMIC STEP FROM
- POLARIZATION
- DISTANCE
- TOLERANCE
- DWELL TIME
- MODULATION
- MODULATION DEPTH
- MODULATION CONSERVATION
- MODULATION FREQUENCY
- MODULATION SIGNAL
- MODULATION DUTY CYCLE
- MAINS VOLTAGE
- FREQUENCY CHANGE MODE
- AVERAGE FIELD
- CURRENT MEASURED
- CALIBRATION METHOD
- FORWARD POWER MEASURED
- REFLECTED POWER MEASURED
- NET POWER MEASURED
- CALCULATED FIELD MEASURED
- RECALCULATE AVERAGE
- DESCRIPTION
- TSF FILENAME
- AMPLIFIER PROTECTION MODE
- AUTODISAPPEAR
- USE VOLTAGE TEST LEVEL
- VOLTAGE TEST LEVEL
- POWER TEST LEVEL
- FIELD TEST LEVEL
- USE CURRENT TEST LEVEL
- CURRENT TEST LEVEL
- STEP DOWN
- USE CURRENT LIMIT
- CURRENT LIMIT
- MODULATION TYPE
- MAGNETIC FIELD TEST LEVEL
When the '||InsertFile(<OTHER WORD DOCUMENT FILENAME>)||' code is used, the contents from the other specified Microsoft Word document will be included in the generated report. The complete contents of the other document will be inserted at the location of the report generator code.
The filename of a Bitmap or JPEG file can also be specified as part of the '||InsertFile(<PICTURE FILENAME>)||' report generator code. The corresponding picture will then be loaded from the specified disk location and will be included at the location of the report generator code. This function is similar to the 'Insert Picture, From File' action of Microsoft Word.
Pictures, Microsoft Word documents and other files can also be attached to the EUT document, on the Attachments tab. These attachments can also be included in the generated report using the code '||EUT|ATTACHMENT <FILENAME> CONTENTS||'. The '<FILENAME>' part should be the name of the attachment. It is also possible to use the code '||EUT|ATTACHMENT 1 CONTENTS||', where the '1' is the number of the attachment, and can be any number. The complete contents of the attachement will be inserted at the location of the report generator code.
Macros[edit]
- A macro name predefined by the Microsoft Word report generator editor
The macro function that is implemented by the Microsoft Word report generator editor is:
- AutoFitThisTable
Table-inclusion modes[edit]
- Row insert mode
- Cell insert mode
- Overwrite mode
- Sheet insert mode
Row insert mode[edit]
Cell insert mode[edit]
Overwrite mode[edit]
Sheet insert mode[edit]
When the '||InsertFile(<OTHER EXCEL WORKBOOK>)||' code is used, the contents from the other specified Microsoft Excel workbook or worksheet will be included in the generated report. The complete contents of the other document will be inserted at the location of the report generator code. The size of the 'to be inserted' data is determined, starting at the 'A1' cell until the last filled cell. This data is then inserted at the location of the keyword, using the selected table-inclusion mode. To prevent that unnecessary worksheets are included in the report, a check is done which prevents that empty worksheets are inserted.
The filename of a Bitmap or JPEG file can also be specified as part of the '||InsertFile(<PICTURE FILENAME>)||' report generator code. The corresponding picture will then be loaded from the specified disk location and will be included at the location of the report generator code. This function is similar to the 'Insert Picture, From File' action of Microsoft Excel.
Pictures, Microsoft Excel documents and other files can also be attached to the EUT document, on the Attachments tab. These attachments can also be included in the generated report using the code '||EUT|ATTACHMENT <FILENAME> CONTENTS||'. The '<FILENAME>' part should be the name of the attachment. It is also possible to use the code '||EUT|ATTACHMENT 1 CONTENTS||', where the '1' is the number of the attachment, and can be any number. The complete contents of the attachement will be inserted at the location of the report generator code.
Macros[edit]
- A macro name predefined by the Microsoft Excel report generator editor
The list of macro functions that are implemented by the Microsoft Excel report generator editor are:
- CELL_PASTE_MODE_CELLS
- CELL_PASTE_MODE_SHEET
- CELL_PASTE_MODE_OVERWRITE
- CELL_PASTE_MODE_ROWS
SQL queries on tables[edit]
SELECT * FROM <TABLE>
Be advised that the names of the columns that can be used in the SQL queries is the same as they appear in the table itself.
- The names of the columns that can be used in the SQL queries is the same as they appear in the table itself. Any column names that include a space, should be included i• Any column name that include a space, should be included in '[ ]'
Examples[edit]
Peak Number | Peak Number_RAW | Frequency | Frequency_RAW | Peak | Peak_RAW | Peak Limit | Peak Limit_RAW | Peak Difference | Peak Difference_RAW | Peak Correction | Peak Correction_RAW | Peak Status | Peak Status_RAW | Status | Status_RAW | Angle | Angle_RAW | Height | Height_RAW | Polarization | Polarization_RAW |
1 | 1 | 829,92 MHz | 829,92 MHz | 80 dBµV | 80 dBµV | 60 dBµV | 60 dBµV | 20 dB | 20 dB | 0,00 dB | 0,00 dB | Fail | Fail | Fail | Fail | 0 Degree | 0 Degree | 1 m | 1 m | Horizontal | Horizontal |
- Peak Number
- Frequency
- Peak
- Peak Limit
- Peak status
- Peak status overall
… which does work.
1 | 829,92 MHz | 80 dBµV | 60 dBµV | 20 dB | Fail | Fail |
Limit the results[edit]
Sorting[edit]
Note[edit]
Internally the SQL statements and the table data itself are passed through the SQLite database engine. Because the SQLite database engine is used, almost all variations of the SQL queries are supported. The SQLite database engine is included in the setup program of RadiMation, and will always be installed.
Keywords Overview[edit]
EUT[edit]
Common[edit]
Keyword | Description |
FILENAME | The filename of the EUT |
BRAND | The brand of the EUT specified in the EUT Window |
CONDITION UPON RECEPTION | The condition of reception of the EUT specified in the EUT Window |
DATE OF RECEIPT | The date of reception of the EUT specified in the EUT Window |
DATE OF MEASUREMENT | The date of measurement of the EUT specified in the EUT Window |
DIMENSIONS | The dimensions of the EUT specified in the EUT Window |
EMISSION STANDARD | The emission standard of the EUT specified in the EUT Window |
ENVIRONMENT | The environment of the EUT specified in the EUT Window |
IMMUNITY STANDARD | The immunity standard of the EUT specified in the EUT Window |
NAME | The name of the EUT specified in the EUT Window |
ORDER NUMBER | The order number of the EUT specified in the EUT Window |
PCB RELEASE | The PCB release of the EUT specified in the EUT Window |
PERIPHERAL EQUIPMENT | The peripheral equipment of the EUT specified in the EUT Window |
PLACE MEASURMENT | The place of measurment of the EUT specified in the EUT Window |
RADIMATION VERSION | The RadiMation® version |
RATED POWER | The rated power of the EUT specified in the EUT Window |
SERIAL NUMBER | The serial number of the EUT specified in the EUT Window |
TEST SETUP TYPE | The test setup type of the EUT specified in the EUT Window |
SOFTWARE RELEASE | The software release of the EUT specified in the EUT Window |
TYPE NUMBER | The type number of the EUT specified in the EUT Window |
VOLTAGE SUPPLY | The voltage supply of the EUT specified in the EUT Window |
CLIENT ADDRESS | The address of the client specified in the EUT Window |
CLIENT CAPACITY | The capacity of the client specified in the EUT Window |
CLIENT CITY | The city of the client specified in the EUT Window |
CLIENT COMPANY | The client name specified in the EUT Window |
CLIENT CONTACT PERSON | The contact person of the client specified in the EUT Window |
CLIENT COUNTRY | The country of the client specified in the EUT Window |
CLIENT EMAIL ADDRESS | The email address of the client specified in the EUT Window |
CLIENT FAX NUMBER | The fax number of the client specified in the EUT Window |
CLIENT STATE | The state of the client specified in the EUT Window |
CLIENT TELEPHONE NUMBER | The telephone number of the client specified in the EUT Window |
CLIENT ZIP CODE | The ZIP code of the client specified in the EUT Window |
MANUFACTURER ADDRESS | The address of the manufacturer specified in the EUT Window |
MANUFACTURER CITY | The city of the manufacturer specified in the EUT Window |
MANUFACTURER COMPANY | The manufacturer name specified in the EUT Window |
MANUFACTURER CONTACT PERSON | The contact person of the manufacturer specified in the EUT Window |
MANUFACTURER COUNTRY | The country of the manufacturer specified in the EUT Window |
MANUFACTURER EMAIL ADDRESS | The email address of the manufacturer specified in the EUT Window |
MANUFACTURER FAX NUMBER | The fax number of the manufacturer specified in the EUT Window |
MANUFACTURER STATE | The state of the manufacturer specified in the EUT Window |
MANUFACTURER TELEPHONE NUMBER | The telephone number of the manufacturer specified in the EUT Window |
MANUFACTURER ZIP CODE | The ZIP code of the manufacturer specified in the EUT Window |
TEST SITE ADDRESS | The address of the test site specified in the EUT Window |
TEST SITE CITY | The city of the test site specified in the EUT Window |
TEST SITE COMPANY | The test site name specified in the EUT Window |
TEST SITE CONTACT PERSON | The contact person of the test site specified in the EUT Window |
TEST SITE COUNTRY | The country of the test site specified in the EUT Window |
TEST SITE EMAIL ADDRESS | The email address of the test site specified in the EUT Window |
TEST SITE FAX NUMBER | The fax number of the test site specified in the EUT Window |
TEST SITE STATE | The state of the test site specified in the EUT Window |
TEST SITE TELEPHONE NUMBER | The telephone number of the test site specified in the EUT Window |
TEST SITE ZIP CODE | The ZIP code of the test site specified in the EUT Window |
Cable information[edit]
CABLE TABLE | A table with the cable information specified in the EUT Window |
CABLE X PORT TYPE | The port type of cable X in the EUT window ( X starting from 1) |
CABLE X CABLE TYPE | The cable type of cable X in the EUT window (X starting from 1) |
CABLE X CABLE LENGTH | The cable length of cable X in the EUT window (X starting from 1) |
CABLE X FIXING SHIELD | The fixing shield of cable X in the EUT window (X starting from 1) |
CABLE X LOAD AT PORT | The load at port of cable X in the EUT window (X starting from 1) |
Standards Information[edit]
Parameter | Description |
REFERRED STANDARDS | A string of referred standards, separated by a "," |
Keyword | Description | |
COMNAME | Internal COM Name selected report generator | |
DESCRIPTION | Selected report generator | |
EXTENSION | Report generator extension | |
CHECKEXTENSION | Report generator should check extension | |
EDITOR | Report generator | |
STARTDELETEBLOCK() | Start block to remove unused keywords | |
ENDDELETEBLOCK() | End block to remove unused keywords | |
RemoveIllegalCodes(TRUE) | Removal illegal keywords |
Directories[edit]
DEVICEDRIVERFILESDIRECTORY | CONFDVDR directory |
CALFILESDIRECTORY | CAL Files directory |
EUTFILESDIRECTORY | EUT Files directory |
TSFFILESDIRECTORY | TSF Files directory |
SEQFILESDIRECTORY | SEQ Files directory |
CORFILESDIRECTORY | COR Files directory |
DIRECTORIES | All directories |
Units[edit]
CURRENT | Current |
EMISSION AMPLITUDE | Emission |
FIELD | Electrical Field |
FREQUENCY | Frequency |
POWER | Power |
VOLTAGE | Voltage |
UNITS | All units |
General[edit]
CURRENTUSERNAME | Current username |
EUT Calibration[edit]
- Attenuation / Gain
- Amplifier linearity
- Gain compression
- Harmonics
- S Parameter
- Antenna scan
- Antenna
- Fixed strength
Amplifier Linearity[edit]
DESCRIPTION | Test level to calibrate with |
TEST-TYPE | Signal gen, forward-power or net-power |
START FREQUENCY | Start frequency of the calibration |
STOP FREQUENCY | End frequency of the calibration |
STEP FREQUENCY | Step size |
CALIBRATION FILE | Used Calibration file |
ALL GRAPHS | Output graph |
Harmonics[edit]
STEP | Step size specifically for using log step in % |
VOLTAGE TEST LEVEL | Test level to calibrate with |
POWER TEST LEVEL | Test level to calibrate with |
CALIBRATION METHOD | Signal gen, forward-power or net-power |
TABLE | Output table |
GRAPH | Output graph |
POWER TABLE | Specific table for harmonics |
Radiated Immunity[edit]
STEP FREQUENCY | The linear frequency step in megahertz [MHz] |
FIELD UNIT | The selected unit for field strength: [dBuV/m], [dBmV/m], [dBV/m], [uV/m], [mV/m] or [V/m] |
POWER UNIT | The selected unit for power: [dBuW], [dBm], [dBmW], [dBW], [uW], [mW] or [W] |
VOLTAGE UNIT | The selected unit for voltage: [dBuVrms], [dBmVrms], [dBVrms], [uVrms], [mVrms] or [Vrms] |
DEVICE DRIVER DLL | N/A |
INI FILE NAME | N/A |
Radiated Immunity Tests[edit]
CURRENT UNIT | The selected current unit: [dBuA],[dBmA], [dBA], [uA], [mA] or [A] |
Conducted Immunity[edit]
CURRENT UNIT | The selected current unit: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] |
6) Substitution method only
Pulsed Immunity[edit]
ESD Test[edit]
Equipment for ESD Tests[edit]
EFT Test[edit]
Equipment for EFT Tests[edit]
Surge Test[edit]
Parameter | Description | ||
BRAND | Brand of Measurement equipment | ||
CALIBRATION EXPIRES | Date when calibration expires of equipment, given by the user | ||
DESCRIPTION | Description of Measurement equipment given by the user | ||
DEVICE DRIVER DLL | N/A | ||
DEVICE DRIVER NAME | The internal name of the device driver | ||
HARDWARE VERSION | Hardware version of Measurement equipment | ||
ID | Internal ID of Measurement equipment | ||
INI FILE NAME | N/A | ||
NAME | Name of Measurement equipment | ||
SERIAL NUMBER | Serial number of Measurement equipment | ||
SOFTWARE VERSION | Software version of Measurement equipment | ||
SUPPORTED DEVICE TYPES | N/A | ||
[edit]
TestRepeatCount | Count to repeat all the sequences (Whole test) |
CABLE CABLELENGTH | Length of the cable |
CABLE DESCRIPTION | User description of the cable |
CABLE FIXINGSHIELD | Fixing shield of the cable |
CABLE LOADATPORT | Data port of the cable |
CABLE MAXCABLELENGTH | Specified maximum cable length |
CABLE NOTE | User note of the cable |
CABLE PORTTYPE | Used Port Type of the cable |
CABLE TYPEOFCABLE | Cable type |
Band Specific Codes[edit]
BAND + X + <Setting>
BAND1 SequenceNumber | The number of the sequence. |
BAND1 TestSiteName | The selected test site with the used devices specified for the first sequence of the test. |
BAND1 TestNote | The user typed note for the first sequence of the test. |
BAND1 TestStatus | The test pass state specified for the first sequence of the test. |
EFT[edit]
BAND1 Line | Selected line coupling. |
BAND1 BurstDuration | The selected burst duration specified in the first sequence of the test. |
BAND1 BurstPeriod | The selected burst period specified in the first sequence of the test. |
BAND1 Polarity | The selected polarity of the VoltageTestLevel. |
BAND1 VoltageTestLevel | The voltage test level specified in the first sequence of the test |
BAND1 PulserepetitionRate | The pulse repetition rate specified in the first sequence of the test. |
BAND1 RepeatCount | The ammount the sequence is repeated. |
BAND1 Test note | The note specified for each band. |
Surge[edit]
BAND1 Line | The selected line specified in the first sequence of the test. |
BAND1 VoltageTestLevel | The voltage test level specified in the first sequence of the test. |
BAND1 Polarity | The polarity specified in the first sequence of the test. |
BAND1 RI | The impedance specified in the first sequence of the test. |
BAND1 CI | The capacitance specified in the first sequence of the test. |
BAND1 PulseRepetitionTime | The pulse repetition time specified in the first sequence of the test. |
BAND1 PulseCount | The pulse count specified in the first sequence of the test. |
BAND1 PhaseAngles | The pulse angles specified in the first sequence of the test. |
VDI[edit]
BAND1 Changedvoltage | The changed voltage for the dips specified for the first sequence of the test |
BAND1 Decreasevoltagetime | The switch time from regular voltage to the dip voltage |
BAND1 Dipcount | The ammount of dips per sequence |
BAND1 Increasevoltagetime | The time to switch from dip voltage to regular voltage |
BAND1 Line | The selected line coupling on which the dip will occur |
BAND1 Locationtypetext | The AC or DC mode |
BAND1 Phaseangles | The angle on which the dip will initiate |
BAND1 Ratedvoltage | The regular voltage |
BAND1 Reducedvoltagetime | The time that the voltage is at reduced voltage |
BAND1 Repetitiontime | The duration of one sequence |
BAND1 Repeatcount | The ammount the sequence will be repeated |
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
Radiated Emission[edit]
Note: | The pass/fail criteria, shown in the tables of the report generator, do not take the measure uncertainty of the test setup into account. |
Radiated Emission Tests[edit]
AVERAGE TRACE | Indication whether or not the Average trace is measured: Yes / No |
CURRENT UNIT | The selected unit for current: [dBuA], [dBmA], [dBA], [uA], [mA] or [A] |
DB/DIV | N/A |
DB/DIV MODE | N/A |
END ANGLE | The stop position in degrees |
NUMBER OF ANGLES | The number of steps between the start position and the stop position |
POLARIZATION | The polarization of the antenna: horizontal, vertical or both |
STEP FREQUENCY | The step frequency in kilo Hertz [kHz] |
USE INTERNAL PREAMP | N/A |
EUT HEIGHT | The height of the EUT in meters [m] |
SEPTUM HEIGHT | The height of the septum in meters [m] |
Radiated Emission Table[edit]
EMISSION TABLE | The selection of the Emission Table |
SELECTED | The selection of the peak(s) |
PEAK NUMBER | The number of the peak(s) |
FREQUENCY | The frequency of the peak(s) |
PEAK | The measurement value of the peak(s) |
PEAK LIMIT | The limit value of the peak(s) |
PEAK DIFFERENCE | The difference value between the measured peak(s) value and the limit value. |
PEAK CORRECTION | The correction value of the peak(s) |
PEAK STATUS | The status of the peak(s): Pass / Fail |
AVERAGE | The average measurement value of the peak(s) |
AVERAGE LIMIT | The average limit value of the peak(s) |
AVERAGE DIFFERENCE | The average difference value between the measured average(s) value and the limit value. |
AVERAGE CORRECTION | The average correction value of the peak(s) |
AVERAGE STATUS | The average status of the peak(s): Pass / Fail |
QUASI-PEAK | The Quasi-peak measurement value of the peak(s) |
QUASI-PEAK LIMIT | The Quasi-peak limit value of the peak(s) |
QUASI-PEAK DIFFERENCE | The Quasi-peak difference value between the measured Quasi-peak(s) value and the limit value. |
QUASI-PEAK CORRECTION | The Quasi-peak correction value of the peak(s) |
QUASI-PEAK STATUS | The Quasi-peak status of the peak(s): Pass / Fail |
RMS | The RMS measurement value of the peak(s) |
RMS LIMIT | The RMS limit value of the peak(s) |
RMS DIFFERENCE | The RMS difference value between the measured RMS(s) value and the limit value. |
RMS CORRECTION | The RMS correction value of the peak(s) |
RMS STATUS | The RMS status of the peak(s): Pass / Fail |
STATUS | The overall status of the peak(s): Pass / Fail |
ANGLE | The angle of the peak(s) in degrees |
HEIGHT | The height of the peak(s) in meters [m] |
POLARIZATION | The polarization of the peak(s) |
Radiated Emission Burst[edit]
BURST GRAPH | The graph plot |
CARRIER FREQUENCY | The carrier frequency in [MHz] |
BURST TRIGGER LEVEL | The burst trigger level in [dBm] |
BURST MEASURE TIME | The burst measure time in seconds [s] |
BURST SAMPLE RATE | The burst sample rate |
BURST MINIMUM GAP TIME | The burst gap time in milli seconds [ms] |
BURST THRESHOLD LEVEL | The burst threshold level in [dBc] |
ANTENNA ASSEMBLY | The Antenna assembly gain in [dBi] |
BEAMFORMING GAIN | The Beamforming gain in [dB] |
MAXIMUM EIRP | The Maximum E.I.R.P. in [dBm] |
MEDIUM UTILISATION | The Medium Utilisation in percentage [%] |
DUTY CYCLE | The Duty Cycle in percentage [%] |
MINIMUM GAP TIME | The minimum Gap time in seconds [s] |
MAXIMUM SEQUENCE TIME | The Maximum Sequency time in seconds [s] |
RMS | The RMS value in [dBm] |
Conducted Emission[edit]
AVERAGE DETECTOR | Indication whether or not the average detector is measured: Yes / No |
AVERAGE TRACE | Indication whether or not the average trace is measured: Yes / No |
DB/DIV | N/A |
DB/DIV MODE | N/A |
Excel Exporter[edit]
Selecting filename[edit]
Changing the test order[edit]
The ‘export module’ exports all the selected tests, starting from the first test at the top and ending with the last test at the bottom of the list. This order is not mandatory and can be changed. To change this order, select the test you want to move and press the ‘arrow up’ or ‘arrow down’ button to move the test up or down in the list.
Export options[edit]
Insert header above data | With ‘insert header above data’ you can insert headers above every exported column. (For example for the signal generator level, forward power and reflected power for immunity test.) |
Export EUT data | ’Export EUT data’ allows you to select the EUT data that is to be exported. Only the EUT information stored in the top test can be exported. |
Include not measured columns | ’Include not measured columns’ allows you to select the ‘not measured’ columns that are to be exported. If you want to run a macro after exporting, it may be easier to export all the columns. You can then use the same macro for a signal level, forward power level and net power level test. It is not possible to have RadiMation® run the macro automatically after the export. Deselecting ‘exporting not measured columns’ will result in faster exporting of the information. |
Performing the export[edit]
Warning: | Microsoft Excel needs to be installed on the PC to export the EUT information. This should be version 'Microsoft Excel 97' or newer. |