| Line 3,169: | Line 3,169: | ||
This can be accomplished by calling "radicli.exe", which is available in the same directory in which RadiMation is also installed. | This can be accomplished by calling "radicli.exe", which is available in the same directory in which RadiMation is also installed. | ||
{{Note|If {{RadiMation}} is not running, the report will be generated in the background with the same version as where the radicli.exe resides. When a RadiMation instance is running which supports the radicli interface, the report will be generated with the running instance | {{Note|If {{RadiMation}} is not running, the report will be generated in the background with the same version as where the radicli.exe resides. When a RadiMation instance is running which supports the radicli interface, the report will be generated with the running instance}} | ||
"C:\Program Files (x86)\Raditeq\RadiMation\Version x.y.z\radicli.exe" | "C:\Program Files (x86)\Raditeq\RadiMation\Version x.y.z\radicli.exe" | ||
Revision as of 11:38, 22 October 2024
Report Generator[edit]
Keywords[edit]
Keep in mind that:
Unrecognized Keywords[edit]
Generating a report[edit]
Making a template file[edit]
Specific Features[edit]
| “ | Example:
Test-specific templates[edit]
| “ | Example of a 'main-template':
| “ | Example:
- SIGNAL POWER GRAPH
- CURRENT GRAPH
- VOLTAGE GRAPH
- FORWARD POWER GRAPH
- REFLECTED POWER GRAPH
- NET POWER GRAPH
- MID POWER GRAPH
- OUT POWER GRAPH
- IMPEDANCE GRAPH
- ERROR GRAPH
- CALCULATED FIELD GRAPH
- AVERAGE FIELD GRAPH
- CALCULATED VOLTAGE GRAPH
- CALCULATED CURRENT GRAPH
- AD GRAPH 1
- AD GRAPH 2
- AD GRAPH 3
- AD GRAPH 4
- AD GRAPH 5
- AD GRAPH 6
- AD GRAPH 7
- AD GRAPH 8
- FIELD GRAPH 1
- FIELD GRAPH 2
- FIELD GRAPH 3
- FIELD GRAPH 4
- FIELD GRAPH 5
- FIELD GRAPH 6
- FIELD GRAPH 7
- FIELD GRAPH 8
- CALIBRATED POWER GRAPH
- EUT ATTENUATION GRAPH
- EUT DYNAMIC RANGE GRAPH
| “ | Example:
- time:
- frequency:
- angle:
- height:
| “ | Example:
| “ | Example:
Export an angle indexed table: ||angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE|| Export an time indexed table: ||time:Signal Power TABLE|| Export a height indexed table: ||height:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE|| |
- Frequency
- Frequency
- Frequency
| “ | Example:
- FREQUENCY
- SIGNAL POWER
- FORWARD POWER
- REFLECTED POWER
- NET POWER
- CURRENT
- VOLTAGE
- CALCULATED FIELD
- AVERAGE FIELD
- FIELD 1
- FIELD 2
- FIELD 3
- FIELD 4
- FIELD 5
- FIELD 6
- FIELD 7
- FIELD 8
- AD CHANNEL 1
- AD CHANNEL 2
- AD CHANNEL 3
- AD CHANNEL 4
- AD CHANNEL 5
- AD CHANNEL 6
- AD CHANNEL 7
- AD CHANNEL 8
- CALIBRATION FILE
- TEST ENGINEER
- TEST DATE
- TEST TIME
- START TIME
- STOP TIME
- TEMPERATURE
- HUMIDITY
- PRESSURE
- START FREQUENCY
- STOP FREQUENCY
- FREQUENCY STEP
- LOGARITHMIC STEP
- LOGARITHMIC STEP FROM
- POLARIZATION
- DISTANCE
- TOLERANCE
- DWELL TIME
- MODULATION
- MODULATION DEPTH
- MODULATION CONSERVATION
- MODULATION FREQUENCY
- MODULATION SIGNAL
- MODULATION DUTY CYCLE
- MAINS VOLTAGE
- FREQUENCY CHANGE MODE
- AVERAGE FIELD
- CURRENT MEASURED
- CALIBRATION METHOD
- FORWARD POWER MEASURED
- REFLECTED POWER MEASURED
- NET POWER MEASURED
- CALCULATED FIELD MEASURED
- RECALCULATE AVERAGE
- DESCRIPTION
- TSF FILENAME
- AMPLIFIER PROTECTION MODE
- AUTODISAPPEAR
- USE VOLTAGE TEST LEVEL
- VOLTAGE TEST LEVEL
- POWER TEST LEVEL
- FIELD TEST LEVEL
- USE CURRENT TEST LEVEL
- CURRENT TEST LEVEL
- STEP DOWN
- USE CURRENT LIMIT
- CURRENT LIMIT
- MODULATION TYPE
- MAGNETIC FIELD TEST LEVEL
Macros[edit]
Table Formatting[edit]
Specific graph sizes[edit]
Table-inclusion modes[edit]
- Row insert mode
- Cell insert mode
- Overwrite mode
- Sheet insert mode
Row insert mode[edit]
Cell insert mode[edit]
Overwrite mode[edit]
Sheet insert mode[edit]
Macros[edit]
- CELL_PASTE_MODE_CELLS
- CELL_PASTE_MODE_SHEET
- CELL_PASTE_MODE_OVERWRITE
- CELL_PASTE_MODE_ROWS
SQL queries on tables[edit]
SELECT * FROM <TABLE>
Examples[edit]
| “ | Example to only include the frequencies above 50 MHz:
| “ | Example to only include the frequencies between 1 Hz and 20 MHz, where the AD-channel was measured out of the specified range ('TABLE'), and the table is reverse-sorted on the frequency column:
| “ | Example to include the 'EQUIPMENT TABLE' including all the columns (also the hidden columns):
| “ | Example to include all columns of a table:
- Peak Number
- Frequency
- Peak
- Peak Limit
- Peak status
- Overall status
… which does work.
| 1 | 829.92 MHz | 80 dBµV | 60 dBµV | 20 dB | Fail | Fail |
| “ | Example to limit the number of rows in the table to 3 rows:
will thus result in:
| “ | Example to sort the rows in the table descending on the frequency:
Keywords Overview[edit]
EUT[edit]
Common[edit]
| Keyword | Description |
| ||EUT|FILENAME|| | The filename of the EUT |
| ||EUT|BRAND|| | The brand of the EUT specified in the EUT Window |
| ||EUT|CONDITION UPON RECEPTION|| | The condition of reception of the EUT specified in the EUT Window |
| ||EUT|DATE OF RECEIPT|| | The date of reception of the EUT specified in the EUT Window |
| ||EUT|DATE OF MEASUREMENT|| | The date of measurement of the EUT specified in the EUT Window |
| ||EUT|DIMENSIONS|| | The dimensions of the EUT specified in the EUT Window |
| ||EUT|EMISSION STANDARD|| | The emission standard of the EUT specified in the EUT Window |
| ||EUT|ENVIRONMENT|| | The environment of the EUT specified in the EUT Window |
| ||EUT|IMMUNITY STANDARD|| | The immunity standard of the EUT specified in the EUT Window |
| ||EUT|NAME|| | The name of the EUT specified in the EUT Window |
| ||EUT|ORDER NUMBER|| | The order number of the EUT specified in the EUT Window |
| ||EUT|PCB RELEASE|| | The PCB release of the EUT specified in the EUT Window |
| ||EUT|PERIPHERAL EQUIPMENT|| | The peripheral equipment of the EUT specified in the EUT Window |
| ||EUT|PLACE MEASURMENT|| | The place of measurment of the EUT specified in the EUT Window |
| ||EUT|RADIMATION VERSION|| | The RadiMation® version |
| ||EUT|RATED POWER|| | The rated power of the EUT specified in the EUT Window |
| ||EUT|SERIAL NUMBER|| | The serial number of the EUT specified in the EUT Window |
| ||EUT|TEST SETUP TYPE|| | The test setup type of the EUT specified in the EUT Window |
| ||EUT|SOFTWARE RELEASE|| | The software release of the EUT specified in the EUT Window |
| ||EUT|TYPE NUMBER|| | The type number of the EUT specified in the EUT Window |
| ||EUT|VOLTAGE SUPPLY|| | The voltage supply of the EUT specified in the EUT Window |
| ||EUT|CLIENT ADDRESS|| | The address of the client specified in the EUT Window |
| ||EUT|CLIENT CAPACITY|| | The capacity of the client specified in the EUT Window |
| ||EUT|CLIENT CITY|| | The city of the client specified in the EUT Window |
| ||EUT|CLIENT COMPANY|| | The client name specified in the EUT Window |
| ||EUT|CLIENT CONTACT PERSON|| | The contact person of the client specified in the EUT Window |
| ||EUT|CLIENT COUNTRY|| | The country of the client specified in the EUT Window |
| ||EUT|CLIENT EMAIL ADDRESS|| | The email address of the client specified in the EUT Window |
| ||EUT|CLIENT FAX NUMBER|| | The fax number of the client specified in the EUT Window |
| ||EUT|CLIENT STATE|| | The state of the client specified in the EUT Window |
| ||EUT|CLIENT TELEPHONE NUMBER|| | The telephone number of the client specified in the EUT Window |
| ||EUT|CLIENT ZIP CODE|| | The ZIP code of the client specified in the EUT Window |
| ||EUT|MANUFACTURER ADDRESS|| | The address of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER CITY|| | The city of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER COMPANY|| | The manufacturer name specified in the EUT Window |
| ||EUT|MANUFACTURER CONTACT PERSON|| | The contact person of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER COUNTRY|| | The country of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER EMAIL ADDRESS|| | The email address of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER FAX NUMBER|| | The fax number of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER STATE|| | The state of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER TELEPHONE NUMBER|| | The telephone number of the manufacturer specified in the EUT Window |
| ||EUT|MANUFACTURER ZIP CODE|| | The ZIP code of the manufacturer specified in the EUT Window |
| ||EUT|TEST SITE ADDRESS|| | The address of the test site specified in the EUT Window |
| ||EUT|TEST SITE CITY|| | The city of the test site specified in the EUT Window |
| ||EUT|TEST SITE COMPANY|| | The test site name specified in the EUT Window |
| ||EUT|TEST SITE CONTACT PERSON|| | The contact person of the test site specified in the EUT Window |
| ||EUT|TEST SITE COUNTRY|| | The country of the test site specified in the EUT Window |
| ||EUT|TEST SITE EMAIL ADDRESS|| | The email address of the test site specified in the EUT Window |
| ||EUT|TEST SITE FAX NUMBER|| | The fax number of the test site specified in the EUT Window |
| ||EUT|TEST SITE STATE|| | The state of the test site specified in the EUT Window |
| ||EUT|TEST SITE TELEPHONE NUMBER|| | The telephone number of the test site specified in the EUT Window |
| ||EUT|TEST SITE ZIP CODE|| | The ZIP code of the test site specified in the EUT Window |
Cable information[edit]
| ||EUT|CABLE TABLE|| | A table with the cable information specified in the EUT Window |
| ||EUT|CABLE X PORT TYPE|| | The port type of cable X in the EUT window ( X starting from 1) |
| ||EUT|CABLE X CABLE TYPE|| | The cable type of cable X in the EUT window (X starting from 1) |
| ||EUT|CABLE X CABLE LENGTH|| | The cable length of cable X in the EUT window (X starting from 1) |
| ||EUT|CABLE X FIXING SHIELD|| | The fixing shield of cable X in the EUT window (X starting from 1) |
| ||EUT|CABLE X LOAD AT PORT|| | The load at port of cable X in the EUT window (X starting from 1) |
Standards Information[edit]
| Parameter | Description |
| REFERRED STANDARDS | A string of referred standards, separated by a "," (Only applicable to product standards) |
| “ | Example:
| EDITOR | Report generator |
Directories[edit]
| ||Configuration|DEVICEDRIVERFILESDIRECTORY|| | CONFDVDR directory |
| ||Configuration|CALFILESDIRECTORY|| | CAL Files directory |
| ||Configuration|EUTFILESDIRECTORY|| | EUT Files directory |
| ||Configuration|TSFFILESDIRECTORY|| | TSF Files directory |
| ||Configuration|SEQFILESDIRECTORY|| | SEQ Files directory |
| ||Configuration|CORFILESDIRECTORY|| | COR Files directory |
| ||Configuration|DIRECTORIES|| | All directories |
Units[edit]
| ||Configuration|CURRENT|| | Current |
| ||Configuration|EMISSION AMPLITUDE|| | Emission |
| ||Configuration|FIELD|| | Electrical Field |
| ||Configuration|FREQUENCY|| | Frequency |
| ||Configuration|POWER|| | Power |
| ||Configuration|VOLTAGE|| | Voltage |
| ||Configuration|UNITS|| | All units |
General[edit]
| ||Configuration|CURRENTUSERNAME|| | Current username |
| ||Configuration|RADIMATION VERSION|| | The RadiMation® version |
EUT Calibration[edit]
- Attenuation / Gain
- Amplifier linearity
- Gain compression
- Harmonics
- S Parameter
- Antenna scan
- Antenna
- Field strength
Amplifier Linearity[edit]
| FREQUENCY STEP | Step size |
| ALL GRAPHS | Output graph |
Gain compression[edit]
| TABLE | Output table |
| GRAPH | Output graph |
Harmonics[edit]
S Parameter[edit]
Antenna scan[edit]
| POWER TABLE | Default table with frequency, signal power and forward power. |
| SELECT ALL FROM [Power Table] | To request all possible columns |
Radiated Immunity[edit]
| FREQUENCY STEP | The linear frequency step in megahertz [MHz] |
| FIELD UNIT | The selected unit for field strength: [dBμV/m], [dBmV/m], [dBV/m], [μV/m], [mV/m] or [V/m] |
| POWER UNIT | The selected unit for power: [dBμW], [dBm], [dBmW], [dBW], [μW], [mW] or [W] |
| VOLTAGE UNIT | The selected unit for voltage: [dBμVrms], [dBmVrms], [dBVrms], [μVrms], [mVrms] or [Vrms] |
| NUMBER OF TOTAL POINTS | The total of calibration points |
| NUMBER OF POINTS TO REMOVE | The number of points removed |
| ACCEPTED TOLERANCE | The accepted tolerance in dB |
| ACCEPTED EXTENDED TOLERANCE | The accepted extended tolerance in dB |
| PERCENTAGE WITHIN TOLERANCE | The percentage of frequencies within the accepted tolerance |
| PERCENTAGE WITHIN EXTENDED TOLERANCE | The percentage of frequencies within the accepted extended tolerance |
| PERCENTAGE OUTSIDE TOLERANCE | The percentage of frequencies outside the tolerance |
| MAX STANDARD DEVIATION MEASURED | The maximum standard deviation |
| FREQUENCY FOR MAX STANDARD DEVIATION | The frequency with the maximum standard deviation |
| MAX TOLERANCE DEVIATION MEASURED | The maximum tolerance deviation |
| FREQUENCY FOR MAX DEVIATION | The frequency with the max tolerance |
| RESULT | UFA pass or fail result |
| STDEV GRAPH | The graph representing the standard deviation per frequency |
| PROBABILITY PERCENTAGE | The probability percentage |
| CROSS POLAR PROBABILITY PERCENTAGE | The cross-polar probability percentage |
| CROSS POLAR ACCEPTED TOLERANCE | The cross-polar accepted tolerance in dB |
| CROSS POLAR ACCEPTED EXTENDED TOLERANCE | The cross-polar accepted extended tolerance in dB |
| CROSS POLAR EXTENDED TOLERANCE PERCENTAGE | The cross-polar extended tolerance percentage |
| CROSS POLAR RESULT | The cross-polar result, pass or fail |
| CROSS POLAR PERCENTAGE WITHIN TOLERANCE | The cross-polar percentage within tolerance |
| CROSS POLAR NUMBER OF POINTS WITHIN TOLERANCE | The cross-polar number of frequency points within tolerance |
| CROSS POLAR PERCENTAGE WITHIN EXTENDED TOLERANCE | The cross-polar percentage of frequency points within the extended tolerance |
| CROSS POLAR NUMBER OF POINTS WITHIN EXTENDED TOLERANCE | The cross-polar number of frequency points within the extended tolerance |
| CROSS POLAR PERCENTAGE OUTSIDE TOLERANCE | The cross-polar percentage outside the tolerance |
| CROSS POLAR NUMBER OF POINTS OUTSIDE TOLERANCE | The cross-polar number of points outside the tolerance |
| CROSS POLAR MAX PQUANTILE | The cross-polar maximum PQuantile |
| CROSS POLAR FREQUENCY FOR MAX PQUANTILE | The cross-polar frequency for the max PQuantile |
| PQUANTILE GRAPH | The graph representing the PQuantile per frequency |
| DEVICE DRIVER DLL | N/A |
| INI FILE NAME | N/A |
| “ | Example:
Radiated Immunity Tests[edit]
| CURRENT UNIT | The selected current unit: [dBμA],[dBmA], [dBA], [μA], [mA] or [A] |
| “ | Example:
| “ | Example:
The report generator code is build as following, where <Test level name> is replaced with the name of the configured test level and <Field> is the actual information that is put in the report. ||Band1 TestLevel: <Test level name> <Field>|| This example puts the 'Description' of the test level named 'Electrical field substitution method' in the report ||Band1 TestLevel: Electrical field substitution method Description|| This example puts the 'Calibration filename' of the test level named 'Electrical field substitution method' in the report ||Band1 TestLevel: Electrical field substitution method Calibration File|| This example puts the 'Field' of the test level named 'Electrical field substitution method' in the report ||Band1 TestLevel: Electrical field substitution method Fixed Field|| This example puts the 'Power' of the test level named 'Power closed loop method' in the report ||Band1 TestLevel: Power closed loop method Fixed Power|| |
| Name | Name of test level |
| Description | Description of test level |
| Tolerance | The tolerance of the test level |
| Calibration File | Calibration file used in a substitution test |
| Calibration method | Calibration method |
| Fixed Field | The field strength to regulate to |
Conducted Immunity[edit]
| CURRENT UNIT | The selected current unit: [dBμA], [dBmA], [dBA], [μA], [mA] or [A] |
| “ | Example:
6) Substitution method only
| “ | Example:
Pulsed Immunity[edit]
ESD Test[edit]
| NEG AIR NOTE POINT x | Note of Negative Air Discharge at point number [x] |
| NEG CONTACT NOTE POINT x | Note of Negative Contact Discharge at point number [x] |
| POS AIR NOTE POINT x | Note of Positive Air Discharge at point number [x] |
| POS CONTACT NOTE POINT x | Note of Positive Contact Discharge at point number [x] |
Equipment for ESD Tests[edit]
| “ | Example:
EFT Test[edit]
| “ | Example:
Equipment for EFT Tests[edit]
| “ | Example:
Surge Test[edit]
| “ | Example:
[edit]
| SEQUENCE OVERVIEW | A table containing the sequence configuration of the test |
Band Specific Codes[edit]
BAND + X + <Setting>
EFT[edit]
| BAND1 SequenceDuration | The selected sequence duration specified in the first sequence of the test. |
Surge[edit]
VDI[edit]
| BAND1 Dipcount | The ammount of dips per sequence |
| BAND1 Repeatcount | The ammount the sequence will be repeated |
| “ | Example:
Radiated Emission[edit]
Radiated Emission Tests[edit]
| CURRENT UNIT | The selected unit for current: [dBμA], [dBmA], [dBA], [μA], [mA] or [A] |
| DB/DIV | N/A |
| DB/DIV MODE | N/A |
| OPTIMIZE ANGLE OPTIMIZE AREA | The angle optimization area. |
| OPTIMIZE ANGLE STEP SIZE | The step size of the angle optimization. |
| OPTIMIZE HEIGHT OPTIMIZE AREA | The height optimization area. |
| OPTIMIZE HEIGHT STEP SIZE | The step size of the height optimization. |
| USE INTERNAL PREAMP | N/A |
Radiated Emission Table[edit]
Radiated Emission Burst[edit]
| “ | Example:
Conducted Emission[edit]
| DB/DIV | N/A |
| DB/DIV MODE | N/A |
| “ | Example:
Excel Exporter[edit]
Selecting filename[edit]
Changing the test order[edit]
or ‘arrow down’
Export options[edit]
Performing the export[edit]
It is possible to automatically generate a report from the command line. This can be accomplished by calling "radicli.exe", which is available in the same directory in which RadiMation is also installed.
| Note: | If RadiMation® is not running, the report will be generated in the background with the same version as where the radicli.exe resides. When a RadiMation instance is running which supports the radicli interface, the report will be generated with the running instance |
| “ | Example for a command to generate a report:
The possible arguments are:
| Argument | Description |
| --testnumbers <1[;2][;3][...]> | Optional argument that specifies a ';' separated list of one or more test numbers (1, 2, 3, ....) for the tests (and their order) that should be included in the generated report. These test numbers correspond to the test numbers as can be seen in the EUT window |
| “ | Example to generate a report for each test:
| “ | Example to only include the tests with the testnumber 3, 2 and 4 in the report:
