==== Removing all 'unrecognized' codes ====
Normally when a keyword is neither correct nor recognized by the Report Generator, it will remain unchanged. The keyword {{ReportGeneratorCode|RemoveIllegalCodes(TRUE)}} can automatically remove all these 'errors'. Keep in mind that using this keyword can be both an advantage and a disadvantage. If you are aware of the error(s), then this is a handy way to remove them from the test report. However, if you are not aware of them, this might lead you to miss errors that need to be addressed!
Normally when a keyword is neither correct nor recognized by the Report Generator, it will remain unchanged, and be left in the generated report.
When the report generator code {{ReportGeneratorCode|RemoveIllegalCodes(TRUE)}} is used in the the template, automatically all these unrecognized and illegal report generator codes will be removed from the generated report. This will only be done for the report generator codes that are present after the {{ReportGeneratorCode|RemoveIllegalCodes(TRUE)}} is used.
==== Removing selected text and codes ====
Line 581: Line 584:
{{Example|header=Example to include all columns of a table:|text=The report generator code <code>{{ReportGeneratorCode|SELECT ALL FROM PEAKTABLE}}</code> will show the table including all the available columns, and could result in:
<!-- <wbr> is a HTML element for a 'word-break opportunity', to allow word wrapping at that point. Used at this location to have a table that fits on the page. -->
{{{!}}class="wikitable"
! Peak Number {{!}}{{!}} Peak Number_RAW {{!}}{{!}} Frequency {{!}}{{!}} Frequency_RAW {{!}}{{!}} Peak {{!}}{{!}} Peak_RAW {{!}}{{!}} Peak Limit {{!}}{{!}} Peak Limit_RAW {{!}}{{!}} Peak Difference {{!}}{{!}} Peak Difference_RAW {{!}}{{!}} Peak Correction {{!}}{{!}} Peak Correction_RAW {{!}}{{!}} Peak Status {{!}}{{!}} Peak Status_RAW {{!}}{{!}} Status {{!}}{{!}} Status_RAW {{!}}{{!}} Angle {{!}}{{!}} Angle_RAW {{!}}{{!}}Height {{!}}{{!}} Height_RAW {{!}}{{!}} Polarization {{!}}{{!}} Polarization_RAW
! Peak Number {{!}}{{!}} Peak Number<wbr>_RAW {{!}}{{!}} Frequency {{!}}{{!}} Frequency<wbr>_RAW {{!}}{{!}} Peak {{!}}{{!}} Peak<wbr>_RAW {{!}}{{!}} Peak Limit {{!}}{{!}} Peak Limit<wbr>_RAW {{!}}{{!}} Peak Difference {{!}}{{!}} Peak Difference<wbr>_RAW {{!}}{{!}} Peak Correction {{!}}{{!}} Peak Correction<wbr>_RAW {{!}}{{!}} Peak Status {{!}}{{!}} Peak Status<wbr>_RAW {{!}}{{!}} Status {{!}}{{!}} Status<wbr>_RAW {{!}}{{!}} Angle {{!}}{{!}} Angle<wbr>_RAW
{{!}} 1 {{!}}{{!}} 1 {{!}}{{!}} 829.92 MHz {{!}}{{!}} 829921205 {{!}}{{!}} 80 dBµV/m {{!}}{{!}} 80 {{!}}{{!}} 60 dBµV/m {{!}}{{!}} 60 {{!}}{{!}} 20 dB {{!}}{{!}} 20 {{!}}{{!}} 0.00 dB {{!}}{{!}} 0.00 {{!}}{{!}} Fail {{!}}{{!}} Fail {{!}}{{!}} Fail {{!}}{{!}} Fail {{!}}{{!}} 0 Degree {{!}}{{!}} 0 {{!}}{{!}} 1 m {{!}}{{!}} 1 {{!}}{{!}} Horizontal {{!}}{{!}}Horizontal
Line 596: Line 600:
*Peak difference to the limit
*Peak status
*Peak status
*Peak status overall
*Overall status
Line 614: Line 618:
<TABLE border=1>
{|class="wikitable"
  <TR>
  <TD>Peak Number</TD>
! Peak Number
  <TD>Frequency</TD>
! Frequency
  <TD>Peak</TD>
! Peak
  <TD>Peak Limit</TD>
! Peak Limit
  <TD>Peak Difference</TD>
! Peak Difference
  <TD>Peak Status</TD>
! Peak Status
  <TD>Status</TD>
! Status
  </TR>
  <TR>
| 1
  <TD>1</TD>
| 829.92 MHz
  <TD>829,92 MHz</TD>
| 80 dBµV
  <TD>80 dBµV</TD>
| 60 dBµV
  <TD>60 dBµV</TD>
| 20 dB
  <TD>20 dB</TD>
| Fail
  <TD>Fail</TD>
| Fail
  <TD>Fail</TD>
</TR>
</TABLE>
==== Limiting the number of rows in the table ====
Line 639: Line 641:
{{Example|header=Example to limit the number of rows in the table to 3 rows:|text=
{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE LIMIT 3}}
<code>{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE '''LIMIT 3'''}}</code>
will thus result in:
Line 659: Line 661:
{{Example|header=Example to sort the rows in the table descending on the frequency:|text=
{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE ORDER BY Frequency DESC}}
<code>{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE '''ORDER BY Frequency DESC'''}}</code>
Line 853: Line 855:
{{Example|text=
{{Example|text=
{{ReportGeneratorCode|EUT|PRODUCT STANDARD REFERRED STANDARDS}}
<code>{{ReportGeneratorCode|EUT|PRODUCT STANDARD REFERRED STANDARDS}}</code>
{{ReportGeneratorCode|EUT|PRODUCT STANDARD DESCRIPTION}}
<code>{{ReportGeneratorCode|EUT|PRODUCT STANDARD DESCRIPTION}}</code>
{{ReportGeneratorCode|EUT|SELECT * FROM [BASIC STANDARD TABLE]}}
<code>{{ReportGeneratorCode|EUT|SELECT * FROM [BASIC STANDARD TABLE]}}</code>
{{ReportGeneratorCode|EUT|SELECT ALL FROM [BASIC STANDARD TABLE]}}
<code>{{ReportGeneratorCode|EUT|SELECT ALL FROM [BASIC STANDARD TABLE]}}</code>
{{ReportGeneratorCode|EUT|SELECT [OFFICIAL NAME] FROM [BASIC STANDARD TABLE]}}
<code>{{ReportGeneratorCode|EUT|SELECT [OFFICIAL NAME] FROM [BASIC STANDARD TABLE]}}</code>
Line 1,165: Line 1,167:
| EVENTS || Table of events which occur during a test
| FIELD UNIT || The selected unit for field strength:  [dBuV/m], [dBmV/m], [dBV/m], [uV/m], [mV/m] or  [V/m]
| FIELD UNIT || The selected unit for field strength:  [dBμV/m], [dBmV/m], [dBV/m], [μV/m], [mV/m] or  [V/m]
| POWER UNIT || The selected unit for power: [dBμW], [dBm], [dBmW], [dBW], [μW], [mW] or  [W]
| POWER UNIT || The selected unit for power: [dBuW], [dBm], [dBmW], [dBW], [uW], [mW] or  [W]
| VOLTAGE UNIT || The selected unit for voltage: [dBμVrms], [dBmVrms], [dBVrms], [μVrms], [mVrms] or  [Vrms]
| VOLTAGE UNIT || The selected unit for voltage:  [dBuVrms], [dBmVrms], [dBVrms], [uVrms], [mVrms] or [Vrms]
| NUMBER OF TOTAL POINTS || The total of calibration points
| NUMBER OF POINTS TO REMOVE || The number of points removed
| ACCEPTED TOLERANCE || The accepted tolerance in dB
| ACCEPTED EXTENDED TOLERANCE || The accepted extended tolerance in dB
| PERCENTAGE WITHIN TOLERANCE || The percentage of frequencies within the accepted tolerance
| PERCENTAGE WITHIN EXTENDED TOLERANCE || The percentage of frequencies within the accepted extended tolerance
| PERCENTAGE OUTSIDE TOLERANCE || The percentage of frequencies outside the tolerance
| MAX STANDARD DEVIATION MEASURED || The maximum standard deviation
| FREQUENCY FOR MAX STANDARD DEVIATION || The frequency with the maximum standard deviation
| MAX TOLERANCE DEVIATION MEASURED || The maximum tolerance deviation
| FREQUENCY FOR MAX DEVIATION || The frequency with the max tolerance
| RESULT || UFA pass or fail result
| STDEV GRAPH || The graph representing the standard deviation per frequency
| PROBABILITY PERCENTAGE || The probability percentage
| CROSS POLAR PROBABILITY PERCENTAGE || The cross-polar probability percentage
| CROSS POLAR ACCEPTED TOLERANCE || The cross-polar accepted tolerance in dB
| CROSS POLAR ACCEPTED EXTENDED TOLERANCE || The cross-polar accepted extended tolerance in dB
| CROSS POLAR EXTENDED TOLERANCE PERCENTAGE || The cross-polar extended tolerance percentage
| CROSS POLAR RESULT || The cross-polar result, pass or fail
| CROSS POLAR PERCENTAGE WITHIN TOLERANCE || The cross-polar percentage within tolerance
| CROSS POLAR NUMBER OF POINTS WITHIN TOLERANCE || The cross-polar number of frequency points within tolerance
| CROSS POLAR PERCENTAGE WITHIN EXTENDED TOLERANCE || The cross-polar percentage of frequency points within the extended tolerance
| CROSS POLAR NUMBER OF POINTS WITHIN EXTENDED TOLERANCE || The cross-polar number of frequency points within the extended tolerance
| CROSS POLAR PERCENTAGE OUTSIDE TOLERANCE || The cross-polar percentage outside the tolerance
| CROSS POLAR NUMBER OF POINTS OUTSIDE TOLERANCE || The cross-polar number of points outside the tolerance
| CROSS POLAR MAX PQUANTILE || The cross-polar maximum PQuantile
| CROSS POLAR FREQUENCY FOR MAX PQUANTILE || The cross-polar frequency for the max PQuantile
| PQUANTILE GRAPH || The graph representing the PQuantile per frequency
Line 1,280: Line 1,338:
| CURRENT MEASURED || Indication whether or not the current is measured: Yes / No
| CURRENT UNIT || The selected current unit: [dBuA],[dBmA], [dBA], [uA], [mA] or [A]
| CURRENT UNIT || The selected current unit: [dBμA],[dBmA], [dBA], [μA], [mA] or [A]
| DESCRIPTION || The description of the Test Setup File (TSF)
Line 1,346: Line 1,404:
| MANUAL MODE || Table containing the information collected during a manual mode session
Line 1,426: Line 1,484:
{{Example|text=
{{Example|text=
{{ReportGeneratorCode|Band1 TestLevel: <Test level name> <Field>}}
The report generator code is build as following, where <Test level name> is replaced with the name of the configured test level and <Field> is the actual information that is put in the report.
{{ReportGeneratorCode|Band1 TestLevel: Electrical field substitution method Description}}
This example puts the 'Description' of the test level named 'Electrical field substitution method' in the report
{{ReportGeneratorCode|Band1 TestLevel: Electrical field substitution method Calibration File}}
This example puts the 'Calibration filename' of the test level named 'Electrical field substitution method' in the report
{{ReportGeneratorCode|Band1 TestLevel: Electrical field substitution method Fixed Field}}
This example puts the 'Field' of the test level named 'Electrical field substitution method' in the report
This example puts the 'Power' of the test level named 'Power closed loop method' in the report
{{ReportGeneratorCode|Band1 TestLevel: Power closed loop method Fixed Power}}
Line 1,481: Line 1,546:
| CURRENT TEST LEVEL || The current test level in Ampere [A]
| CURRENT UNIT || The selected current unit: [dBuA], [dBmA], [dBA], [uA], [mA] or  [A]
| CURRENT UNIT || The selected current unit: [dBμA], [dBmA], [dBA], [μA], [mA] or  [A]
Line 1,533: Line 1,598:
Line 1,660: Line 1,725:
Line 1,730: Line 1,795:
| POWER TEST LEVEL (3) || The power test level in the unit selected by the user
Line 2,345: Line 2,410:
| CABLE TYPEOFCABLE || Cable type
| SEQUENCE OVERVIEW || A table containing the sequence configuration of the test
Line 2,374: Line 2,441:
=== EFT ===
=== EFT ===
Line 2,536: Line 2,602:
| CONFIGURATION TIME || The date and time when the TSF file is saved
| CURRENT UNIT || The selected unit for current: [dBuA], [dBmA], [dBA], [uA], [mA] or  [A]
| CURRENT UNIT || The selected unit for current: [dBμA], [dBmA], [dBA], [μA], [mA] or  [A]
| DB/DIV || N/A
| DB/DIV || N/A
Line 2,894: Line 2,960:
| DB/DIV || N/A
| DB/DIV || N/A
Line 3,103: Line 3,169:
== Command Line Interface (CLI) ==
It is possible to automatically generate a report from the command line.
With the command line interface it is possible to automatically generate a report from the command line.
This can be accomplished by calling "radicli.exe", which is available in the same directory in which RadiMation is also installed.
{{Note|When a RadiMation instance is active which supports the radimation cli interface, the report will be generated with the running instance. If {{RadiMation}} is not running, the report will be generated in the background with the same version as where the radicli.exe resides. }}
Line 3,139: Line 3,207:
{{Example|header=Example to generate a report for each test:
|text=radicli.exe generatereport --eut eutfile.eut --template template.txt --report report||testnumber||.txt --reportforeachtest
|text=radicli.exe generatereport --eut eutfile.eut --template template.txt --report report{{ReportGeneratorCode|testnumber}}.txt --reportforeachtest
Latest revision as of 10:54, 8 November 2024
Report Generator[edit]
Keywords[edit]
Keep in mind that:
  • In a 'test-specific' template the Main Keywords are not used, (only the second keywords).
Unrecognized Keywords[edit]
Generating a report[edit]
Making a template file[edit]
Specific Features[edit]
Example:
Client: ||EUT|CLIENT CONTACT PERSON||
||ENDDELETEBLOCK()||
  • If the 'Client Contact Person' data has been entered in the EUT file, the code above will be replaced (with the corresponding data) and the text 'Client: ' and the actual data will remain visible in the report.
  • If no data was entered for the 'Client Contact Person' in the EUT file, the code will not be replaced and the complete block including the text 'Client: ' and the not replace report generator code will be removed from the final report.
  • If multiple report generator keywords are specified within the ||STARTDELETEBLOCK()|| and ||ENDDELETEBLOCK()||, the block will only be removed if all of the report generator codes are not replaced. The complete block will remain in the report if one or more of the report generator codes are replaced.
  • It is allowed and possible to use a ||STARTDELETEBLOCK()|| and ||ENDDELETEBLOCK()|| combination within another ||STARTDELETEBLOCK()|| and ||ENDDELETEBLOCK()|| set.
Test-specific templates[edit]
Example of a 'main-template':
Introduction
||EUT|ALL TESTS||
Conclusion
Example:
||SIGNAL POWER GRAPH; FORWARD POWER GRAPH|| Will create a graph with the signal power and the forward power graph lines.
||FIELD GRAPH 1; FIELD GRAPH 2; AVERAGE FIELD GRAPH|| Will create a field strength graph with the field strength of the probes 1 and 2, and the average field graph.
||CURRENT GRAPH; NET POWER GRAPH|| Will create a graph with the current and the net power. This graph will have two y-axes to indicate the different units.
  • SIGNAL POWER GRAPH
  • CURRENT GRAPH
  • VOLTAGE GRAPH
  • FORWARD POWER GRAPH
  • REFLECTED POWER GRAPH
  • NET POWER GRAPH
  • MID POWER GRAPH
  • OUT POWER GRAPH
  • IMPEDANCE GRAPH
  • ERROR GRAPH
  • CALCULATED FIELD GRAPH
  • AVERAGE FIELD GRAPH
  • CALCULATED VOLTAGE GRAPH
  • CALCULATED CURRENT GRAPH
  • AD GRAPH 1
  • AD GRAPH 2
  • AD GRAPH 3
  • AD GRAPH 4
  • AD GRAPH 5
  • AD GRAPH 6
  • AD GRAPH 7
  • AD GRAPH 8
  • FIELD GRAPH 1
  • FIELD GRAPH 2
  • FIELD GRAPH 3
  • FIELD GRAPH 4
  • FIELD GRAPH 5
  • FIELD GRAPH 6
  • FIELD GRAPH 7
  • FIELD GRAPH 8
  • CALIBRATED POWER GRAPH
  • EUT ATTENUATION GRAPH
  • EUT DYNAMIC RANGE GRAPH
Example:
||POWER; MEASURED POWER LIMIT; Amplifier forward power; Amplifier input||
will result in the following graph:
  • time:
  • frequency:
  • angle:
  • height:
Example:
Export an angle graph: ||angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1||
Export a time graph: ||time:Signal Power;Channel1;Channel2||
Export a height graph: ||height:RBW: 1000 kHz, Horizontal Max Peak for peak 1||
Example:
Export an angle indexed table: ||angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE||
Export an time indexed table: ||time:Signal Power TABLE||
Export a height indexed table: ||height:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE||
  • Frequency
  • Frequency
  • Frequency
Example:
||SIGNAL POWER, FORWARD POWER, AD CHANNEL 5 LIST||
||NET POWER, AD CHANNEL 2, CALCULATED FIELD TABLE||
||VOLTAGE, CALCULATED CURRENT, FORWARD POWER LIST||
  • FREQUENCY
  • SIGNAL POWER
  • FORWARD POWER
  • REFLECTED POWER
  • NET POWER
  • CURRENT
  • VOLTAGE
  • CALCULATED FIELD
  • AVERAGE FIELD
  • FIELD 1
  • FIELD 2
  • FIELD 3
  • FIELD 4
  • FIELD 5
  • FIELD 6
  • FIELD 7
  • FIELD 8
  • AD CHANNEL 1
  • AD CHANNEL 2
  • AD CHANNEL 3
  • AD CHANNEL 4
  • AD CHANNEL 5
  • AD CHANNEL 6
  • AD CHANNEL 7
  • AD CHANNEL 8
  • CALIBRATION FILE
  • TEST ENGINEER
  • TEST DATE
  • TEST TIME
  • START TIME
  • STOP TIME
  • TEMPERATURE
  • HUMIDITY
  • PRESSURE
  • START FREQUENCY
  • STOP FREQUENCY
  • FREQUENCY STEP
  • LOGARITHMIC STEP
  • LOGARITHMIC STEP FROM
  • POLARIZATION
  • DISTANCE
  • TOLERANCE
  • DWELL TIME
  • MODULATION
  • MODULATION DEPTH
  • MODULATION CONSERVATION
  • MODULATION FREQUENCY
  • MODULATION SIGNAL
  • MODULATION DUTY CYCLE
  • MAINS VOLTAGE
  • FREQUENCY CHANGE MODE
  • AVERAGE FIELD
  • CURRENT MEASURED
  • CALIBRATION METHOD
  • FORWARD POWER MEASURED
  • REFLECTED POWER MEASURED
  • NET POWER MEASURED
  • CALCULATED FIELD MEASURED
  • RECALCULATE AVERAGE
  • DESCRIPTION
  • TSF FILENAME
  • AMPLIFIER PROTECTION MODE
  • AUTODISAPPEAR
  • USE VOLTAGE TEST LEVEL
  • VOLTAGE TEST LEVEL
  • POWER TEST LEVEL
  • FIELD TEST LEVEL
  • USE CURRENT TEST LEVEL
  • CURRENT TEST LEVEL
  • STEP DOWN
  • USE CURRENT LIMIT
  • CURRENT LIMIT
  • MODULATION TYPE
  • MAGNETIC FIELD TEST LEVEL
Macros[edit]
  • A macro name predefined by the Microsoft Word report generator editor. These macro's are:
Table Formatting[edit]
Specific graph sizes[edit]
Table-inclusion modes[edit]
  • Row insert mode
  • Cell insert mode
  • Overwrite mode
  • Sheet insert mode
Row insert mode[edit]
Cell insert mode[edit]
Overwrite mode[edit]
Sheet insert mode[edit]
Macros[edit]
    • CELL_PASTE_MODE_CELLS
    • CELL_PASTE_MODE_SHEET
    • CELL_PASTE_MODE_OVERWRITE
    • CELL_PASTE_MODE_ROWS
SQL queries on tables[edit]
SELECT * FROM <TABLE>
Examples[edit]
Example to only include the frequencies above 50 MHz:
Example to only include the frequencies between 1 Hz and 20 MHz, where the AD-channel was measured out of the specified range ('TABLE'), and the table is reverse-sorted on the frequency column:
||SELECT * FROM AD CHANNEL 8,7,6,5,4,3,2,1 TABLE WHERE FREQUENCY >= 1 and FREQUENCY <= 20E6 ORDER BY Frequency desc||
Example to include the 'EQUIPMENT TABLE' including all the columns (also the hidden columns):
The report generator code ||SELECT ALL FROM PEAKTABLE|| will show the table including all the available columns, and could result in:
Peak Number Peak Number_RAW Frequency Frequency_RAW Peak Peak_RAW Peak Limit Peak Limit_RAW Peak Difference Peak Difference_RAW Peak Correction Peak Correction_RAW Peak Status Peak Status_RAW Status Status_RAW Angle Angle_RAW
1 1 829.92 MHz 829921205 80 dBµV/m 80 60 dBµV/m 60 20 dB 20 0.00 dB 0.00 Fail Fail Fail Fail 0 Degree 0
  • Peak Number
  • Frequency
  • Peak
  • Peak Limit
  • Peak status
  • Overall status
… which does work.
1 829.92 MHz 80 dBµV 60 dBµV 20 dB Fail Fail
||SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE LIMIT 3||
1 829.92 MHz 80 dBµV/m 60 dBµV/m 20 dB Fail Fail
2 749.42 MHz 68.8 dBµV/m 60 dBµV/m 8.8 dB Fail Fail
3 951.44 MHz 73.2 dBµV/m 60 dBµV/m 12.2 dB Fail Fail
||SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE ORDER BY Frequency DESC||
4 805.11 MHz 62.1 dBµV/m 60 dBµV/m 1.2 dB Fail Fail
Keywords Overview[edit]
EUT[edit]
Common[edit]
Keyword Description
||EUT|FILENAME|| The filename of the EUT
||EUT|BRAND|| The brand of the EUT specified in the EUT Window
||EUT|CONDITION UPON RECEPTION|| The condition of reception of the EUT specified in the EUT Window
||EUT|DATE OF RECEIPT|| The date of reception of the EUT specified in the EUT Window
||EUT|DATE OF MEASUREMENT|| The date of measurement of the EUT specified in the EUT Window
||EUT|DIMENSIONS|| The dimensions of the EUT specified in the EUT Window
||EUT|EMISSION STANDARD|| The emission standard of the EUT specified in the EUT Window
||EUT|ENVIRONMENT|| The environment of the EUT specified in the EUT Window
||EUT|IMMUNITY STANDARD|| The immunity standard of the EUT specified in the EUT Window
||EUT|NAME|| The name of the EUT specified in the EUT Window
||EUT|ORDER NUMBER|| The order number of the EUT specified in the EUT Window
||EUT|PCB RELEASE|| The PCB release of the EUT specified in the EUT Window
||EUT|PERIPHERAL EQUIPMENT|| The peripheral equipment of the EUT specified in the EUT Window
||EUT|PLACE MEASURMENT|| The place of measurment of the EUT specified in the EUT Window
||EUT|RADIMATION VERSION|| The RadiMation® version
||EUT|RATED POWER|| The rated power of the EUT specified in the EUT Window
||EUT|SERIAL NUMBER|| The serial number of the EUT specified in the EUT Window
||EUT|TEST SETUP TYPE|| The test setup type of the EUT specified in the EUT Window
||EUT|SOFTWARE RELEASE|| The software release of the EUT specified in the EUT Window
||EUT|TYPE NUMBER|| The type number of the EUT specified in the EUT Window
||EUT|VOLTAGE SUPPLY|| The voltage supply of the EUT specified in the EUT Window
Keyword Description
||EUT|CLIENT ADDRESS|| The address of the client specified in the EUT Window
||EUT|CLIENT CAPACITY|| The capacity of the client specified in the EUT Window
||EUT|CLIENT CITY|| The city of the client specified in the EUT Window
||EUT|CLIENT COMPANY|| The client name specified in the EUT Window
||EUT|CLIENT CONTACT PERSON|| The contact person of the client specified in the EUT Window
||EUT|CLIENT COUNTRY|| The country of the client specified in the EUT Window
||EUT|CLIENT EMAIL ADDRESS|| The email address of the client specified in the EUT Window
||EUT|CLIENT FAX NUMBER|| The fax number of the client specified in the EUT Window
||EUT|CLIENT STATE|| The state of the client specified in the EUT Window
||EUT|CLIENT TELEPHONE NUMBER|| The telephone number of the client specified in the EUT Window
||EUT|CLIENT ZIP CODE|| The ZIP code of the client specified in the EUT Window
Keyword Description
||EUT|MANUFACTURER ADDRESS|| The address of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER CITY|| The city of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER COMPANY|| The manufacturer name specified in the EUT Window
||EUT|MANUFACTURER CONTACT PERSON|| The contact person of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER COUNTRY|| The country of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER EMAIL ADDRESS|| The email address of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER FAX NUMBER|| The fax number of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER STATE|| The state of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER TELEPHONE NUMBER|| The telephone number of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER ZIP CODE|| The ZIP code of the manufacturer specified in the EUT Window
Keyword Description
||EUT|TEST SITE ADDRESS|| The address of the test site specified in the EUT Window
||EUT|TEST SITE CITY|| The city of the test site specified in the EUT Window
||EUT|TEST SITE COMPANY|| The test site name specified in the EUT Window
||EUT|TEST SITE CONTACT PERSON|| The contact person of the test site specified in the EUT Window
||EUT|TEST SITE COUNTRY|| The country of the test site specified in the EUT Window
||EUT|TEST SITE EMAIL ADDRESS|| The email address of the test site specified in the EUT Window
||EUT|TEST SITE FAX NUMBER|| The fax number of the test site specified in the EUT Window
||EUT|TEST SITE STATE|| The state of the test site specified in the EUT Window
||EUT|TEST SITE TELEPHONE NUMBER|| The telephone number of the test site specified in the EUT Window
||EUT|TEST SITE ZIP CODE|| The ZIP code of the test site specified in the EUT Window
Cable information[edit]
Keyword Description
||EUT|CABLE TABLE|| A table with the cable information specified in the EUT Window
||EUT|CABLE X PORT TYPE|| The port type of cable X in the EUT window ( X starting from 1)
||EUT|CABLE X CABLE TYPE|| The cable type of cable X in the EUT window (X starting from 1)
||EUT|CABLE X CABLE LENGTH|| The cable length of cable X in the EUT window (X starting from 1)
||EUT|CABLE X FIXING SHIELD|| The fixing shield of cable X in the EUT window (X starting from 1)
||EUT|CABLE X LOAD AT PORT|| The load at port of cable X in the EUT window (X starting from 1)
Standards Information[edit]
Parameter Description
REFERRED STANDARDS A string of referred standards, separated by a "," (Only applicable to product standards)
Example:
||EUT|PRODUCT STANDARD REFERRED STANDARDS||
||EUT|PRODUCT STANDARD DESCRIPTION||
||EUT|SELECT * FROM [BASIC STANDARD TABLE]||
||EUT|SELECT ALL FROM [BASIC STANDARD TABLE]||
||EUT|SELECT [OFFICIAL NAME] FROM [BASIC STANDARD TABLE]||
Keyword Description
EDITOR Report generator
Directories[edit]
Keyword Description
||Configuration|DEVICEDRIVERFILESDIRECTORY|| CONFDVDR directory
||Configuration|CALFILESDIRECTORY|| CAL Files directory
||Configuration|EUTFILESDIRECTORY|| EUT Files directory
||Configuration|TSFFILESDIRECTORY|| TSF Files directory
||Configuration|SEQFILESDIRECTORY|| SEQ Files directory
||Configuration|CORFILESDIRECTORY|| COR Files directory
||Configuration|DIRECTORIES|| All directories
Units[edit]
Keyword Description
||Configuration|CURRENT|| Current
||Configuration|EMISSION AMPLITUDE|| Emission
||Configuration|FIELD|| Electrical Field
||Configuration|FREQUENCY|| Frequency
||Configuration|POWER|| Power
||Configuration|VOLTAGE|| Voltage
||Configuration|UNITS|| All units
General[edit]
Keyword Description
||Configuration|CURRENTUSERNAME|| Current username
||Configuration|RADIMATION VERSION|| The RadiMation® version
EUT Calibration[edit]
  • Attenuation / Gain
  • Amplifier linearity
  • Gain compression
  • Harmonics
  • S Parameter
  • Antenna scan
  • Antenna
  • Field strength
Amplifier Linearity[edit]
Keyword Description
FREQUENCY STEP Step size
ALL GRAPHS Output graph
Gain compression[edit]
Keyword Description
TABLE Output table
GRAPH Output graph
POWER TABLE Table with all the compression power levels for every frequency
Harmonics[edit]
Keyword Description
TABLE Output table
GRAPH Output graph
S Parameter[edit]
Keyword Description
ALL GRAPHS All available graphs plotted in one graph
Sxx PRESCAN GRAPH Prescan Graph
Sxx MAGNITUDE AVG GRAPH Magnitude Average Graph
Sxx MAGNITUDE MAX GRAPH Magnitude Maximum Graph
Sxx MAGNITUDE MIN GRAPH Magnitude Minimum Graph
Sxx MAGNITUDE STDDEV GRAPH Magnitude Standard Deviation Graph
Sxx IMPEDANCE AVG GRAPH Graph with the average of the measured impedance
Sxx IMPEDANCE MAX GRAPH Graph with the max-hold of the measured impedance
Sxx IMPEDANCE MIN GRAPH Graph with the min-hold of the measured impedance
Sxx IMPEDANCE STDDEV GRAPH Graph with the Standard Deviation of the measured impedance
Keyword Description
S11 TABLE All columns default visible
Select * FROM S11 TABLE Using sql to retrieve all values
Select IMPEDANCE FROM S11 TABLE Because we added impedance later, it is not visible by default
Antenna scan[edit]
Keyword Description
POWER TABLE Default table with frequency, signal power and forward power.
SELECT ALL FROM [Power Table] To request all possible columns
Radiated Immunity[edit]
Keyword Description
Keyword Description
FREQUENCY STEP The linear frequency step in megahertz [MHz]
STEP A textual description of the used frequency step, also containing if it is linear or logarithmic step size
FIELD UNIT The selected unit for field strength: [dBμV/m], [dBmV/m], [dBV/m], [μV/m], [mV/m] or [V/m]
POWER UNIT The selected unit for power: [dBμW], [dBm], [dBmW], [dBW], [μW], [mW] or [W]
VOLTAGE UNIT The selected unit for voltage: [dBμVrms], [dBmVrms], [dBVrms], [μVrms], [mVrms] or [Vrms]
NUMBER OF TOTAL POINTS The total of calibration points
NUMBER OF POINTS TO REMOVE The number of points removed
ACCEPTED TOLERANCE The accepted tolerance in dB
ACCEPTED EXTENDED TOLERANCE The accepted extended tolerance in dB
PERCENTAGE WITHIN TOLERANCE The percentage of frequencies within the accepted tolerance
PERCENTAGE WITHIN EXTENDED TOLERANCE The percentage of frequencies within the accepted extended tolerance
PERCENTAGE OUTSIDE TOLERANCE The percentage of frequencies outside the tolerance
MAX STANDARD DEVIATION MEASURED The maximum standard deviation
FREQUENCY FOR MAX STANDARD DEVIATION The frequency with the maximum standard deviation
MAX TOLERANCE DEVIATION MEASURED The maximum tolerance deviation
FREQUENCY FOR MAX DEVIATION The frequency with the max tolerance
RESULT UFA pass or fail result
STDEV GRAPH The graph representing the standard deviation per frequency
PROBABILITY PERCENTAGE The probability percentage
CROSS POLAR PROBABILITY PERCENTAGE The cross-polar probability percentage
CROSS POLAR ACCEPTED TOLERANCE The cross-polar accepted tolerance in dB
CROSS POLAR ACCEPTED EXTENDED TOLERANCE The cross-polar accepted extended tolerance in dB
CROSS POLAR EXTENDED TOLERANCE PERCENTAGE The cross-polar extended tolerance percentage
CROSS POLAR RESULT The cross-polar result, pass or fail
CROSS POLAR PERCENTAGE WITHIN TOLERANCE The cross-polar percentage within tolerance
CROSS POLAR NUMBER OF POINTS WITHIN TOLERANCE The cross-polar number of frequency points within tolerance
CROSS POLAR PERCENTAGE WITHIN EXTENDED TOLERANCE The cross-polar percentage of frequency points within the extended tolerance
CROSS POLAR NUMBER OF POINTS WITHIN EXTENDED TOLERANCE The cross-polar number of frequency points within the extended tolerance
CROSS POLAR PERCENTAGE OUTSIDE TOLERANCE The cross-polar percentage outside the tolerance
CROSS POLAR NUMBER OF POINTS OUTSIDE TOLERANCE The cross-polar number of points outside the tolerance
CROSS POLAR MAX PQUANTILE The cross-polar maximum PQuantile
CROSS POLAR FREQUENCY FOR MAX PQUANTILE The cross-polar frequency for the max PQuantile
PQUANTILE GRAPH The graph representing the PQuantile per frequency
Keyword Description
DEVICE DRIVER DLL N/A
INI FILE NAME N/A
SUPPORTED DEVICE TYPES N/A
Example:
||AMPLIFIER BRAND|| returns the brand name of the amplifier.
Radiated Immunity Tests[edit]
Keyword Description
CURRENT UNIT The selected current unit: [dBμA],[dBmA], [dBA], [μA], [mA] or [A]
Keyword Description
INI FILE NAME N/A
Example:
||AMPLIFIER SERIAL NUMBER|| returns the serial number of the amplifier.
Example:
The report generator code is build as following, where <Test level name> is replaced with the name of the configured test level and <Field> is the actual information that is put in the report.
||Band1 TestLevel: <Test level name> <Field>||
This example puts the 'Description' of the test level named 'Electrical field substitution method' in the report
||Band1 TestLevel: Electrical field substitution method Description||
This example puts the 'Calibration filename' of the test level named 'Electrical field substitution method' in the report
||Band1 TestLevel: Electrical field substitution method Calibration File||
This example puts the 'Field' of the test level named 'Electrical field substitution method' in the report
||Band1 TestLevel: Electrical field substitution method Fixed Field||
This example puts the 'Power' of the test level named 'Power closed loop method' in the report
||Band1 TestLevel: Power closed loop method Fixed Power||
Other fields than "Description" and "Calibration File" can be accessed, keep in mind that it depends on the test level which fields are exposed. Some examples of other fields:
Name Name of test level
Description Description of test level
Tolerance The tolerance of the test level
Calibration File Calibration file used in a substitution test
Calibration method Calibration method
Fixed Field The field strength to regulate to
Conducted Immunity[edit]
Keyword Description
CURRENT UNIT The selected current unit: [dBμA], [dBmA], [dBA], [μA], [mA] or [A]
Keyword Description
INI FILE NAME N/A
Example:
||COUPLER TYPE|| returns the type of the coupler.
Keyword Description
Keyword Description
INI FILE NAME N/A
Example:
||ANTENNA DESCRIPTION|| returns the description of the antenna.
Pulsed Immunity[edit]
ESD Test[edit]
Keyword Description
NEG AIR NOTE POINT x Note of Negative Air Discharge at point number [x]
NEG CONTACT NOTE POINT x Note of Negative Contact Discharge at point number [x]
POS AIR NOTE POINT x Note of Positive Air Discharge at point number [x]
POS CONTACT NOTE POINT x Note of Positive Contact Discharge at point number [x]
POS HCP MEASURED POINT X Indication a Positive HCP Discharge is performed at point number [x]: Yes / No
NEG HCP MEASURED POINT X Indication a Negative HCP Discharge is performed at point number [x]: Yes / No
POS HCP RESULT POINT X Results of Positive HCP Discharge at point number [x]: pass or fail
NEG HCP RESULT POINT X Results of Negative HCP Discharge at point number [x]: pass or fail
POS HCP LEVEL POINT X Level of the Positive HCP Discharge at point number [x] in [V]
NEG HCP LEVEL POINT X Level of the Negative HCP Discharge at point number [x] in [V]
POS HCP NOTE POINT X Note or remark for the Positive HCP Discharge at point number [x]
NEG HCP NOTE POINT X Note or remark for the Negative HCP Discharge at point number [x]
POS VCP MEASURED POINT X Indication a Positive VCP Discharge is performed at point number [x]: Yes / No
NEG VCP MEASURED POINT X Indication a Negative VCP Discharge is performed at point number [x]: Yes / No
POS VCP RESULT POINT X Results of Positive VCP Discharge at point number [x]: pass or fail
NEG VCP RESULT POINT X Results of Negative VCP Discharge at point number [x]: pass or fail
POS VCP LEVEL POINT X Level of the Positive VCP Discharge at point number [x] in [V]
NEG VCP LEVEL POINT X Level of the Negative VCP Discharge at point number [x] in [V]
POS VCP NOTE POINT X Note or remark for the Positive VCP Discharge at point number [x]
NEG VCP NOTE POINT X Note or remark for the Negative VCP Discharge at point number [x]
Equipment for ESD Tests[edit]
Keyword Description
INI FILE NAME N/A
Example:
||ESDGUN BRAND|| returns the brand name of the ESD gun.
EFT Test[edit]
Keyword Description
INI FILE NAME N/A
Keyword Description
Example:
||L1 / L2 RESULT|| is a valid code, and will return the actual result of the L1/L2 line.
Equipment for EFT Tests[edit]
Keyword Description
INI FILE NAME N/A
Example:
||EFTGENERATOR TYPE|| returns the type of the EFT generator.
Surge Test[edit]
Keyword Description
Keyword Description
Keyword Description
INI FILE NAME N/A
Example:
||SURGEGENERATOR ID|| returns the internal ID of the Surge generator.
Keyword Description
CABLE TYPEOFCABLE Cable type
SEQUENCE OVERVIEW A table containing the sequence configuration of the test
Band Specific Codes[edit]
BAND + X + <Setting>
Keyword Description
EFT[edit]
Keyword Description
BAND1 SequenceDuration The selected sequence duration specified in the first sequence of the test.
Surge[edit]
Keyword Description
VDI[edit]
Keyword Description
BAND1 Dipcount The ammount of dips per sequence
BAND1 Repeatcount The ammount the sequence will be repeated
Example:
||STARTDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
Radiated Emission[edit]
Radiated Emission Tests[edit]
Keyword Description
CURRENT UNIT The selected unit for current: [dBμA], [dBmA], [dBA], [μA], [mA] or [A]
DB/DIV N/A
DB/DIV MODE N/A
OPTIMIZE ANGLE OPTIMIZE AREA The angle optimization area.
OPTIMIZE ANGLE STEP SIZE The step size of the angle optimization.
OPTIMIZE HEIGHT OPTIMIZE AREA The height optimization area.
OPTIMIZE HEIGHT STEP SIZE The step size of the height optimization.
Keyword Description
FREQUENCY STEP The step frequency in kilo Hertz [kHz]
USE INTERNAL PREAMP N/A
Keyword Description
Radiated Emission Table[edit]
Keyword Description
Radiated Emission Burst[edit]
Keyword Description
BURST GRAPH The graph plot
RMS The RMS value in [dBm]
Keyword Description
INI FILE NAME N/A
Example:
||ANTENNA TOWER BRAND|| returns the brand name of the antenna tower.
Conducted Emission[edit]
Keyword Description
DB/DIV N/A
DB/DIV MODE N/A
Keyword Description
Keyword Description
INI FILE NAME N/A
Example:
||CABLE 5 BRAND|| returns the brand name of cable nr. 5.
Excel Exporter[edit]
Selecting filename[edit]
Changing the test order[edit]
Export options[edit]
Performing the export[edit]
Note: When a RadiMation instance is active which supports the radimation cli interface, the report will be generated with the running instance. If RadiMation® is not running, the report will be generated in the background with the same version as where the radicli.exe resides.
Example for a command to generate a report:
cd "C:\Program Files (x86)\Raditeq\RadiMation\Version x.y.z\"
radicli.exe generatereport --eut "C:\EUT_Files\eutfile.eut" --template "C:\templates\template.txt" --report "C:\reports\report.txt"
The possible arguments are:
Argument Description
generatereport Required verb that specifies that a report should be generated from the command line interface.
--eut <eutfilename> Required argument that specifies the EUT filename that should be used to generate the report.
--template <templatefilename> Required argument that specifies the template filename to use.
--report <reportfilename> Required argument that specifies the report filename to use.
--reportforeachtest Optional argument that specifies that a report is generated for each test individually. This will result in multiple reports being generated. A report generator code can be specified in the <reportfilename> to have a dynamic report filename. See the example below this table.
--testnumbers <1[;2][;3][...]> Optional argument that specifies a ';' separated list of one or more test numbers (1, 2, 3, ....) for the tests (and their order) that should be included in the generated report. These test numbers correspond to the test numbers as can be seen in the EUT window
--removeillegalcodes Optional argument that specifies that report generator codes that are not replaced to an actual value will be removed from the generated report.
Example to generate a report for each test:
radicli.exe generatereport --eut eutfile.eut --template template.txt --report report||testnumber||.txt --reportforeachtest
Example to only include the tests with the testnumber 3, 2 and 4 in the report:
radicli.exe generatereport --eut eutfile.eut --template template.txt --report report.txt --testnumbers "3;2;4"