(341 intermediate revisions by 10 users not shown)
__NOTOC__
__NOTOC__
=Report Generator=
= Report Generator =
The [[Report Generator]] is a software module to automatically generate a test report from tests that are carried out by {{RadiMation}}. Once {{RadiMation}} had carried out a test, all collected test data is available to the {{RadiMation}} report generator, and this data can be merged into a test report.
Using the Report Generator module, a test engineer can save a huge amount of time. The Report Generator only needs a few minutes to generate a complete test report.
Test reports can be written in Microsoft Word, Microsoft Excel or Text format.
== Introduction ==
The Report Generator is a software module that can automatically generate a test report from tests carried out by {{RadiMation}}. This function can save a test engineer a lot of time, since the Report Generator only needs a few minutes to generate a complete test report. These test reports can be written in Microsoft Word, Microsoft Excel or Text format.
==Generating a report==
The basic information needed to generate a test report, as well as the more specific features of the Report Generator, are explained in the following paragraphs.
Open an [[EUT]] and press the {{ScreenElement|Report}} button on the {{ScreenElement|Reports}} tab of the EUT window. This button shows the screen in which the settings for the generation of the report can be selected.
== About the Report Generator ==
[[Image:Report Generator Settings Window.png]]
=== Template files and Report files ===
The '''Report file''' is the 'output' of the Report Generator and the actual test report. It contains all the desired data from the tests performed in the form of text, graphs, tables, etc. . Depending on the Template file, this report will be written in Microsoft Word, Microsoft Excel or Text format.
The '''Template file''' is the 'input' of the Report Generator. It contains the data for the Report file in the form of specific keywords. The Report Generator then replaces these keywords with test data to produce the actual test report (i.e. the Report file). Depending on the wishes of the test engineer, this file can be written in Microsoft Word, Microsoft Excel or Text format.
To select a template file, select the folder button for the template or click on the path. The red cross is for deleting the file selection, not the file itself. The magnifying glass is for viewing the template. If you want to print the report directly after it has been generated then check the check box for 'Print Report'. If you want to edit the report then check the 'Edit Report' check box. By default all tests are included for generating a report. If you want to alter the selection press the 'Advanced Option' button. An additional screen will be displayed like as shown below.
[[Image:Report Generator Settings Window Advanced Options.png]]
All the tests with the 'Include' checks selected will be included in the final report. Selecting the check box of a test will alter the include/exclude of this specific test. With the "Select all" button, all tests will be included. With the "Unselect all" button, all tests will be excluded from the final report. "Invert selection" will invert all the including/excluding. Selecting the "Generate" button will start the report generation process. This process can not be paused or cancelled be the end user. The report will be finished in a few moments. The generation time varies between a couple of seconds for a text file to several minutes for an extensive report containing many tables and graphs in MS Word.
===Template files and Report files===
The report Generator uses two files to generate a report. These two files are the Template file and the Report file. The Template file is the complete Report containing all the necessary keywords. The test Data is represented by unique keywords, which tell the report generator which data to insert at the position of the code. The Report Generator changes (replaces) all keywords into test data and will create the actual report. This Report file is the complete test report including all test data and is thus the output file. The Template file is not altered. It is recommended to make this Template file read-only, so no one can accidentally delete or modify this file. The Template file must exist, the Report file will be created, if it doesn't exist. The Template file and report file may not be the same file!
===Keywords and Main Keywords===
=== Keywords ===
A template file contains one or more unique keywords. These keywords consist of a Main Keyword and a second Keyword. The codes are of the following type: ||Main keyword | second keyword|| The Main Keyword indicates the part of {{RadiMation}} from which the data must be retrieved. The Keyword specifies the data to be merged into the report file. The Main Keyword can be:
* EUT
In the (generic) Template file these keywords are coded as followed:  {{ReportGeneratorCode| Main keyword | second keyword }}
* "Configuration" - The name of a test specified in the list in the EUT Window
Every Main Keyword has a set of Keywords that can only be used in combination with the used Main Keyword.
The '''Main Keyword''' indicates the part of {{RadiMation}} from which the data must be retrieved, such as; EUT, Configuration, Radiated Immunity, etc. (A list of Main Keywords and their complementing Secondary Keywords is visible further in this chapter.)
The '''Second Keyword''' indicates the data that must be retrieved, such as; filename, test start time, frequency, etc. (A list of Main Keywords and their complementing Secondary Keywords is visible further in this chapter.)
The EUT Main Keyword can use the keyword "Serial Number", but the "Serial Number" keyword is not supported by the "Configuration" Main Keyword. A list of Keywords is given in the succeeding paragraphs of in this manual.
The keywords must be spelled correctly, including spaces and other symbols. To separate multiple words, only one space is used. Keywords and Main keywords are not case sensitive. When a Keyword or Main Keyword is not recognised, or when there is no test data available for this Keyword or Main keyword, the code will not be altered. The test engineer will be given a message in the settings screen of the Report Generator. One failure will be added to the total amount of failures. If a template file is altered, or the test-name has changed, the test engineer should search a generated report for codes from the template file for he may have used codes that contain no data or are not recognised by the Report Generator.
===Available Report Generator Codes for EUT's===
The Main keyword for EUT data is "EUT". All data entered in the EUT Window in {{RadiMation}} can be merged into a report file. All data can be merged as single lines of text or in tables. Tables are supported for Client data, Manufacturer data and Test site data. In addition, a table with all information can be merged.
Keep in mind that:
* The Keywords must be spelled correctly, including spaces and other symbols.
* To separate multiple words, only one space is used.
* Keywords are not case sensitive.
* Keywords that are not recognized cannot be replaced with data.
* Unrecognized Keywords will appear in the Report file 'unchanged' (as codes).
* In a 'test-specific' template the Main Keywords are not used, (only the second keywords).
===Available Report Generator Keywords===
==== Unrecognized Keywords ====
As different tests use different equipment and perform different actions each test has its own set of keywords. In many cases tests share the same keywords although this will refer to different data as the combination of main keyword and secondary keyword will always result in a unique combination.
Errors are easily located by searching the report for the characters "||". Once found these can be deleted, or changed into an existing code. If the Report Generator is executed once again, using the modified file as the template, the Report Generator will finish the report without errors.
The ||STARTDELETEBLOCK()|| and ||ENDDELETEBLOCK()|| keyword are removing all the text and keywords in between these two tags, provided that nothing is replaced by any report generator keyword. This helps keeping your report automatically clean when certain parameters are not used when generating the report.
* The Keyword is spelled incorrectly.
* The data it refers to does not exist/no longer exists.
* The data it refers to does not exist/no longer exists in the specified location.
* The data it refers to does not exist/no longer exists under the specified name.
Normally when a keyword is not correct or recognized by the report generator, it will remain unchanged. With the keyword ||RemoveIllegalCodes(TRUE)|| it is possible to automatically remove all these keywords. Be aware that when you are searching for incorrect keywords and missing data, it can be very convenient when this function is not enabled. Then you are able to find the 'not replaced' keywords easily.
==EUT==
== Generating a report ==
===Common===
!Keyword||Description
|FILENAME ||The filename of the EUT
|BRAND || The brand of the EUT specified in the EUT Window
|CONDITION RECEPTION || The condition of reception of the EUT specified in the EUT Window
|DATE RECEIPT || The date of reception of the EUT specified in the EUT Window
|DATE MEASUREMENT || The date of measurement of the EUT specified in the EUT Window
|DIMENSIONS || The dimensions of the EUT specified in the EUT Window
|EMISSION STANDARD || The emission standard of the EUT specified in the EUT Window
|ENVIRONMENT || The environment of the EUT specified in the EUT Window
|IMMUNITY STANDARD || The immunity standard of the EUT specified in the EUT Window
|NAME || The name of the EUT specified in the EUT Window
|ORDER NUMBER || The order number of the EUT specified in the EUT Window
|PCB RELEASE || The PCB release of the EUT specified in the EUT Window
|PERIPHERAL EQUIPMENT || The peripheral equipment of the EUT specified in the EUT Window
|PLACE MEASURMENT || The place of measurment of the EUT specified in the EUT Window
|RADIMATION VERSION || The {{RadiMation}} version
|RATED POWER || The rated power of the EUT specified in the EUT Window
|SERIAL NUMBER || The serial number of the EUT specified in the EUT Window
|SETUP TYPE || The test setup type of the EUT specified in the EUT Window
|SOFTWARE RELEASE || The software release of the EUT specified in the EUT Window
|TYPE NUMBER || The type number of the EUT specified in the EUT Window
|VOLTAGE SUPPLY || The voltage supply of the EUT specified in the EUT Window
===Client Address Information===
!Keyword || Description
|CLIENT ADDRESS || The address of the client specified in the EUT Window
|CLIENT CAPACITY || The capacity of the client specified in the EUT Window
=== Step 1. Making and selecting a Template file ===
|CLIENT CITY || The city of the client specified in the EUT Window
|CLIENT COMPANY || The client name specified in the EUT Window
|CLIENT CONTACT PERSON || The contact person of the client specified in the EUT Window
|CLIENT COUNTRY || The country of the client specified in the EUT Window
|CLIENT EMAIL ADDRESS || The email address of the client specified in the EUT Window
|CLIENT FAX NUMBER || The fax number of the client specified in the EUT Window
|CLIENT STATE || The state of the client specified in the EUT Window
|CLIENT TELEPHONE NUMBER || The telephone number of the client specified in the EUT Window
|CLIENT ZIP CODE || The ZIP code of the client specified in the EUT Window
===Manufacturer Address Information===
==== Making a template file ====
|MANUFACTURER ADDRESS || The address of the manufacturer specified in the EUT Window
|MANUFACTURER CITY || The city of the manufacturer specified in the EUT Window
|MANUFACTURER COMPANY || The manufacturer name specified in the EUT Window
|MANUFACTURER CONTACT PERSON || The contact person of the manufacturer specified in the EUT Window
|MANUFACTURER COUNTRY || The country of the manufacturer specified in the EUT Window
|MANUFACTURER EMAIL ADDRESS || The email address of the manufacturer specified in the EUT Window
|MANUFACTURER FAX NUMBER || The fax number of the manufacturer specified in the EUT Window
|MANUFACTURER STATE || The state of the manufacturer specified in the EUT Window
|MANUFACTURER TELEPHONE NUMBER || The telephone number of the manufacturer specified in the EUT Window
|MANUFACTURER ZIP CODE || The ZIP code of the manufacturer specified in the EUT Window
===Test Site Address Information===
In a template file made in '''Microsoft Word''':
* Report generator codes can be included 'as is' (meaning, without context).
* Report generator codes can be included in a table.
* Report generator codes can be included in a header or footer.
* Report generator codes can be included in a heading.
* Report generator codes can be combined with other text and report generator codes in a cell of a table.
* Report generator codes that include texts into the final report are allowed.
* Report generator codes that include tables into the final report are allowed.
* Report generator codes that include pictures into the final report are allowed.
In a template file made in '''Microsoft Excel''':
* Report generator codes can be included 'as is' (meaning, without context) in a single cell.
* Report generator codes can be combined with other text in a cell.
|TEST SITE ADDRESS || The address of the test site specified in the EUT Window
* Multiple report generator codes (and also text) can be combined in a cell.
* (Multiple) Report generator codes can be included in a note attached to a cell.
|TEST SITE CITY || The city of the test site specified in the EUT Window
* Multiple worksheets are allowed, all worksheets will be checked for report generator codes.
* Report generator codes can be used as part of a formula of a cell. However it might be necessary to convert the contents from a text to numerical value. The simplest form of a formula is: "=VALUE("{{ReportGeneratorCode|START FREQUENCY}}")".
|TEST SITE COMPANY || The test site name specified in the EUT Window
|TEST SITE CONTACT PERSON || The contact person of the test site specified in the EUT Window
|TEST SITE COUNTRY || The country of the test site specified in the EUT Window
|TEST SITE EMAIL ADDRESS || The email address of the test site specified in the EUT Window
==== Selecting a template file in RadiMation ====
|TEST SITE FAX NUMBER || The fax number of the test site specified in the EUT Window
To generate a report, open the ''' 'EUT' window''' and go to the ''' 'Reports' tab'''. Click ''' 'Report' ''' to open the ''' 'Report Generator settings' window'''.
|TEST SITE STATE || The state of the test site specified in the EUT Window
|TEST SITE TELEPHONE NUMBER || The telephone number of the test site specified in the EUT Window
|TEST SITE ZIP CODE || The ZIP code of the test site specified in the EUT Window
===Cable information===
|CABLE TABLE || A table with the cable information specified in the EUT Window
|CABLE X PORT TYPE || The port type of cable X in the EUT window ( X starting from 1)
|CABLE X CABLE TYPE || The cable type of cable X in the EUT window (X starting from 1)
|CABLE X CABLE LENGTH || The cable length of cable X in the EUT window (X starting from 1)
|CABLE X FIXING SHIELD || The fixing shield of cable X in the EUT window (X starting from 1)
|CABLE X LOAD AT PORT || The load at port of cable X in the EUT window (X starting from 1)
===Standards Information===
Keyword Description
=== Step 2. Viewing and/or altering the report content ===
PRODUCT STANDARD ... Specifications of the used Product Standard
BASIC STANDARD ... Specifications of the used Basic Standard
By default all the tests performed on the EUT are included in the report. To view these tests, click the ''' 'Advanced Options' button''' in the ''' 'Report Generator settings' window'''.
The tests that are included in the report, as well as the order in which they are displayed, is visible in the ''' 'Included Tests' section''' of the ''' 'Report Generator settings' window'''. Excluding tests from the report and/or changing the order of the tests can be achieved in this section as well.
!Parameter || Description
Click the ''' 'Advanced Options' button''' in the ''' 'Report Generator settings' window''' to open this field.
|DATEOFRELEASE || Release date of the standard
To '''change the order of the list''', select a test from the list and change its position by using the arrows on the right side of the field.
|DESCRIPTION || String containing official name / scope
To '''include/exclude tests from the list''', use the buttons just below the field. Only the tests that are marked be included in the final report.
|FILELOCATION || The location of the file associated with the standard
|LANGUAGE || The language in which the standard is described
=== Step 3. Generating the report ===
|NOTE || Additional note to the standard
The report can be generated once the template file has been selected and the content of the report has been determined.
|OFFICIALNAME || The official name of the standard
The report will be generated when you click the ''' 'Generate' button'''. If you want to print the report directly after it has been generated, check the ''' 'Print Report' checkbox''' on the bottom left of the window. If you want to open the report in your screen (and possibly edit it as well), check the ''' 'Edit Report' checkbox'''.
|REFERRED STANDARDS || A string of referred standards, separated by a ","
|SCOPE || The scope of the standard
|STANDARDISATIONINSTITUTE || Description of Measurement equipment given by the user
|TABLE || Table of the used standards
==Configuration ==
The report generation process cannot be paused or cancelled by the end user. Depending on the content, this process can take a few seconds (for a text file) up to several minutes (for an extensive report containing many tables and graphs in MS Word).
=== Report Generator ===
== Specific Features ==
=== Available report generator codes for EUT's ===
The Main Keyword for EUT data is "EUT". All the data entered in the EUT Window in {{RadiMation}} can be placed in one Report file. This data can be added as single lines of text or in tables. Tables are supported for Client data, Manufacturer data and Test site data. In addition, a table with all the information can be inserted as well.
|COMNAME || Internal COM Name selected report generator
=== Removing unrecognized codes 'automatically' ===
|DESCRIPTION || Selected report generator
|EXTENSION || Report generator extension
==== Removing all 'unrecognized' codes ====
Normally when a keyword is neither correct nor recognized by the Report Generator, it will remain unchanged, and be left in the generated report.
|CHECKEXTENSION || Report generator should check extension
When the report generator code {{ReportGeneratorCode|RemoveIllegalCodes(TRUE)}} is used in the the template, automatically all these unrecognized and illegal report generator codes will be removed from the generated report. This will only be done for the report generator codes that are present after the {{ReportGeneratorCode|RemoveIllegalCodes(TRUE)}} is used.
|EDITOR || Report generator
|STARTDELETEBLOCK() || Start block to remove unused keywords
==== Removing selected text and codes ====
The special {{ReportGeneratorCode|STARTDELETEBLOCK()}} and {{ReportGeneratorCode|ENDDELETEBLOCK()}} keywords identify a block of text that will be removed from the report if none of the report generator codes between the tags are replaced.
|ENDDELETEBLOCK() || End block to remove unused keywords
|RemoveIllegalCodes(TRUE) || Removal illegal keywords
=== Directories ===
{{Example|text={{ReportGeneratorCode|STARTDELETEBLOCK()}}
|DEVICEDRIVERFILESDIRECTORY || CONFDVDR directory
|CALFILESDIRECTORY || CAL Files directory
|EUTFILESDIRECTORY || EUT Files directory
|TSFFILESDIRECTORY || TSF Files directory
|SEQFILESDIRECTORY || SEQ Files directory
|CORFILESDIRECTORY || COR Files directory
|DIRECTORIES || All directories
Client: {{ReportGeneratorCode|EUT|CLIENT CONTACT PERSON}}
=== Units ===
{{ReportGeneratorCode|ENDDELETEBLOCK()}} }}
* If the 'Client Contact Person' data has been entered in the EUT file, the code above will be replaced (with the corresponding data) and the text 'Client: ' and the actual data will remain visible in the report.
* If no data was entered for the 'Client Contact Person' in the EUT file, the code will '''not''' be replaced and the complete block including the text 'Client: ' and the not replaced report generator code will be removed from the final report.
|CURRENT || Current
* If multiple report generator keywords are specified within the {{ReportGeneratorCode|STARTDELETEBLOCK()}} and {{ReportGeneratorCode|ENDDELETEBLOCK()}}, the block will only be removed if all of the report generator codes are not replaced. The complete block will remain in the report if one or more of the report generator codes are replaced.
* It is allowed and possible to use a {{ReportGeneratorCode|STARTDELETEBLOCK()}} and {{ReportGeneratorCode|ENDDELETEBLOCK()}} combination within another {{ReportGeneratorCode|STARTDELETEBLOCK()}} and {{ReportGeneratorCode|ENDDELETEBLOCK()}} set.
|EMISSION AMPLITUDE || Emission
|FIELD || Electrical Field
=== Test-specific templates ===
|FREQUENCY || Frequency
The [[report generator]] has been extended in such a way, that it is no longer necessary to create one single template that includes all the codes for all the (performed) tests. Instead, it is now possible to link a 'test-specific' template to each test and then combine these into one 'main-template'. (This is an additional method to create a template file and does NOT replace the original method. Templates made with the original method will still work correctly.)
|POWER || Power
* Test-specific template files contain simpler codes, are more compact and easier to 're-use'.
|VOLTAGE || Voltage
* The main-template is easier to set up and more accessible because its content is divided into more manageable chunks.
|UNITS || All units
==== Example of a 'main-template' ====
=== General ===
{{Example|header=Example of a 'main-template':
|text=
Introduction
| CURRENTUSERNAME || Current username
{{ReportGeneratorCode|EUT|ALL TESTS}}
==Radiated Immunity==
Conclusion
===Calibration===
During the report generation, the '''{{ReportGeneratorCode|EUT|ALL TESTS}}''' code will be replaced with the codes of the ‘test-specific’ templates of all the (different) tests performed on a EUT. It is also possible generate a report based on a specific set of tests, for example '''{{ReportGeneratorCode|EUT|ALL IMMUNITY TESTS}}'''. (See the [[Test specific report generator inclusion codes]] for a complete list of all the available codes.)
|NUMBER OF EVENTS || The number of events that occur during a test
|EVENTS || Table of events which occur during a test
=== Uniform Field Area Calculation ===
[[Image:RadiMationReportGeneratorSettings.JPG]]
==== Making a 'test-specifc' template ====
* In 'generic' templates the Main Keyword is used to indicate the location of the data that needs to be retrieved. For example: '''{{ReportGeneratorCode|TEST 1|MODULATION}}''', where 'TEST 1' refers to the test from which the used modulation should be included.
* With 'test-specific' templates it is sufficient to specify '''{{ReportGeneratorCode|MODULATION}}''', because the Report Generator already knows that the template (and the codes within) are linked to that test.
A 'test-specific' template can be linked to a test by using the ''' 'Reporting' button''' of the test. Each TSF configuration window, and each test-information window have this button. By pressing the button the ''' 'Reporting Settings' window''' will open. More detailed reporting settings can be modified in this configuration window. The top entry field can be used to specify the name of the 'test-specific' template that should be linked to the test.
[[Image:RadiMationReportingSettings.JPG]]
=== Combining graphs in a report ===
The [[report generator]] has been extended in such a way that it is more flexible and powerful in its ability to include graphs in a report. Not only is it easier to specify which graphs should be included, but it is even possible to include multiple 'lines' in one single graph.
==== Combining multiple graph lines ====
The code '{{ReportGeneratorCode|SIGNAL POWER GRAPH}}' will include a graph with the signal power in a report. The same code can be extended with other 'graph lines' which are semi-colon separated. All the possible graph report generator codes can be combined into one single report generator code, as long as codes are ';' separated. If a specified graph line is not measured, the graph line will not be included in the final graph. It is thus acceptable to include the code '{{ReportGeneratorCode|SIGNAL POWER GRAPH; FORWARD POWER GRAPH}}', even if the forward power is not measured. The graph will still be inserted but only contain the graph line for the signal power. If one or more of the graph lines are using a different unit, additional y-axes will be displayed.
{{Example|text=
{{ReportGeneratorCode|SIGNAL POWER GRAPH; FORWARD POWER GRAPH}} Will create a graph with the signal power and the forward power graph lines.
{{ReportGeneratorCode|FIELD GRAPH 1; FIELD GRAPH 2; AVERAGE FIELD GRAPH}} Will create a field strength graph with the field strength of the probes 1 and 2, and the average field graph.
{{ReportGeneratorCode|CURRENT GRAPH; NET POWER GRAPH}} Will create a graph with the current and the net power. This graph will have two y-axes to indicate the different units.
A special report generator code '{{ReportGeneratorCode|ALL GRAPHS}}' is available, which will include a graph in the report which includes all the measured graph lines.
==== Single band immunity tests ====
* SIGNAL POWER GRAPH
* CURRENT GRAPH
* VOLTAGE GRAPH
* FORWARD POWER GRAPH
* REFLECTED POWER GRAPH
* NET POWER GRAPH
* MID POWER GRAPH
* OUT POWER GRAPH
* IMPEDANCE GRAPH
* ERROR GRAPH
* CALCULATED FIELD GRAPH
* AVERAGE FIELD GRAPH
* CALCULATED VOLTAGE GRAPH
* CALCULATED CURRENT GRAPH
* AD GRAPH 1
* AD GRAPH 2
* AD GRAPH 3
* AD GRAPH 4
* AD GRAPH 5
* AD GRAPH 6
* AD GRAPH 7
* AD GRAPH 8
* FIELD GRAPH 1
* FIELD GRAPH 2
* FIELD GRAPH 3
* FIELD GRAPH 4
* FIELD GRAPH 5
* FIELD GRAPH 6
* FIELD GRAPH 7
* FIELD GRAPH 8
* POINT 16 POWER GRAPH <n>    where <n> is a number between 1 and 16 inclusive.
* POINT 16 FIELD GRAPH <n>    where <n> is a number between 1 and 16 inclusive.
* POINT 16 ERROR GRAPH <n>    where <n> is a number between 1 and 16 inclusive.
* CALIBRATED POWER GRAPH
* GENERATED MAGNETIC FIELD GRAPH
* MEASURED MAGNETIC FIELD GRAPH
* EUT ATTENUATION GRAPH
* EUT ATTENUATION STD DEV GRAPH
* EUT DYNAMIC RANGE GRAPH
* EUT DYNAMIC RANGE STD DEV GRAPH
* SIGNAL POWER ATTENUATION GRAPH
==== Multiband immunity and Multiband emission ====
The list of known graph report generator codes for multiband tests is not a fixed list, because an unlimited number of graph lines can be created in a multiband test. However, the report generator can include any graph that is also available in the multiband test. The report generator code to include a graph in the report for a multiband test is the same as the name of the graph as it is shown in the legend above the graph in the multiband test results. A combination of the available graphs can also be included as a single graph in the report by combining the desired graph names in a single report generator code. As long as the labels of the graph are ';' separated, any number of graph lines can be included in the graph in the report.
{{Example|text={{ReportGeneratorCode|POWER; MEASURED POWER LIMIT; Amplifier forward power; Amplifier input}}
will result in the following graph:
[[Image:MultibandGraphInReport.jpg]]
* time:
* frequency:
* angle:
* height:
{{Example|text=
Export an angle graph: {{ReportGeneratorCode|angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1}}
Export a time graph: {{ReportGeneratorCode|time:Signal Power;Channel1;Channel2}}
Export a height graph: {{ReportGeneratorCode|height:RBW: 1000 kHz, Horizontal Max Peak for peak 1}}
{{Example|text=
Export an angle indexed table: {{ReportGeneratorCode|angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE}}
Export a time indexed table: {{ReportGeneratorCode|time:Signal Power TABLE}}
Export a height indexed table: {{ReportGeneratorCode|height:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE}}
=== Customizable AD-Channel measurement table ===
The [[report generator]] is extended in such a way that it is more flexible and powerful in its ability to specify the columns that should be included in a 'frequency sorted table' that includes the values of one or more measured channels.
==== Implementation in version 5.2 ====
The code '{{ReportGeneratorCode|AD CHANNEL 1 LIST}}' will generate a table with the columns:
* Frequency
* Measured value of AD-Channel 1
The code '{{ReportGeneratorCode|AD CHANNEL 3,2,4,1 LIST}}' will generate a table with the columns:
* Frequency
* Measured value of AD-Channel 3
* Measured value of AD-Channel 2
* Measured value of AD-Channel 4
! Keyword || Description
| TEST START TIME || The date and time at which the test is started
| TEST STOP TIME || The date and time at which the test is stopped
| TEST DATE || The date at which the test is performed
| TEST ENGINEER || The name of the test engineer
| TEST TIME || The time at which the test is performed
| TEST SITE NAME || The name of the test equipment list
| AVERAGE FIELD GRAPH || The graph representing the average field graph
| CALCULATED FIELD GRAPH || The graph representing the calculated field graph
| ERROR GRAPH || The graph representing the error graph
| FIELD GRAPH x || The graph representing the field graph of field strength sensor number [x], where x is a number between 1 and 8
| FORWARD POWER GRAPH || The graph representing the forward power graph
| NET POWER GRAPH || The graph representing the net-power power graph
| POINT 16 ERROR GRAPH x || The graph representing the error graph of point number [x], where x is a number between 1 and 16
| POINT 16 FIELD GRAPH x || The graph representing the field graph of point number [x], where x is a number between 1 and 16
| POINT 16 POWER GRAPH x || The graph representing the power graph of point number [x], where x is a number between 1 and 16
| REFLECTED POWER GRAPH || The graph representing the reflected power graph
| SIGNAL POWER GRAPH || The graph representing the signal power graph
| AMPLIFIER PROTECTION MODE || The Amplifier Protection mode: Signal Power, Forward Power and / or Reflected Power.
| AVERAGE FIELD || Which probes are selected for the calculation of the average field
| CURRENT MEASURED || Indication whether or not the current is measured: Yes / No
| DESCRIPTION || The description of the Test Setup File (TSF)
| DISTANCE || The distance between the antenna and the EUT in meters [m]
| DWELL TIME || The dwell time in seconds [s]
| FIELD TEST LEVEL || The field strength level in the unit selected by the user
| FORWARD POWER MEASURED || Indication whether or not the forward power is measured: Yes / No
| FREQUENCY CHANGE MODE || The method used to change frequency: Stress Method, Ramp Method, Constant Method or Fast Constant Method.
The code '{{ReportGeneratorCode|AD CHANNEL 5,8 TABLE}}' will generate a table with the columns:
* Frequency
* Measured value of AD-Channel 5
* Measured value of AD-Channel 8
| HUMIDITY || Relative Humidity during the test in percentage relative Humidity [%RH]
| LOGARITHMIC STEP || The logarithmic frequency step in a percentage of the previous step
| LOGARITHMIC STEP FROM || The frequency the test was started from to calculate the proper logarithmic step size
| MAIN VOLTAGE || The mains voltage during the test in Volt [V]
| NET POWER MEASURED || Indication whether or not the net-power power is measured: Yes / No
| NOTE || Note or remark made during the test
| POLARIZATION || Antenna polarisation: Horizontal or Vertical
| PRESSURE || Air Pressure during the test in milliBar [mbar]
| RECALCULATE AVERAGE || Indication whether or not the average is recalculated: Yes / No
| REFLECTED POWER MEASURED || Indication whether or not the reflected power is measured: Yes / No
| START FREQUENCY || The start frequency in megahertz [MHz]
| STEP FREQUENCY || The linear frequency step in megahertz [MHz]
| STOP FREQUENCY || The stop frequency in megahertz [MHz]
| TEMPERATURE || Temperature during the test in degrees Celsius [oC]
| TSF FILENAME || The filename of the TSF file
| EQUIPMENT TABLE || A table containing all used equipment
| FIELD UNIT || The selected unit for field strength:  [dBuV/m], [dBmV/m], [dBV/m], [uV/m], [mV/m] or  [V/m]
| POWER UNIT || The selected unit for power: [dBuW], [dBm], [dBmW], [dBW], [uW], [mW] or  [W]
| VOLTAGE UNIT || The selected unit for voltage:  [dBuVrms], [dBmVrms], [dBVrms], [uVrms], [mVrms] or  [Vrms]
===Equipment for Uniform Field Area calculation===
Please keep in mind that there is a significant difference between the '{{ReportGeneratorCode|.... LIST}}' and '{{ReportGeneratorCode|.... TABLE}}' codes.
* The '{{ReportGeneratorCode|.... TABLE}}' code will only include the frequencies for which the values were out of the specified range for that AD-Channel.
* The '{{ReportGeneratorCode|.... LIST}}' code will include all the frequencies.
==== Implementation in version 5.3 ====
{{Example|text={{ReportGeneratorCode|SIGNAL POWER, FORWARD POWER, AD CHANNEL 5 LIST}}
{{ReportGeneratorCode|NET POWER, AD CHANNEL 2, CALCULATED FIELD TABLE}}
{{ReportGeneratorCode|VOLTAGE, CALCULATED CURRENT, FORWARD POWER LIST}}
The '''frequency dependent''' measurement values that can be used in the codes are:
* FREQUENCY
* SIGNAL POWER
* FORWARD POWER
* REFLECTED POWER
* NET POWER
* CURRENT
* VOLTAGE
* CALCULATED FIELD
* AVERAGE FIELD
* FIELD 1
* FIELD 2
* FIELD 3
* FIELD 4
* FIELD 5
* FIELD 6
* FIELD 7
* FIELD 8
* AD CHANNEL 1
* AD CHANNEL 2
* AD CHANNEL 3
* AD CHANNEL 4
* AD CHANNEL 5
* AD CHANNEL 6
* AD CHANNEL 7
* AD CHANNEL 8
All the 'measurement values' mentioned above are dependent on the frequency that is used for that row of the table. It is also possible to include '''frequency independent data''' as part of the table. If one or more of these frequency independent codes are included in the report generator code, an additional column will be added to the table, where the value is repeated on each row. This feature can for example be used in combination with the Microsoft Excel report generator, where each row contains all the needed information.
The '''frequency independent''' measurement values that can be used in the codes are:
* CALIBRATION FILE
* TEST ENGINEER
* TEST DATE
* TEST TIME
* START TIME
* STOP TIME
* TEMPERATURE
* HUMIDITY
* PRESSURE
* START FREQUENCY
* STOP FREQUENCY
* FREQUENCY STEP
* LOGARITHMIC STEP
* LOGARITHMIC STEP FROM
* POLARIZATION
* DISTANCE
* TOLERANCE
* DWELL TIME
* MODULATION
* MODULATION DEPTH
* MODULATION CONSERVATION
* MODULATION FREQUENCY
* MODULATION SIGNAL
* MODULATION DUTY CYCLE
* MAINS VOLTAGE
* FREQUENCY CHANGE MODE
* AVERAGE FIELD
* CURRENT MEASURED
* CALIBRATION METHOD
* FORWARD POWER MEASURED
* REFLECTED POWER MEASURED
* NET POWER MEASURED
* CALCULATED FIELD MEASURED
* RECALCULATE AVERAGE
* DESCRIPTION
* TSF FILENAME
* AMPLIFIER PROTECTION MODE
* AUTODISAPPEAR
* USE VOLTAGE TEST LEVEL
* VOLTAGE TEST LEVEL
* POWER TEST LEVEL
* FIELD TEST LEVEL
* USE CURRENT TEST LEVEL
* CURRENT TEST LEVEL
* STEP DOWN
* USE CURRENT LIMIT
* CURRENT LIMIT
* MODULATION TYPE
* MAGNETIC FIELD TEST LEVEL
=== Microsoft Word report generator editor ===
==== Inclusion of other files ====
When the "{{ReportGeneratorCode|InsertFile(<OTHER WORD DOCUMENT FILENAME>)}}" code is used, the contents from the other specified Microsoft Word document will be included in the generated report. The complete contents of the other document will be inserted at the location of the report generator code.
The filename of a Bitmap, JPEG or PNG file can also be specified as part of the "{{ReportGeneratorCode|InsertFile(<PICTURE FILENAME>)}}" report generator code. The corresponding picture will then be loaded from the specified disk location and will be included at the location of the report generator code. This function is similar to the 'Insert Picture, From File' action of Microsoft Word.
The filename of a text or ASCII file can also be specified as part of the "{{ReportGeneratorCode|InsertFile(<TXT FILENAME>)}}" report generator code. The contents of the file be loaded from the specified disk location and will be included at the location of the report generator code.
==== Inclusion of EUT attachments ====
Pictures, Microsoft Word documents and other files can also be attached to the EUT document, on the {{ScreenElement|Attachments}} tab. These attachments can also be included in the generated report using the code "{{ReportGeneratorCode|EUT|ATTACHMENT <FILENAME> CONTENTS}}". The '<FILENAME>' part should be the name of the attachment. It is also possible to use the code "{{ReportGeneratorCode|EUT|ATTACHMENT 1 CONTENTS}}", where the "1" is the number of the attachment, but can be any number. The complete contents of the attachment will be inserted at the location of the report generator code.
[[File:Attachments.png]]
==== Macros ====
With the report generator code "{{ReportGeneratorCode|RunMacro(<MACRO NAME>)}}" it is possible to start a predefined macro function. The '<MACRO NAME>' may be:
* A macro defined by Microsoft Word
* The name of a public function in a public macro module
* A macro name predefined by the Microsoft Word report generator editor. These macros are:
** AutoFitThisTable: Will perform an 'Auto-Fit' on the table in which the "{{ReportGeneratorCode|RunMacro(AutoFitThisTable)}}" is present.
==== Table Formatting ====
The report generator will also use the formatting and settings that are present in the template document. This allows defining how inserted tables are formatted in the final report. To make this possible, the table code should be placed in a table in the template. In Microsoft Word, you can then set the border styles of the outer table.
==== Specific graph sizes ====
The equipment keyword consists of two parts: the keyword followed by a parameter. For each keyword all parameters are valid.
[[File:AutoResize.png]]
=== Microsoft Excel report generator editor ===
The Microsoft Excel report generator editor allows generating Microsoft Excel reports. A normal Microsoft Excel file can be specified as the template file for the report generator, and all the available {{RadiMation}} report generator codes can be used in that template file.
Values that are inserted in the generated Microsoft Excel report are always using the US English formatting, to prevent problems with the interpretation of the values if Microsoft Excel is using different Culture settings. If additional formulas and calculations are performed in the generated Microsoft Excel report, the formulas should be aware that the values are inserted in the US English format, and may need to be converted to the local format.
==== Table-inclusion modes ====
| AMPLIFIER ... || Specifications of the used Amplifier
| ANTENNA ... || Specifications of the used Antenna
* Row insert mode
* Cell insert mode
| COUPLER ... || Specifications of the used Coupler
* Overwrite mode
* Sheet insert mode
| FIELD SENSOR x ... || Specifications of the used Field Sensor number [x], where x is a number between 1 and 8
| FORWARD POWER METER ... || Specifications of the used Forward Power Meter
==== Row insert mode ====
| REFLECTED POWER METER ... || Specifications of the used Reflected Power Meter
| SIGNAL GENERATOR ... || Specifications of the used Signal Generator
* An Excel-sheet contains a report generator code for a table with 10 rows and 5 columns.
* When the code is detected, 9 rows are inserted below the cell containing the code. (Together creating a '10 row area'.)
* After the new rows have been added, the table is inserted at the cell where the code is detected, overwriting the '10 rows and 5 columns area' connected to this starting cell.
This mode can be activated by including the following report generator code: "{{ReportGeneratorCode|RunMacro(CELL_PASTE_MODE_ROWS)}}"
==== Cell insert mode ====
The 'cell insert mode' is very similar to the 'row insert mode' in that new cells are inserted before the table is included. The difference however is that the surrounding content is not just moved down, but also to the right. Meaning that, once the table is inserted, any content that was placed next to the starting cell will now be present next to the table. This can be useful if calculations should be performed on the data that is present in the inserted table.
* When the code is detected, a block of 9 rows and 4 columns is inserted to the right and below the cell containing the code. (Together creating a '10 row, 5 column area'.)
* After the new rows and columns have been added, the table is inserted at the cell where the code is detected, overwriting the '10 rows and 5 columns area' connected to this starting cell.
This mode can be activated by including the following report generator code: "{{ReportGeneratorCode|RunMacro(CELL_PASTE_MODE_CELLS)}}"
==== Overwrite mode ====
! Parameter || Description
| BRAND || Brand of Measurement equipment
| CALIBRATION EXPIRES || Date when calibration expires of equipment, given by the user
* When the code is detected, the table is inserted at the cell where the code is detected, overwriting the '10 rows and 5 columns area' connected to this starting cell.
| DESCRIPTION || Description of Measurement equipment given by the user
| DEVICE DRIVER DLL || N/A
This mode can be activated by including the following report generator code: "{{ReportGeneratorCode|RunMacro(CELL_PASTE_MODE_OVERWRITE)}}"
| DEVICE DRIVER NAME || The internal name of the device driver
==== Sheet insert mode  ====
| HARDWARE VERSION || Hardware version of Measurement equipment
| ID || Internal ID of Measurement equipment
This mode can be activated by including the following report generator code: "{{ReportGeneratorCode|RunMacro(CELL_PASTE_MODE_SHEET)}}"
| INI FILE NAME || N/A
| NAME || Name of Measurement equipment
When the "{{ReportGeneratorCode|InsertFile(<OTHER EXCEL WORKBOOK>)}}" code is used, the contents from the other specified Microsoft Excel workbook or worksheet will be included in the generated report. The complete contents of the other document will be inserted at the location of the report generator code.
The size of the 'to be inserted' data is determined, starting at the 'A1' cell until the last filled cell. This data is then inserted at the location of the keyword, using the selected table-inclusion mode. To prevent that unnecessary worksheets are included in the report, a check is done which prevents that empty worksheets are inserted.
| SERIAL NUMBER || Serial number of Measurement equipment
| SOFTWARE VERSION || Software version of Measurement equipment
| SUPPORTED DEVICE TYPES || N/A
| TYPE || Type of Measurement equipment
===Radiated Immunity Tests===
The filename of a Bitmap, JPEG or PNG file can also be specified as part of the "{{ReportGeneratorCode|InsertFile(<PICTURE FILENAME>)}}" report generator code. The corresponding picture will then be loaded from the specified disk location and will be included at the location of the report generator code. This function is similar to the 'Insert Picture, From File' action of Microsoft Excel.
Pictures, Microsoft Excel documents and other files can also be attached to the EUT document, on the {{ScreenElement|Attachments}} tab. These attachments can also be included in the generated report using the code "{{ReportGeneratorCode|EUT|ATTACHMENT <FILENAME> CONTENTS}}". The '<FILENAME>' part should be the name of the attachment. It is also possible to use the code "{{ReportGeneratorCode|EUT|ATTACHMENT 1 CONTENTS}}", where the "1" is the number of the attachment, but can be any number. The complete contents of the attachment will be inserted at the location of the report generator code.
| "keyword or name of photo" || The name or keyword of the photo
==== Macros ====
* A macro defined by Microsoft Excel
* A macro name predefined by the Microsoft Excel report generator editor, which can be:
** CELL_PASTE_MODE_CELLS
** CELL_PASTE_MODE_SHEET
** CELL_PASTE_MODE_OVERWRITE
** CELL_PASTE_MODE_ROWS
=== SQL queries on tables ===
The [[report generator]] of [[RadiMation]] supports the usage of [[wikipedia:SQL|SQL Queries]] on tables that are provided by the report generator. This makes it possible to include a customised table in the final report.
* Changing the order of the columns in the table
* Hiding selected columns from the table
* Changing the sorting of the rows, based on a specific column
* Filtering of the rows, based on the data that is included
SELECT * FROM &lt;TABLE&gt;
Where '&lt;TABLE&gt;' is an existing report generator code that includes a table in the report. This simple SQL statement can be extended with:
* An optional 'WHERE' clause that defines an additional filter that should be used on the rows.
* An optional 'ORDER BY' clause that specifies any specific ordering of the rows.
The names of the columns that can be used in the SQL queries are the same as they names that appear in the header row of the generated table itself. Any column name that includes a space, should be included in '[    ]' brackets.
The general keywords can't be used for selection within a table. For an example of keywords used within the table selection see the [[Chapter_13#Radiated_Emission_Table | Radiated Emission Table]]
The report generator is passing the SQL statements and the table data itself through an internal SQLite database engine. Because the [[wikipedia:SQLite|SQLite]] database engine is used, almost all variations of the SQL queries are supported.
==== Examples ====
{{Example|header=Example to only include the frequencies above 50 MHz:
|text={{ReportGeneratorCode|SELECT * FROM AD CHANNEL 4,3,2,1 LIST WHERE Frequency  > 50E6}}
{{Example|header=Example to only include the frequencies between 1 Hz and 20 MHz, where the AD-channel was measured out of the specified range ('TABLE'), and the table is reverse-sorted on the frequency column:
|text={{ReportGeneratorCode|1=SELECT * FROM AD CHANNEL 8,7,6,5,4,3,2,1 TABLE WHERE FREQUENCY >= 1 and FREQUENCY <= 20E6 ORDER BY Frequency desc}}
{{Example|header=Example to include the 'EQUIPMENT TABLE' including all the columns (also the hidden columns):
|text={{ReportGeneratorCode|SELECT ALL FROM EQUIPMENT TABLE}}
==== Including additional columns in the table ====
The report generator code <code>{{ReportGeneratorCode|SELECT ALL FROM <nowiki><TABLE></nowiki>}}</code> (where <nowiki><TABLE></nowiki> is the actual report generator code that inserts a table), will insert the table including all available (and by default hidden) columns.
{{Example|header=Example to include all columns of a table:|text=The report generator code <code>{{ReportGeneratorCode|SELECT ALL FROM PEAKTABLE}}</code> will show the table including all the available columns, and could result in:
<!-- <wbr> is a HTML element for a 'word-break opportunity', to allow word wrapping at that point. Used at this location to have a table that fits on the page. -->
{{{!}}class="wikitable"
! Peak Number {{!}}{{!}} Peak Number<wbr>_RAW {{!}}{{!}} Frequency {{!}}{{!}} Frequency<wbr>_RAW {{!}}{{!}} Peak {{!}}{{!}} Peak<wbr>_RAW {{!}}{{!}} Peak Limit {{!}}{{!}} Peak Limit<wbr>_RAW {{!}}{{!}} Peak Difference {{!}}{{!}} Peak Difference<wbr>_RAW {{!}}{{!}} Peak Correction {{!}}{{!}} Peak Correction<wbr>_RAW {{!}}{{!}} Peak Status {{!}}{{!}} Peak Status<wbr>_RAW {{!}}{{!}} Status {{!}}{{!}} Status<wbr>_RAW {{!}}{{!}} Angle {{!}}{{!}} Angle<wbr>_RAW
{{!}} 1 {{!}}{{!}} 1 {{!}}{{!}} 829.92 MHz {{!}}{{!}} 829921205 {{!}}{{!}} 80 dBµV/m {{!}}{{!}} 80 {{!}}{{!}} 60 dBµV/m {{!}}{{!}} 60 {{!}}{{!}} 20 dB {{!}}{{!}} 20 {{!}}{{!}} 0.00 dB {{!}}{{!}} 0.00 {{!}}{{!}} Fail {{!}}{{!}} Fail {{!}}{{!}} Fail {{!}}{{!}} Fail {{!}}{{!}} 0 Degree {{!}}{{!}} 0
*Peak Number
*Frequency
*Peak
*Peak Limit
*Peak difference to the limit
*Peak status
*Overall status
{{ReportGeneratorCode|SELECT Peak Number,Frequency,Peak,Peak Limit,Peak Difference,Peak Status,Status FROM PEAKTABLE}}
The reason for this, as stated earlier, is that any column name that includes a space, should be included in '[ ]'.
{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE}}
… which does work.
{|class="wikitable"
! Peak Number
! Frequency
! Peak
! Peak Limit
! Peak Difference
! Peak Status
! Status
| 1
| 829.92 MHz
| 80 dBµV
| 60 dBµV
| 20 dB
| Fail
| Fail
==== Limiting the number of rows in the table ====
{{Example|header=Example to limit the number of rows in the table to 3 rows:|text=
<code>{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE '''LIMIT 3'''}}</code>
will thus result in:
! Peak Number {{!}}{{!}} Frequency {{!}}{{!}} Peak {{!}}{{!}} Peak Limit {{!}}{{!}} Peak Difference {{!}}{{!}} Peak Status {{!}}{{!}} Status
{{!}}  1 {{!}}{{!}} 829.92 MHz {{!}}{{!}} 80 dBµV/m {{!}}{{!}} 60 dBµV/m {{!}}{{!}} 20 dB {{!}}{{!}} Fail {{!}}{{!}} Fail
{{!}}  2 {{!}}{{!}} 749.42 MHz {{!}}{{!}} 68.8 dBµV/m {{!}}{{!}} 60 dBµV/m {{!}}{{!}} 8.8 dB {{!}}{{!}} Fail {{!}}{{!}} Fail
{{!}}  3 {{!}}{{!}} 951.44 MHz {{!}}{{!}} 73.2 dBµV/m {{!}}{{!}} 60 dBµV/m {{!}}{{!}} 12.2 dB {{!}}{{!}} Fail {{!}}{{!}} Fail
==== Sorting the rows in another order ====
{{Example|header=Example to sort the rows in the table descending on the frequency:|text=
<code>{{ReportGeneratorCode|SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE '''ORDER BY Frequency DESC'''}}</code>
{{!}}  4 {{!}}{{!}} 805.11 MHz {{!}}{{!}} 62.1 dBµV/m {{!}}{{!}} 60 dBµV/m {{!}}{{!}} 1.2 dB {{!}}{{!}} Fail {{!}}{{!}} Fail
== Keywords Overview ==
== EUT ==
=== Common ===
{|class="wikitable"
|{{ReportGeneratorCode|EUT|FILENAME}} || The filename of the EUT
|{{ReportGeneratorCode|EUT|BRAND}} || The brand of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|CONDITION UPON RECEPTION}} || The condition of reception of the EUT specified in the EUT Window
| AD GRAPH x || The graph representing the A/D graph of channel number [x], where x is a number between 1 and 8
|{{ReportGeneratorCode|EUT|DATE OF RECEIPT}} || The date of reception of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|DATE OF MEASUREMENT}} || The date of measurement of the EUT specified in the EUT Window
| CALCULATED FIELD GRAPH (1) || The graph representing the calculated field graph
|{{ReportGeneratorCode|EUT|DIMENSIONS}} || The dimensions of the EUT specified in the EUT Window
| CURRENT GRAPH || The graph representing the current graph
|{{ReportGeneratorCode|EUT|EMISSION STANDARD}} || The emission standard of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|ENVIRONMENT}} || The environment of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|IMMUNITY STANDARD}} || The immunity standard of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|NAME}} || The name of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|ORDER NUMBER}} || The order number of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|PCB RELEASE}} || The PCB release of the EUT specified in the EUT Window
| AMPLIFIER PROTECTION MODE || The Amplifier Protection mode: Signal Power, Forward Power and  or Reflected Power.
|{{ReportGeneratorCode|EUT|PERIPHERAL EQUIPMENT}} || The peripheral equipment of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|PLACE MEASURMENT}} || The place of measurment of the EUT specified in the EUT Window
| CALCULATED FIELD MEASURED(1) || Indication whether or not the calculated power is measured: Yes / No
|{{ReportGeneratorCode|EUT|RADIMATION VERSION}} || The {{RadiMation}} version
| CALIBRATION FILE (1) || The name of the calibration file used for the test
|{{ReportGeneratorCode|EUT|RATED POWER}} || The rated power of the EUT specified in the EUT Window
| CALIBRATION METHOD || The calibration method used: Signal Generator Level, Forward Power or Net-Power
|{{ReportGeneratorCode|EUT|SERIAL NUMBER}} || The serial number of the EUT specified in the EUT Window
| CURRENT LIMIT || The Current limit in Ampere [A]
|{{ReportGeneratorCode|EUT|TEST SETUP TYPE}} || The test setup type of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|SOFTWARE RELEASE}} || The software release of the EUT specified in the EUT Window
| CURRENT UNIT || The selected current unit: [dBuA],[dBmA], [dBA], [uA], [mA] or [A]
|{{ReportGeneratorCode|EUT|TYPE NUMBER}} || The type number of the EUT specified in the EUT Window
|{{ReportGeneratorCode|EUT|VOLTAGE SUPPLY}} || The voltage supply of the EUT specified in the EUT Window
=== Client Address Information ===
{|class="wikitable"
|{{ReportGeneratorCode|EUT|CLIENT ADDRESS}} || The address of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT CAPACITY}} || The capacity of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT CITY}} || The city of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT COMPANY}} || The client name specified in the EUT Window
| FIELD TEST LEVEL (2) || The field strength level in the unit selected by the user
|{{ReportGeneratorCode|EUT|CLIENT CONTACT PERSON}} || The contact person of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT COUNTRY}} || The country of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT EMAIL ADDRESS}} || The email address of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT FAX NUMBER}} || The fax number of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT STATE}} || The state of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT TELEPHONE NUMBER}} || The telephone number of the client specified in the EUT Window
|{{ReportGeneratorCode|EUT|CLIENT ZIP CODE}} || The ZIP code of the client specified in the EUT Window
=== Manufacturer Address Information ===
{|class="wikitable"
| MODULATION || The type of modulation: none, Internal AM, External AM or Pulse
|{{ReportGeneratorCode|EUT|MANUFACTURER ADDRESS}} || The address of the manufacturer specified in the EUT Window
| MODULATION CONSERVATION || Indication whether peak conservation is used: Power, Voltage, none
|{{ReportGeneratorCode|EUT|MANUFACTURER CITY}} || The city of the manufacturer specified in the EUT Window
| MODULATION DEPTH || Modulation depth in percentage of the carrier [%]
|{{ReportGeneratorCode|EUT|MANUFACTURER COMPANY}} || The manufacturer name specified in the EUT Window
| MODULATION DUTY CYCLE || The duty cycle of the pulse modulation as a percentage [%]
|{{ReportGeneratorCode|EUT|MANUFACTURER CONTACT PERSON}} || The contact person of the manufacturer specified in the EUT Window
| MODULATION FREQUENCY || The modulation frequency in Hertz [Hz]
|{{ReportGeneratorCode|EUT|MANUFACTURER COUNTRY}} || The country of the manufacturer specified in the EUT Window
| MODULATION SIGNAL || The type of modulation signal: Sine Wave or Square Wave
|{{ReportGeneratorCode|EUT|MANUFACTURER EMAIL ADDRESS}} || The email address of the manufacturer specified in the EUT Window
|{{ReportGeneratorCode|EUT|MANUFACTURER FAX NUMBER}} || The fax number of the manufacturer specified in the EUT Window
|{{ReportGeneratorCode|EUT|MANUFACTURER STATE}} || The state of the manufacturer specified in the EUT Window
|{{ReportGeneratorCode|EUT|MANUFACTURER TELEPHONE NUMBER}} || The telephone number of the manufacturer specified in the EUT Window
|{{ReportGeneratorCode|EUT|MANUFACTURER ZIP CODE}} || The ZIP code of the manufacturer specified in the EUT Window
=== Test Site Address Information ===
{|class="wikitable"
|{{ReportGeneratorCode|EUT|TEST SITE ADDRESS}} || The address of the test site specified in the EUT Window
|{{ReportGeneratorCode|EUT|TEST SITE CITY}} || The city of the test site specified in the EUT Window
|{{ReportGeneratorCode|EUT|TEST SITE COMPANY}} || The test site name specified in the EUT Window
|{{ReportGeneratorCode|EUT|TEST SITE CONTACT PERSON}} || The contact person of the test site specified in the EUT Window
|{{ReportGeneratorCode|EUT|TEST SITE COUNTRY}} || The country of the test site specified in the EUT Window
|{{ReportGeneratorCode|EUT|TEST SITE EMAIL ADDRESS}} || The email address of the test site specified in the EUT Window
| TAKE START PICTURE || Indication whether or not to take a photo at the start of the test: Yes/No
|{{ReportGeneratorCode|EUT|TEST SITE FAX NUMBER}} || The fax number of the test site specified in the EUT Window
|{{ReportGeneratorCode|EUT|TEST SITE STATE}} || The state of the test site specified in the EUT Window
|{{ReportGeneratorCode|EUT|TEST SITE TELEPHONE NUMBER}} || The telephone number of the test site specified in the EUT Window
| USE CURRENT LIMIT || Is the current limit used: Yes / No
|{{ReportGeneratorCode|EUT|TEST SITE ZIP CODE}} || The ZIP code of the test site specified in the EUT Window
=== Cable information ===
{|class="wikitable"
|{{ReportGeneratorCode|EUT|CABLE TABLE}} || A table with the cable information specified in the EUT Window
|{{ReportGeneratorCode|EUT|CABLE X PORT TYPE}} || The port type of cable X in the EUT window ( X starting from 1)
| MANUAL MODE || Table containing the information collected during a manual mode session
|{{ReportGeneratorCode|EUT|CABLE X CABLE TYPE}} || The cable type of cable X in the EUT window (X starting from 1)
|{{ReportGeneratorCode|EUT|CABLE X CABLE LENGTH}} || The cable length of cable X in the EUT window (X starting from 1)
|{{ReportGeneratorCode|EUT|CABLE X FIXING SHIELD}} || The fixing shield of cable X in the EUT window (X starting from 1)
|{{ReportGeneratorCode|EUT|CABLE X LOAD AT PORT}} || The load at port of cable X in the EUT window (X starting from 1)
1) Substitution Method only
=== Standards Information ===
2) Substitution Method and Fixed Field Method only
===Equipment for Radiated Immunity Tests===
Keyword Description
{{ReportGeneratorCode|EUT|PRODUCT STANDARD ...}} Specifications of the used Product Standard
{{ReportGeneratorCode|EUT|BASIC STANDARD ...}} Specifications of the used Basic Standard
{|class="wikitable"
| AD CONVERTOR ... || Specifications of the used A/D Converter
| CABLE x ... || Specifications of the used Cable number [x], where x is a number between 1 to 8
| CURRENT SENSOR ... || Specifications of the used Current Sensor
|REFERRED STANDARDS || A string of referred standards, separated by a "," (Only applicable to product standards)
| SENSOR POWER METER ... || Specifications of the used Sensor Power Meter
{{Example|text=
<code>{{ReportGeneratorCode|EUT|PRODUCT STANDARD REFERRED STANDARDS}}</code>
<code>{{ReportGeneratorCode|EUT|PRODUCT STANDARD DESCRIPTION}}</code>
<code>{{ReportGeneratorCode|EUT|SELECT * FROM [BASIC STANDARD TABLE]}}</code>
<code>{{ReportGeneratorCode|EUT|SELECT ALL FROM [BASIC STANDARD TABLE]}}</code>
<code>{{ReportGeneratorCode|EUT|SELECT [OFFICIAL NAME] FROM [BASIC STANDARD TABLE]}}</code>
== Configuration ==
{|class="wikitable"
=== Directories ===
{|class="wikitable"
|{{ReportGeneratorCode|Configuration|DEVICEDRIVERFILESDIRECTORY}} || CONFDVDR directory
|{{ReportGeneratorCode|Configuration|CALFILESDIRECTORY}} || CAL Files directory
|{{ReportGeneratorCode|Configuration|EUTFILESDIRECTORY}} || EUT Files directory
|{{ReportGeneratorCode|Configuration|TSFFILESDIRECTORY}} || TSF Files directory
|{{ReportGeneratorCode|Configuration|SEQFILESDIRECTORY}} || SEQ Files directory
|{{ReportGeneratorCode|Configuration|CORFILESDIRECTORY}} || COR Files directory
|{{ReportGeneratorCode|Configuration|DIRECTORIES}} || All directories
==Conducted Immunity==
=== Units ===
===Calibration===
{|class="wikitable"
|{{ReportGeneratorCode|Configuration|CURRENT}} || Current
|{{ReportGeneratorCode|Configuration|EMISSION AMPLITUDE}} || Emission
|{{ReportGeneratorCode|Configuration|FIELD}} || Electrical Field
|{{ReportGeneratorCode|Configuration|FREQUENCY}} || Frequency
|{{ReportGeneratorCode|Configuration|POWER}} || Power
|{{ReportGeneratorCode|Configuration|VOLTAGE}} || Voltage
|{{ReportGeneratorCode|Configuration|UNITS}} || All units
=== General ===
{|class="wikitable"
|{{ReportGeneratorCode|Configuration|CURRENTUSERNAME}} || Current username
|{{ReportGeneratorCode|Configuration|RADIMATION VERSION}} || The {{RadiMation}} version
== EUT Calibration ==
* Attenuation / Gain
* Amplifier linearity
* Gain compression
* Harmonics
* S Parameter
* Antenna scan
* Antenna
* Field strength
=== Amplifier Linearity ===
{|class="wikitable"
|DESCRIPTION|| Test level to calibrate with
| CURRENT TEST LEVEL || The current test level in Ampere [A]
|TEST-TYPE|| Signal gen, forward-power or net-power
| CURRENT UNIT || The selected current unit: [dBuA], [dBmA], [dBA], [uA], [mA] or  [A]
|START FREQUENCY || Start frequency of the calibration
|STOP FREQUENCY|| End frequency of the calibration
|FREQUENCY STEP|| Step size
|CALIBRATION FILE|| Used Calibration file
|ALL GRAPHS|| Output graph
=== Gain compression ===
{|class="wikitable"
| HUMIDITY || Relative Humidity during the test in percentage rel.  Humidity [%RH]
|STEP || Step size specifically for using log step in %
|CALIBRATION METHOD || Signal gen, forward-power or net-power
|TABLE || Output table
|GRAPH|| Output graph
|POWER TABLE|| Table with all the compression power levels for every frequency
=== Harmonics ===
{|class="wikitable"
|POWER TABLE|| Specific table for harmonics
=== S Parameter ===
{|class="wikitable"
!Keyword || Description
|ALL GRAPHS || All available graphs plotted in one graph
|Sxx PRESCAN GRAPH || Prescan Graph
|Sxx MAGNITUDE AVG GRAPH || Magnitude Average Graph
|Sxx MAGNITUDE MAX GRAPH || Magnitude Maximum Graph
| USE CURRENT TEST LEVEL || Indication the current test level is used: Yes / No
|Sxx MAGNITUDE MIN GRAPH || Magnitude Minimum Graph
| USE VOLTAGE TEST LEVEL || Indication the voltage test level is used: Yes / No
|Sxx MAGNITUDE STDDEV GRAPH || Magnitude Standard Deviation Graph
| VOLTAGE TEST LEVEL || The voltage test level in the unit selected by the user
|Sxx IMPEDANCE AVG GRAPH || Graph with the average of the measured impedance
|Sxx IMPEDANCE MAX GRAPH || Graph with the max-hold of the measured impedance
|Sxx IMPEDANCE MIN GRAPH || Graph with the min-hold of the measured impedance
|Sxx IMPEDANCE STDDEV GRAPH || Graph with the Standard Deviation of the measured impedance
===Equipment for Calibration===
{|class="wikitable"
|S11 TABLE || All columns default visible
|Select * FROM S11 TABLE || Using sql to retrieve all values
|Select IMPEDANCE FROM S11 TABLE || Because we added impedance later, it is not visible by default
=== Antenna scan ===
{|class="wikitable"
|POWER TABLE || Default table with frequency, signal power and forward power.
|SELECT ALL FROM [Power Table]|| To request all possible columns
== Radiated Immunity ==
=== Calibration ===
{|class="wikitable"
{|class="wikitable"
| CLAMP ... || Specifications of the used Clamp
| CLAMP RESISTANCE || The resistance in Ohm ({{Ohm}}) of the clamp
| JIG ... || Specifications of the used Calibration Jig
===Conducted Immunity Tests===
| CALCULATED CURRENT GRAPH (1) || The graph representing the current graph
| CALCULATED VOLTAGE GRAPH (1) || The graph representing the calculated voltage graph
| FREQUENCY STEP || The linear frequency step in megahertz [MHz]
| VOLTAGE GRAPH || The graph representing the voltage graph
| STEP || A textual description of the used frequency step, also containing if it is linear or logarithmic step size
| CALIBRATED POWER GRAPH || The graph representing the calibrated power limitation graph
| CALIBRATED POWER LIMIT GRAPH || The graph representing the calibrated power limitation graph
| FIELD UNIT || The selected unit for field strength: [dBμV/m], [dBmV/m], [dBV/m], [μV/m], [mV/m] or [V/m]
| POWER UNIT || The selected unit for power: [dBμW], [dBm], [dBmW], [dBW], [μW], [mW] or  [W]
| VOLTAGE UNIT || The selected unit for voltage:  [dBμVrms], [dBmVrms], [dBVrms], [μVrms], [mVrms] or [Vrms]
| NUMBER OF TOTAL POINTS || The total of calibration points
| NUMBER OF POINTS TO REMOVE || The number of points removed
| ACCEPTED TOLERANCE || The accepted tolerance in dB
| ACCEPTED EXTENDED TOLERANCE || The accepted extended tolerance in dB
| PERCENTAGE WITHIN TOLERANCE || The percentage of frequencies within the accepted tolerance
| PERCENTAGE WITHIN EXTENDED TOLERANCE || The percentage of frequencies within the accepted extended tolerance
| PERCENTAGE OUTSIDE TOLERANCE || The percentage of frequencies outside the tolerance
| MAX STANDARD DEVIATION MEASURED || The maximum standard deviation
| FREQUENCY FOR MAX STANDARD DEVIATION || The frequency with the maximum standard deviation
| MAX TOLERANCE DEVIATION MEASURED || The maximum tolerance deviation
| FREQUENCY FOR MAX DEVIATION || The frequency with the max tolerance
| RESULT || UFA pass or fail result
| STDEV GRAPH || The graph representing the standard deviation per frequency
| PROBABILITY PERCENTAGE || The probability percentage
| CROSS POLAR PROBABILITY PERCENTAGE || The cross-polar probability percentage
| CROSS POLAR ACCEPTED TOLERANCE || The cross-polar accepted tolerance in dB
| CROSS POLAR ACCEPTED EXTENDED TOLERANCE || The cross-polar accepted extended tolerance in dB
| CROSS POLAR EXTENDED TOLERANCE PERCENTAGE || The cross-polar extended tolerance percentage
| CROSS POLAR RESULT || The cross-polar result, pass or fail
| CROSS POLAR PERCENTAGE WITHIN TOLERANCE || The cross-polar percentage within tolerance
| CROSS POLAR NUMBER OF POINTS WITHIN TOLERANCE || The cross-polar number of frequency points within tolerance
| CROSS POLAR PERCENTAGE WITHIN EXTENDED TOLERANCE || The cross-polar percentage of frequency points within the extended tolerance
| CROSS POLAR NUMBER OF POINTS WITHIN EXTENDED TOLERANCE || The cross-polar number of frequency points within the extended tolerance
| CROSS POLAR PERCENTAGE OUTSIDE TOLERANCE || The cross-polar percentage outside the tolerance
| CROSS POLAR NUMBER OF POINTS OUTSIDE TOLERANCE || The cross-polar number of points outside the tolerance
| CROSS POLAR MAX PQUANTILE || The cross-polar maximum PQuantile
| TAKE START PICTURE || Indication whether or not a photo is taken at the start of the test: Yes/No
| CROSS POLAR FREQUENCY FOR MAX PQUANTILE || The cross-polar frequency for the max PQuantile
| PQUANTILE GRAPH || The graph representing the PQuantile per frequency
=== Equipment for Uniform Field Area calculation ===
{|class="wikitable"
| USE CURRENT TEST LEVEL (2) || Indication the current test level is used: Yes / No
| USE VOLTAGE TEST LEVEL (6) || Indication the voltage test level is used: Yes / No
| VOLTAGE TEST LEVEL (6) || The voltage test level in the unit selected by the user
{|class="wikitable"
| POWER TEST LEVEL (3) || The power test level in the unit selected by the user
| STEP DOWN (4) || The number of dB's the test level is stepped down during a minimum performance test
| TOLERANCE (5) || The maximum allowable tolerance of the test-level
{{Example|text={{ReportGeneratorCode|AMPLIFIER BRAND}} returns the brand name of the amplifier.}}
1) Fixed Generator method and Minimum Performance Level Method Excluded
=== Radiated Immunity Tests ===
{|class="wikitable"
6) Substitution method only
===Equipment for Conducted Immunity Tests===
| CURRENT UNIT || The selected current unit: [dBμA],[dBmA], [dBA], [μA], [mA] or [A]
==Pulsed Immunity==
===ESD Test===
| NEG AIR MEASURED POINT x || Indication a Negative Air Discharge is performed at point number [x]: Yes / No
| NEG AIR RESULT POINT x || Results of Negative Air Discharge at point number [x]: pass or fail
| NEG CONTACT MEASURED POINT x || Indication a Negative Contact Discharge is performed at point number [x]: Yes / No
| NEG CONTACT RESULT POINT x || Results of Negative Contact Discharge at point number [x]: pass or fail
| POS AIR MEASURED POINT x || Indication a Positive Air Discharge is performed at point number [x]: Yes / No
| POS AIR RESULT POINT x || Results of Positive Air Discharge at point number [x]: pass or fail
| POS CONTACT MEASURED POINT x || Indication a Positive Contact Discharge is performed at point number [x]: Yes / No
| POS CONTACT RESULT POINT x || Results of Positive Contact Discharge at point number [x]: pass or fail
| TEST START TIME || The time at which the test is stopped
| TEST STOP TIME || The time at which the test is started
| AMPLITUDE POINT x || Amplitude of pulse performed at point number [x] in [V]
| CONFIGURATION TIME || The date and time when the TSF file is saved
| DESCRIPTION || Description of the Test Setup File (TSF)
| DISCHARGE INTERVAL || The interval of the discharge in seconds.
| MAIN VOLTAGE || Mains Voltage during the test in Volt [V]
| NEG AIR LEVEL POINT x || Level of the Negative Air Discharge at point number [x] in [V]
| NEG CONTACT LEVEL POINT x || Level of the Negative Contact Discharge at point number [x] in [V]
| NOTE POINT x || Note or remark for point number [x]
| STEP FREQUENCY|| The linear frequency step in megahertz [MHz]
| NUMBER OF DISCHARGES || The number of discharges
| PERFORMANCE CRITERIA CUSTOMER || The performance criteria according to the customer
| PERFORMANCE CRITERIA PRODUCT STANDARD || The performance criteria according to the product standard
| POS AIR LEVEL POINT x || Level of the Positive Air Discharge at point number [x] in [V]
| POS CONTACT LEVEL POINT x || Level of the Positive Contact Discharge at point number [x] in [V]
| PRESSURE || Air Pressure during the test in MilliBar [mbar]
| TDURATION POINT x || Duration of pulse at point number [x] in milliseconds [ms]
| TESTPOINT DESCRIPTION POINT nr || Position at EUT where discharge was applied
| TRISE POINT x || Rise time of pulse at point number [x] in milliseconds [ms]
===Equipment for ESD Tests===
=== Equipment for Radiated Immunity Tests ===
{|class="wikitable"
| ESDGUN ... || Specifications of the used ESD Gun
{|class="wikitable"
Line 1,150: Line 1,476:
===EFT Test===
{{Example|text={{ReportGeneratorCode|AMPLIFIER SERIAL NUMBER}} returns the serial number of the amplifier.}}
== Radiated Immunity Multiband ==
It is possible to include the test levels of a multi band immunity test.
The name of a test level can be adjusted in the TSF and therefore needs to be looked up first.
| (L1-L2)/PE RESULT || Results on (L1-L2)/PE: Pass or Fail
{{Example|text=
The report generator code is built as following, where <Test level name> is replaced with the name of the configured test level and <Field> is the actual information that is put in the report.
{{ReportGeneratorCode|Band1 TestLevel: <Test level name> <Field>}}
This example puts the 'Description' of the test level named 'Electrical field substitution method' in the report
{{ReportGeneratorCode|Band1 TestLevel: Electrical field substitution method Description}}
This example puts the 'Calibration filename' of the test level named 'Electrical field substitution method' in the report
{{ReportGeneratorCode|Band1 TestLevel: Electrical field substitution method Calibration File}}
This example puts the 'Field' of the test level named 'Electrical field substitution method' in the report
{{ReportGeneratorCode|Band1 TestLevel: Electrical field substitution method Fixed Field}}
This example puts the 'Power' of the test level named 'Power closed loop method' in the report
{{ReportGeneratorCode|Band1 TestLevel: Power closed loop method Fixed Power}}
{|class="wikitable"
| CAPCLAMP RESULT || Results on Capacitive Clamp: Pass or Fail
| Name || Name of test level
| L1 RESULT || Results on L1: Pass or Fail
| Description || Description of test level
| L1/L2 RESULT || Results on L1/L2: Pass or Fail
| Tolerance || The tolerance of the test level
| L1/PE RESULT || Results on L1/PE: Pass or Fail
| Calibration File || Calibration file used in a substitution test
| L2 RESULT || Results on L2: Pass or Fail
| Calibration method || Calibration method
| L2/PE RESULT || Results on L2/PE: Pass or Fail
| Fixed Field || The field strength to regulate to
== Conducted Immunity ==
=== Calibration ===
{|class="wikitable"
| PE RESULT || Results on PE (Protective Earth): Pass or Fail
| CURRENT UNIT || The selected current unit: [dBμA], [dBmA], [dBA], [μA], [mA] or  [A]
| (L1-L2)/PE MEASURED || (L1-L2)/PE Measured: Yes / No
| AMPLITUDE || Amplitude of burst in Volt [V]
| CAPCLAMP MEASURED || Capacitive Clamp Measured: Yes / No
| DURATION || Duration of test in milliseconds [ms]
| FREQUENCY || Frequency of burst in Hertz [Hz]
| L1 MEASURED || L1 Measured: Yes / No
| L1/L2 MEASURED || L1/L2 Measured: Yes / No
| L1/PE MEASURED || L1/PE Measured: Yes / No
| L2 MEASURED || L2 Measured: Yes / No
| L2/PE MEASURED || L2/PE Measured: Yes / No
=== Equipment for Calibration ===
{|class="wikitable"
| NOTE (L1-L2)/PE || Notes or remarks for (L1-L2)/PE
| NOTE CAPCLAMP || Notes or remarks for Capacitive Clamp
| NOTE L1 || Notes or remarks for L1
| NOTE L1/L2 || Notes or remarks for L1/L2
| NOTE L1/PE || Notes or remarks for L1/PE
| NOTE L2 || Notes or remarks for L2
| NOTE L2/PE || Notes or remarks for L2/PE
| NOTE PE || Notes or remarks for PE
| PE MEASURED || PE Measured: Yes / No
{|class="wikitable"
| TBURST || Duration of burst in milliseconds [ms]
| TPERIOD || Duration of period in milliseconds [ms]
| MEASURED || If the line was measured
| RESULT || The result on this line
| AMPLITUDE || The amplitude on this line
| NOTE || The note that is related to this line
{{Example|text={{ReportGeneratorCode|COUPLER TYPE}} returns the type of the coupler.}}
=== Conducted Immunity Tests ===
{|class="wikitable"
===Equipment for EFT Tests===
| EFTGENERATOR ... || Specifications of the used EFT Generator
===Surge Test===
| (L1-L2)/PE MEASURED || (L1-L2)/PE measured: Yes / No
| (L1-L2)/PE RESULT || Results of Surge Test at (L1-L2)/PE: pass or fail
| L1/L2 MEASURED || Indication whether or not L1/L2 is measured: Yes / No
| L1/L2 RESULT || Results of Surge Test at L1/L2: pass or fail
| L1/PE MEASURED || Indication whether or not L1/PE is measured: Yes / No
| L1/PE RESULT || Results of Surge Test at L1/PE: pass or fail
| L2/PE MEASURED || Indication whether or not L2/PE is measured: Yes / No
| L2/PE RESULT || Results of Surge Test at L2/PE: pass or fail
| CINTSETTINGS (L1-L2)/PE || Internal Capacitance for (L1-L2)/PE
| CINTSETTINGS L1/L2 || Internal Capacitance for L1/L2
| CINTSETTINGS L1/PE || Internal Capacitance for L1/PE
| CINTSETTINGS L2/PE || Internal Capacitance for L2/PE
| NOTE (L1-L2)/PE || Note or remark for Surge Test at (L1-L2)/PE
| NOTE L1/L2 || Note or remark for Surge Test at L1/L2
| NOTE L1/PE || Note or remark for Surge Test at L1/PE
| NOTE L2/PE || Note or remark for Surge Test at L2/PE
| PHASE (L1-L2)/PE || Phase angle between L1 - L2 and PE in degrees: 0, 90, 270, random
| PHASE L1/L2 || Phase angle between L1 and L2 in degrees: 0, 90, 270, random
| PHASE L1/PE || Phase angle between L1 and PE in degrees: 0, 90, 270, random
| PHASE L2/PE || Phase angle between L2 and PE in degrees: 0, 90, 270, random
| POLARISATION OF (L1-L2)/PE TESTLEVEL || Polarisation of test voltage at (L1-L2)/PE: Positive, Negative or both
| POLARISATION OF L1/L2 TESTLEVEL || Polarisation of test voltage at L1/L2: Positive, Negative or both
| POLARISATION OF L1/PE TESTLEVEL || Polarisation of test voltage at L1/PE: Positive, Negative or both
| POLARISATION OF L2/PE TESTLEVEL || Polarisation of test voltage at L2/PE: Positive, Negative or both
| RINTSETTINGS (L1-L2)/PE || Internal Resistance for (L1-L2)/PE
| RINTSETTINGS L1/L2 || Internal Resistance for L1/L2
| RINTSETTINGS L1/PE || Internal Resistance for L1/PE
| RINTSETTINGS L2/PE || Internal Resistance for L2/PE
| TESTLEVEL (L1-L2)/PE || Test level at (L1-L2)/PE in kilo Volt [kV]
| TESTLEVEL L1/L2 || Test level at L1/L2 in kilo Volt [kV]
| TESTLEVEL L1/PE || Test level at L1/PE in kilo Volt [kV]
| TESTLEVEL L2/PE || Test level at L2/PE in kilo Volt [kV]
| EVENTS || Table of events which occur during a test
=== Equipment for Conducted Immunity Tests ===
{|class="wikitable"
| MEASURED || If the selected line was measured
| RESULT || The result of the measurement
| POSITIVE TEST LEVEL || The positive test level(Yes / No)
| NEGATIVE TEST LEVEL || The negative testlevel(Yes / No)
| CI || The capacity used for this line configuration
| RI || The resistance used for this line configuration
| Test Level || The test level used for this line configuration
| Note || The note that applies to this line configuration
| PHASE 0 || Phase 0 measured (Yes / No)
| PHASE 90 || Phase 90 measured (Yes / No)
| PHASE 270 || Phase 270 measured (Yes / No)
| PHASE RANDOM || Phase random measured (Yes / No)
===Equipment for Surge Tests===
| SURGEGENERATOR ... || Specifications of the used Surge Generator
{|class="wikitable"
Line 1,516: Line 1,877:
== Radiated Emission ==
{{Example|text={{ReportGeneratorCode|ANTENNA DESCRIPTION}} returns the description of the antenna.}}
{{Note|The pass/fail criteria, shown in the tables of the report generator do not take the measure uncertainty of the test setup into account.}}
===Radiated Emission Tests===
== Pulsed Immunity ==
=== ESD Test ===
{|class="wikitable"
| NUMBER OF EVENTS || The number of events that occur during a test
| PEAK DETECTION COUNT || The number of detected peaks
| NEG AIR NOTE POINT x || Note of Negative Air Discharge at point number [x]
| MAX HOLD GRAPH || The max hold graph of the all the traces
| NEG CONTACT NOTE POINT x || Note of Negative Contact Discharge at point number [x]
| MAX HOLD PEAKS GRAPH || The max hold graph of all the traces and markers
| MAX PEAK GRAPH || The max hold graph of the Peak trace
| MAX AVERAGE GRAPH || The max hold graph of the Average trace
| MAX QUASI PEAK GRAPH || The max hold graph of the Quasi Peak  trace
| POS AIR NOTE POINT x || Note of Positive Air Discharge at point number [x]
| MAX RMS GRAPH || The max hold graph of the RMS trace
| ANGLE STEPS || The number of steps between the start position and the stop position
| ATTENUATOR || The attenuation of the attenuator in dB
| POS CONTACT NOTE POINT x || Note of Positive Contact Discharge at point number [x]
| ATTENUATOR MODE || The Attenuator Mode: Auto, Coupled or Manual
| AVERAGE DETECTOR || Indication whether or not the Average detector is measured: Yes / No
| AVERAGE MEASURE TIME || The measurement time of the average peak detector in seconds [s]
| AVERAGE OBSERVATION TIME || The observation time per average peak in seconds [s]
| AVERAGE TRACE || Indication whether or not the Average trace is measured: Yes / No
| BELOW LIMITLINE OFFSET || The level of peaks to detect in dB below the limit line
| CURRENT UNIT || The selected unit for current: [dBuA], [dBmA], [dBA], [uA], [mA] or  [A]
| DB/DIV || N/A
| DB/DIV MODE || N/A
| EMISSION OPTIMIZING METHOD || N/A
| HEIGHT STEPS || The number of steps between the minimum height and the maximum height
| HIGHEST HEIGHT || The maximum height of the antenna tower in meters [m]
| INTERNAL PREAMPLIFIER || The amplification of the internal pre amplifier in [dB]
| INTERNAL PREAMPLIFIER MODE || The Internal Pre Amplifier Mode: Auto, Coupled or Manual
| LOWEST HEIGHT || The minimum height of the antenna tower in meters [m]
| MAXIMUM PEAK COUNT || The maximum number of peaks to be detected
| MEASURE TIME || The measure time in milliseconds [ms]
| MEASURE TIME MODE || The Mode: Auto, Coupled or Manual
| PEAK DETECTOR || Indication whether or not the Peak detector is measured: Yes / No
| PEAK MEASURE TIME || The measurement time of the peak detector in seconds [s]
| PEAK OBSERVATION TIME || The observation time per peak in seconds [s]
| POS HCP MEASURED POINT X|| Indication a Positive HCP Discharge is performed at point number [x]: Yes / No
| PEAK SEARCH || Indication whether or not a peak search is performed: Yes / No
| NEG HCP MEASURED POINT X|| Indication a Negative HCP Discharge is performed at point number [x]: Yes / No
| PEAK TRACE || Indication whether or not the Peak trace is measured: Yes / No
| POS HCP RESULT POINT X|| Results of Positive HCP Discharge at point number [x]: pass or fail
| PEAKSWEEPS || Number of peak sweeps performed
| NEG HCP RESULT POINT X|| Results of Negative HCP Discharge at point number [x]: pass or fail
| POS HCP LEVEL POINT X|| Level of the Positive HCP Discharge at point number [x] in [V]
| POINT X FREQUENCY || The frequency of point X in the peak list ( X starting from 1)
| NEG HCP LEVEL POINT X|| Level of the Negative HCP Discharge at point number [x] in [V]
| POINT X PEAK || The peak value of point X in the peak list (X starting from 1)
| POS HCP NOTE POINT X|| Note or remark for the Positive HCP Discharge at point number [x]
| POINT X QUASI PEAK || The quasi peak value of point X in the peak list (X starting from 1)
| NEG HCP NOTE POINT X|| Note or remark for the Negative HCP Discharge at point number [x]
| POINT X AVERAGE || The average value of point X in the peak list (X starting from 1)
| POS VCP MEASURED POINT X|| Indication a Positive VCP Discharge is performed at point number [x]: Yes / No
| POLARIZATION || The polarization of the antenna: horizontal, vertical or both
| NEG VCP MEASURED POINT X|| Indication a Negative VCP Discharge is performed at point number [x]: Yes / No
| POS VCP RESULT POINT X|| Results of Positive VCP Discharge at point number [x]: pass or fail
| QUASIPEAK DETECTOR || Indication whether or not the Quasi peak detector is measured: Yes/No
| NEG VCP RESULT POINT X|| Results of Negative VCP Discharge at point number [x]: pass or fail
| QUASIPEAK MEASURE TIME || The measurement time of the quasi peak detector in seconds [s]
| POS VCP LEVEL POINT X|| Level of the Positive VCP Discharge at point number [x] in [V]
| QUASIPEAK OBSERVATION TIME || The observation time per quasi peak in seconds [s]
| NEG VCP LEVEL POINT X|| Level of the Negative VCP Discharge at point number [x] in [V]
| QUASIPEAK TRACE || Indication whether or not the Quasi peak trace is measured: Yes / No
| POS VCP NOTE POINT X||   Note or remark for the Positive VCP Discharge at point number [x]
| RBW || The Radio Bandwidth in kilo Hertz [kHz]
| NEG VCP NOTE POINT X||   Note or remark for the Negative VCP Discharge at point number [x]
| RBW MODE || The RBW Mode: Auto, Coupled or Manual
=== Equipment for ESD Tests ===
{|class="wikitable"
| RECEIVER DATA SAVE MODE || Indication whether all data is saved or only the test results
{|class="wikitable"
| REFERENCE LEVEL || The reference level in the unit selected by the user
| REFERENCE LEVEL MODE || The Reference Level Mode: Auto, Coupled or Manual
| RMS DETECTOR || Indication whether or not the RMS detector is measured: Yes / No
| RMS MEASURE TIME || The measurement time of the RMS peak detector in seconds [s]
| RMS OBSERVATION TIME || The observation time per RMS peak in seconds [s]
| RMS TRACE || Indication whether or not the RMS trace is measured: Yes / No
| START ANGLE || The start position in degrees
| STEP FREQUENCY || The step frequency in kilo Hertz [kHz]
| STEP FREQUENCY MODE || The Step frequency Mode: Auto, Coupled or Manual
| STOP ANGLE || The stop position in degrees
| SWEEP TIME || The sweep time in milliseconds [ms]
{{Example|text={{ReportGeneratorCode|ESDGUN BRAND}} returns the brand name of the ESD gun.}}
=== EFT Test ===
{|class="wikitable"
| SWEEP TIME MODE || The Sweep Time Mode: Auto, Coupled or Manual
| TAKE START PICTURE || Indication whether or not a picture is taken at the start of the test: Yes / No
| USE INTERNAL PREAMP || N/A
| USE START STOP FREQUENCY || Yes / No, if Yes: Start/Stop if No: Frequency list is used.
| VBW || The Video Bandwidth in kilo Hertz [kHz]
| VBW MODE || The Video Bandwidth Mode: Auto, Coupled or Manual
| EMISSION TABLE || Table of detected peaks and per peak: frequency, level(s), limit value, pass/fail, location
| TEST SITE NAME || The name of the test equipment list
! Keyword || Description
| EUT HEIGHT || The height of the EUT in meters [m]
| SEPTUM HEIGHT || The height of the septum in meters [m]
===Radiated Emission Table===
Each keyword can also be accessed in RAW format, which generally implies that only the RAW data (without unit) is retrieved.
To retrieve the RAW value extend the keyword with applying "_RAW" (without any white space) to this keyword. E.g. "PEAK_RAW".
The LIMIT, DIFFERENCE and STATUS keywords are only available when a valid LLF is present for those peaks. Otherwise, no calculations of these values can be made.
| EMISSION TABLE || The selection of the Emission Table
| SELECTED || The selection of the peak(s)
| PEAK NUMBER || The number of the peak(s)
| FREQUENCY || The frequency of the peak(s)
| PEAK || The measurement value of the peak(s)
| PEAK LIMIT || The limit value of the peak(s)
| PEAK DIFFERENCE || The difference value between the measured peak(s) value and the limit value.
| PEAK CORRECTION || The correction value of the peak(s)
| PEAK STATUS || The status of the peak(s): Pass / Fail
| AVERAGE || The average measurement value of the peak(s)
| AVERAGE LIMIT || The average limit value of the peak(s)
| AVERAGE DIFFERENCE || The average difference value between the measured average(s) value and the limit value.
| AVERAGE CORRECTION || The average correction value of the peak(s)
| AVERAGE STATUS || The average status of the peak(s): Pass / Fail
| QUASI-PEAK || The Quasi-peak measurement value of the peak(s)
| QUASI-PEAK LIMIT || The Quasi-peak limit value of the peak(s)
| QUASI-PEAK DIFFERENCE || The Quasi-peak difference value between the measured Quasi-peak(s) value and the limit value.
| QUASI-PEAK CORRECTION || The Quasi-peak correction value of the peak(s)
| HUMIDITY || Relative Humidity during the test in percentage relative Humidity [%RH]
| QUASI-PEAK STATUS || The Quasi-peak status of the peak(s): Pass / Fail
| RMS || The RMS measurement value of the peak(s)
| RMS LIMIT || The RMS limit value of the peak(s)
| RMS DIFFERENCE || The RMS difference value between the measured RMS(s) value and the limit value.
| RMS CORRECTION || The RMS correction value of the peak(s)
| RMS STATUS || The RMS status of the peak(s): Pass / Fail
| STATUS || The overall status of the peak(s): Pass / Fail
| NOTE || Note or remark made during the test
| ANGLE|| The angle of the peak(s) in degrees
| HEIGHT || The height of the peak(s) in meters [m]
| POLARIZATION || The polarization of the peak(s)
=== Radiated Emission Burst ===
| BURST GRAPH || The graph plot
| CARRIER FREQUENCY || The carrier frequency in [MHz]
| BURST TRIGGER LEVEL || The burst trigger level in [dBm]
| BURST MEASURE TIME || The burst measure time in seconds [s]
| BURST SAMPLE RATE || The burst sample rate
| BURST MINIMUM GAP TIME || The burst gap time in milli seconds [ms]
| PRESSURE || Air Pressure during the test in milliBar [mbar]
| BURST THRESHOLD LEVEL || The burst threshold level in [dBc]
| ANTENNA ASSEMBLY || The Antenna assembly gain in [dBi]
| TEMPERATURE || Temperature during the test in degrees Celsius [oC]
| BEAMFORMING GAIN || The Beamforming gain in [dB]
| TEST ENGINEER || The name of the test engineer
| MAXIMUM EIRP || The Maximum E.I.R.P. in [dBm]
| MEDIUM UTILISATION || The Medium Utilisation in percentage [%]
| EQUIPMENT TABLE || A table containing all used equipment
{|class="wikitable"
| DUTY CYCLE || The Duty Cycle in percentage [%]
| MINIMUM GAP TIME || The minimum Gap time in seconds [s]
| MAXIMUM SEQUENCE TIME || The Maximum Sequency time in seconds [s]
| RMS || The RMS value in [dBm]
===Equipment for Radiated Emission Tests===
{{Example|text={{ReportGeneratorCode|L1 / L2 RESULT}} is a valid code, and will return the actual result of the L1/L2 line.}}
=== Equipment for EFT Tests ===
{|class="wikitable"
{|class="wikitable"
| ANTENNA TOWER ... || Specifications of the used Antenna Tower
| NUMBER OF DETECTORS || The number of detectors the Receiver is able to handle simultaneously
| PRE AMPLIFIER ... || Specifications of the used Pre Amplifier
| RECEIVER TYPE || The type of receiver: Spectrum Analyser, Measurement Receiver or Scanning Receiver
| SPECTRUM ANALYZER ... || Specifications of the used Spectrum Analyser or Measurement Receiver
| TURN TABLE ... || Specifications of the used Turntable
Line 1,866: Line 2,213:
== Conducted Emission==
{{Example|text={{ReportGeneratorCode|EFTGENERATOR TYPE}} returns the type of the EFT generator.}}
===Conducted Emission Tests===
=== Surge Test ===
{|class="wikitable"
| AVERAGE DETECTOR || Indication whether or not the average detector is measured: Yes / No
| AVERAGE TRACE || Indication whether or not the average trace is measured: Yes / No
| DB/DIV || N/A
| DB/DIV MODE || N/A
| LISN LINE || Which LISN line is used: Neutral, Phase 1, Phase 2 or Phase 3
| QUASIPEAK DETECTOR || Indication whether or not the Quasi peak detector is measured: Yes / No
{|class="wikitable"
| START FREQUENCY || The start frequency in megahertz [MHz]
| STOP FREQUENCY || The stop frequency in megahertz [MHz]
=== Equipment for Surge Tests ===
{|class="wikitable"
{|class="wikitable"
{{Example|text={{ReportGeneratorCode|SURGEGENERATOR ID}} returns the internal ID of the Surge generator.}}
== Pulsed Immunity Multi Band Tests ==
===All Pulsed Immunity MB Tests share the following codes:===
{|class="wikitable"
| TestRepeatCount || Count to repeat all the sequences (Whole test)
| CABLE CABLELENGTH || Length of the cable
| CABLE DESCRIPTION || User description of the cable
| CABLE FIXINGSHIELD || Fixing shield of the cable
| CABLE LOADATPORT || Data port of the cable
| CABLE MAXCABLELENGTH || Specified maximum cable length
| CABLE NOTE || User note of the cable
| CABLE PORTTYPE || Used Port Type of the cable
| CABLE TYPEOFCABLE || Cable type
| SEQUENCE OVERVIEW || A table containing the sequence configuration of the test
==== Band Specific Codes ====
BAND + X + <Setting>
{|class="wikitable"
| BAND1 SequenceNumber || The number of the sequence.
| BAND1 TestSiteName || The selected test site with the used devices specified for the first sequence of the test.
| BAND1 TestNote || The user typed note for the first sequence of the test.
| BAND1 TestStatus || The test pass state specified for the first sequence of the test.
=== EFT ===
{|class="wikitable"
| BAND1 Line || Selected line coupling.
| BAND1 BurstDuration || The selected burst duration specified in the first sequence of the test.
| BAND1 BurstPeriod || The selected burst period specified in the first sequence of the test.
| BAND1 SequenceDuration|| The selected sequence duration specified in the first sequence of the test.
| BAND1 Polarity || The selected polarity of the VoltageTestLevel.
| BAND1 VoltageTestLevel || The voltage test level specified in the first sequence of the test
| BAND1 PulserepetitionRate || The pulse repetition rate specified in the first sequence of the test.
| BAND1 RepeatCount || The amount the sequence is repeated.
| BAND1 Test note || The note specified for each band.
=== Surge ===
{|class="wikitable"
| BAND1 Line || The selected line specified in the first sequence of the test.
| BAND1 VoltageTestLevel || The voltage test level specified in the first sequence of the test.
| BAND1 Polarity || The polarity specified in the first sequence of the test.
| BAND1 RI || The impedance specified in the first sequence of the test.
| BAND1 CI || The capacitance specified in the first sequence of the test.
| BAND1 PulseRepetitionTime || The pulse repetition time specified in the first sequence of the test.
| BAND1 PulseCount || The pulse count specified in the first sequence of the test.
| BAND1 PhaseAngles || The pulse angles specified in the first sequence of the test.
=== VDI ===
{|class="wikitable"
| BAND1 Changedvoltage|| The changed voltage for the dips specified for the first sequence of the test
| BAND1 Decreasevoltagetime|| The switch time from regular voltage to the dip voltage
| BAND1 Dipcount|| The amount of dips per sequence
| BAND1 Increasevoltagetime|| The time to switch from dip voltage to regular voltage
| BAND1 Line|| The selected line coupling on which the dip will occur
| BAND1 Locationtypetext|| The AC or DC mode
| BAND1 Phaseangles|| The angle on which the dip will initiate
| BAND1 Ratedvoltage|| The regular voltage
| BAND1 Reducedvoltagetime|| The time that the voltage is at reduced voltage
| BAND1 Repetitiontime|| The duration of one sequence
| BAND1 Repeatcount|| The amount the sequence will be repeated
=== Pulsed Immunity MB Example ===
To remove keywords of unused sequences, the {{ReportGeneratorCode|StartDeleteBlock()}} and {{ReportGeneratorCode|EndDeleteBlock()}} codes can be used:
{{Example|text=
{{ReportGeneratorCode|STARTDELETEBLOCK()}}
Band 1 notes: {{ReportGeneratorCode|BAND1 TestNote}}
{{ReportGeneratorCode|ENDDELETEBLOCK()}}
Band 2 notes: {{ReportGeneratorCode|BAND2 TestNote}}
Band 3 notes: {{ReportGeneratorCode|BAND3 TestNote}}
Band 4 notes: {{ReportGeneratorCode|BAND4 TestNote}}
If only 2 sequences are available in the test, only 2 sequences' bands will be seen in the Report output.
{{Note|The pass/fail criteria, shown in the tables of the report generator, do not take the measurement uncertainty of the test setup into account.}}
=== Radiated Emission Tests ===
{|class="wikitable"
| CURRENT UNIT || The selected unit for current: [dBμA], [dBmA], [dBA], [μA], [mA] or  [A]
| DB/DIV || N/A
| DB/DIV MODE || N/A
| DISTANCE || The distance between the antenna and the EUT in meters [m]
| OPTIMIZE ANGLE OPTIMIZE AREA || The angle optimization area.
| OPTIMIZE ANGLE STEP SIZE || The step size of the angle optimization.
| OPTIMIZE HEIGHT OPTIMIZE AREA || The height optimization area.
| OPTIMIZE HEIGHT STEP SIZE || The step size of the height optimization.
| END ANGLE || The stop position in degrees
| NUMBER OF ANGLES || The number of steps between the start position and the stop position
| FREQUENCY STEP || The step frequency in kilo Hertz [kHz]
| ABSORBING CLAMP ... (1) || Specifications of the used Absorbing Clamp
| CLAMP MOVER ... (1) || Specifications of the used Clamp Mover
{|class="wikitable"
| LARGEST DISTANCE || The maximum distance of the clamp in meters [m]
| SMALLEST DISTANCE || The minimum distance of the clamp in meters [m]
| STEPS || The number of steps between the minimum distance and the maximum distance
=== Radiated Emission Table ===
To retrieve the RAW value, extend the keyword by applying "_RAW" (without any white space) to this keyword. E.g. "PEAK_RAW".
{|class="wikitable"
{|class="wikitable"
=== Equipment for Radiated Emission Tests ===
{|class="wikitable"
{|class="wikitable"
{{Example|text={{ReportGeneratorCode|ANTENNA TOWER BRAND}} returns the brand name of the antenna tower.}}
== Conducted Emission ==
=== Conducted Emission Tests ===
{|class="wikitable"
| DB/DIV || N/A
| DB/DIV MODE || N/A
| EMISSION TABLE || Table of detected peaks and per peak: frequency, level(s), limit value, location
=== Equipment for Conducted Emission Tests ===
{|class="wikitable"
| LISN ... || Specifications of the used LISN
{|class="wikitable"
{{Example|text={{ReportGeneratorCode|CABLE 5 BRAND}} returns the brand name of cable nr. 5.}}
== Excel Exporter ==
The EUT data can be exported to Microsoft Excel with the Excel Exporter. Press the {{ScreenElement|Export}} button on the {{ScreenElement|Reports}} tab of the EUT window, and the {{WindowTitle|Excel Export}} window will open:
[[Image:Excel Export.png]]
=== Selecting filename ===
=== Selecting tests to be exported ===
You can select or deselect a test (that is to be exported) by checking or un-checking the checkbox in front of the name of the test. Press {{ScreenElement|Select all}} to select all the tests, press {{ScreenElement|Unselect all}} to deselect all the tests and press {{ScreenElement|Invert selection}} to invert the selection. (This means that all the selected tests will be deselected, and all the deselect tests will be selected.)
=== Changing the test order ===
The ‘export module’ exports all the selected tests, starting from the first test at the top and ending with the last test at the bottom of the list. This order is not mandatory and can be changed. To change this order, select the test you want to move and press the ‘arrow up’ [[Image:Button arrow up.png]] or ‘arrow down’ [[Image:Button arrow down.png]] button to move the test up or down in the list.
=== Export options ===
{{ScreenElementDescription|Insert header above data|With ‘insert header above data’ you can insert headers above every exported column. (For example for the signal generator level, forward power and reflected power for immunity test.)}}
{{ScreenElementDescription|Export EUT data|’Export EUT data’ allows you to select the EUT data that is to be exported. Only the EUT information stored in the top test can be exported.}}
{{ScreenElementDescription|Include not measured columns|’Include not measured columns’ allows you to select the ‘not measured’ columns that are to be exported. If you want to run a macro after exporting, it may be easier to export all the columns. You can then use the same macro for a signal level, forward power level and net power level test. It is not possible to have {{RadiMation}} run the macro automatically after the export. Deselecting ‘exporting not measured columns’ will result in faster exporting of the information.}}
=== Performing the export ===
Once the {{ScreenElement|Export}} button is pressed, all the raw data of the selected tests will be exported to the selected Excel file.
The exported data, its format and the amount of data that is exported, depends on the selected tests. When the data has been exported, the {{WindowTitle|Excel Export}} window will close.
{{Warning|Microsoft Excel needs to be installed on the PC to export the EUT information. This should be version 'Microsoft Excel 97' or newer.}}
== Command Line Interface (CLI) ==
With the command line interface it is possible to automatically generate a report from the command line.
This can be accomplished by calling "radicli.exe", which is available in the same directory in which RadiMation is also installed.
===Equipment for Conducted Emission Tests===
{{Note|When a RadiMation instance is active which supports the radimation cli interface, the report will be generated with the running instance. If {{RadiMation}} is not running, the report will be generated in the background with the same version as where the radicli.exe resides. }}
The equipment keyword consists of two parts: the keyword followed by a parameter. For each keyword all parameters are valid.
{{Example|header=Example for a command to generate a report:
|text=cd "C:\Program Files (x86)\Raditeq\RadiMation\Version x.y.z\"
radicli.exe generatereport --eut "C:\EUT_Files\eutfile.eut" --template "C:\templates\template.txt" --report "C:\reports\report.txt"
The possible arguments are:
{|class="wikitable"
! Parameter || Description
!style="column-width: 350px;" | Argument
! Description
| BRAND || Brand of Measurement equipment
| CALIBRATION EXPIRES || Date when calibration expires of equipment, given by the user
| DESCRIPTION || Description of Measurement equipment given by the user
| DEVICE DRIVER DLL || N/A
| generatereport || Required verb that specifies that a report should be generated from the command line interface.
| DEVICE DRIVER NAME || The internal name of the device driver
| --eut <eutfilename> || Required argument that specifies the EUT filename that should be used to generate the report.
|- style="break-inside:avoid-column"
| --template <templatefilename> || Required argument that specifies the template filename to use.
| HARDWARE VERSION || Hardware version of Measurement equipment
| --report <reportfilename> || Required argument that specifies the report filename to use.
| ID || Internal ID of Measurement equipment
| --reportforeachtest || Optional argument that specifies that a report is generated for each test individually. This will result in multiple reports being generated. A report generator code can be specified in the <reportfilename> to have a dynamic report filename. See the example below this table.
| INI FILE NAME || N/A
| --testnumbers <1[;2][;3][...]>|| Optional argument that specifies a ';' separated list of one or more test numbers (1, 2, 3, ....) for the tests (and their order) that should be included in the generated report. These test numbers correspond to the test numbers as can be seen in the EUT window
| NAME || Name of Measurement equipment
| --removeillegalcodes || Optional argument that specifies that report generator codes that are not replaced to an actual value will be removed from the generated report.
| SERIAL NUMBER || Serial number of Measurement equipment
| --help || Shows all the possible command line arguments that can be used.
| SOFTWARE VERSION || Software version of Measurement equipment
| SUPPORTED DEVICE TYPES || N/A
| TYPE || Type of Measurement equipment
{{Example|header=Example to generate a report for each test:
|text=radicli.exe generatereport --eut eutfile.eut --template template.txt --report report{{ReportGeneratorCode|testnumber}}.txt --reportforeachtest
= Excel Exporter =
The EUT data can be exported using the Excel Exporter to Microsoft Excel. Press the {{ScreenElement|Export}} button on the {{ScreenElement|Reports}} tab of the EUT window, and the {{WindowTitle|Excel Export}} window will open:
To select the filename you want to use for saving the information press the select file button next to the red cross. A save window will appear allowing you to select the path and filename. After closing the save window the name will be displayed on the export window.
You can select or deselect a test for export by checking or un-checking the checkbox in front of the name of the test. Pressing {{ScreenElement|Select all}} will select all tests, pressing {{ScreenElement|Unselect all}} will deselect all tests and pressing {{ScreenElement|Invert selection}} will invert the selected. Invert means that all selected tests will be deselected and all deselect tests will be selected.
The export module exports all the selected tests from the first test at the top to the last test at the bottom of the list. This order of the tests can be changed. Select the test you want to move, press the arrow up [[Image:Button arrow up.png]] or down [[Image:Button arrow down.png]] button and the test is moved one place up or one place down.
{{ScreenElementDescription|Insert header above data|With insert header above data you can select, if you want, headers above every exported column, for example signal generator level, forward power and reflected power for immunity test. }}
{{ScreenElementDescription|Export EUT data|Export EUT data allows you to select, if you want, the EUT data to be exported. Only the EUT information stored in the top test can be exported.}}
{{ScreenElementDescription|Include not measured columns|Include not measured columns allows you to select, if you want, not measured columns to be exported. If you want to run a macro after exporting, it may be easier to export all the columns. You can then use the same macro for a signal level, forward power level and net power level test. It is not possible to have {{RadiMation}} run the macro automatically after the export. Deselecting exporting not measured columns will result in faster exporting of the information.}}
The exported data, it's format and the amount of data that is exported is depending on the selected tests. When the export of the data is finished, the {{WindowTitle|Excel Export}} window will close.
{{Warning|Microsoft Excel needs to be installed on the PC to export the EUT information. The version of Microsoft Excel should be greater then or equal to version 'Microsoft Excel 97'.}}
{{Example|header=Example to only include the tests with the testnumber 3, 2 and 4 in the report:
|text=radicli.exe generatereport --eut eutfile.eut --template template.txt --report report.txt --testnumbers "3;2;4"
[[Category:Manual]]
[[Category:Manual]]
Latest revision as of 10:12, 23 July 2025
Report Generator[edit]
The basic information needed to generate a test report, as well as the more specific features of the Report Generator, are explained in the following paragraphs.
The Report file is the 'output' of the Report Generator and the actual test report. It contains all the desired data from the tests performed in the form of text, graphs, tables, etc. . Depending on the Template file, this report will be written in Microsoft Word, Microsoft Excel or Text format.
Keywords[edit]
Keep in mind that:
  • In a 'test-specific' template the Main Keywords are not used, (only the second keywords).
Unrecognized Keywords[edit]
Generating a report[edit]
Making a template file[edit]
To change the order of the list, select a test from the list and change its position by using the arrows on the right side of the field.
The report can be generated once the template file has been selected and the content of the report has been determined.
The report generation process cannot be paused or cancelled by the end user. Depending on the content, this process can take a few seconds (for a text file) up to several minutes (for an extensive report containing many tables and graphs in MS Word).
Specific Features[edit]
Example:
Client: ||EUT|CLIENT CONTACT PERSON||
||ENDDELETEBLOCK()||
  • If the 'Client Contact Person' data has been entered in the EUT file, the code above will be replaced (with the corresponding data) and the text 'Client: ' and the actual data will remain visible in the report.
  • If no data was entered for the 'Client Contact Person' in the EUT file, the code will not be replaced and the complete block including the text 'Client: ' and the not replaced report generator code will be removed from the final report.
  • If multiple report generator keywords are specified within the ||STARTDELETEBLOCK()|| and ||ENDDELETEBLOCK()||, the block will only be removed if all of the report generator codes are not replaced. The complete block will remain in the report if one or more of the report generator codes are replaced.
  • It is allowed and possible to use a ||STARTDELETEBLOCK()|| and ||ENDDELETEBLOCK()|| combination within another ||STARTDELETEBLOCK()|| and ||ENDDELETEBLOCK()|| set.
Test-specific templates[edit]
Example of a 'main-template':
Introduction
||EUT|ALL TESTS||
Conclusion
  • With 'test-specific' templates it is sufficient to specify ||MODULATION||, because the Report Generator already knows that the template (and the codes within) are linked to that test.
Example:
||SIGNAL POWER GRAPH; FORWARD POWER GRAPH|| Will create a graph with the signal power and the forward power graph lines.
||FIELD GRAPH 1; FIELD GRAPH 2; AVERAGE FIELD GRAPH|| Will create a field strength graph with the field strength of the probes 1 and 2, and the average field graph.
||CURRENT GRAPH; NET POWER GRAPH|| Will create a graph with the current and the net power. This graph will have two y-axes to indicate the different units.
  • SIGNAL POWER GRAPH
  • CURRENT GRAPH
  • VOLTAGE GRAPH
  • FORWARD POWER GRAPH
  • REFLECTED POWER GRAPH
  • NET POWER GRAPH
  • MID POWER GRAPH
  • OUT POWER GRAPH
  • IMPEDANCE GRAPH
  • ERROR GRAPH
  • CALCULATED FIELD GRAPH
  • AVERAGE FIELD GRAPH
  • CALCULATED VOLTAGE GRAPH
  • CALCULATED CURRENT GRAPH
  • AD GRAPH 1
  • AD GRAPH 2
  • AD GRAPH 3
  • AD GRAPH 4
  • AD GRAPH 5
  • AD GRAPH 6
  • AD GRAPH 7
  • AD GRAPH 8
  • FIELD GRAPH 1
  • FIELD GRAPH 2
  • FIELD GRAPH 3
  • FIELD GRAPH 4
  • FIELD GRAPH 5
  • FIELD GRAPH 6
  • FIELD GRAPH 7
  • FIELD GRAPH 8
  • CALIBRATED POWER GRAPH
  • EUT ATTENUATION GRAPH
  • EUT DYNAMIC RANGE GRAPH
The list of known graph report generator codes for multiband tests is not a fixed list, because an unlimited number of graph lines can be created in a multiband test. However, the report generator can include any graph that is also available in the multiband test. The report generator code to include a graph in the report for a multiband test is the same as the name of the graph as it is shown in the legend above the graph in the multiband test results. A combination of the available graphs can also be included as a single graph in the report by combining the desired graph names in a single report generator code. As long as the labels of the graph are ';' separated, any number of graph lines can be included in the graph in the report.
Example:
||POWER; MEASURED POWER LIMIT; Amplifier forward power; Amplifier input||
will result in the following graph:
  • time:
  • frequency:
  • angle:
  • height:
Example:
Export an angle graph: ||angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1||
Export a time graph: ||time:Signal Power;Channel1;Channel2||
Export a height graph: ||height:RBW: 1000 kHz, Horizontal Max Peak for peak 1||
Example:
Export an angle indexed table: ||angle:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE||
Export a time indexed table: ||time:Signal Power TABLE||
Export a height indexed table: ||height:RBW: 1000 kHz, Horizontal Max Peak for peak 1 TABLE||
  • Frequency
  • Frequency
  • Frequency
Example:
||SIGNAL POWER, FORWARD POWER, AD CHANNEL 5 LIST||
||NET POWER, AD CHANNEL 2, CALCULATED FIELD TABLE||
||VOLTAGE, CALCULATED CURRENT, FORWARD POWER LIST||
  • FREQUENCY
  • SIGNAL POWER
  • FORWARD POWER
  • REFLECTED POWER
  • NET POWER
  • CURRENT
  • VOLTAGE
  • CALCULATED FIELD
  • AVERAGE FIELD
  • FIELD 1
  • FIELD 2
  • FIELD 3
  • FIELD 4
  • FIELD 5
  • FIELD 6
  • FIELD 7
  • FIELD 8
  • AD CHANNEL 1
  • AD CHANNEL 2
  • AD CHANNEL 3
  • AD CHANNEL 4
  • AD CHANNEL 5
  • AD CHANNEL 6
  • AD CHANNEL 7
  • AD CHANNEL 8
  • CALIBRATION FILE
  • TEST ENGINEER
  • TEST DATE
  • TEST TIME
  • START TIME
  • STOP TIME
  • TEMPERATURE
  • HUMIDITY
  • PRESSURE
  • START FREQUENCY
  • STOP FREQUENCY
  • FREQUENCY STEP
  • LOGARITHMIC STEP
  • LOGARITHMIC STEP FROM
  • POLARIZATION
  • DISTANCE
  • TOLERANCE
  • DWELL TIME
  • MODULATION
  • MODULATION DEPTH
  • MODULATION CONSERVATION
  • MODULATION FREQUENCY
  • MODULATION SIGNAL
  • MODULATION DUTY CYCLE
  • MAINS VOLTAGE
  • FREQUENCY CHANGE MODE
  • AVERAGE FIELD
  • CURRENT MEASURED
  • CALIBRATION METHOD
  • FORWARD POWER MEASURED
  • REFLECTED POWER MEASURED
  • NET POWER MEASURED
  • CALCULATED FIELD MEASURED
  • RECALCULATE AVERAGE
  • DESCRIPTION
  • TSF FILENAME
  • AMPLIFIER PROTECTION MODE
  • AUTODISAPPEAR
  • USE VOLTAGE TEST LEVEL
  • VOLTAGE TEST LEVEL
  • POWER TEST LEVEL
  • FIELD TEST LEVEL
  • USE CURRENT TEST LEVEL
  • CURRENT TEST LEVEL
  • STEP DOWN
  • USE CURRENT LIMIT
  • CURRENT LIMIT
  • MODULATION TYPE
  • MAGNETIC FIELD TEST LEVEL
Macros[edit]
  • A macro name predefined by the Microsoft Word report generator editor. These macros are:
Table Formatting[edit]
The report generator will also use the formatting and settings that are present in the template document. This allows defining how inserted tables are formatted in the final report. To make this possible, the table code should be placed in a table in the template. In Microsoft Word, you can then set the border styles of the outer table.
Specific graph sizes[edit]
The Microsoft Excel report generator editor allows generating Microsoft Excel reports. A normal Microsoft Excel file can be specified as the template file for the report generator, and all the available RadiMation® report generator codes can be used in that template file.
Values that are inserted in the generated Microsoft Excel report are always using the US English formatting, to prevent problems with the interpretation of the values if Microsoft Excel is using different Culture settings. If additional formulas and calculations are performed in the generated Microsoft Excel report, the formulas should be aware that the values are inserted in the US English format, and may need to be converted to the local format.
Table-inclusion modes[edit]
  • Row insert mode
  • Cell insert mode
  • Overwrite mode
  • Sheet insert mode
Row insert mode[edit]
Cell insert mode[edit]
The 'cell insert mode' is very similar to the 'row insert mode' in that new cells are inserted before the table is included. The difference however is that the surrounding content is not just moved down, but also to the right. Meaning that, once the table is inserted, any content that was placed next to the starting cell will now be present next to the table. This can be useful if calculations should be performed on the data that is present in the inserted table.
Overwrite mode[edit]
Sheet insert mode[edit]
Macros[edit]
    • CELL_PASTE_MODE_CELLS
    • CELL_PASTE_MODE_SHEET
    • CELL_PASTE_MODE_OVERWRITE
    • CELL_PASTE_MODE_ROWS
SQL queries on tables[edit]
SELECT * FROM <TABLE>
Examples[edit]
Example to only include the frequencies above 50 MHz:
Example to only include the frequencies between 1 Hz and 20 MHz, where the AD-channel was measured out of the specified range ('TABLE'), and the table is reverse-sorted on the frequency column:
||SELECT * FROM AD CHANNEL 8,7,6,5,4,3,2,1 TABLE WHERE FREQUENCY >= 1 and FREQUENCY <= 20E6 ORDER BY Frequency desc||
Example to include the 'EQUIPMENT TABLE' including all the columns (also the hidden columns):
Example to include all columns of a table:
The report generator code ||SELECT ALL FROM PEAKTABLE|| will show the table including all the available columns, and could result in:
Peak Number Peak Number_RAW Frequency Frequency_RAW Peak Peak_RAW Peak Limit Peak Limit_RAW Peak Difference Peak Difference_RAW Peak Correction Peak Correction_RAW Peak Status Peak Status_RAW Status Status_RAW Angle Angle_RAW
1 1 829.92 MHz 829921205 80 dBµV/m 80 60 dBµV/m 60 20 dB 20 0.00 dB 0.00 Fail Fail Fail Fail 0 Degree 0
  • Peak Number
  • Frequency
  • Peak
  • Peak Limit
  • Peak status
  • Overall status
The reason for this, as stated earlier, is that any column name that includes a space, should be included in '[ ]'.
… which does work.
1 829.92 MHz 80 dBµV 60 dBµV 20 dB Fail Fail
Example to limit the number of rows in the table to 3 rows:
||SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE LIMIT 3||
will thus result in:
1 829.92 MHz 80 dBµV/m 60 dBµV/m 20 dB Fail Fail
2 749.42 MHz 68.8 dBµV/m 60 dBµV/m 8.8 dB Fail Fail
3 951.44 MHz 73.2 dBµV/m 60 dBµV/m 12.2 dB Fail Fail
Example to sort the rows in the table descending on the frequency:
||SELECT [Peak Number],Frequency,Peak,[Peak Limit],[Peak Difference],[Peak Status],Status FROM PEAKTABLE ORDER BY Frequency DESC||
4 805.11 MHz 62.1 dBµV/m 60 dBµV/m 1.2 dB Fail Fail
Keywords Overview[edit]
EUT[edit]
Common[edit]
Keyword Description
||EUT|FILENAME|| The filename of the EUT
||EUT|BRAND|| The brand of the EUT specified in the EUT Window
||EUT|CONDITION UPON RECEPTION|| The condition of reception of the EUT specified in the EUT Window
||EUT|DATE OF RECEIPT|| The date of reception of the EUT specified in the EUT Window
||EUT|DATE OF MEASUREMENT|| The date of measurement of the EUT specified in the EUT Window
||EUT|DIMENSIONS|| The dimensions of the EUT specified in the EUT Window
||EUT|EMISSION STANDARD|| The emission standard of the EUT specified in the EUT Window
||EUT|ENVIRONMENT|| The environment of the EUT specified in the EUT Window
||EUT|IMMUNITY STANDARD|| The immunity standard of the EUT specified in the EUT Window
||EUT|NAME|| The name of the EUT specified in the EUT Window
||EUT|ORDER NUMBER|| The order number of the EUT specified in the EUT Window
||EUT|PCB RELEASE|| The PCB release of the EUT specified in the EUT Window
||EUT|PERIPHERAL EQUIPMENT|| The peripheral equipment of the EUT specified in the EUT Window
||EUT|PLACE MEASURMENT|| The place of measurment of the EUT specified in the EUT Window
||EUT|RADIMATION VERSION|| The RadiMation® version
||EUT|RATED POWER|| The rated power of the EUT specified in the EUT Window
||EUT|SERIAL NUMBER|| The serial number of the EUT specified in the EUT Window
||EUT|TEST SETUP TYPE|| The test setup type of the EUT specified in the EUT Window
||EUT|SOFTWARE RELEASE|| The software release of the EUT specified in the EUT Window
||EUT|TYPE NUMBER|| The type number of the EUT specified in the EUT Window
||EUT|VOLTAGE SUPPLY|| The voltage supply of the EUT specified in the EUT Window
Keyword Description
||EUT|CLIENT ADDRESS|| The address of the client specified in the EUT Window
||EUT|CLIENT CAPACITY|| The capacity of the client specified in the EUT Window
||EUT|CLIENT CITY|| The city of the client specified in the EUT Window
||EUT|CLIENT COMPANY|| The client name specified in the EUT Window
||EUT|CLIENT CONTACT PERSON|| The contact person of the client specified in the EUT Window
||EUT|CLIENT COUNTRY|| The country of the client specified in the EUT Window
||EUT|CLIENT EMAIL ADDRESS|| The email address of the client specified in the EUT Window
||EUT|CLIENT FAX NUMBER|| The fax number of the client specified in the EUT Window
||EUT|CLIENT STATE|| The state of the client specified in the EUT Window
||EUT|CLIENT TELEPHONE NUMBER|| The telephone number of the client specified in the EUT Window
||EUT|CLIENT ZIP CODE|| The ZIP code of the client specified in the EUT Window
Keyword Description
||EUT|MANUFACTURER ADDRESS|| The address of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER CITY|| The city of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER COMPANY|| The manufacturer name specified in the EUT Window
||EUT|MANUFACTURER CONTACT PERSON|| The contact person of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER COUNTRY|| The country of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER EMAIL ADDRESS|| The email address of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER FAX NUMBER|| The fax number of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER STATE|| The state of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER TELEPHONE NUMBER|| The telephone number of the manufacturer specified in the EUT Window
||EUT|MANUFACTURER ZIP CODE|| The ZIP code of the manufacturer specified in the EUT Window
Keyword Description
||EUT|TEST SITE ADDRESS|| The address of the test site specified in the EUT Window
||EUT|TEST SITE CITY|| The city of the test site specified in the EUT Window
||EUT|TEST SITE COMPANY|| The test site name specified in the EUT Window
||EUT|TEST SITE CONTACT PERSON|| The contact person of the test site specified in the EUT Window
||EUT|TEST SITE COUNTRY|| The country of the test site specified in the EUT Window
||EUT|TEST SITE EMAIL ADDRESS|| The email address of the test site specified in the EUT Window
||EUT|TEST SITE FAX NUMBER|| The fax number of the test site specified in the EUT Window
||EUT|TEST SITE STATE|| The state of the test site specified in the EUT Window
||EUT|TEST SITE TELEPHONE NUMBER|| The telephone number of the test site specified in the EUT Window
||EUT|TEST SITE ZIP CODE|| The ZIP code of the test site specified in the EUT Window
Cable information[edit]
Keyword Description
||EUT|CABLE TABLE|| A table with the cable information specified in the EUT Window
||EUT|CABLE X PORT TYPE|| The port type of cable X in the EUT window ( X starting from 1)
||EUT|CABLE X CABLE TYPE|| The cable type of cable X in the EUT window (X starting from 1)
||EUT|CABLE X CABLE LENGTH|| The cable length of cable X in the EUT window (X starting from 1)
||EUT|CABLE X FIXING SHIELD|| The fixing shield of cable X in the EUT window (X starting from 1)
||EUT|CABLE X LOAD AT PORT|| The load at port of cable X in the EUT window (X starting from 1)
Standards Information[edit]
Parameter Description
REFERRED STANDARDS A string of referred standards, separated by a "," (Only applicable to product standards)
Example:
||EUT|PRODUCT STANDARD REFERRED STANDARDS||
||EUT|PRODUCT STANDARD DESCRIPTION||
||EUT|SELECT * FROM [BASIC STANDARD TABLE]||
||EUT|SELECT ALL FROM [BASIC STANDARD TABLE]||
||EUT|SELECT [OFFICIAL NAME] FROM [BASIC STANDARD TABLE]||
Keyword Description
EDITOR Report generator
Directories[edit]
Keyword Description
||Configuration|DEVICEDRIVERFILESDIRECTORY|| CONFDVDR directory
||Configuration|CALFILESDIRECTORY|| CAL Files directory
||Configuration|EUTFILESDIRECTORY|| EUT Files directory
||Configuration|TSFFILESDIRECTORY|| TSF Files directory
||Configuration|SEQFILESDIRECTORY|| SEQ Files directory
||Configuration|CORFILESDIRECTORY|| COR Files directory
||Configuration|DIRECTORIES|| All directories
Units[edit]
Keyword Description
||Configuration|CURRENT|| Current
||Configuration|EMISSION AMPLITUDE|| Emission
||Configuration|FIELD|| Electrical Field
||Configuration|FREQUENCY|| Frequency
||Configuration|POWER|| Power
||Configuration|VOLTAGE|| Voltage
||Configuration|UNITS|| All units
General[edit]
Keyword Description
||Configuration|CURRENTUSERNAME|| Current username
||Configuration|RADIMATION VERSION|| The RadiMation® version
EUT Calibration[edit]
  • Attenuation / Gain
  • Amplifier linearity
  • Gain compression
  • Harmonics
  • S Parameter
  • Antenna scan
  • Antenna
  • Field strength
Amplifier Linearity[edit]
Keyword Description
FREQUENCY STEP Step size
ALL GRAPHS Output graph
Gain compression[edit]
Keyword Description
TABLE Output table
GRAPH Output graph
POWER TABLE Table with all the compression power levels for every frequency
Harmonics[edit]
Keyword Description
TABLE Output table
GRAPH Output graph
S Parameter[edit]
Keyword Description
ALL GRAPHS All available graphs plotted in one graph
Sxx PRESCAN GRAPH Prescan Graph
Sxx MAGNITUDE AVG GRAPH Magnitude Average Graph
Sxx MAGNITUDE MAX GRAPH Magnitude Maximum Graph
Sxx MAGNITUDE MIN GRAPH Magnitude Minimum Graph
Sxx MAGNITUDE STDDEV GRAPH Magnitude Standard Deviation Graph
Sxx IMPEDANCE AVG GRAPH Graph with the average of the measured impedance
Sxx IMPEDANCE MAX GRAPH Graph with the max-hold of the measured impedance
Sxx IMPEDANCE MIN GRAPH Graph with the min-hold of the measured impedance
Sxx IMPEDANCE STDDEV GRAPH Graph with the Standard Deviation of the measured impedance
Keyword Description
S11 TABLE All columns default visible
Select * FROM S11 TABLE Using sql to retrieve all values
Select IMPEDANCE FROM S11 TABLE Because we added impedance later, it is not visible by default
Antenna scan[edit]
Keyword Description
POWER TABLE Default table with frequency, signal power and forward power.
SELECT ALL FROM [Power Table] To request all possible columns
Radiated Immunity[edit]
Keyword Description
Keyword Description
FREQUENCY STEP The linear frequency step in megahertz [MHz]
STEP A textual description of the used frequency step, also containing if it is linear or logarithmic step size
FIELD UNIT The selected unit for field strength: [dBμV/m], [dBmV/m], [dBV/m], [μV/m], [mV/m] or [V/m]
POWER UNIT The selected unit for power: [dBμW], [dBm], [dBmW], [dBW], [μW], [mW] or [W]
VOLTAGE UNIT The selected unit for voltage: [dBμVrms], [dBmVrms], [dBVrms], [μVrms], [mVrms] or [Vrms]
NUMBER OF TOTAL POINTS The total of calibration points
NUMBER OF POINTS TO REMOVE The number of points removed
ACCEPTED TOLERANCE The accepted tolerance in dB
ACCEPTED EXTENDED TOLERANCE The accepted extended tolerance in dB
PERCENTAGE WITHIN TOLERANCE The percentage of frequencies within the accepted tolerance
PERCENTAGE WITHIN EXTENDED TOLERANCE The percentage of frequencies within the accepted extended tolerance
PERCENTAGE OUTSIDE TOLERANCE The percentage of frequencies outside the tolerance
MAX STANDARD DEVIATION MEASURED The maximum standard deviation
FREQUENCY FOR MAX STANDARD DEVIATION The frequency with the maximum standard deviation
MAX TOLERANCE DEVIATION MEASURED The maximum tolerance deviation
FREQUENCY FOR MAX DEVIATION The frequency with the max tolerance
RESULT UFA pass or fail result
STDEV GRAPH The graph representing the standard deviation per frequency
PROBABILITY PERCENTAGE The probability percentage
CROSS POLAR PROBABILITY PERCENTAGE The cross-polar probability percentage
CROSS POLAR ACCEPTED TOLERANCE The cross-polar accepted tolerance in dB
CROSS POLAR ACCEPTED EXTENDED TOLERANCE The cross-polar accepted extended tolerance in dB
CROSS POLAR EXTENDED TOLERANCE PERCENTAGE The cross-polar extended tolerance percentage
CROSS POLAR RESULT The cross-polar result, pass or fail
CROSS POLAR PERCENTAGE WITHIN TOLERANCE The cross-polar percentage within tolerance
CROSS POLAR NUMBER OF POINTS WITHIN TOLERANCE The cross-polar number of frequency points within tolerance
CROSS POLAR PERCENTAGE WITHIN EXTENDED TOLERANCE The cross-polar percentage of frequency points within the extended tolerance
CROSS POLAR NUMBER OF POINTS WITHIN EXTENDED TOLERANCE The cross-polar number of frequency points within the extended tolerance
CROSS POLAR PERCENTAGE OUTSIDE TOLERANCE The cross-polar percentage outside the tolerance
CROSS POLAR NUMBER OF POINTS OUTSIDE TOLERANCE The cross-polar number of points outside the tolerance
CROSS POLAR MAX PQUANTILE The cross-polar maximum PQuantile
CROSS POLAR FREQUENCY FOR MAX PQUANTILE The cross-polar frequency for the max PQuantile
PQUANTILE GRAPH The graph representing the PQuantile per frequency
Keyword Description
DEVICE DRIVER DLL N/A
INI FILE NAME N/A
SUPPORTED DEVICE TYPES N/A
Example:
||AMPLIFIER BRAND|| returns the brand name of the amplifier.
Radiated Immunity Tests[edit]
Keyword Description
CURRENT UNIT The selected current unit: [dBμA],[dBmA], [dBA], [μA], [mA] or [A]
STEP FREQUENCY The linear frequency step in megahertz [MHz]
Keyword Description
INI FILE NAME N/A
Example:
||AMPLIFIER SERIAL NUMBER|| returns the serial number of the amplifier.
It is possible to include the test levels of a multi band immunity test. The name of a test level can be adjusted in the TSF and therefore needs to be looked up first.
Example:
The report generator code is built as following, where <Test level name> is replaced with the name of the configured test level and <Field> is the actual information that is put in the report.
||Band1 TestLevel: <Test level name> <Field>||
This example puts the 'Description' of the test level named 'Electrical field substitution method' in the report
||Band1 TestLevel: Electrical field substitution method Description||
This example puts the 'Calibration filename' of the test level named 'Electrical field substitution method' in the report
||Band1 TestLevel: Electrical field substitution method Calibration File||
This example puts the 'Field' of the test level named 'Electrical field substitution method' in the report
||Band1 TestLevel: Electrical field substitution method Fixed Field||
This example puts the 'Power' of the test level named 'Power closed loop method' in the report
||Band1 TestLevel: Power closed loop method Fixed Power||
Other fields than "Description" and "Calibration File" can be accessed, keep in mind that it depends on the test level which fields are exposed. Some examples of other fields:
Name Name of test level
Description Description of test level
Tolerance The tolerance of the test level
Calibration File Calibration file used in a substitution test
Calibration method Calibration method
Fixed Field The field strength to regulate to
Conducted Immunity[edit]
Keyword Description
CURRENT UNIT The selected current unit: [dBμA], [dBmA], [dBA], [μA], [mA] or [A]
Keyword Description
INI FILE NAME N/A
Example:
||COUPLER TYPE|| returns the type of the coupler.
Keyword Description
Keyword Description
INI FILE NAME N/A
Example:
||ANTENNA DESCRIPTION|| returns the description of the antenna.
Pulsed Immunity[edit]
ESD Test[edit]
Keyword Description
NEG AIR NOTE POINT x Note of Negative Air Discharge at point number [x]
NEG CONTACT NOTE POINT x Note of Negative Contact Discharge at point number [x]
POS AIR NOTE POINT x Note of Positive Air Discharge at point number [x]
POS CONTACT NOTE POINT x Note of Positive Contact Discharge at point number [x]
POS HCP MEASURED POINT X Indication a Positive HCP Discharge is performed at point number [x]: Yes / No
NEG HCP MEASURED POINT X Indication a Negative HCP Discharge is performed at point number [x]: Yes / No
POS HCP RESULT POINT X Results of Positive HCP Discharge at point number [x]: pass or fail
NEG HCP RESULT POINT X Results of Negative HCP Discharge at point number [x]: pass or fail
POS HCP LEVEL POINT X Level of the Positive HCP Discharge at point number [x] in [V]
NEG HCP LEVEL POINT X Level of the Negative HCP Discharge at point number [x] in [V]
POS HCP NOTE POINT X Note or remark for the Positive HCP Discharge at point number [x]
NEG HCP NOTE POINT X Note or remark for the Negative HCP Discharge at point number [x]
POS VCP MEASURED POINT X Indication a Positive VCP Discharge is performed at point number [x]: Yes / No
NEG VCP MEASURED POINT X Indication a Negative VCP Discharge is performed at point number [x]: Yes / No
POS VCP RESULT POINT X Results of Positive VCP Discharge at point number [x]: pass or fail
NEG VCP RESULT POINT X Results of Negative VCP Discharge at point number [x]: pass or fail
POS VCP LEVEL POINT X Level of the Positive VCP Discharge at point number [x] in [V]
NEG VCP LEVEL POINT X Level of the Negative VCP Discharge at point number [x] in [V]
POS VCP NOTE POINT X Note or remark for the Positive VCP Discharge at point number [x]
NEG VCP NOTE POINT X Note or remark for the Negative VCP Discharge at point number [x]
Equipment for ESD Tests[edit]
Keyword Description
INI FILE NAME N/A
Example:
||ESDGUN BRAND|| returns the brand name of the ESD gun.
EFT Test[edit]
Keyword Description
INI FILE NAME N/A
Keyword Description
Example:
||L1 / L2 RESULT|| is a valid code, and will return the actual result of the L1/L2 line.
Equipment for EFT Tests[edit]
Keyword Description
INI FILE NAME N/A
Example:
||EFTGENERATOR TYPE|| returns the type of the EFT generator.
Surge Test[edit]
Keyword Description
Keyword Description
Keyword Description
INI FILE NAME N/A
Example:
||SURGEGENERATOR ID|| returns the internal ID of the Surge generator.
Keyword Description
CABLE TYPEOFCABLE Cable type
SEQUENCE OVERVIEW A table containing the sequence configuration of the test
Band Specific Codes[edit]
BAND + X + <Setting>
Keyword Description
EFT[edit]
Keyword Description
BAND1 SequenceDuration The selected sequence duration specified in the first sequence of the test.
BAND1 RepeatCount The amount the sequence is repeated.
Surge[edit]
Keyword Description
VDI[edit]
Keyword Description
BAND1 Dipcount The amount of dips per sequence
BAND1 Repeatcount The amount the sequence will be repeated
Example:
||STARTDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
||ENDDELETEBLOCK()||
If only 2 sequences are available in the test, only 2 sequences' bands will be seen in the Report output.
Radiated Emission[edit]
Note: The pass/fail criteria, shown in the tables of the report generator, do not take the measurement uncertainty of the test setup into account.
Radiated Emission Tests[edit]
Keyword Description
CURRENT UNIT The selected unit for current: [dBμA], [dBmA], [dBA], [μA], [mA] or [A]
DB/DIV N/A
DB/DIV MODE N/A
OPTIMIZE ANGLE OPTIMIZE AREA The angle optimization area.
OPTIMIZE ANGLE STEP SIZE The step size of the angle optimization.
OPTIMIZE HEIGHT OPTIMIZE AREA The height optimization area.
OPTIMIZE HEIGHT STEP SIZE The step size of the height optimization.
Keyword Description
FREQUENCY STEP The step frequency in kilo Hertz [kHz]
USE INTERNAL PREAMP N/A
Keyword Description
Radiated Emission Table[edit]
Keyword Description
Radiated Emission Burst[edit]
Keyword Description
BURST GRAPH The graph plot
RMS The RMS value in [dBm]
Keyword Description
INI FILE NAME N/A
Example:
||ANTENNA TOWER BRAND|| returns the brand name of the antenna tower.
Conducted Emission[edit]
Keyword Description
DB/DIV N/A
DB/DIV MODE N/A
Keyword Description
Keyword Description
INI FILE NAME N/A
Example:
||CABLE 5 BRAND|| returns the brand name of cable nr. 5.
Excel Exporter[edit]
Selecting filename[edit]
Changing the test order[edit]
Export options[edit]
Performing the export[edit]
Note: When a RadiMation instance is active which supports the radimation cli interface, the report will be generated with the running instance. If RadiMation® is not running, the report will be generated in the background with the same version as where the radicli.exe resides.
Example for a command to generate a report:
cd "C:\Program Files (x86)\Raditeq\RadiMation\Version x.y.z\"
radicli.exe generatereport --eut "C:\EUT_Files\eutfile.eut" --template "C:\templates\template.txt" --report "C:\reports\report.txt"
The possible arguments are:
Argument Description
generatereport Required verb that specifies that a report should be generated from the command line interface.
--eut <eutfilename> Required argument that specifies the EUT filename that should be used to generate the report.
--template <templatefilename> Required argument that specifies the template filename to use.
--report <reportfilename> Required argument that specifies the report filename to use.
--reportforeachtest Optional argument that specifies that a report is generated for each test individually. This will result in multiple reports being generated. A report generator code can be specified in the <reportfilename> to have a dynamic report filename. See the example below this table.
--testnumbers <1[;2][;3][...]> Optional argument that specifies a ';' separated list of one or more test numbers (1, 2, 3, ....) for the tests (and their order) that should be included in the generated report. These test numbers correspond to the test numbers as can be seen in the EUT window
--removeillegalcodes Optional argument that specifies that report generator codes that are not replaced to an actual value will be removed from the generated report.
Example to generate a report for each test:
radicli.exe generatereport --eut eutfile.eut --template template.txt --report report||testnumber||.txt --reportforeachtest
Example to only include the tests with the testnumber 3, 2 and 4 in the report:
radicli.exe generatereport --eut eutfile.eut --template template.txt --report report.txt --testnumbers "3;2;4"