This page contains all kind of Frequently Asked Questions related to {{RadiMation}} | This page contains all kind of Frequently Asked Questions related to [[RadiMation|{{RadiMation}}]] | ||
{{RadiMationFAQ}} | {{RadiMationFAQ}} | ||
=== Does RadiMation support the linearity check that is described in the EN 61000-4-3 standard? === | === Does RadiMation support the linearity check that is described in the EN 61000-4-3 standard? === |
Revision as of 19:45, 21 January 2009
This page contains all kind of Frequently Asked Questions related to RadiMation®
Why are the Uniform Field Area calculations of expired standards not removed?[edit]
Is RadiMation® using the correct method for the Uniform Field Area calibration?[edit]
How does a G-TEM compare to an anechoic chamber?[edit]
A G-TEM cell is in principle not an accepted (full compliant) EMC test environment and compared to an anechoic chamber the cables will be routed and placed differently.
Sometimes the error at higher frequencies becomes smaller through the band. This is what normally is expected in case of correlation measurements between a small G-TEM in relation to a large Anechoic chamber or OATS (due to near field effects smaller test rooms will have larger deviations at low frequencies).
Beside this, one should realize that even full compliant radiated emission measurements still have relatively high measurement uncertainties. To give you some concrete figures:
- The absolute accuracy for an OATS measurement (including all uncertainties of equipment, site attenuation and antenna factor) are about +/- 5 dB.
- For emission measurements in anechoic chambers this figure is about +/- 6 dB for a SAR and even +/- 10 to 20 dB for a FAR.
So one can imagine that you can expect even higher uncertainties in a (G)TEM cell
Generally when comparing two sites with each other it is important to make the correct calculations for both uncertainties (root sum of the squares). So if both sites have 5 dB uncertainty (which is not feasible with a TEM cell) the total expected difference will be maximum 7 dB. In other words a measured 10 dB difference is not even bad at all!
What is the difference between the Report Generator and the Excel Exporter?[edit]
The report generator and the Excel exporter are using two different ways to export the information from RadiMation®.
- The report generator is fully configurable by the report generator template which specifies the information that should be exported, and how it should be represented in the final report. The formatting, generated information and amount of information is completely configurable.
- The Excel exporter can not be configured, it just dumps all the raw data into a Microsoft Excel worksheet. By design, all the data is exported without any correction factors applied to the values. The order and the location on which the information is exported cannot be modified, and is always the same.
The report generator is also capable of generating customizable reports in Microsoft Excel. The most common usage of the Excel exporter is to have all the information available in Microsoft Excel to perform further calculations.
Why are the TXT and the HTML formats not supported anymore by the report generator?[edit]
The report generator is being using more than in the past, for the inclusion of tables and pictures in the final report. The support in the report generator for the TXT and HTML files has been removed in version 5.3 and higher because those file types do not support embedded tables and pictures within the file. For the same reason the file types XML and RTF are not supported anymore.
The decision to stop the support of these file types in the report generator will not have a large impact, because to our knowledge, these file types are not use by any customer.
Signal generator is going to -100 dBm and RadiMation is reporting that it can not go any lower[edit]
If the signal generator is going to -100 dBm during a test, then there is something wrong in the hardware configuration or the software configuration.
Possible causes and solutions:
- Verify that the hardware configuration is the same as the software configuration. It is possible that an attenuation is not included in the test-site configuration. It is also possible that a coupler with another attenuation was selected in the test-site configuration. Adjust the configuration of the test-site in such a way that the hardware configuration and the software configuration are equal.
- Check the noise level of the powermeter, so the minimum power level which can be measured with the used power meter is known. The problem can be caused by a combination of the noise level of the used powermeter and the attenuation in front of the powermeter, which is in total higher then the requested power level during the test.
Auto attenuator is going to 10 dB even with low emission values[edit]
When auto attenuation is selected, RadiMation® sends a command to the receiver or analyser to select auto attenuation. The attenuation that will be selected by the receiver or analyser is then depending on the used measurement device itself. When an analyser is used, it can help to decrease the reference level to for example 50 dbuV. Verify if the attenuator has been decreased at the lower reference level setting. A lot of analysers can only use a 0 dB attenuation value, if the reference level is lower or equal to -20 dbm (or 87 dBuV).
Why is there a big drop between the value during and after a G-tem measurement?[edit]
For a G-tem measurement you need to change the EUT to all three axis (X,Y and Z). During the measurement of one side you will see the emission coming out of that side. After measuring all three side the G-tem correction to the OATS can be applied, including the normalised site attenuation. For frequencies smaller then 1 Ghz this can mean that the total value is lower then the measured on one side. For example the attenuation for a horizontal 3 meter site at 30 MHz is 15.8. For the horizontal 10 meter site at 30 MHz this is 29.8. See the standard for the rest of the values.
- In the RadiMation® CONFDVDR folder (most often: C:\Program Files\DARE Development\RadiMation\CONFDVDR\) the file SETTINGS.INI is present. Open this file, and search for the keyword [Engineers]. After this line, search for the engineer that is named Administrator. (for example: Engineer1=Administrator).
Why is a power delivery test performed, even if Forward power levelling is not selected?[edit]
When RadiMation® is instructed to use the forward power meter it will always perform a power delivery test.
The following items that may cause a forward power measurement:
- Forward power amplifier protection was activated.
- The forward power or net power is measured as an input.
- Forward power or Net power calibration method is selected.
In all these cases, also a power delivery test is performed.
Why is the signal generator carrier level limited to 0 dBm?[edit]
By default the input of an amplifier is limited to 0 dBm. This is a level that most amplifier can handle without breaking. If a higher or lower amplifier input level is needed, the amplifier input protection should be configured to that specific level. More information regarding the configuration of amplifier power protection can be found in Chapter 14 of the RadiMation® end-user manual.
If you have found a problem in the RadiMation® software, it is best to report it to us. We will categorize the reported issue, register it in our defect database, and investigate the problem. When a workaround or solution is available, we will report this back to you.
