Difference between revisions of "FAQ/Immunity"
(→I get a peak on the oscilloscope after the amplitude modulation is turned off) |
Line 77: | Line 77: | ||
Please read the explanation blow. | Please read the explanation blow. | ||
The screen shot of the oscilloscope looks like: | The screen shot of the oscilloscope looks like: | ||
Revision as of 11:08, 12 June 2009
This page contains all the Frequently Asked Questions related to Radiated Immunity and Conducted Immunity module of RadiMation®
Contents
- 1 Does RadiMation support the linearity check that is described in the EN 61000-4-3 standard?
- 2 Why are the Uniform Field Area calculations of expired standards not removed?
- 3 Is RadiMation® using the correct method for the Uniform Field Area calibration?
- 4 Why is an immunity test trying to control a field sensor, if no field sensor is selected in the used testsite?
- 5 Signal generator is going to -100 dBm and RadiMation is reporting that it can not go any lower
- 6 Why is a power delivery test performed, even if Forward power levelling is not selected?
- 7 Why is the signal generator carrier level limited to 0 dBm?
- 8 Can I get the average or calculated field strength on the Can bus?
- 9 I get a peak on the oscilloscope after the amplitude modulation is turned off
- 10 I have found a problem. What should I do?
- 11 I have another question, that is not answered yet
If the signal generator is going to -100 dBm during a test, then there is something wrong in the hardware configuration or the software configuration.
Possible causes and solutions:
- Verify that the hardware configuration is the same as the software configuration. It is possible that an attenuation is not included in the test-site configuration. It is also possible that a coupler with another attenuation was selected in the test-site configuration. Adjust the configuration of the test-site in such a way that the hardware configuration and the software configuration are equal.
- Check the noise level of the powermeter, so the minimum power level which can be measured with the used power meter is known. The problem can be caused by a combination of the noise level of the used powermeter and the attenuation in front of the powermeter, which is in total higher then the requested power level during the test.
When RadiMation® is instructed to use the forward power meter it will always perform a power delivery test.
The following items that may cause a forward power measurement:
- Forward power amplifier protection was activated.
- The forward power or net power is measured as an input.
- Forward power or Net power calibration method is selected.
In all these cases, also a power delivery test is performed.
By default the input of an amplifier is limited to 0 dBm. This is a level that most amplifier can handle without breaking. If a higher or lower amplifier input level is needed, the amplifier input protection should be configured to that specific level. More information regarding the configuration of amplifier power protection can be found in Chapter 14 of the RadiMation® end-user manual.
There is an interface between CANoe/CANalyser where you can find more information here.
When using the RadiMationInterface.can in the calpl interface you need to implement the function:
- Test_Calculated_Field, for the calculated field.
- Test_Average_Field, for the average field.
These functions will be called when the information is changing inside RadiMation®.
This effect has been reported before, the problem is caused by measurement errors with a D.S.O. Due to aliasing problems.
In order to understand the problem we put in a green line to determine where the modulation was turned off. At first sight it looks like the part on the left and the first part of the right are correct. The amplitude during modulation should be 1.8 times higher then the CW value with 80% depth.
First check if the amplitude is correct, in a 50 Ohm system this can be calculated very easily. If the amplitude values are not what is expect there may be an aliasing problem with the oscilloscope. There is a very easy set up to check if this is the problem.
The 50 Ohm load is to make sure that you are measuring in 50 Ohm. When the oscilloscope has an 50 Ohm input impedance select this impedance and connect directly to the signal generator.
Follow the steps and compair the values by hand:
- set the signal generator to 0 dBm, no modulation at 100 MHZ.
- measure the peak voltage.
- set the signal generator to 0 dBm, with 80% modulation at 100 MHz.
For the AM modulation the peak is 1.8 times higher so:
When the measured values are not what is expected, try increasing the number of samples on the oscilloscope, when is it no possible lower the carrier frequency until you do get the correct reading.
If you have found a problem in the RadiMation® software, it is best to report it to us (for example through this online form). We will categorize the reported issue, register it in our defect database, and investigate the problem. When a workaround or solution is available, we will report this back to you.
Just submit your question on RadiMation® forum or the online web-form. And we will give you an answer.