FAQ/Immunity: Difference between revisions
Line 107: Line 107:
When the measured values are not what is expected, try increasing the number of samples on the oscilloscope, when is it no possible lower the carrier frequency until you do get the correct reading.
===The definition of measured field, measured voltage and measured current===
The definition of measured field, measured voltage and measured current changed. From version 4.3.9 and upward, the specified inputs are always displayed as corresponding non-modulated test-levels. The measured values are corrected for the applied modulation and the possible applied modulation conservation. In previous versions these corrections were not applied.
An example: in a radiated immunity substitution test with no modulation with a test-level of 10 V/m, the measured field strength will be 10 V/m, and this value will also be displayed.
If the same test is performed with 80% AM Modulation, a field strength of 18 V/m will be measured. In this case the software will however display 10 V/m, because that is the requested test-level. The displayed values are still measured values, but the measured values are corrected for the applied modulation.
As a result of these changes, in an optimal situation, the values of calculated field, measured field and specified test-level should be the same.
The power measurements (signal, forward, reflected and net power) are not corrected for the applied modulation and conservation.
The corrections on the measured values are depending on the ability of the used measuring device to correctly measure a value when modulation is applied. If for example a field strength sensor makes an error during the measurement of the electrical field when modulation is also applied, the corrected and displayed value of the measured field strength will also not be accurate. This situation only applies to situations where a modulated signal is measured, and that is only possible when the 'fast constant' frequency change mode is used.
{{RadiMationFAQ_AnotherQuestion}}
Revision as of 12:25, 15 April 2010
RadiMation® FAQ: General | Immunity | Emission | Report Generator
This page contains all the Frequently Asked Questions related to Radiated Immunity and Conducted Immunity module of RadiMation®
  1. Configure a Radiated Immunity substitution test with the calibration file that was the result of the Uniform Field Area calculation. Ensure that the substitution test is configured with the correct settings:
    • use the signal generator calibration method.
    • the forward power input should be activated, because the forward power should be measured.
    • level to an electrical field strength that is exactly 1.8 times lower than the field strength that was calibrated. This will result in a 5.1 dB lower signal generator power.
    • use the same frequency range and frequency steps, as was used during the calibration.
    • use the same testsite as was used during the calibration.
  2. Start the substitution test and wait until it is finished.
  3. Now the forward power that is used during the substitution test, and the forward power during the calibration should be compared. It is the most easiest to do this in Microsoft Excel. Start Microsoft Excel, and create an empty sheet.
  4. Open the calibration file that was the result of the Uniform Field Area calculation.
  5. Open the forward power graph, and copy the data (table) of the forward power to the clipboard.
  6. Paste the forward power of the calibration file into the Microsoft Excel worksheet (for example on cell A6)
  7. Open the results of the substitution test, open the forward power graph, and copy the data (table) of the forward power to the clipboard.
  8. Paste the forward power of the substitution test into the Microsoft Excel worksheet (for example on cell D6)
  9. Compare the data of the forward power of the calibration and the forward power during the substitution. The forward power of the substitution should be 5.1 dB lower then the forward power of the calibration file. The comparison is very easy, by inserting the formula '=B6-E6' in cell H6. The formula can then be copied down for all other frequencies.
  10. The EN 61000-4-3 standard assumes that the amplifier is not in saturation if the difference between the forward power values on all frequencies is within +3.1 dB to +5.1 dB.
Why are the Uniform Field Area calculations of expired standards not removed?[edit]
  1. A number of product standards do mention basic standards and there year of issue, for the way of performing radiated immunity tests. As long as the product standards are not changed one should use the 'old' basic standard as well. Although many lab's do use the latest version of the basic standard for all product standards (while it is much more convenient to use only one version of the EN61000-4-3), when there is doubt about the test results, one should use the harmonized standard and it's associated basic standard as mentioned in the product standard.
  2. A number of customers simply require the old methods, for the reasons mentioned above, or while they want to be able to perform verification tests on old products
Is RadiMation® using the correct method for the Uniform Field Area calibration?[edit]
Signal generator is going to -100 dBm and RadiMation is reporting that it can not go any lower[edit]
If the signal generator is going to -100 dBm during a test, then there is something wrong in the hardware configuration or the software configuration.
Possible causes and solutions:
  1. Verify that the hardware configuration is the same as the software configuration. It is possible that an attenuation is not included in the test-site configuration. It is also possible that a coupler with another attenuation was selected in the test-site configuration. Adjust the configuration of the test-site in such a way that the hardware configuration and the software configuration are equal.
  2. Check the noise level of the powermeter, so the minimum power level which can be measured with the used power meter is known. The problem can be caused by a combination of the noise level of the used powermeter and the attenuation in front of the powermeter, which is in total higher then the requested power level during the test.
Why is a power delivery test performed, even if Forward power levelling is not selected?[edit]
When RadiMation® is instructed to use the forward power meter it will always perform a power delivery test.
The following items that may cause a forward power measurement:
  • Forward power amplifier protection was activated.
  • The forward power or net power is measured as an input.
  • Forward power or Net power calibration method is selected.
In all these cases, also a power delivery test is performed.
Why is the signal generator carrier level limited to 0 dBm?[edit]
By default the input of an amplifier is limited to 0 dBm. This is a level that most amplifier can handle without breaking. If a higher or lower amplifier input level is needed, the amplifier input protection should be configured to that specific level. More information regarding the configuration of amplifier power protection can be found in Chapter 14 of the RadiMation® end-user manual.
Can I get the average or calculated field strength on the Can bus?[edit]
There is an interface between CANoe/CANalyser where you can find more information here.
When using the RadiMationInterface.can in the calpl interface you need to implement the function:
  • Test_Calculated_Field, for the calculated field.
  • Test_Average_Field, for the average field.
These functions will be called when the information is changing inside RadiMation®.
I get a peak on the oscilloscope after the amplitude modulation is turned off[edit]
This effect has been reported before, the problem is caused by measurement errors with a D.S.O. Due to aliasing problems.
Please read the explanation below.
The screen shot of the oscilloscope looks like:
In order to understand the problem we put in a green line to determine where the modulation was turned off. At first sight it looks like the part on the left and the first part of the right are correct. The amplitude during modulation should be 1.8 times higher then the CW value with 80% depth.
First check if the amplitude is correct, in a 50 Ohm system this can be calculated very easily. If the amplitude values are not what is expect there may be an aliasing problem with the oscilloscope. There is a very easy set up to check if this is the problem.
The 50 Ohm load is to make sure that you are measuring in 50 Ohm. When the oscilloscope has an 50 Ohm input impedance select this impedance and connect directly to the signal generator.
Follow the steps and compair the values by hand:
  1. set the signal generator to 0 dBm, no modulation at 100 MHZ.
  2. measure the peak voltage.
  3. set the signal generator to 0 dBm, with 80% modulation at 100 MHz.
  4. measure the peak voltage.
For the AM modulation the peak is 1.8 times higher so:
The definition of measured field, measured voltage and measured current[edit]
The definition of measured field, measured voltage and measured current changed. From version 4.3.9 and upward, the specified inputs are always displayed as corresponding non-modulated test-levels. The measured values are corrected for the applied modulation and the possible applied modulation conservation. In previous versions these corrections were not applied. An example: in a radiated immunity substitution test with no modulation with a test-level of 10 V/m, the measured field strength will be 10 V/m, and this value will also be displayed. If the same test is performed with 80% AM Modulation, a field strength of 18 V/m will be measured. In this case the software will however display 10 V/m, because that is the requested test-level. The displayed values are still measured values, but the measured values are corrected for the applied modulation. As a result of these changes, in an optimal situation, the values of calculated field, measured field and specified test-level should be the same. The power measurements (signal, forward, reflected and net power) are not corrected for the applied modulation and conservation. The corrections on the measured values are depending on the ability of the used measuring device to correctly measure a value when modulation is applied. If for example a field strength sensor makes an error during the measurement of the electrical field when modulation is also applied, the corrected and displayed value of the measured field strength will also not be accurate. This situation only applies to situations where a modulated signal is measured, and that is only possible when the 'fast constant' frequency change mode is used.
