Line 3: | Line 3: | ||
− | + | {{RadiMation}} consists of 7 separate software modules to perform 5 different types of EMC tests, calibrations, automatic report generation and [[sequence]] testing. | |
− | # Radiated immunity testing (RI) | + | # [[Radiated immunity]] testing (RI) |
− | # Conducted immunity testing (CI) | + | # [[Conducted immunity]] testing (CI) |
− | # Pulsed immunity (ESD, EFT, Surge) | + | # [[Pulsed immunity]] (ESD, EFT, Surge) |
− | # Radiated emission testing (RE) | + | # [[Radiated emission]] testing (RE) |
− | # Conducted emission testing (CE) | + | # [[Conducted emission]] testing (CE) |
− | # Calibration | + | # [[Calibration]] |
− | # Report generator | + | # [[Report generator]] |
The combination of the different software modules in one software package has a number of advantages compared to other software packages: | The combination of the different software modules in one software package has a number of advantages compared to other software packages: | ||
Line 21: | Line 21: | ||
− | The figure below represents an overview of the | + | The figure below represents an overview of the {{RadiMation}} software: |
[[Image:Modular overview.png]] | [[Image:Modular overview.png]] | ||
− | The | + | The {{RadiMation}} Core consists of the following components: |
* The Visual and Graph System | * The Visual and Graph System | ||
* The EUT data management System | * The EUT data management System | ||
Line 34: | Line 34: | ||
* The Device Driver System | * The Device Driver System | ||
− | Depending on the requirement, one ore more modules can be added to the | + | Depending on the requirement, one ore more modules can be added to the {{RadiMation}} Core. |
− | Version {{LatestMajorMinorVersion}} of the | + | Version {{LatestMajorMinorVersion}} of the {{RadiMation}} package contains the following modules: |
# Radiated immunity | # Radiated immunity | ||
# Conducted immunity | # Conducted immunity | ||
Line 46: | Line 46: | ||
− | The | + | The {{RadiMation}} software will operate (and is tested) with the Microsoft Windows 2000 and Windows XP operating systems. |
=== Hardware requirements === | === Hardware requirements === | ||
− | The | + | The {{RadiMation}} software extensively makes use of data calculations and screen interactions (for the use of graphs). For this reason, processor speed is quite important. To use the software to its full extent, we require the following hardware configuration: |
* Processor type: 2 GHz minimum, > 3 GHz recommended | * Processor type: 2 GHz minimum, > 3 GHz recommended | ||
* Free RAM memory: 256 MB minimal, > 512 MB recommended | * Free RAM memory: 256 MB minimal, > 512 MB recommended | ||
Line 58: | Line 58: | ||
== Conventions == | == Conventions == | ||
− | + | {{RadiMation}} makes use of the Windows structure of pull down menu’s. In this manual, the pull down menu’s are described in the following way: | |
{{Menu|File|Open|EUT}} | {{Menu|File|Open|EUT}} | ||
Line 67: | Line 67: | ||
− | To install the | + | To install the {{RadiMation}} software please insert the {{RadiMation}} CD in the CD ROM drive of your computer. The setup program of {{RadiMation}} should now start automatically. If the program doesn't start automatically, run E:\setup.exe (where E:\ is the CD-ROM drive) from the CD. The Installation Wizard will guide you through the installation process. After installation, the computer must be restarted before running {{RadiMation}}. Please see User rights and installation for rights needed for first time installation. |
− | In a network environment multiple computers are connected with each other with a shared communication medium. This had advantages because data can be shared between the connected computers. The | + | In a network environment multiple computers are connected with each other with a shared communication medium. This had advantages because data can be shared between the connected computers. The {{RadiMation}} software is designed to optimally use this advantage. All data that will be used or generated by {{RadiMation}} can be shared on the network. |
− | To optimally use this sharing of data the configuration of | + | To optimally use this sharing of data the configuration of {{RadiMation}} should be slightly changed. Use the following steps to do this. |
# Create a shared network directory on one of the Network Servers. Your Network Administrator can help you with this. This directory will be called <NetworkDir> from now on. | # Create a shared network directory on one of the Network Servers. Your Network Administrator can help you with this. This directory will be called <NetworkDir> from now on. | ||
# Create a sub-directory called Setup in the <NetworkDir> | # Create a sub-directory called Setup in the <NetworkDir> | ||
− | # Place the | + | # Place the {{RadiMation}} setup files from the {{RadiMation}} CD (SETUP.EXE and all SETUP.W?? files) in the Setup sub-directory. |
− | # Install | + | # Install {{RadiMation}} from the Setup directory on one of the connected PC's. This can be done by starting the SETUP.EXE program. Use the installation procedure as described in the previous section. |
# In the <NetworkDir>, create the following sub-directories: | # In the <NetworkDir>, create the following sub-directories: | ||
#* EUT_File | #* EUT_File | ||
Line 89: | Line 89: | ||
− | # Start the installed | + | # Start the installed {{RadiMation}} program, and open the Configuration window. On the {{ScreenElement|Directories}} tab, change the location of the {{ScreenElement|Device Driver Configuration}} directory to the CONFDVDR directory in the <NetworkDir>. |
− | # Restart the RadiMation program | + | # Restart the {{RadiMation}} program |
# Open the Configuration window again and on the {{ScreenElement|Directories} tab, change the location of all the directories to the corresponding directories in the <NetworkDir>. | # Open the Configuration window again and on the {{ScreenElement|Directories} tab, change the location of all the directories to the corresponding directories in the <NetworkDir>. | ||
− | # Restart the RadiMation program again. | + | # Restart the {{RadiMation}} program again. |
− | # From now on the | + | # From now on the {{RadiMation}} program will use the <NetworkDir> as its main data directory. |
− | # Start the Configuration of the | + | # Start the Configuration of the {{RadiMation}} software as described in the remaining part of this chapter. The complete configuration will be written to the <NetworkDir>. |
− | When the | + | When the {{RadiMation}} software should be installed on another PC in the Network, the following steps can be used. |
− | # Install the | + | # Install the {{RadiMation}} software from the Setup sub-directory in the <NetworkDir>. Use the installation procedure as described in the previous section. |
− | # Restart the | + | # Restart the {{RadiMation}} program |
− | The newly installed | + | The newly installed {{RadiMation}} will now use the already configured <NetworkDir>. |
− | Software updates are distributed on a CD-ROM. To install the new | + | Software updates are distributed on a CD-ROM. To install the new {{RadiMation}} version, please insert the CD-ROM in the CD-ROM drive of the computer. Run the installation program (i.e. E:\setup.exe, where E:\ is the CD-ROM drive). The setup program will detect that a version of {{RadiMation}} is already installed on your PC. The setup program will install the new {{RadiMation}} version at the same location as the already installed version. A back up of the previously installed version of {{RadiMation}} will be copied to the backup directory automatically. All the settings of the previous version will also be used for the new installed {{RadiMation}} version. When problems do occur with the newly installed {{RadiMation}} version, the old version of {{RadiMation}} can be recovered. |
{{Note|Before installing the update it is advised that you backup your TSF/Sequence/Correction/Calibration and EUT files.}} | {{Note|Before installing the update it is advised that you backup your TSF/Sequence/Correction/Calibration and EUT files.}} | ||
− | {{Warning|'''Do not uninstall | + | {{Warning|'''Do not uninstall {{RadiMation}}!!!''' This will cause removal of all TSF, sequence, calibration and correction files. This is unrecoverable.}} |
− | If the already installed | + | If the already installed {{RadiMation}} version was installed from a network directory, the new setup files can be copied from the CD-ROM to the Setup sub-directory of the <NetworkDir>. On each PC the new {{RadiMation}} version can then be installed from the Setup sub-directory |
− | This chapter describes what rights a user must have in order to perform an installation or an upgrade of the RadiMation software. | + | This chapter describes what rights a user must have in order to perform an installation or an upgrade of the {{RadiMation}} software. |
=== Upgrade === | === Upgrade === | ||
− | For an upgrade the user must have the rights to read and write in the | + | For an upgrade the user must have the rights to read and write in the {{RadiMation}} folder in the program files directory. And must have rights to add and check items to the HKEY_CLASSES_ROOT tree in the registry as well as the Dare Development folder in HKEY_LOCAL_MACHINE --> SOFTWARE tree please note that upgrading like described here can only be performed with {{RadiMation}} version greater then version 4.3.20. And upgrading from a version earlier or equal to version 4.3.20 can only be performed by an administrator. |
Line 152: | Line 152: | ||
When a new EUT file is made the name and address information of the producer, test house and client can be entered separately. | When a new EUT file is made the name and address information of the producer, test house and client can be entered separately. | ||
In many situations one or more of these will be the same for every EUT (i.e.: when a test house owns the software, the test house data will always be the same). | In many situations one or more of these will be the same for every EUT (i.e.: when a test house owns the software, the test house data will always be the same). | ||
− | In order to prevent retyping of often-used information | + | In order to prevent retyping of often-used information {{RadiMation}} allows entering the default address information for: |
Line 167: | Line 167: | ||
− | + | {{RadiMation}} allows the user to define the default units. These units are used throughout the package although they can be changed at any time during measurements and viewing. | |
The units for the following parameters can be defined: | The units for the following parameters can be defined: | ||
Line 185: | Line 185: | ||
− | The | + | The {{RadiMation}} software can use different directories to store different data file types. The location of the following file types can be specified: |
Line 195: | Line 195: | ||
− | Under | + | Under “{{RadiMation}} file types and locations” the different file types are described. |
For each data type the default path name can be entered through the configuration, directories pull down menu. | For each data type the default path name can be entered through the configuration, directories pull down menu. | ||
Line 225: | Line 225: | ||
Device drivers are used to let your software know which equipment has to be used and controlled during EMC tests. | Device drivers are used to let your software know which equipment has to be used and controlled during EMC tests. | ||
− | In | + | In {{RadiMation}}, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information like the IEEE address or com-port has to be entered. |
Configuration of device drivers can be carried out by selecting: | Configuration of device drivers can be carried out by selecting: | ||
Line 232: | Line 232: | ||
[[Image:Configuration DeviceDrivers.png]] | [[Image:Configuration DeviceDrivers.png]] | ||
− | {{ScreenElementDescription|Device Driver Type|Different [[Device Types|types]] of devices are available in RadiMation. Only the created device drivers of the type that is selected for {{ScreenElement|Device Driver Type}} will be shown in the list}} | + | {{ScreenElementDescription|Device Driver Type|Different [[Device Types|types]] of devices are available in {{RadiMation}}. Only the created device drivers of the type that is selected for {{ScreenElement|Device Driver Type}} will be shown in the list}} |
{{ScreenElementDescription|Available Device Drivers|Shows a list of all the available device drivers for the selected {{ScreenElement|Device Driver Type}}}} | {{ScreenElementDescription|Available Device Drivers|Shows a list of all the available device drivers for the selected {{ScreenElement|Device Driver Type}}}} | ||
− | {{ScreenElementDescription|New|Creates a new [[RadiMation]] device driver for a specific device. It is only possible to create a new device driver for the selected {{ScreenElement|Device Driver Type}}.}} | + | {{ScreenElementDescription|New|Creates a new [[RadiMation|{{RadiMation}}]] device driver for a specific device. It is only possible to create a new device driver for the selected {{ScreenElement|Device Driver Type}}.}} |
{{ScreenElementDescription|Edit|Changes the configuration of the selected device driver from the list of {{ScreenElement|Availalbe Device Drivers}}}} | {{ScreenElementDescription|Edit|Changes the configuration of the selected device driver from the list of {{ScreenElement|Availalbe Device Drivers}}}} | ||
{{ScreenElementDescription|Remove|Removes the configuration of the selected device driver from the list of {{ScreenElement|Availalbe Device Drivers}}}} | {{ScreenElementDescription|Remove|Removes the configuration of the selected device driver from the list of {{ScreenElement|Availalbe Device Drivers}}}} | ||
Line 251: | Line 251: | ||
− | + | {{RadiMation}} allows you to customise the way graphics are displayed on your screen. | |
To configure the looks of the shown graphs, please select: | To configure the looks of the shown graphs, please select: | ||
Line 277: | Line 277: | ||
{{ScreenElementDescription|Scaling|Determines if the initial x-axis scaling of the graph should be shown linear or logarithmic}} | {{ScreenElementDescription|Scaling|Determines if the initial x-axis scaling of the graph should be shown linear or logarithmic}} | ||
− | The selected settings in this {{WindowTitle|Graph Customization Configuration}} window are used as the [[RadiMation]] default settings. It is still possible to overrule these settings in each test to more specific settings. | + | The selected settings in this {{WindowTitle|Graph Customization Configuration}} window are used as the [[RadiMation|{{RadiMation}}]] default settings. It is still possible to overrule these settings in each test to more specific settings. |
Line 283: | Line 283: | ||
− | + | {{RadiMation}} allows interfacing to different database structures. These interfaces can be used to for example have easy access to the customer database that is already in use in your company. This will reduce the need for retyping customer information in {{RadiMation}}. | |
A number of drivers are available for the most commonly used databases. | A number of drivers are available for the most commonly used databases. | ||
==== Changing the customer database driver ==== | ==== Changing the customer database driver ==== | ||
− | To configure | + | To configure {{RadiMation}} to support your customer database, please select |
{{Menu|Configuration|Configuration}} | {{Menu|Configuration|Configuration}} | ||
Line 304: | Line 304: | ||
{{ScreenElementDescription|Location|In this field, the location of the database must be entered. It is also possible to press on the small button behind the textbox, after which a popup window will be shown that allows the end-user to select the correct customer database.}} | {{ScreenElementDescription|Location|In this field, the location of the database must be entered. It is also possible to press on the small button behind the textbox, after which a popup window will be shown that allows the end-user to select the correct customer database.}} | ||
{{ScreenElementDescription|Table and Fields| | {{ScreenElementDescription|Table and Fields| | ||
− | It is necessary to specify in the customer database configuration, which table and column(s) should be used, to correctly map the information in the customer database to the data in RadiMation. There for the following selections must be made. It is not necessary to specify a selection for each and every field, but the {{ScreenElement|Table}} and {{ScreenElement|Company}} has to be specified. | + | It is necessary to specify in the customer database configuration, which table and column(s) should be used, to correctly map the information in the customer database to the data in {{RadiMation}}. There for the following selections must be made. It is not necessary to specify a selection for each and every field, but the {{ScreenElement|Table}} and {{ScreenElement|Company}} has to be specified. |
{{ScreenElementDescription|Table|The table that should be used to retrieve the information from}} | {{ScreenElementDescription|Table|The table that should be used to retrieve the information from}} | ||
{{ScreenElementDescription|Company|The database column name from which the 'company' name should be retrieved.}} | {{ScreenElementDescription|Company|The database column name from which the 'company' name should be retrieved.}} | ||
Line 327: | Line 327: | ||
− | RadiMation supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require. | + | {{RadiMation}} supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require. |
− | The default language of | + | The default language of {{RadiMation}} is English. If you want to change the language it must be selected from the list of available languages. To configure the language, please select: |
Line 336: | Line 336: | ||
− | The currently selected language is marked with the text ' '''(current)''' ' appended. If another language should be used in RadiMation, one of the other languages can just easily be selected. The selected language will only be active after the | + | The currently selected language is marked with the text ' '''(current)''' ' appended. If another language should be used in {{RadiMation}}, one of the other languages can just easily be selected. The selected language will only be active after the {{RadiMation}} package is restarted. |
Line 355: | Line 355: | ||
− | To add a Product Standard to | + | To add a Product Standard to {{RadiMation}}, please select |
Line 366: | Line 366: | ||
− | Please note that the {{ScreenElement|Official name}} can only be inserted the first time and has to be unique. All fields speak for themselves, except for the {{ScreenElement|File location}}. In the {{ScreenElement|File location}} field, it is possible to associate a file to this standard. For example a PDF, DOC or HTML file that describes the standard. RadiMation will open the file, with the default reader for this file type, when the engineer selects the magnifying glass. | + | Please note that the {{ScreenElement|Official name}} can only be inserted the first time and has to be unique. All fields speak for themselves, except for the {{ScreenElement|File location}}. In the {{ScreenElement|File location}} field, it is possible to associate a file to this standard. For example a PDF, DOC or HTML file that describes the standard. {{RadiMation}} will open the file, with the default reader for this file type, when the engineer selects the magnifying glass. |
Line 375: | Line 375: | ||
It is possible to select multiple items by selecting the top item, then hold the 'shift' key and selecting the lowest items. Holding the 'Ctrl' key allows you to select or deselect one or more items. Double clicking on an item will display the item details in a new window. | It is possible to select multiple items by selecting the top item, then hold the 'shift' key and selecting the lowest items. Holding the 'Ctrl' key allows you to select or deselect one or more items. Double clicking on an item will display the item details in a new window. | ||
− | {{Note|When on the EUT window, a product standard is selected, or added, the data of the standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that if the configuration of a standard in the Configuration window of RadiMation is changed, the data in the EUT file will still have the old data of the standard!!! This behaviour is correct, because it is the information of the standard that is used at the time that the EUT was tested.}} | + | {{Note|When on the EUT window, a product standard is selected, or added, the data of the standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that if the configuration of a standard in the Configuration window of {{RadiMation}} is changed, the data in the EUT file will still have the old data of the standard!!! This behaviour is correct, because it is the information of the standard that is used at the time that the EUT was tested.}} |
Line 388: | Line 388: | ||
− | The test equipment configuration contains a list of all test equipment, used during a test. From this list, the software knows which device drivers to use to control the test equipment. | + | The test equipment configuration contains a list of all test equipment, used during a test. From this list, the software knows which device drivers to use to control the test equipment. {{RadiMation}} allows you to configure different test equipment lists for each type of test or even different sets of test equipment for the same kind of tests. (This will be useful if you have, for example, more than one signal generator to perform conducted immunity tests). |
In the test configuration window, one can select which set of test equipment will be used. | In the test configuration window, one can select which set of test equipment will be used. | ||
During the generation of a calibration file, the software records the test equipment configuration together with the calibration data. When a substitution test is started, the software checks the currently selected equipment (from the selected test equipment list) with the test equipment used during calibration. If the equipment does not match an error message will be displayed. | During the generation of a calibration file, the software records the test equipment configuration together with the calibration data. When a substitution test is started, the software checks the currently selected equipment (from the selected test equipment list) with the test equipment used during calibration. If the equipment does not match an error message will be displayed. | ||
Line 406: | Line 406: | ||
− | + | {{RadiMation}} allows you to protect a number of configurations through a password protection. | |
To activate, modify or de-activate passwords, select: | To activate, modify or de-activate passwords, select: | ||
Line 423: | Line 423: | ||
− | + | {{RadiMation}} allows the user to create its own device drivers for test equipment, which is (not yet) implemented in the standard device driver list. | |
User configurable device drivers are available for nearly all types of test equipment. However, device drivers for spectrum analysers and measurement receivers can not be made with user configurable device drivers because the complexity (and differences between suppliers) of these devices is too high. | User configurable device drivers are available for nearly all types of test equipment. However, device drivers for spectrum analysers and measurement receivers can not be made with user configurable device drivers because the complexity (and differences between suppliers) of these devices is too high. | ||
Line 444: | Line 444: | ||
The custom-made device driver can be used as any other device driver by selecting the driver in the Equipment list. | The custom-made device driver can be used as any other device driver by selecting the driver in the Equipment list. | ||
− | {{Note|Passive equipment such as antenna’s, current probes, cables etc. do also need a device driver. The reason for this is that | + | {{Note|Passive equipment such as antenna’s, current probes, cables etc. do also need a device driver. The reason for this is that {{RadiMation}} has to know a number of parameters of these devices. Among others the following information is relevant: |
Line 452: | Line 452: | ||
− | During the development of | + | During the development of {{RadiMation}} it became clear to us that not the EMC tests itself, but the Equipment Under Test (EUT) plays a central role for EMC automation software. |
− | Because of this, in contradiction to most other software packages, in the | + | Because of this, in contradiction to most other software packages, in the {{RadiMation}} software the EUT must be defined prior to any test. Defining an EUT consists of “Creating a new EUT file”, specifying EUT information, client information and producer information. |
After an EUT has been defined, a test window can be opened and a test can be started. Because of this approach, different types of tests can be performed from the same EUT window which all test data for the EUT will be started in the same EUT file. | After an EUT has been defined, a test window can be opened and a test can be started. Because of this approach, different types of tests can be performed from the same EUT window which all test data for the EUT will be started in the same EUT file. | ||
Line 489: | Line 489: | ||
=== Cable information of the EUT === | === Cable information of the EUT === | ||
− | RadiMation is capable of storing a lot of EUT related information. One of these items is the cables connected to the EUT. You can specify them by selecting the {{ScreenElement|EUT Information}} tab in the EUT window. On the bottom of the window you will find a control looking like the one below. | + | {{RadiMation}} is capable of storing a lot of EUT related information. One of these items is the cables connected to the EUT. You can specify them by selecting the {{ScreenElement|EUT Information}} tab in the EUT window. On the bottom of the window you will find a control looking like the one below. |
[[Image:Cables control.png]] | [[Image:Cables control.png]] | ||
Line 522: | Line 522: | ||
{{ScreenElementDescription|Logarithmic Interpolation|In between the specified frequency points, the correct value of the limit line will be calculated by using a logarithmic interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.}} | {{ScreenElementDescription|Logarithmic Interpolation|In between the specified frequency points, the correct value of the limit line will be calculated by using a logarithmic interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.}} | ||
{{ScreenElementDescription|Linear Interpolation|In between the specified frequency points, the correct value of the limit line will be calculated by using a linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.}} | {{ScreenElementDescription|Linear Interpolation|In between the specified frequency points, the correct value of the limit line will be calculated by using a linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.}} | ||
− | {{ScreenElementDescription|Detectors|During emission tests, RadiMation will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. In the {{ScreenElement|Detectors}} menu the operator can select for which detector(s) the limit line is valid.}} | + | {{ScreenElementDescription|Detectors|During emission tests, {{RadiMation}} will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. In the {{ScreenElement|Detectors}} menu the operator can select for which detector(s) the limit line is valid.}} |
Line 529: | Line 529: | ||
{{Menu|File|Save Limit line}} | {{Menu|File|Save Limit line}} | ||
− | RadiMation will prompt you for a filename for the new limit line file. This can be the same or a different filename as the limit line {{ScreenElement|Description}} field. | + | {{RadiMation}} will prompt you for a filename for the new limit line file. This can be the same or a different filename as the limit line {{ScreenElement|Description}} field. |
Line 552: | Line 552: | ||
{{Menu|File|Save correction}} | {{Menu|File|Save correction}} | ||
− | + | {{RadiMation}} will prompt you to enter a filename for the correction file. | |
== Using Test Setup Files (T.S.F.) == | == Using Test Setup Files (T.S.F.) == | ||
− | + | {{RadiMation}} is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes Radiated immunity tests [R.I.], Conducted immunity tests [C.I.] as well as Radiated emission tests [R.E.] and Conducted emission tests [C.E.]. (See also "software overview"). | |
For each kind of test, the test configuration data (start/stop frequency, test level, dwell time etc.) can be saved in a Test Configuration File (*.TSF). Separate from the test configuration data, the TSF file also stores the calibration file and test site information. The description of the TSF file can be 40 characters max. When the same kind of test must be performed, one can simply load the TSF file and start the test. | For each kind of test, the test configuration data (start/stop frequency, test level, dwell time etc.) can be saved in a Test Configuration File (*.TSF). Separate from the test configuration data, the TSF file also stores the calibration file and test site information. The description of the TSF file can be 40 characters max. When the same kind of test must be performed, one can simply load the TSF file and start the test. | ||
− | In order to organise the different TSF files | + | In order to organise the different TSF files {{RadiMation}} enables the user to create a directory tree thus enabling the user to store the TSF files in a structured manner. |
Line 566: | Line 566: | ||
− | + | {{RadiMation}} stores all test results for the corresponding EUT in the EUT file. To view test data, open the corresponding EUT file through the File, Open, EUT pull down menu. | |
The EUT window will show a list of all performed tests for the selected EUT. To view the extended data from a single test, select the test from the list and press the {{ScreenElement|view}} button. The following data will be displayed: | The EUT window will show a list of all performed tests for the selected EUT. To view the extended data from a single test, select the test from the list and press the {{ScreenElement|view}} button. The following data will be displayed: | ||
Line 595: | Line 595: | ||
* When you want to print the data of more than one test, select the tests to print while holding down the shift button. All selected tests will be printed when the {{ScreenElement|print}} button is pressed. | * When you want to print the data of more than one test, select the tests to print while holding down the shift button. All selected tests will be printed when the {{ScreenElement|print}} button is pressed. | ||
− | {{Warning|All print jobs from | + | {{Warning|All print jobs from {{RadiMation}} will be printed on the default Windows printer. To use another printer, please change the default printer in your Windows environment.}} |
== Exporting data using Microsoft Excel == | == Exporting data using Microsoft Excel == | ||
Line 613: | Line 613: | ||
{{ScreenElementDescription|Insert header above data|With insert header above data you can select, if you want, headers above every exported column, for example signal generator level, forward power and reflected power for immunity test. }} | {{ScreenElementDescription|Insert header above data|With insert header above data you can select, if you want, headers above every exported column, for example signal generator level, forward power and reflected power for immunity test. }} | ||
{{ScreenElementDescription|Export EUT data|Export EUT data allows you to select, if you want, the EUT data to be exported. Only the EUT information stored in the top test can be exported.}} | {{ScreenElementDescription|Export EUT data|Export EUT data allows you to select, if you want, the EUT data to be exported. Only the EUT information stored in the top test can be exported.}} | ||
− | {{ScreenElementDescription|Include not measured columns|Include not measured columns allows you to select, if you want, not measured columns to be exported. If you want to run a macro after exporting, it may be easier to export all the columns. You can then use the same macro for a signal level, forward power level and net power level test. It is not possible to have | + | {{ScreenElementDescription|Include not measured columns|Include not measured columns allows you to select, if you want, not measured columns to be exported. If you want to run a macro after exporting, it may be easier to export all the columns. You can then use the same macro for a signal level, forward power level and net power level test. It is not possible to have {{RadiMation}} run the macro automatically after the export. Deselecting exporting not measured columns will result in faster exporting of the information.}} |
{{Warning|Microsoft Excel needs to be installed on the PC to export the EUT information. The version of Microsoft Excel should be greater then or equal to version 'Microsoft Excel 97'.}} | {{Warning|Microsoft Excel needs to be installed on the PC to export the EUT information. The version of Microsoft Excel should be greater then or equal to version 'Microsoft Excel 97'.}} |
Revision as of 20:19, 25 January 2009
Getting started
Software overview
RadiMation® consists of 7 separate software modules to perform 5 different types of EMC tests, calibrations, automatic report generation and sequence testing.
- Calibration
- Report generator
- Easy to use software by implementation same "look & feel" over all modules.
- Seamless data exchange between the different software modules.
- Calibration functionality included in the package enables easy exchange of calibration files.
- Different types of tests can be performed, fully automatic after each other (Sequence testing).
The figure below represents an overview of the RadiMation® software:
- The Visual and Graph System
- Sequence Testing
- Calibration
- Data Logging
- Database Functionality
- The Device Driver System
Version 2023.2 of the RadiMation® package contains the following modules:
- Radiated immunity
- Pulsed immunity
- Radiated emission
- Report generator
The RadiMation® software will operate (and is tested) with the Microsoft Windows 2000 and Windows XP operating systems.
The RadiMation® software extensively makes use of data calculations and screen interactions (for the use of graphs). For this reason, processor speed is quite important. To use the software to its full extent, we require the following hardware configuration:
- Free RAM memory: 256 MB minimal, > 512 MB recommended
- Disk drives: CDROM drive
- Monitor: Super VGA (1024 x 768)
Conventions
RadiMation® makes use of the Windows structure of pull down menu’s. In this manual, the pull down menu’s are described in the following way:
The above example shows that first the File item from the menu bar should be selected, and after that the Open item, and the EUT item to show the file open selection window.
To install the RadiMation® software please insert the RadiMation® CD in the CD ROM drive of your computer. The setup program of RadiMation® should now start automatically. If the program doesn't start automatically, run E:\setup.exe (where E:\ is the CD-ROM drive) from the CD. The Installation Wizard will guide you through the installation process. After installation, the computer must be restarted before running RadiMation®. Please see User rights and installation for rights needed for first time installation.
In a network environment multiple computers are connected with each other with a shared communication medium. This had advantages because data can be shared between the connected computers. The RadiMation® software is designed to optimally use this advantage. All data that will be used or generated by RadiMation® can be shared on the network.
To optimally use this sharing of data the configuration of RadiMation® should be slightly changed. Use the following steps to do this.
- Create a shared network directory on one of the Network Servers. Your Network Administrator can help you with this. This directory will be called <NetworkDir> from now on.
- Create a sub-directory called Setup in the <NetworkDir>
- Place the RadiMation® setup files from the RadiMation® CD (SETUP.EXE and all SETUP.W?? files) in the Setup sub-directory.
- Install RadiMation® from the Setup directory on one of the connected PC's. This can be done by starting the SETUP.EXE program. Use the installation procedure as described in the previous section.
- EUT_File
- CAL_File
- COR_File
- CONFDVDR
- LLF_File
- RGT_File
- RGO_File
- Printout
- SEQ_File
- TSF_File
- Start the installed RadiMation® program, and open the Configuration window. On the Directories tab, change the location of the Device Driver Configuration directory to the CONFDVDR directory in the <NetworkDir>.
- Restart the RadiMation® program
- Open the Configuration window again and on the {{ScreenElement|Directories} tab, change the location of all the directories to the corresponding directories in the <NetworkDir>.
- From now on the RadiMation® program will use the <NetworkDir> as its main data directory.
When the RadiMation® software should be installed on another PC in the Network, the following steps can be used.
- Install the RadiMation® software from the Setup sub-directory in the <NetworkDir>. Use the installation procedure as described in the previous section.
Software updates are distributed on a CD-ROM. To install the new RadiMation® version, please insert the CD-ROM in the CD-ROM drive of the computer. Run the installation program (i.e. E:\setup.exe, where E:\ is the CD-ROM drive). The setup program will detect that a version of RadiMation® is already installed on your PC. The setup program will install the new RadiMation® version at the same location as the already installed version. A back up of the previously installed version of RadiMation® will be copied to the backup directory automatically. All the settings of the previous version will also be used for the new installed RadiMation® version. When problems do occur with the newly installed RadiMation® version, the old version of RadiMation® can be recovered.
If the already installed RadiMation® version was installed from a network directory, the new setup files can be copied from the CD-ROM to the Setup sub-directory of the <NetworkDir>. On each PC the new RadiMation® version can then be installed from the Setup sub-directory
This chapter describes what rights a user must have in order to perform an installation or an upgrade of the RadiMation® software.
Upgrade
For an upgrade the user must have the rights to read and write in the RadiMation® folder in the program files directory. And must have rights to add and check items to the HKEY_CLASSES_ROOT tree in the registry as well as the Dare Development folder in HKEY_LOCAL_MACHINE --> SOFTWARE tree please note that upgrading like described here can only be performed with RadiMation® version greater then version 4.3.20. And upgrading from a version earlier or equal to version 4.3.20 can only be performed by an administrator.
New installation
Software configuration
Password
To prevent unauthorised persons to change system configurations, a password protection can be used. The password is standard turned on with the default password Radimation. This password protection can be switched off, changed and configured under:
Test engineers
Engineers | Shows the currently configured list of engineer's. |
Add | The Add button can be used to add another engineer's name to the list of available engineers. After the Add button is pressed, the following window will be shown, and another engineer's name can be specified. |
Edit | The Edit button can be used to change the name of the selected engineer. |
Delete | The Delete button can be used to remove the name of the selected engineer from the list of available engineers. |
Ok | The Ok button will close the dialog. |
- The default client
- The default manufacturer
- The default test site
Data entered under default address information, will automatically be entered in the EUT window and therefore only needs to be entered once.
Default Units
- Power: dBuW, dBm, dBmW, dBW, uW, mW, W
- Field Strength: dBuV/m, dBmV/m, dBV/m, uV/m, mV/m, V/m
- Voltage: dBuVrms, dBmVrms, dBVrms, uVrms, mVrms, Vrms
- Current: dBuA, dBmA, dBA, uA, mA, A
- Emission: dBuV, dBmv, dBV, V, W, dBuV/m, dBmV/m, V/m
Change | Will show the Unit configuration dialog, for the selected usage of the unit, so more specific details of the used unit, and the formatting of the values can be configured. |
The path name can be either a local directory or a directory on a network drive. If a network drive is selected, it makes it possible to store for example EUT data central on a computer network.
To modify the path name select:
And then select the Directories tab.
A directory location can be changed by selecting the directory from the list and then pressing the Change button.
Configuring the Report Generator
The Report generator must be configured for the type of word-processor you are using. To configure the report generator please select:
Now select the Report Generator tab and select the correct word-processor that should be used to the generate the reports in.
Device driver configuration
In RadiMation®, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information like the IEEE address or com-port has to be entered.
Device Driver Type | Different types of devices are available in RadiMation®. Only the created device drivers of the type that is selected for Device Driver Type will be shown in the list |
Available Device Drivers | Shows a list of all the available device drivers for the selected Device Driver Type |
New | Creates a new RadiMation® device driver for a specific device. It is only possible to create a new device driver for the selected Device Driver Type. |
Edit | Changes the configuration of the selected device driver from the list of Availalbe Device Drivers |
Remove | Removes the configuration of the selected device driver from the list of Availalbe Device Drivers |
The required device driver can be selected from this list. The new device driver will then be created when New button is pressed. At that moment it is possible to use the default name of the device driver, or to change it to another (eg. more specific) name.
Virtual device drivers
Virtual device drivers do not have limitations like normal device drivers (frequency Band, max power, etc.)
Then select the Graphs tab.
An example graph is shown that is created with the currently selected graph settings. When the Customize button is pressed, the Graph Customization Configuration window is shown, which allows to change the look of all graphs.
Back Color | Shows the currently selected color for the back(ground) of the graph. The color can be changed by clicking on the color itself. |
Grid Color | Shows the currently selected color for the grid of the graph. The color can be changed by clicking on the color itself. |
Text Color | Shows the currently selected color for the text around the graph. The color can be changed by clicking on the color itself. |
Line Style | Shows the currently selected type of style that is used for the grid of the graph. Possible line styles are:
Scaling | Determines if the initial x-axis scaling of the graph should be shown linear or logarithmic |
The selected settings in this Graph Customization Configuration window are used as the RadiMation® default settings. It is still possible to overrule these settings in each test to more specific settings.
Introduction
RadiMation® allows interfacing to different database structures. These interfaces can be used to for example have easy access to the customer database that is already in use in your company. This will reduce the need for retyping customer information in RadiMation®. A number of drivers are available for the most commonly used databases.
To configure RadiMation® to support your customer database, please select
Then select the Database tab.
Database Type | Specifies the database driver that should be used to connect to the customer database |
Configuration | Shows a more detailed configuration dialog of the selected Database Type to allow the configuration of specific customer database settings. |
When the database Configuration button is pressed, the Customer database configuration window will appear.
Advanced | Shows some more advanced settings for the customer database configuration:
Language Selection
RadiMation® supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require.
Then select the Language tab:
The currently selected language is marked with the text ' (current) ' appended. If another language should be used in RadiMation®, one of the other languages can just easily be selected. The selected language will only be active after the RadiMation® package is restarted.
Product Standard
Introduction
- Washing machine
- Hair dryer
- Car radio
- Television
And then select the Product standards tab:
After pressing Add the end-user can insert all necessary information.
Please note that the Official name can only be inserted the first time and has to be unique. All fields speak for themselves, except for the File location. In the File location field, it is possible to associate a file to this standard. For example a PDF, DOC or HTML file that describes the standard. RadiMation® will open the file, with the default reader for this file type, when the engineer selects the magnifying glass.
Adding a basic standard to the product standard can be done simply by pressing the Add button. The following window will appear:
Basic Standard
The error message window will ask you to abort the test or to ignore the error message and continue the test with different test equipment. Of course, to achieve maximum accuracy, one should use the same test equipment during the calibration and during the substitution tests. However, in some circumstances, changing test equipment between calibration and test will not have any influence on test results. This is for example true when a RI substitution list is made using FWD power calibration file. Changing the amplifier won’t change test results as long as the new amplifier covers the required frequency range and can deliver the required output power.
Password protection
In the Change Password window, the password can be enabled, disabled and changed (to disable or change the password, the current password is required). Furthermore, in the password window, for the following items, the protection can be enabled or disabled:
Equipment | prevents unauthorised changes to test equipment configuration. |
Engineer | prevents unauthorised changes to test engineers configuration. |
TSF Files | prevents unauthorised changes to Test Setup Files (TSF) files. |
Limit Line | prevents unauthorised changes to limit line files. |
Correction File | prevents unauthorised changes to correction files. |
To make your own device driver, open the configuration window:
and select the Device drivers tab.
- Press the New button.
- Select the driver called Configurable <XXX> (i.e. if you want to make a signal generator device driver you would select the Configurable signal generator device driver).
- Enter a description for the device driver (for example the type number of the generator) and press Ok.
- Press the Edit button.
EUT files
Opening an existing EUT file
The first column displays the testnumber. The second column displays the description of the test. The third column shows the notes you inserted. The three columns can be sorted one at a time. This can be done by pressing the desired column header. If you wish to sort on testnumbers, press the first column header. If you wish to sort by note press the third column header. Pressing a column header twice will reverse the sorting order. For example the first time you press the third column header the notes will be sorted ascending, the second time you press the header the notes will be sorted descending. This pertains to all the columns.
RadiMation® is capable of storing a lot of EUT related information. One of these items is the cables connected to the EUT. You can specify them by selecting the EUT Information tab in the EUT window. On the bottom of the window you will find a control looking like the one below.
You can specify every cable of the EUT, pressing Add or Edit will popup the following window:
Limit line files
Creating limit lines
The window below will be displayed:
The user can enter a file description to the limit line in the Description field.
Entering data points
To enter data point, press the Insert button. A new line will appear and you can enter frequency and level. When entering data points, the limit line will be drawn directly in the graph. This makes it easy to check if the limit line is entered correctly.
When a data point in incorrect, it can be removed by selecting the data point and pressing the Remove button.
Color | Clicking on the coloured box behind the Color selection box will change the line color. The selected line colour will be used during emission tests. |
Logarithmic Interpolation | In between the specified frequency points, the correct value of the limit line will be calculated by using a logarithmic interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. |
Linear Interpolation | In between the specified frequency points, the correct value of the limit line will be calculated by using a linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. |
Detectors | During emission tests, RadiMation® will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. In the Detectors menu the operator can select for which detector(s) the limit line is valid. |
RadiMation® will prompt you for a filename for the new limit line file. This can be the same or a different filename as the limit line Description field.
Correction files
Creating correction files
Entering data points | To enter data points, press the Add Row button. A new line will appear and you can enter frequency and level. |
Remove Row | When a data point is incorrect, it can be removed by selecting the data point and pressing the Remove Row button. |
Saving correction files
RadiMation® is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes Radiated immunity tests [R.I.], Conducted immunity tests [C.I.] as well as Radiated emission tests [R.E.] and Conducted emission tests [C.E.]. (See also "software overview").
In order to organise the different TSF files RadiMation® enables the user to create a directory tree thus enabling the user to store the TSF files in a structured manner.
A more sophisticated use of TSF files can be found during sequence testing. In a sequence, different TSF files (even from different types of tests) can be placed after each other, and started without the intervention of a test engineer (See also “Sequence testing”)
Viewing test data
RadiMation® stores all test results for the corresponding EUT in the EUT file. To view test data, open the corresponding EUT file through the File, Open, EUT pull down menu.
The EUT window will show a list of all performed tests for the selected EUT. To view the extended data from a single test, select the test from the list and press the view button. The following data will be displayed:
- Time and date of the test.
- Equipment used during the test (Including serial numbers).
- Test engineer.
- Test configuration data.
- All graphs of AD channels used Field graphs (If used).
- Current graph (If used).
- Notes.
- Signal generator level graph (Used for fault diagnoses only).
- Forward power level graph (If applicable).
- Reflected power level graph (If applicable).
- Net power level graph (If applicable).
- Ambient information:
- Temperature graph (If used).
- Pressure graph (If used).
- Relative humidity graph (If used).
- Calibration file used during the test (If applicable).
Printing test data
Test data can be printed in several different ways:
- When only a part of the test data needs to be printed, select the desired test, then select view and select the graph you want to print. The graph will be displayed on the screen. Above the graph you will find a print button. This print button will only print the current selected graph.
- From the list of performed tests, a single test can be selected. When all data of this test must be printed, simply press the print button at the right side of the list.
- When you want to print the data of more than one test, select the tests to print while holding down the shift button. All selected tests will be printed when the print button is pressed.
The EUT data can be exported using Microsoft Excel. Press the Export button on the Reports tab of the EUT window, and a new screen will open:
Selecting filename
To select the filename you want to use for saving the information press the select file button next to the red cross. A save window will appear allowing you to select the path and filename. After closing the save window the name will be displayed on the export window.
Selecting tests to be exported
You can select or deselect a test for export by checking or un-checking the checkbox in front of the name of the test. Pressing Select all will select all tests, pressing Unselect all will deselect all tests and pressing Invert selection will invert the selected. Invert means that all selected tests will be deselected and all deselect tests will be selected.
Changing the test order
The export module exports all the selected tests from the first test at the top to the last test at the bottom of the list. This order of the tests can be changed. Select the test you want to move, press the arrow up or down button and the test is moved one place up or one place down.
Export options
Insert header above data | With insert header above data you can select, if you want, headers above every exported column, for example signal generator level, forward power and reflected power for immunity test. |
Export EUT data | Export EUT data allows you to select, if you want, the EUT data to be exported. Only the EUT information stored in the top test can be exported. |
Include not measured columns | Include not measured columns allows you to select, if you want, not measured columns to be exported. If you want to run a macro after exporting, it may be easier to export all the columns. You can then use the same macro for a signal level, forward power level and net power level test. It is not possible to have RadiMation® run the macro automatically after the export. Deselecting exporting not measured columns will result in faster exporting of the information. |
Ambient information