m (Fix spelling)
m (→‎Powermeter: Fix grammar)
Line 61: Line 61:
'''Speed up(2):''' When using a spectrum analyser a small amount of samples may be sufficient to determine the right amount of power. It is up to the engineer to find the right accuracy vs. time setting.
'''Measuring:''' Powermeter who doesn't have a RMS detector, should not be used in Fast constant mode.
'''Measuring:''' Powermeters which do not have a RMS detector, should not be used in Fast constant mode.
Revision as of 15:13, 10 February 2025
Device drivers[edit]
Virtual device drivers[edit]
  • Usable frequency range
  • Maximum input power
Signal generator[edit]
Amplifier[edit]
Antenna[edit]
Coupler[edit]
Powermeter[edit]
Measuring: Powermeters which do not have a RMS detector, should not be used in Fast constant mode.
Field Sensor[edit]
Axis setting panel[edit]
The Axis configuration is a generic driver setting panel. There are several (old) field sensors which don’t support the read out of an isotropic value: instead they provide the measurements values of all three axes separately. Using the ‘Software isotropic’ setting in the device driver can then activate that the measurement data of all three axes are being interpret and calculate to the isotropic value and have this isotropic value being return as the measured field strength. When the field sensor already supports the retrieval of the isotropic measurement value, often this value is also returned when the software option is being selected in the device configuration: both isotropic as software isotropic will return the isotropic value of the field sensor. This is depending on the device driver and used equipment.
AD converter[edit]
Injection device[edit]
Calibration Jigs[edit]
Current Sensor[edit]
Pre Amplifiers[edit]
LISN[edit]
Turn Table[edit]
Antenna Tower[edit]
Absorbing Clamps[edit]
Clamp positioner[edit]
Cables[edit]
Switch matrix[edit]
EUT Controller[edit]
Communication settings[edit]
RS-232 Setting[edit]
The RS-232 Communication settings are used to specify the RS-232 communication parameters, for the communication with a measurement device over RS-232. A lot of modern measurement devices are connected by USB, however very often then a Virtual RS-232 COMPort (VCP) is generated over the USB connection. In that situation also the RS-232 communication settings should configured.
ScreenElementDescription.svg COM Port The COM port which RadiMation® has to use to be communicate with the measurement device.
ScreenElementDescription.svg Baudrate Allows to configure the baudrate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.
GPIB Setting[edit]
ScreenElementDescription.svg Primary Address The primary address specifies the address of the device, so if the device you want to configure is on address is on 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device.
ScreenElementDescription.svg Use advanced configuration Allows the end-user to enables the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient enough.
ScreenElementDescription.svg Secondary Address The second address is default 0, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.
ScreenElementDescription.svg GPIB Delay GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PC's (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and some times gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see of the problem as disappeared. Is the problem has disappear then your problem was timing, if not please contact your reseller and report the problems you are having.
ScreenElementDescription.svg Clear device during initialisation When checked, a low level GPIB command to reset the device will be send to the device during initialisation.
ScreenElementDescription.svg Readdress device When checked, a low level GPIB command to select the correct address will be transmitted every time a command is send to the device.
VISA Settings[edit]
The VISA Communication settings are used to specify the VISA communication parameters, for the communication with a measurement device. The VISA library is a higher level communication library that supports different kind of communication methods. RadiMation® doesn't provide a VISA library itself, and thus requires that a VISA library from another supplier like National Instruments or Keysight Technologies is installed. The VISA manager that is provided by that VISA library can be used to determine the correct VISA resource. The selected VISA configuration thus also has influence if a measurement device is controlled by GPIB, RS-232, LAN, USB-TMC, VXI11, network-socket or another communication method .
ScreenElementDescription.svg Alias Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. National Instruments MAX) that is installed on the PC.
ScreenElementDescription.svg LAN Allows to specify that a VXI11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.
  • TCPIP[board]::<IP-address>::INSTR: VXI11 or LXI ethernet communication with the device with '<IP-address>'.
  • TCPIP[board]::<IP-address>::<port-number>::SOCKET: Raw socket based ethernet communication with on the port '<port-number>' with the the device '<IP-address>'
USB Settings[edit]
The USB Communication settings are used to specify that an USB connection to a DARE!! Instruments measurement device is used. It is not possible to use this USB communication setting for devices that simulate a Virtual COMPort (VCP) over an USB connection. The RS232 Settings should be used for such a kind of measurement device.
ScreenElementDescription.svg Device Identifier The device identifier (which is an unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.
TCPIP Settings[edit]
The TCPIP Communication settings are used to specify that a socket based TCPIP connection to a measurement device should be used.
RadiMation® allows the user to create its own device drivers for test and measurement equipment, which is (not yet) implemented in the standard device driver list.
  2. Press the “New” button.
  7. Press the “Edit” button.
Generic settings[edit]
ScreenElementDescription.svg Start and stop frequency The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, RadiMation® can warn the test engineer when he/she want to use the device out of its valid frequency range.
ScreenElementDescription.svg Get ID In the Get ID window you need to specify the string that the software needs to send when it want to get the ID string of the device. For example *IDN? is commonly used Get ID string.
ScreenElementDescription.svg Returned ID In the Returned ID window you need to specify the string that the software will receive so that it know that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.
ScreenElementDescription.svg Set Frequency Set Frequency is the string that needs to send to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz.
Example: “FRQ__freq__MHZ”.
Example: “FRQ__freqHz__HZ”.
Example: “FRQ__freqkHz__KHZ”.
Example: “FRQ__freqMHz__MHZ”.
Example: “FRQ__freqGHz__GHZ”.
ScreenElementDescription.svg Set Carrier Set Carrier is the string that needs to send to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.
ScreenElementDescription.svg Carrier on Carrier on is the string that needs to send to set the signal generator carrier on.
Example: “OUTPUT ON”.
ScreenElementDescription.svg Carrier off Carrier off is the string that needs to send to set the signal generator carrier off.
Example: “OUTPUT OFF”.
ScreenElementDescription.svg AM on AM on is the string that needs to send to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.
ScreenElementDescription.svg AM off AM off is the string that needs to send to set the internal am modulation of the signal generator off.
ScreenElementDescription.svg External on External on is the string that needs to send to set the modulator to the external input.
ScreenElementDescription.svg External off External off is the string that needs to send to set the modulator to the internal input.
ScreenElementDescription.svg PM on PM on is the string that needs to send to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.
ScreenElementDescription.svg PM off PM off is the string that needs to send to set the internal PM modulation of the signal generator off.
ScreenElementDescription.svg Sine Wave Sine wave is the string that needs to be send when RadiMation® wants to set the wave form of the internal source to sine.
Example: “SOURCE2:FUNC SIN”
ScreenElementDescription.svg Square Wave Square wave is the string that needs to be send when RadiMation® wants to set the wave form of the internal source to square.
Example: “SOURCE2:FUNC SQU”
Configurable Amplifier[edit]
Main[edit]
Operation[edit]
Example[edit]
1 = RF ON
1 = BAND 2
1 = LINE ON
Configurable Antenna[edit]
Configurable Powermeter[edit]
ScreenElementDescription.svg Set frequency Set Frequency is the string that needs to send to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.
Example: “P1,U1”.
Example: “X1”
ScreenElementDescription.svg Result Format Result Format is the string that RadiMation® needs to use to decode the value from the string send by the powermeter. The return value is interpreted in dBm.
Example: “__result__”
Example: “O1”
ScreenElementDescription.svg Duration Duration is the time that RadiMation® waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable result.
Channels[edit]
Read Back examples[edit]
Communication[edit]
Configurable LISN[edit]
Configurable Turn Table[edit]
Configurable commands[edit]
Check device[edit]
Device is initialised[edit]
Device is deinitialised[edit]
Some examples are:
Communication[edit]
Identification[edit]
Antenna Tower[edit]
Antenna Tower Advanced[edit]
Currently unavailable.
Configurable Cable[edit]
sending the next command.
keyword inserted value
__calculatedvoltage__ The calculated current expressed in dBA.
Messages[edit]
Unknown Device Driver[edit]
Device not connected[edit]
How to Report an Error[edit]
When encountering a problem with the software you might would like to report it to the RadiMation support. The RadiMation error popups, contains detailed section. This can be expanded with details button on the error popup. In the expanded detailed error popup, a Report Error button is present. This allows Error Reporting to the RadiMation support within RadiMation. More information about Error Reporting can be found here: https://wiki.radimation.com/wiki/index.php/Error_Report
AD convertors[edit]
Type of communication: IEEE.
Channels 1:
Hewlett Packard 34401A[edit]
Channel 1
Type of measurement: AC Volt
Minimum value: 0Volt
Maximum value: 1 kVolt.
Channel 2:
Minimum value: 0 Amp.
Maximum value: 3 Amp.
Channel 3
Type of measurement: DC Volt
Channel 4:
Maximum value: 3 Amp. (rms)
Channel 5
Minimum value: 0 Ω
Maximum value: 100 MΩ
Channel 6
Minimum value: 0 Hz.
Maximum value: 300 kHz
Channel 7 and 8
Not used
Hewlett Packard 54600[edit]
Channel 1:
Type of measurement: V Max.
Maximum value: 1000 Volt.
Channel 2:
Type of measurement: V Min
Minimum value: 0 Volt.
Channel 3
Channel 4:
Channel 5
Minimum value: 0 Hz
Maximum value: 100 kHz
Channel 6
Type of measurement: Period
Minimum value: 0 ms
Maximum value: 10000 ms
Channel 7
Minimum value: 0 ms
Channel 8
Minimum value: 0 ms
Hewlett Packard 3562A[edit]
Type of communication: GPIB
Channels
Minimum value: 0 dB
Maximum value: 100 dB
Hewlett Packard 59313[edit]
Channel 1
Minimum value: 0
Maximum value: 2048
Channel 2:
Minimum value: 0
Maximum value: 2048
Channel 3
Minimum value: 0
Maximum value: 2048
Channel 4:
Minimum value: 0
Maximum value: 2048
Channel 5 to 8
Not used.
Channels
Marconi 2305[edit]
Channel 1
Minimum value: 0
Maximum value: 1000 MHz
Channel 2:
Minimum value: 0
Maximum value: 1000
Channel 3 to 8
Not used.
Fluke 45 AC Current[edit]
Channels
Minimum value: 0 mA
Maximum value: 10.000 mA
Fluke 45 DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 45 AC Voltage[edit]
Channels
Minimum value: 0 mV
Maximum value: 1.000.000 mV
Fluke 45 DC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 45 Frequency[edit]
Channels
Minimum value: 0 Hz
Maximum value: 1.000.000 Hz
Fluke 45 Resistance[edit]
Channels
Minimum value: 0 Ω
Maximum value: 100.000.000 Ω
Channel 1
Maximum value: 353.55 Volt.
Channel 2:
Channel 3
Channel 4:
Maximum value: 707.1 Volt
Channel 5
Type of measurement: RMS
Maximum value: 250 Volt
Channel 6
Maximum value: 200 MHz
Channel 7 and 8
Not used
Fluke 8840A AC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A AC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 8840A DC Voltage[edit]
Channels
Minimum value: 0 mV
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mV
Maximum value: 300.000 mV
Channels
Minimum value: 0 mV
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 8
Minimum value: 0
Maximum value: 16383
EIP 575[edit]
Channel 1
Minimum value: 0 Hz
Maximum value: 10 kHz
Channel 2:
Channel 3
Minimum value: 0 Hz
Maximum value: 1 MHz.
Channel 4:
Maximum value: 10 MHz.
Channel 5
Minimum value: 0 Hz
Maximum value: 100 MHz
Channel 6
Minimum value: 0 Hz
Maximum value: 1 GHz
Channel 7
Minimum value: 0 Hz
Maximum value: 10 GHz
Channel 8
Minimum value: 0 Hz
Maximum value: 100 GHz
Channel 1 to 8