Line 146: Line 146:
| [[wikipedia:High Speed LAN Instrument Protocol|HiSLIP]] || [[#VISA_Setting|VISA]] (with LAN sub option) ||
| [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option met ::SOCKET identifier)
| [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier)
| HTTP || [[#HTTP_Setting|HTTP]] ||
Line 162: Line 162:
| [[wikipedia:D-subminiature|Sub-D9]] / [[wikipedia:RS-232|RS-232]] || RS-232 || [[#RS-232_Setting|RS-232]] || [[#VISA_Setting|VISA]] (with RS-232 sub option)
|rowspan="2"| other software || [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option met ::SOCKET identifier)
|rowspan="2"| other software || [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier)
| VXI-11 || [[#VISA_Setting|VISA]] (with LAN sub option) ||
Line 212: Line 212:
{{ScreenElementDescription|COM Port|The COM port which {{RadiMation}} has to be used to communicate with the measurement device.}}
{{ScreenElementDescription|Baudrate|Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Baud rate|Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Data bits|Allows to configure the number of data bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Parity|Allows to configure the parity bit that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
Revision as of 11:15, 4 November 2025
Device drivers[edit]
Virtual device drivers[edit]
  • Usable frequency range
  • Maximum input power
Signal generator[edit]
Amplifier[edit]
Antenna[edit]
Coupler[edit]
Powermeter[edit]
Measuring: Powermeters which do not have a RMS detector, should not be used in Fast constant mode.
Field Sensor[edit]
Axis setting panel[edit]
AD converter[edit]
Injection device[edit]
Calibration Jigs[edit]
Current Sensor[edit]
Pre Amplifiers[edit]
LISN[edit]
Turn Table[edit]
Antenna Tower[edit]
Absorbing Clamps[edit]
Clamp positioner[edit]
Cables[edit]
Switch matrix[edit]
EUT Controller[edit]
Different test and measurement instrumentation devices, can be controlled by different electronic communication buses and connectors, with on top of that different communication protocols. Which communication stream setting should be used to control a measurement instruments, cannot be easily based alone on the connector that is used to connect the measurement instrument. The following list shows the possible communication interfaces and protocols that are possible with a specific connector. This table can help to identify the actual communication protocol that is used. The manual of the test and measurement instrument also often lists the communication protocol that should be used, with similar terms, as is shown in the table. Based on the actual communication protocol the best communication stream to use is suggested.
Connector Communication protocol Suggested communication stream Alternative communication stream
GPIB / IEEE-488 GPIB GPIB VISA (with GPIB sub option)
Ethernet / RJ45 VXI-11 VISA (with LAN sub option)
HiSLIP VISA (with LAN sub option)
Raw socket TCPIP VISA (with Visa-ID sub option with ::SOCKET identifier)
HTTP HTTP
UDP UDP
USB Virtual COM Port (VCP) (RS-232) RS-232 VISA (with RS-232 sub option)
USBTMC (Test and Measurement Class) VISA (with Visa-ID sub option)
Raditeq USB devices USB RS-232
Device specific USB should be implemented device specifically
Sub-D9 / RS-232 RS-232 RS-232 VISA (with RS-232 sub option)
other software Raw socket TCPIP VISA (with Visa-ID sub option with ::SOCKET identifier)
The details for the configuration of each possible communication stream is described in the next section.
Communication settings[edit]
GPIB Setting[edit]
ScreenElementDescription.svg Primary Address The primary address specifies the address of the device, so if the device you want to configure is on address 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device.
ScreenElementDescription.svg Use advanced configuration Allows the end-user to enable the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient.
ScreenElementDescription.svg Secondary Address The second address is 0 by default, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.
ScreenElementDescription.svg GPIB Delay GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PCs (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and sometimes gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see if the problem has disappeared. If the problem has disappeared then your problem was timing, if not please contact your reseller and report the problems you are having.
ScreenElementDescription.svg Clear device during initialisation When checked, a low level GPIB command to reset the device will be sent to the device during initialisation.
ScreenElementDescription.svg Readdress device When checked, a low level GPIB command to select the correct address will be transmitted every time a command is sent to the device.
HTTP Setting[edit]
ScreenElementDescription.svg Address Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. By default the socket port number is 80, but this can be changed to another port if desired.
Manual Setting[edit]
The manual communication stream can be used to show message boxes on the screen for the end-user, to manually perform some operation on a device. The commands that are specified in the device driver that is using the manual communication stream, are thus the messages that are shown to the end-user in a message box, for example with the instruction to change a setting on a test and measurement device.
No further configuration in the manual communication stream is needed.
RS-232 Setting[edit]
The RS-232 communication settings specify the serial port parameters when communicating with a measurement device over RS-232. Although many modern devices connect via USB, they often appear as a virtual COM port (VCP). In that case you still configure the RS-232 settings as below.
ScreenElementDescription.svg COM Port The COM port which RadiMation® has to be used to communicate with the measurement device.
ScreenElementDescription.svg Baud rate Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.
RSIB Setting[edit]
ScreenElementDescription.svg RSIB Address Allows to specify RSIB identifier of the device that should be controlled.
TCPIP Setting[edit]
UDP Setting[edit]
The UPD communication settings are used to specify that an UDP based communication over ethernet should be used to control the measurement.
USB Setting[edit]
ScreenElementDescription.svg Device Identifier The device identifier (which is a unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.
VISA Setting[edit]
ScreenElementDescription.svg Alias Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. National Instruments MAX or Keysight Technologies connection expert) that is installed on the PC.
ScreenElementDescription.svg LAN Allows to specify that a VXI-11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.
  • TCPIP[board]::<IP-address>::INSTR: VXI-11 or LXI ethernet communication with the device with '<IP-address>'.
  • TCPIP[board]::<IP-address>::<port-number>::SOCKET: Raw socket based ethernet communication on the port '<port-number>' with the the device '<IP-address>'
  2. Press the “New” button.
  5. The name of the device driver will be added in the available device drivers list.
  6. Select the new device driver from the available device drivers list.
  7. Press the “Edit” button.
Generic settings[edit]
ScreenElementDescription.svg Start and stop frequency The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, RadiMation® can warn the test engineer when they want to use the device out of its valid frequency range.
ScreenElementDescription.svg Get ID In the Get ID window you need to specify the string that the software needs to send when it wants to get the ID string of the device. For example *IDN? is commonly used Get ID string.
ScreenElementDescription.svg Returned ID In the Returned ID window you need to specify the string that the software will receive so that it knows that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.
ScreenElementDescription.svg Set Frequency Set Frequency is the string that needs to be sent to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz.
Example: “FRQ__freq__MHZ”.
Example: “FRQ__freqHz__HZ”.
Example: “FRQ__freqkHz__KHZ”.
Example: “FRQ__freqMHz__MHZ”.
Example: “FRQ__freqGHz__GHZ”.
ScreenElementDescription.svg Set Carrier Set Carrier is the string that needs to be sent to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.
ScreenElementDescription.svg Carrier on Carrier on is the string that needs to be sent to set the signal generator carrier on.
Example: “OUTPUT ON”.
ScreenElementDescription.svg Carrier off Carrier off is the string that needs to be sent to set the signal generator carrier off.
Example: “OUTPUT OFF”.
ScreenElementDescription.svg AM on AM on is the string that needs to be sent to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.
ScreenElementDescription.svg AM off AM off is the string that needs to be sent to set the internal am modulation of the signal generator off.
ScreenElementDescription.svg External on External on is the string that needs to be sent to set the modulator to the external input.
ScreenElementDescription.svg External off External off is the string that needs to be sent to set the modulator to the internal input.
ScreenElementDescription.svg PM on PM on is the string that needs to be sent to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.
ScreenElementDescription.svg PM off PM off is the string that needs to be sent to set the internal PM modulation of the signal generator off.
ScreenElementDescription.svg Sine Wave Sine wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to sine.
Example: “SOURCE2:FUNC SIN”
ScreenElementDescription.svg Square Wave Square wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to square.
Example: “SOURCE2:FUNC SQU”
Configurable Amplifier[edit]
Main[edit]
Operation[edit]
Example[edit]
1 = RF ON
1 = BAND 2
1 = LINE ON
Configurable Antenna[edit]
Configurable Powermeter[edit]
ScreenElementDescription.svg Set frequency Set Frequency is the string that needs to be sent to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.
Example: “P1,U1”.
Example: “X1”
ScreenElementDescription.svg Result Format Result Format is the string that RadiMation® needs to use to decode the value from the string sent by the powermeter. The return value is interpreted in dBm.
Example: “__result__”
Example: “O1”
ScreenElementDescription.svg Duration Duration is the time that RadiMation® waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable results.
Channels[edit]
Read Back examples[edit]
Communication[edit]
Configurable LISN[edit]
Configurable Turn Table[edit]
Configurable commands[edit]
Check device[edit]
Device is initialised[edit]
Device is deinitialised[edit]
Some examples are:
Communication[edit]
Identification[edit]
Antenna Tower[edit]
Antenna Tower Advanced[edit]
Currently unavailable.
Configurable Cable[edit]
sending the next command.
keyword inserted value
__calculatedvoltage__ The calculated voltage expressed in V.
Messages[edit]
Unknown Device Driver[edit]
Device not connected[edit]
How to Report an Error[edit]
AD convertors[edit]
Type of communication: IEEE.
Channels 1:
Hewlett Packard 34401A[edit]
Channel 1
Type of measurement: AC Volt
Minimum value: 0Volt
Maximum value: 1 kVolt.
Channel 2:
Minimum value: 0 Amp.
Maximum value: 3 Amp.
Channel 3
Type of measurement: DC Volt
Channel 4:
Maximum value: 3 Amp. (rms)
Channel 5
Minimum value: 0 Ω
Maximum value: 100 MΩ
Channel 6
Minimum value: 0 Hz.
Maximum value: 300 kHz
Channel 7 and 8
Not used
Hewlett Packard 54600[edit]
Channel 1:
Type of measurement: V Max.
Maximum value: 1000 Volt.
Channel 2:
Type of measurement: V Min
Minimum value: 0 Volt.
Channel 3
Channel 4:
Channel 5
Minimum value: 0 Hz
Maximum value: 100 kHz
Channel 6
Type of measurement: Period
Minimum value: 0 ms
Maximum value: 10000 ms
Channel 7
Minimum value: 0 ms
Channel 8
Minimum value: 0 ms
Hewlett Packard 3562A[edit]
Type of communication: GPIB
Channels
Minimum value: 0 dB
Maximum value: 100 dB
Hewlett Packard 59313[edit]
Channel 1
Minimum value: 0
Maximum value: 2048
Channel 2:
Minimum value: 0
Maximum value: 2048
Channel 3
Minimum value: 0
Maximum value: 2048
Channel 4:
Minimum value: 0
Maximum value: 2048
Channel 5 to 8
Not used.
Channels
Marconi 2305[edit]
Channel 1
Minimum value: 0
Maximum value: 1000 MHz
Channel 2:
Minimum value: 0
Maximum value: 1000
Channel 3 to 8
Not used.
Fluke 45 AC Current[edit]
Channels
Minimum value: 0 mA
Maximum value: 10.000 mA
Fluke 45 DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 45 AC Voltage[edit]
Channels
Minimum value: 0 mV
Maximum value: 1.000.000 mV
Fluke 45 DC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 45 Frequency[edit]
Channels
Minimum value: 0 Hz
Maximum value: 1.000.000 Hz
Fluke 45 Resistance[edit]
Channels
Minimum value: 0 Ω
Maximum value: 100.000.000 Ω
Channel 1
Maximum value: 353.55 Volt.
Channel 2:
Channel 3
Channel 4:
Maximum value: 707.1 Volt
Channel 5
Type of measurement: RMS
Maximum value: 250 Volt
Channel 6
Maximum value: 200 MHz
Channel 7 and 8
Not used
Fluke 8840A AC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A AC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 8840A DC Voltage[edit]
Channels
Minimum value: 0 mV
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mV
Maximum value: 300.000 mV
Channels
Minimum value: 0 mV
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 8
Minimum value: 0
Maximum value: 16383
EIP 575[edit]
Channel 1
Minimum value: 0 Hz
Maximum value: 10 kHz
Channel 2:
Channel 3
Minimum value: 0 Hz
Maximum value: 1 MHz.
Channel 4:
Maximum value: 10 MHz.
Channel 5
Minimum value: 0 Hz
Maximum value: 100 MHz
Channel 6
Minimum value: 0 Hz
Maximum value: 1 GHz
Channel 7
Minimum value: 0 Hz
Maximum value: 10 GHz
Channel 8
Minimum value: 0 Hz
Maximum value: 100 GHz
Channel 1 to 8