| (110 intermediate revisions by 9 users not shown) | |||
| Line 2: | Line 2: | ||
= Device drivers = | = Device drivers = | ||
Device drivers are | Device drivers are available in {{Radimation}} to make it possible to communicate with thousands of different test and measurement equipment devices. They make the system flexible to use different devices, while the {{RadiMation}} Core stays generic. The {{RadiMation}} Core does not even know how to control specific devices it is controlling, only the type of device, and which device driver to use. The device driver is responsible for the communication between the {{RadiMation}} core and the specific test and measurement equipment, by sending the correct commands to the devices. If for example a new spectrum analyser or signal generator has been acquired, it is only needed to configure the corresponding device driver in the {{RadiMation}} software, and it will be able to use the new device. All the configurations of all the tests will remain the same. | ||
This chapter only describes the device driver specific configuration. All the configuration of the device drivers that are managed by the {{RadiMation}} Core, are described in {{ManualLink|14|3. Device Driver configuration}}. Also the common device driver settings are described in {{ManualLink|14|Device Driver Settings}}. | |||
{{RadiMation}} | |||
== | == Virtual device drivers == | ||
For testing and demonstration purposes, virtual device drivers can be used and added as well. | |||
Virtual device drivers are devices that act like normal devices but are emulated by software. These virtual devices do not physically exist, and thus do not control any real hardware. If a test site with only virtual devices is created, it can be used to perform completely virtual tests. | |||
Virtual device drivers do not have limitations like normal device drivers (like frequency band, maximum power, sweep times, etc.) | |||
There are several advantages to including and using virtual device drivers. It is mainly used for debugging, investigating and solving software problems, but it has proven to be very useful when used as a temporarily workaround. For example a test site can have an amplifier which is IEEE controlled and one day somebody accidentally destroys the communication between computer and amplifier. To allow the test engineer to still perform some tests, he can select the virtual amplifier device driver as the amplifier to be used in the test site. The test engineer then only has to set the amplifier in operate manually and will be able to perform the tests that day. When the IEEE communication is repaired the original amplifier device driver can be selected in the testsite again. | |||
==Active device drivers and passive device drivers== | |||
{{RadiMation}} supports two types of device drivers: active and passive device drivers. | |||
Active devices are test and measurement equipment like [[powermeter]], [[signal generator]]s, [[spectrum analyser]]s, [[field sensor]]s, etc. For these devices the communication settings (like the IEEE address or the serial port) can be set which {{RadiMation}} has to use to be able to communicate with these devices. | |||
Passive devices are test and measurement equipment like [[coupler]]s, [[antenna]], [[cable]]s, [[calibration jig]]s, [[current sensor]]s, etc. Passive equipment do also need a device driver. | |||
The reason for this is that {{RadiMation}} also has to know a number of details for these devices. Among others, the following information is relevant: | |||
* Usable frequency range | |||
* Maximum input power | |||
* General information of the device, to include a list of used equipment during a test in the report | |||
* Correction files that are relevant for the device | |||
With this information {{RadiMation}} can prevent the test engineer from making large mistakes, like for example using a [[coupler]] in an unsupported frequency range. | |||
==Device specific configuration== | ==Device specific configuration== | ||
===Signal generator=== | ===Signal generator=== | ||
Pressing advanced will open an IEEE configuration screen. Please view chapter IEEE setting for a complete description. See | Pressing advanced will open an IEEE configuration screen. Please view chapter IEEE setting for a complete description. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device. | ||
===Amplifier=== | ===Amplifier=== | ||
When an amplifier can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the amplifier can not be remotely controlled the message This Device cannot be configured will appear. See | When an amplifier can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the amplifier can not be remotely controlled the message This Device cannot be configured will appear. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting and protecting this device. | ||
===Antenna=== | ===Antenna=== | ||
When pressing advanced the message This Device cannot be configured will appear because this device can not be controlled remotely. See | When pressing advanced the message This Device cannot be configured will appear because this device can not be controlled remotely. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device. | ||
===Coupler=== | ===Coupler=== | ||
===Powermeter=== | ===Powermeter=== | ||
[[Image:Power Meter Configuration Window.png]] | [[Image:Power Meter Configuration Window.png]] | ||
{{ScreenElementDescription|Wait | {{ScreenElementDescriptionStart}} | ||
{{ScreenElementDescription|Pre Wait time| The minimum time that should be waited before the measurements starts.}} | |||
{{ScreenElementDescription|Measure| The minimum amount of measurements that {{RadiMation}} has to perform to determine if the <Max. Difference> condition is met.}} | |||
{{ScreenElementDescription|Wait time| The minimum time that {{RadiMation}} should be waiting between each measurement.}} | |||
{{ScreenElementDescription|Max. Difference| The maximum difference that is allowed between the <Measure> measurement value(s); the highest and the lowest values are compared.}} | |||
{{ScreenElementDescription|Max. Measure| The maximum amount of measurements that ''could'' be performed by {{RadiMation}}, before there is a final measurement value available (which could meet these conditions, or not).}} | |||
If Measure is set to a value of 1 the Wait time, Max. Difference and Max. Measure settings will be disabled, because only one measurement will be performed. | |||
Now that all windows have been generally explained, the procedure {{RadiMation}} uses is the following. {{RadiMation}} takes the amount of measurements as defined in the Measure window. After that {{RadiMation}} determines the minimum, maximum and difference. When the difference is equal or smaller than defined in Max difference {{RadiMation}} determines the power. If the difference is greater than defined, {{RadiMation}} takes one new measurement. Replaces the oldest value with the | |||
Now that all windows have been generally explained, the procedure {{RadiMation}} uses is the following. {{RadiMation}} takes the amount of measurements as defined in the Measure window. After that {{RadiMation}} determines the minimum, maximum and difference. When the difference is equal or smaller | |||
new one. Determines the minimum, maximum and difference again. This continues as long as the difference is larger then defined and the maximum amount of measurement has not yet been reached. If the maximum amount of measurement has been reached the last measurement is taken as the measured value. | new one. Determines the minimum, maximum and difference again. This continues as long as the difference is larger then defined and the maximum amount of measurement has not yet been reached. If the maximum amount of measurement has been reached the last measurement is taken as the measured value. | ||
See | See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device. | ||
{{Note|'''Speed up(1):''' Sometimes slower means faster. For instance, if you have a slower powermeter (resistor head) it may be quicker to have a longer waiting time. Because the powermeter has more time to determine the right value, {{RadiMation}} needs less measurements to determine the value. It is up to the engineer to find the right accuracy vs. time setting. | {{Note|'''Speed up(1):''' Sometimes slower means faster. For instance, if you have a slower powermeter (resistor head) it may be quicker to have a longer waiting time. Because the powermeter has more time to determine the right value, {{RadiMation}} needs less measurements to determine the value. It is up to the engineer to find the right accuracy vs. time setting. | ||
| Line 42: | Line 61: | ||
'''Speed up(2):''' When using a spectrum analyser a small amount of samples may be sufficient to determine the right amount of power. It is up to the engineer to find the right accuracy vs. time setting. | '''Speed up(2):''' When using a spectrum analyser a small amount of samples may be sufficient to determine the right amount of power. It is up to the engineer to find the right accuracy vs. time setting. | ||
'''Measuring:''' | '''Measuring:''' Powermeters which do not have a RMS detector, should not be used in Fast constant mode. | ||
===Field Sensor=== | ===Field Sensor=== | ||
==== Measurement setting panel ==== | |||
[[Image:Field Sensor Configuration Window.png]] | [[Image:Field Sensor Configuration Window.png]] | ||
{{ | Now that all windows and buttons have been generally explained, the procedure {{RadiMation}} uses is the following. {{RadiMation}} takes the amount of measurements as defined in the Measure window. After that {{RadiMation}} determines the minimum, maximum and difference. When the difference is equal or smaller than defined in Max difference {{RadiMation}} determines the field. If the difference is greater than defined, {{RadiMation}} takes one new measurement. Replaces the oldest value with the | ||
new one. Determines the minimum, maximum and difference again. This continues as long as the difference is larger then defined and the maximum amount of measurement has not yet been reached. If the maximum amount of measurement has been reached the last measurement is taken as the measured value. | |||
{{ | {{Note|'''Speed up:''' Sometimes slower means faster. For instance, if you have a slower field sensor it may be quicker to have a longer waiting time. Because the field sensor has more time to determine the right value, {{RadiMation}} needs less measurements to determine the value. It is up to the engineer to find the right accuracy vs. time setting.}} | ||
==== Axis setting panel ==== | |||
[[Image:Field sensor Advanced settings Axis Configuration.png]] | |||
The Axis configuration is a generic driver setting panel. | |||
There are several (old) field sensors which don’t support the readout of an isotropic value: instead they provide the measurement values of all three axes separately. | |||
Using the ‘Software isotropic’ setting in the device driver can then activate that the measurement data of all three axes are being interpreted and used to calculate the isotropic value and have this isotropic value being returned as the measured field strength. | |||
When the field sensor already supports the retrieval of the isotropic measurement value, often this value is also returned when the software option is being selected in the device configuration: both isotropic as well as software isotropic will return the isotropic value of the field sensor. This is depending on the device driver and used equipment. | |||
Normally, this setting is set to ‘Isotropic’ in case the isotropic value should be returned. | |||
===AD converter=== | ===AD converter=== | ||
| Line 67: | Line 96: | ||
===Injection device=== | ===Injection device=== | ||
===Calibration Jigs=== | ===Calibration Jigs=== | ||
===Current Sensor=== | ===Current Sensor=== | ||
===Pre Amplifiers=== | ===Pre Amplifiers=== | ||
===Receivers / Spectrum analyser=== | ===Receivers / Spectrum analyser=== | ||
Pressing advanced will open an IEEE configuration screen. Please view chapter IEEE setting for a complete description. | Pressing advanced will open an IEEE configuration screen. Please view chapter IEEE setting for a complete description. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device. | ||
===LISN=== | ===LISN=== | ||
When an LISN can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the LISN can not be remotely controlled the message This Device cannot be configured will appear. See | When an LISN can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the LISN can not be remotely controlled the message This Device cannot be configured will appear. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device. | ||
===Turn Table=== | ===Turn Table=== | ||
| Line 91: | Line 120: | ||
===Absorbing Clamps=== | ===Absorbing Clamps=== | ||
===Clamp positioner=== | ===Clamp positioner=== | ||
| Line 97: | Line 126: | ||
===Cables=== | ===Cables=== | ||
===Switch matrix=== | ===Switch matrix=== | ||
| Line 104: | Line 133: | ||
When a EUT Controller can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the EUT Controller can not be remotely controlled the message This Device cannot be configured will appear. | When a EUT Controller can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the EUT Controller can not be remotely controlled the message This Device cannot be configured will appear. | ||
===Auxiliary equipment=== | |||
When a Auxiliary equipment can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the Auxiliary equipment can not be remotely controlled the message This Device cannot be configured will appear. | |||
== Communication interfaces and protocols == | |||
Different test and measurement instrumentation devices, can be controlled by different electronic communication buses and connectors, with on top of that different communication protocols. Which communication stream setting should be used to control a measurement instruments, cannot be easily based alone on the connector that is used to connect the measurement instrument. The following list shows the possible communication interfaces and protocols that are possible with a specific connector. This table can help to identify the actual communication protocol that is used. The manual of the test and measurement instrument also often lists the communication protocol that should be used, with similar terms, as is shown in the table. Based on the actual communication protocol the best communication stream to use is suggested. | |||
{|class="wikitable" | |||
! Connector || Communication protocol || Suggested communication stream || Alternative communication stream | |||
| [[GPIB]] / [[wikipedia:IEEE-488|IEEE-488]] || GPIB || [[#GPIB_Setting|GPIB]] || [[#VISA_Setting|VISA]] (with GPIB sub option) | |||
|rowspan="5"| Ethernet / [[wikipedia:Rj45|RJ45]] || VXI-11 || [[#VISA_Setting|VISA]] (with LAN sub option) || | |||
| [[wikipedia:High Speed LAN Instrument Protocol|HiSLIP]] || [[#VISA_Setting|VISA]] (with LAN sub option) || | |||
| [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier) | |||
| HTTP || [[#HTTP_Setting|HTTP]] || | |||
| UDP || [[#UDP_Setting|UDP]] || | |||
|rowspan="4"| [[wikipedia:USB|USB]]|| [[wikipedia:Virtual COM port|Virtual COM Port (VCP)]] (RS-232) || [[#RS-232_Setting|RS-232]] || [[#VISA_Setting|VISA]] (with RS-232 sub option) | |||
| USBTMC (Test and Measurement Class) || [[#VISA_Setting|VISA]] (with Visa-ID sub option) || | |||
| Raditeq USB devices || [[#USB_Setting|USB]] || [[#RS-232_Setting|RS-232]]<BR/>[[#VISA_Setting|VISA]] (with RS-232 sub option) | |||
| Device specific USB || || should be implemented device specifically | |||
| [[wikipedia:D-subminiature|Sub-D9]] / [[wikipedia:RS-232|RS-232]] || RS-232 || [[#RS-232_Setting|RS-232]] || [[#VISA_Setting|VISA]] (with RS-232 sub option) | |||
|rowspan="2"| other software || [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier) | |||
The details for the configuration of each possible communication stream is described in the next section. | |||
== Communication settings == | == Communication settings == | ||
Depending on the brand and model of the device, it may be necessary to specify the parameters for the communication with the device. The IEEE 488 and | Depending on the brand and model of the device, it may be necessary to specify the parameters for the communication with the device. The GPIB/IEEE-488 and RS-232 buses are most commonly used. This section describes which communication buses are supported and which communication parameters can be configured. Not all devices support all the communication buses, so only the applicable communication buses can be selected in the device driver configuration. | ||
=== | === GPIB Setting === | ||
The GPIB Communication settings are used to specify the GPIB communication parameters, for the communication with a measurement device over GPIB. | |||
[[Image:GPIB DeviceStream Configuration.png]] | |||
{{ScreenElementDescription|Primary Address|The primary address specifies the address of the device, so if the device you want to configure is on address 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device.}} | |||
{{ScreenElementDescription|Use advanced configuration|Allows the end-user to enable the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient.}} | |||
{{ScreenElementDescription|GPIB Board|With GPIB Board you can specify the GPIB board that is used. The default value is 0 and can only be different when multiple GPIB boards are present.}} | |||
{{ScreenElementDescription|Secondary Address|The second address is 0 by default, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.}} | |||
{{ScreenElementDescription|GPIB Delay|GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PCs (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and sometimes gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see if the problem has disappeared. If the problem has disappeared then your problem was timing, if not please contact your reseller and report the problems you are having.}} | |||
{{ScreenElementDescription|Clear device during initialisation|When checked, a low level GPIB command to reset the device will be sent to the device during initialisation.}} | |||
{{ScreenElementDescription|Readdress device|When checked, a low level GPIB command to select the correct address will be transmitted every time a command is sent to the device.}} | |||
=== | === HTTP Setting === | ||
The HTTP communication settings are used to specify the HTTP communication parameters, for the communication with a measurement device over HTTP. In this HTTP communication mode, the [[wikipedia:Hypertext Transfer Protocol|HyperText Transfer Protocol (HTTP)]] is used, and actual GET requests are performed. | |||
[[ | [[File:HTTPDeviceStreamConfiguration.png]] | ||
{{ScreenElementDescription| | {{ScreenElementDescriptionStart}} | ||
{{ScreenElementDescription|Address|Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. By default the socket port number is 80, but this can be changed to another port if desired.}} | |||
If for example in a configurable device driver this HTTP communication stream is used, and the specified command is for example "SETA=1" (which set relay A of a Mini-Circuits relay to position 1), an actual "GET /SETA=1" HTTP request is transmitted to the specified hostname and socket port that is configured for the {{ScreenElement|Address}}. | |||
=== Manual Setting === | |||
The manual communication stream can be used to show message boxes on the screen for the end-user, to manually perform some operation on a device. The commands that are specified in the device driver that is using the manual communication stream, are thus the messages that are shown to the end-user in a message box, for example with the instruction to change a setting on a test and measurement device. | |||
No further configuration in the manual communication stream is needed. | |||
[[File:ManualDeviceStreamConfiguration.png]] | |||
===RS-232 Setting=== | |||
The RS-232 communication settings specify the serial port parameters when communicating with a measurement device over RS-232. Although many modern devices connect via USB, they often appear as a virtual COM port (VCP). In that case you still configure the RS-232 settings as below. | |||
[[Image:RS-232 DeviceStream Configuration.png]] | |||
{{ScreenElementDescription|COM Port|The COM port which {{RadiMation}} has to be used to communicate with the measurement device.}} | |||
{{ScreenElementDescription|Baud rate|Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}} | |||
{{ScreenElementDescription|Data bits|Allows to configure the number of data bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}} | |||
{{ScreenElementDescription|Parity|Allows to configure the parity bit that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}} | |||
{{ScreenElementDescription|Stop bits|Allows to configure the number of stop bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}} | |||
{{ScreenElementDescription|Send termination|Allows to configure if and which terminator should be used during data transmissions to the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}} | |||
{{ScreenElementDescription|Receive termination|Allows to configure if and which terminator should be used during the receiving of data from the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}} | |||
{{ScreenElementDescription|Receiving a NewLine ends read|Allows to configure if a read should be ended as soon as the receive terminator is received from the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}} | |||
== | === RSIB Setting === | ||
The RSIB communication settings are used to communicate with an ethernet connected measurement device through the [[Rohde & Schwarz]] RSISB32.DLL, which implements a VISA VXI-11 compatible layer. | |||
[[File:RSIBDeviceStreamConfiguration.png]] | |||
{{ScreenElementDescription|RSIB Address|Allows to specify RSIB identifier of the device that should be controlled.}} | |||
=== TCPIP Setting === | |||
The TCPIP communication settings are used to specify that a (raw-)socket based [[wikipedia:TCP/IP|TCP/IP]] connection to a measurement device should be used. | |||
[[Image:TCPIP DeviceStream Configuration.png]] | |||
{{ScreenElementDescription|Address|Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. If the socket port number is a fixed port number, it is already shown as the default value, and it will be automatically added when no socket port number is specified.}} | |||
The | === UDP Setting === | ||
The UPD communication settings are used to specify that an UDP based communication over ethernet should be used to control the measurement. | |||
[[Image:UDPDeviceStreamConfiguration.png]] | |||
{{ | {{ScreenElementDescriptionStart}} | ||
=== USB Setting === | |||
The USB Communication settings are used to specify that a USB connection to a [[Raditeq]] or [[DARE!! Instruments]] measurement device is used. For devices that appear as a Virtual COM port (VCP) over USB, configure the [[Chapter_15#RS-232_Setting|RS-232 setting]] rather than the USB setting. | |||
[[Image:USB DeviceStream Configuration.png]] | |||
{{ScreenElementDescription|Device Identifier|The device identifier (which is a unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.}} | |||
{{ScreenElementDescription|Detect|Will automatically determine the correct {{ScreenElement|Device Identifier}} for the measurement device that is connected.}} | |||
The | === VISA Setting === | ||
The VISA Communication settings are used to specify the VISA communication parameters, for the communication with a measurement device. The VISA library is a higher level communication library that supports different kinds of communication methods. {{RadiMation}} doesn't provide a VISA library itself, and thus requires that a VISA library from another supplier like [[National Instruments]] or [[Keysight Technologies]] is installed. The VISA manager that is provided by that VISA library can be used to determine the correct VISA resource. The selected VISA configuration thus also has an influence if a measurement device is controlled by GPIB, RS-232, LAN, USBTMC, VXI-11, network-socket or another communication method . | |||
[[Image:VISA DeviceStream Configuration.png]] | |||
{{ScreenElementDescription|Alias|Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. [[National Instruments]] MAX or [[Keysight Technologies]] connection expert) that is installed on the PC.}} | |||
{{ScreenElementDescription|GPIB|Allows to specify that GPIB should be used for the communication with the measurement device. The GPIB address of the measurement device should be specified.}} | |||
{{ScreenElementDescription|LAN|Allows to specify that a VXI-11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.}} | |||
{{ScreenElementDescription|RS-232|Allows to specify that a RS232 (ASRL in VISA terms) connection should be used for the communication with the measurement device. The COM port of the measurement device should be specified.}} | |||
{{ScreenElementDescription|Visa-ID|Allows to specify a VISA Identifier that should be used for the communication with the measurement device. Any VISA Identifier that is supported by VISA is accepted. If one of the other communication methods is selected the corresponding VISA Identifier is also shown in this setting. Often used VISA identifiers are: | |||
* <code>TCPIP[board]::<IP-address>::INSTR</code>: VXI-11 or LXI ethernet communication with the device with '<IP-address>'. | |||
* <code>TCPIP[board]::<IP-address>::<port-number>::SOCKET</code>: Raw socket based ethernet communication on the port '<port-number>' with the the device '<IP-address>' | |||
* <code>GPIB[board]::<primary address>::INSTR</code>: GPIB communication to the GPIB device with the address '<primary address>' | |||
* <code>ASRL<port>::INSTR</code>: Serial communcation on COM-port '<port>' | |||
==Configurable device drivers== | |||
{{RadiMation}} allows the user to create their own device drivers for test and measurement equipment, which is (not yet) implemented in the standard device driver list. | |||
User configurable device drivers are available for nearly all types of test and measurement equipment. However, device drivers for spectrum analysers and (scanning) receivers can not be made with user configurable device drivers because the complexity (and differences between suppliers) of these devices is too high. | |||
To make your own device driver, use the “Device drivers” tab in the “Configuration” > “Configuration” menu and follow the steps below: | |||
# In the device driver’s menu, select the required device driver type. | |||
# Press the “New” button. | |||
# Select the driver called “Configurable xx” (i.e. if you want to make a signal generator device driver you would select the “Configurable signal generator” device driver). | |||
# Enter a description for the device driver (for example the type number of the generator) and press OK. | |||
# The name of the device driver will be added in the available device drivers list. | |||
# Select the new device driver from the available device drivers list. | |||
# Press the “Edit” button. | |||
# A configuration screen for the device driver will appear. For active controlled equipment all required control commands for the device must be entered. Refer to the operating manual of the equipment for these codes. | |||
After all codes are entered, the device driver is ready for use. The custom-made device driver can then be used as any other device driver by selecting the driver in the Equipment list. | |||
In the Returned ID window you need to specify the string that the software will receive so that it | ===Generic settings=== | ||
{{ScreenElementDescription|Start and stop frequency|The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, {{RadiMation}} can warn the test engineer when they want to use the device out of its valid frequency range.}} | |||
{{ScreenElementDescription|Reset|In the reset window you need to specify the string that the software needs to send when it want to reset the device. For example *RST is commonly used reset string. If you don’t know the string then leave this window blank, and make sure that the device in a neutral state.}} | |||
{{ScreenElementDescription|Init|In the init window you need to specify the string that the software needs to send when it want to Init the device. For example *RST is commonly used reset string.}} | |||
{{ScreenElementDescription|Get ID|In the Get ID window you need to specify the string that the software needs to send when it wants to get the ID string of the device. For example *IDN? is commonly used Get ID string.}} | |||
{{ScreenElementDescription|Returned ID|In the Returned ID window you need to specify the string that the software will receive so that it knows that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.}} | |||
===Signal generator=== | === Configurable Signal generator === | ||
[[Image:Configurable Signal Generator Window.png]] | [[Image:Configurable Signal Generator Window.png]] | ||
{{ScreenElementDescription|Set Frequency|Set Frequency is the string that needs to | {{ScreenElementDescription|Set Frequency|Set Frequency is the string that needs to be sent to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz. }} | ||
'''Example:''' “FRQ__freq__MHZ”. | '''Example:''' “FRQ__freq__MHZ”. | ||
'''Example:''' “FRQ__freqHz__HZ”. | |||
'''Example:''' “FRQ__freqkHz__KHZ”. | |||
'''Example:''' “FRQ__freqMHz__MHZ”. | |||
'''Example:''' “FRQ__freqGHz__GHZ”. | |||
__freq__ will be replaced by the value that {{RadiMation}} want to set the signal generator to. The “__” of “__freq__” are two “_”. | __freq__ will be replaced by the value that {{RadiMation}} want to set the signal generator to. The “__” of “__freq__” are two “_”. | ||
{{ScreenElementDescription|Set Carrier|Set Carrier is the string that needs to | All the __freq__ keywords below can also use the unit specifier: Hz,kHz,MHz and GHz. | ||
{{ScreenElementDescription|Set Carrier|Set Carrier is the string that needs to be sent to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.}} | |||
'''Example:''' “:SOURCE:POWER __carrier__ DBM”. | '''Example:''' “:SOURCE:POWER __carrier__ DBM”. | ||
__carrier__ will be replaced by the value that {{RadiMation}} | __carrier__ will be replaced by the value that {{RadiMation}} wants to set the signal generator to. The “__” of “__carrier__ ” are two “_”. | ||
{{ScreenElementDescription|Carrier on|Carrier on is the string that needs to | {{ScreenElementDescription|Carrier on|Carrier on is the string that needs to be sent to set the signal generator carrier on.}} | ||
'''Example:''' “OUTPUT ON”. | '''Example:''' “OUTPUT ON”. | ||
{{ScreenElementDescription|Carrier off|Carrier off is the string that needs to | {{ScreenElementDescription|Carrier off|Carrier off is the string that needs to be sent to set the signal generator carrier off.}} | ||
'''Example:''' “OUTPUT OFF”. | '''Example:''' “OUTPUT OFF”. | ||
{{ScreenElementDescription|AM on|AM on is the string that needs to | {{ScreenElementDescription|AM on|AM on is the string that needs to be sent to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.}} | ||
'''Example:''' “SOURCE2:FREQ __freq__ KHZ;SOURCE:AM:DEPTH __depth__ PCT;SOURCE:AM:STATE ON”. | '''Example:''' “SOURCE2:FREQ __freq__ KHZ;SOURCE:AM:DEPTH __depth__ PCT;SOURCE:AM:STATE ON”. | ||
| Line 205: | Line 343: | ||
“__freq__ ” are two “_”. | “__freq__ ” are two “_”. | ||
{{ScreenElementDescription|AM off|AM off is the string that needs to | {{ScreenElementDescription|AM off|AM off is the string that needs to be sent to set the internal am modulation of the signal generator off.}} | ||
'''Example:''' “SOURCE:AM:STATE OFF”. | '''Example:''' “SOURCE:AM:STATE OFF”. | ||
{{ScreenElementDescription|External on|External on is the string that needs to | {{ScreenElementDescription|External on|External on is the string that needs to be sent to set the modulator to the external input.}} | ||
'''Example:''' “SOURCE:AM:SOURCE EXT”. | '''Example:''' “SOURCE:AM:SOURCE EXT”. | ||
{{ScreenElementDescription|External off|External off is the string that needs to | {{ScreenElementDescription|External off|External off is the string that needs to be sent to set the modulator to the internal input.}} | ||
'''Example:''' “SOURCE:AM:SOURCE INT”. | '''Example:''' “SOURCE:AM:SOURCE INT”. | ||
{{ScreenElementDescription|PM on|PM on is the string that needs to | {{ScreenElementDescription|PM on|PM on is the string that needs to be sent to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.}} | ||
'''Example:''' “SOURCE2:FREQ __freq__ KHZ;SOURCE:PM:DUTY __duty__ PCT;SOURCE:PM:STATE ON”. | '''Example:''' “SOURCE2:FREQ __freq__ KHZ;SOURCE:PM:DUTY __duty__ PCT;SOURCE:PM:STATE ON”. | ||
| Line 224: | Line 362: | ||
{{ScreenElementDescription|PM off|PM off is the string that needs to | {{ScreenElementDescription|PM off|PM off is the string that needs to be sent to set the internal PM modulation of the signal generator off.}} | ||
'''Example:''' “SOURCE:PM:STATE OFF”. | '''Example:''' “SOURCE:PM:STATE OFF”. | ||
{{ScreenElementDescription|Sine Wave|Sine wave is the string that needs to be | {{ScreenElementDescription|Sine Wave|Sine wave is the string that needs to be sent when {{RadiMation}} wants to set the wave form of the internal source to sine.}} | ||
'''Example:''' “SOURCE2:FUNC SIN” | '''Example:''' “SOURCE2:FUNC SIN” | ||
{{ScreenElementDescription|Square Wave|Square wave is the string that needs to be | {{ScreenElementDescription|Square Wave|Square wave is the string that needs to be sent when {{RadiMation}} wants to set the wave form of the internal source to square.}} | ||
'''Example:''' “SOURCE2:FUNC SQU” | '''Example:''' “SOURCE2:FUNC SQU” | ||
===Amplifier=== | === Configurable Amplifier === | ||
{{:Configurable Amplifier}} | |||
=== Configurable Antenna === | |||
{{:Configurable Antenna}} | |||
=== Configurable Powermeter === | |||
===Powermeter=== | |||
[[Image:Configurable Power Meter Window.png]] | [[Image:Configurable Power Meter Window.png]] | ||
{{ScreenElementDescription|Set frequency|Set Frequency is the string that needs to | {{ScreenElementDescription|Set frequency|Set Frequency is the string that needs to be sent to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.}} | ||
| Line 266: | Line 398: | ||
'''Example:''' “X1” | '''Example:''' “X1” | ||
{{ScreenElementDescription|Result Format|Result Format is the string that {{RadiMation}} needs to use to decode the value from the string | {{ScreenElementDescription|Result Format|Result Format is the string that {{RadiMation}} needs to use to decode the value from the string sent by the powermeter. The return value is interpreted in dBm.}} | ||
'''Example:''' “__result__” | '''Example:''' “__result__” | ||
| Line 277: | Line 409: | ||
'''Example:''' “O1” | '''Example:''' “O1” | ||
{{ScreenElementDescription|Duration|Duration is the time that {{RadiMation}} waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable | {{ScreenElementDescription|Duration|Duration is the time that {{RadiMation}} waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable results.}} | ||
===AD converter=== | === Configurable AD converter === | ||
{{:Configurable_AD_Convertor}} | |||
=== Configurable Calibration Jig === | |||
{{:Configurable Calibration Jig}} | |||
===Calibration | |||
=== Configurable Current Sensor === | |||
Only the start and stop frequency can be set for the amplifier. Chapter generic settings will give more information about the start and stop frequency | Only the start and stop frequency can be set for the amplifier. Chapter generic settings will give more information about the start and stop frequency | ||
=== | === Configurable Pre Amplifier === | ||
=== | === Configurable Receivers / Spectrum analyser === | ||
Due to the complexity of controlling analysers and (scanning) receivers, no configurable device driver is available for analyser and (scanning) receivers. | |||
=== Configurable LISN === | |||
{{:Configurable LISN}} | |||
=== | === Configurable Turn Table=== | ||
{{:Configurable TurnTable}} | |||
=== Configurable Antenna Tower=== | |||
{{:Configurable Antenna Tower}} | |||
=== Configurable Absorbing Clamps=== | |||
=== | === Configurable Clamp positioner=== | ||
Currently unavailable. | Currently unavailable. | ||
=== | === Configurable Cable=== | ||
===Switch matrix=== | === Configurable Switch matrix=== | ||
{{:Configurable Switch Matrix}} | |||
{{:Configurable EUT Controller}} | |||
{{:Configurable equipment}} | |||
==Messages == | ==Messages == | ||
| Line 336: | Line 458: | ||
[[Image:Amplifier Device Driver Can Not Be Configured Window.png]] | [[Image:Amplifier Device Driver Can Not Be Configured Window.png]] | ||
This message box is displayed when you want to edit a device driver that cannot be configured, like a coupler or calibration jig. This does not mean that the device driver is useless. Please see chapter Configurable device drivers vs. | This message box is displayed when you want to edit a device driver that cannot be configured, like a coupler or calibration jig. This does not mean that the device driver is useless. Please see chapter Configurable device drivers vs. non configurable device drivers for explanation. Nowadays also information about the used configuration is shown in this message to the end-user. | ||
===Unknown Device Driver=== | ===Unknown Device Driver=== | ||
| Line 344: | Line 466: | ||
This message box is displayed when {{RadiMation}} is trying to locate a device driver and was unable to find it. If you see this message please contact your reseller and tell him which driver you are trying to use. The reseller will take action so that you will receive the right device driver. | This message box is displayed when {{RadiMation}} is trying to locate a device driver and was unable to find it. If you see this message please contact your reseller and tell him which driver you are trying to use. The reseller will take action so that you will receive the right device driver. | ||
=== GPIB: Device is not connected=== | ===GPIB: Device is not connected=== | ||
[[Image:GPIB Not Connected Window.png]] | [[Image:GPIB Not Connected Window.png]] | ||
| Line 355: | Line 477: | ||
This message box is displayed when {{RadiMation}} is unable to connect to the device. Please check cables and device driver settings. | This message box is displayed when {{RadiMation}} is unable to connect to the device. Please check cables and device driver settings. | ||
==How to | ==How to Report an Error== | ||
When encountering a problem with the software you might | When encountering a problem with the software you might like to report it to the RadiMation support. | ||
The RadiMation error popups, contains a detailed section. This can be expanded with the details button on the error popup. | |||
In the expanded detailed error popup, a Report Error button is present. This allows Error Reporting to the RadiMation support within RadiMation. | |||
More information about Error Reporting can be found here: | |||
https://wiki.radimation.com/wiki/index.php/Error_Report | |||
[[File:ErrorPopupWindow.png]] | |||
==AD convertors== | ==AD convertors== | ||
| Line 922: | Line 1,030: | ||
Not used | Not used | ||
=== | ===DARE!! Development Radimate 2 and 3=== | ||
Type of communication: RS 232 | Type of communication: RS 232 | ||
| Line 934: | Line 1,042: | ||
Maximum value: 16383 | Maximum value: 16383 | ||
=== | ===EIP 575=== | ||
Type of communication: IEEE | Type of communication: IEEE | ||
Device drivers[edit]
Virtual device drivers[edit]
- Usable frequency range
- Maximum input power
Signal generator[edit]
Amplifier[edit]
Antenna[edit]
Coupler[edit]
Powermeter[edit]
| The minimum time that should be waited before the measurements starts. |
| The minimum amount of measurements that RadiMation® has to perform to determine if the <Max. Difference> condition is met. |
| The minimum time that RadiMation® should be waiting between each measurement. |
| The maximum difference that is allowed between the <Measure> measurement value(s); the highest and the lowest values are compared. |
| The maximum amount of measurements that could be performed by RadiMation®, before there is a final measurement value available (which could meet these conditions, or not). |
Speed up(2): When using a spectrum analyser a small amount of samples may be sufficient to determine the right amount of power. It is up to the engineer to find the right accuracy vs. time setting. Measuring: Powermeters which do not have a RMS detector, should not be used in Fast constant mode. |
Field Sensor[edit]
Axis setting panel[edit]
AD converter[edit]
Injection device[edit]
Calibration Jigs[edit]
Current Sensor[edit]
Pre Amplifiers[edit]
LISN[edit]
Turn Table[edit]
Antenna Tower[edit]
Absorbing Clamps[edit]
Clamp positioner[edit]
Cables[edit]
Switch matrix[edit]
EUT Controller[edit]
Auxiliary equipment[edit]
| Connector | Communication protocol | Suggested communication stream | Alternative communication stream |
| GPIB / IEEE-488 | GPIB | GPIB | VISA (with GPIB sub option) |
| Ethernet / RJ45 | VXI-11 | VISA (with LAN sub option) | |
| HiSLIP | VISA (with LAN sub option) | ||
| Raw socket | TCPIP | VISA (with Visa-ID sub option with ::SOCKET identifier)
| HTTP | HTTP | ||
| UDP | UDP | ||
| USB | Virtual COM Port (VCP) (RS-232) | RS-232 | VISA (with RS-232 sub option) |
| USBTMC (Test and Measurement Class) | VISA (with Visa-ID sub option) | ||
| Raditeq USB devices | USB | RS-232 VISA (with RS-232 sub option) | |
| Device specific USB | should be implemented device specifically | ||
| Sub-D9 / RS-232 | RS-232 | RS-232 | VISA (with RS-232 sub option) |
| other software | Raw socket | TCPIP | VISA (with Visa-ID sub option with ::SOCKET identifier)
Communication settings[edit]
GPIB Setting[edit]
| The primary address specifies the address of the device, so if the device you want to configure is on address 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device. |
| Allows the end-user to enable the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient. |
| The second address is 0 by default, and should only be changed when needed. For further information please look in the help of National Instruments 488.2. |
| GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PCs (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and sometimes gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see if the problem has disappeared. If the problem has disappeared then your problem was timing, if not please contact your reseller and report the problems you are having. |
| When checked, a low level GPIB command to reset the device will be sent to the device during initialisation. |
| When checked, a low level GPIB command to select the correct address will be transmitted every time a command is sent to the device. |
HTTP Setting[edit]
| Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. By default the socket port number is 80, but this can be changed to another port if desired. |
Manual Setting[edit]
RS-232 Setting[edit]
| The COM port which RadiMation® has to be used to communicate with the measurement device. |
| Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled. |
| Allows to configure the number of data bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled. |
| Allows to configure the parity bit that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled. |
| Allows to configure the number of stop bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled. |
| Allows to configure if and which terminator should be used during data transmissions to the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled. |
| Allows to configure if and which terminator should be used during the receiving of data from the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled. |
| Allows to configure if a read should be ended as soon as the receive terminator is received from the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled. |
RSIB Setting[edit]
| Allows to specify RSIB identifier of the device that should be controlled. |
TCPIP Setting[edit]
| Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. If the socket port number is a fixed port number, it is already shown as the default value, and it will be automatically added when no socket port number is specified. |
UDP Setting[edit]
USB Setting[edit]
| The device identifier (which is a unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB. |
| Will automatically determine the correct Device Identifier for the measurement device that is connected. |
VISA Setting[edit]
| Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. National Instruments MAX or Keysight Technologies connection expert) that is installed on the PC. |
| Allows to specify that GPIB should be used for the communication with the measurement device. The GPIB address of the measurement device should be specified. |
| Allows to specify that a VXI-11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device. |
| Allows to specify that a RS232 (ASRL in VISA terms) connection should be used for the communication with the measurement device. The COM port of the measurement device should be specified. |
- Press the “New” button.
- The name of the device driver will be added in the available device drivers list.
- Select the new device driver from the available device drivers list.
- Press the “Edit” button.
Generic settings[edit]
| The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, RadiMation® can warn the test engineer when they want to use the device out of its valid frequency range. |
| In the reset window you need to specify the string that the software needs to send when it want to reset the device. For example *RST is commonly used reset string. If you don’t know the string then leave this window blank, and make sure that the device in a neutral state. |
| In the init window you need to specify the string that the software needs to send when it want to Init the device. For example *RST is commonly used reset string. |
| In the Get ID window you need to specify the string that the software needs to send when it wants to get the ID string of the device. For example *IDN? is commonly used Get ID string. |
| In the Returned ID window you need to specify the string that the software will receive so that it knows that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted. |
| Set Frequency is the string that needs to be sent to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz. |
Example: “FRQ__freq__MHZ”.
Example: “FRQ__freqHz__HZ”.
Example: “FRQ__freqkHz__KHZ”.
Example: “FRQ__freqMHz__MHZ”.
Example: “FRQ__freqGHz__GHZ”.
| Set Carrier is the string that needs to be sent to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM. |
| Carrier on is the string that needs to be sent to set the signal generator carrier on. |
Example: “OUTPUT ON”.
| Carrier off is the string that needs to be sent to set the signal generator carrier off. |
Example: “OUTPUT OFF”.
| AM on is the string that needs to be sent to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %. |
| AM off is the string that needs to be sent to set the internal am modulation of the signal generator off. |
| External on is the string that needs to be sent to set the modulator to the external input. |
| External off is the string that needs to be sent to set the modulator to the internal input. |
| PM on is the string that needs to be sent to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %. |
| PM off is the string that needs to be sent to set the internal PM modulation of the signal generator off. |
| Sine wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to sine. |
Example: “SOURCE2:FUNC SIN”
| Square wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to square. |
Example: “SOURCE2:FUNC SQU”
Configurable Amplifier[edit]
Main[edit]
Operation[edit]
Example[edit]
Configurable Antenna[edit]
Configurable Powermeter[edit]
| Set Frequency is the string that needs to be sent to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz. |
Example: “P1,U1”.
Example: “X1”
| Result Format is the string that RadiMation® needs to use to decode the value from the string sent by the powermeter. The return value is interpreted in dBm. |
Example: “__result__”
Example: “O1”
| Duration is the time that RadiMation® waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable results. |
Channels[edit]
Read Back examples[edit]
Communication[edit]
| The lowest usable frequency of the device. |
| The highest usable frequency of the device. |
The impedance of the injection system itself, for which the calibration jig is used to calibrate it.
Configurable LISN[edit]
| The reset command that needs to be send to device. When left blank, no command will be send. |
| The command that is send to retrieve the identity of the device. This is used to check if the device is connected. A commonly used SCPI command is: *IDN?. When left blank, no command will be send, and the device check is skipped. |
| A string that is used to check if the correct device driver is selected and the device is connected. The device check consists of a few steps. First the Get ID command is transmitted to the device, and the response from the device is then received. The received response is then searched for the string specified at Returned ID, and if the string is found it is confirmed that the correct device is connected. The string specified can also be a Regular expression, which is then used to find a match in the response from the device. Leaving this blank, will skip the device check, and it is silently assumed (but not checked) that the correct device is connected. |
| The command that should be transmitted when the 'Neutral' line of the LISN should be selected. If no command is specified, nothing will be transmitted. |
| The command that should be transmitted when the 'Line 1' line of the LISN should be selected. If no command is specified, nothing will be transmitted. |
| The command that should be transmitted when the 'Line 2' line of the LISN should be selected. If no command is specified, nothing will be transmitted. |
| The command that should be transmitted when the 'Line 3' line of the LISN should be selected. If no command is specified, nothing will be transmitted. |
| Closes the dialog, where all the changed settings are stored. |
| Closes the dialog, where all the changed settings are discarded. |
| Opens an additional dialog which allows to set the correct communication medium to communicate with the device, including any relevant settings. |
Configurable Turn Table[edit]
Configurable commands[edit]
Check device[edit]
Device is initialised[edit]
Device is deinitialised[edit]
Some examples are:
| Received information | Regular expression | RadiMation® Readout | Note |
Communication[edit]
Identification[edit]
Antenna Tower[edit]
| The command that is send to stop the antenna tower. When left blank, no command will be send. |
| The command that is send to request the current height. When left blank no command will be send. |
| A regular expression to parse the response from the Current Height request. The received value will be interpreted as being the the unit in Meter, Centimeter or Millimeter, as selected by the dropdownbox. |
| The command that is send to move the antenna tower to a certain height. The command must include a keyword, __meter__, __centimeter__ or __millimeter__ which will be replaced with the corresponding height for the antenna tower. When left blank, no command will be sent. |
| The command that is send to request the status if the antenna tower is moving. When left blank, no command will be send. |
| A regular expression which will be matched against the response of the Position stopped command. When the result is matching with the response, the tower is stopped. |
| The command to request the current polarisation. It is mandatory to also supply the format of the polarization in the next fields. When left blank, no command will be sent. |
| A regular expression which will be matched against the response of the Current Polarization command. When the result is matching the driver reports Horizontal back to RadiMation. |
| A regular expression which will be matched against the response of the Current Polarization command. When the result is matching the driver reports Vertical back to RadiMation. |
| The command that is send to change the polarization to horizontal. When left blank, no command will be send. |
| The command that is send to change the polarization to vertical. When left blank, no command will be send. |
| The command that is send after the command Hor. Polarization is sent, with a delay of Waiting time between the two commands. When left blank, no command will be send. |
| The command that is send after the command Vert. Polarization is sent, with a delay of Waiting time between the two commands. When left blank, no command will be send. |
| The wait time, specified in milli seconds, between the Polarization command and Polarization stopped commands. |
Antenna Tower Advanced[edit]
| When checked, no height related commands and changes will be send to the antenna tower. This can for example be checked if the the antenna is mounted on a fixed height tripod. |
| When checked, no polarisation related commands and changes will be send to the Antenna Tower. This can for example be checked if the antenna tower does not have a automatic polariser. |
| When checked, the antenna will be moved to the Start position before the actual test is started. |
| The height, specified in centimeters, to which the antenna is moved, before the actual test is started. |
| When checked, the antenna will be moved to the End position after the test has finished. |
| The height, specified in centimeters, to which the antenna is moved, after the test has finished. |
| The time, specified in seconds, to wait after a command is issued to change the polarisation. This can for example be used to include an additional delay to ensure that the antenna polariser has reached the final location before the actual test continues. |
Configurable Cable[edit]
| keyword | inserted value |
| __freq__ | The signal generator frequency expressed in MHz |
| __freqHz__ | The signal generator frequency expressed in Hz |
| __freqkHz__ | The signal generator frequency expressed in kHz |
| __freqMHz__ | The signal generator frequency expressed in MHz |
| __freqGHz__ | The signal generator frequency expressed in GHz |
| __carrier__ | The signal generator carrier level expressed in dBm |
| __carrierdBm__ | The signal generator carrier level expressed in dBm |
| __carrierW__ | The signal generator carrier level expressed in Watt |
| __carriermW__ | The signal generator carrier level expressed in milli-Watt |
| __forward__ | The forward power expressed in dBm |
| __forwarddBm__ | The forward power expressed in dBm |
| __forwardW__ | The forward power expressed in Watt |
| __forwardmW__ | The forward power expressed in milli-Watt |
| __reflected__ | The reflected power expressed in dBm |
| __reflecteddBm__ | The reflected power expressed in dBm |
| __reflectedW__ | The reflected power expressed in Watt |
| __reflectedmW__ | The reflected power expressed in milli-Watt |
| __height__ | The height where the antenna is moving to expressed in metre |
| __heightm__ | The height where the antenna is moving to expressed in metre |
| __heightcm__ | The height where the antenna is moving to expressed in centimetre |
| __angle__ | The angle where the turntable is turning to expressed in degrees |
| __degree__ | The angle where the turntable is turning to expressed in degrees |
| __radian__ | The angle where the turntable is turning to expressed in radians |
| __calculatedvoltage__ | The calculated voltage expressed in V. |
| __<InformationItemName>__ | The value of the corresponding EUT or General information item. |
Messages[edit]
Unknown Device Driver[edit]
Device not connected[edit]
How to Report an Error[edit]
AD convertors[edit]
Type of communication: IEEE.
Channels 1:
Hewlett Packard 34401A[edit]
Channel 1
Type of measurement: AC Volt
Minimum value: 0Volt
Maximum value: 1 kVolt.
Channel 2:
Minimum value: 0 Amp.
Maximum value: 3 Amp.
Channel 3
Type of measurement: DC Volt
Channel 4:
Maximum value: 3 Amp. (rms)
Channel 5
Minimum value: 0 Ω
Maximum value: 100 MΩ
Channel 6
Minimum value: 0 Hz.
Maximum value: 300 kHz
Channel 7 and 8
Not used
Hewlett Packard 54600[edit]
Channel 1:
Type of measurement: V Max.
Maximum value: 1000 Volt.
Channel 2:
Type of measurement: V Min
Minimum value: 0 Volt.
Channel 3
Channel 4:
Channel 5
Minimum value: 0 Hz
Maximum value: 100 kHz
Channel 6
Type of measurement: Period
Minimum value: 0 ms
Maximum value: 10000 ms
Channel 7
Minimum value: 0 ms
Channel 8
Minimum value: 0 ms
Hewlett Packard 3562A[edit]
Type of communication: GPIB
Channels
Minimum value: 0 dB
Maximum value: 100 dB
Hewlett Packard 59313[edit]
Channel 1
Minimum value: 0
Maximum value: 2048
Channel 2:
Minimum value: 0
Maximum value: 2048
Channel 3
Minimum value: 0
Maximum value: 2048
Channel 4:
Minimum value: 0
Maximum value: 2048
Channel 5 to 8
Not used.
Channels
Marconi 2305[edit]
Channel 1
Minimum value: 0
Maximum value: 1000 MHz
Channel 2:
Minimum value: 0
Maximum value: 1000
Channel 3 to 8
Not used.
Fluke 45 AC Current[edit]
Channels
Minimum value: 0 mA
Maximum value: 10.000 mA
Fluke 45 DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 45 AC Voltage[edit]
Channels
Minimum value: 0 mV
Maximum value: 1.000.000 mV
Fluke 45 DC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 45 Frequency[edit]
Channels
Minimum value: 0 Hz
Maximum value: 1.000.000 Hz
Fluke 45 Resistance[edit]
Channels
Minimum value: 0 Ω
Maximum value: 100.000.000 Ω
Channel 1
Maximum value: 353.55 Volt.
Channel 2:
Channel 3
Channel 4:
Maximum value: 707.1 Volt
Channel 5
Type of measurement: RMS
Maximum value: 250 Volt
Channel 6
Maximum value: 200 MHz
Channel 7 and 8
Not used
Fluke 8840A AC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A AC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 8840A DC Voltage[edit]
Channels
Minimum value: 0 mV
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mV
Maximum value: 300.000 mV
Channels
Minimum value: 0 mV
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 8
Minimum value: 0
Maximum value: 16383
EIP 575[edit]
Channel 1
Minimum value: 0 Hz
Maximum value: 10 kHz
Channel 2:
Channel 3
Minimum value: 0 Hz
Maximum value: 1 MHz.
Channel 4:
Maximum value: 10 MHz.
Channel 5
Minimum value: 0 Hz
Maximum value: 100 MHz
Channel 6
Minimum value: 0 Hz
Maximum value: 1 GHz
Channel 7
Minimum value: 0 Hz
Maximum value: 10 GHz
Channel 8
Minimum value: 0 Hz
Maximum value: 100 GHz
Channel 1 to 8