(New page: Radimation® EMC software ------- Device drivers ------- Table of contents: 15 Device drivers 54 15.1 Introduction 54 15.2 Configurable device drivers vs. non configurable device dri...)
(180 intermediate revisions by 9 users not shown)
__NOTOC__
= Device drivers =
==Introduction==
Device drivers are available in {{Radimation}} to make it possible to communicate with thousands of different test and measurement equipment devices. They make the system flexible to use different devices, while the {{RadiMation}} Core stays generic. The {{RadiMation}} Core does not even know how to control specific devices it is controlling, only the type of device, and which device driver to use. The device driver is responsible for the communication between the {{RadiMation}} core and the specific test and measurement equipment, by sending the correct commands to the devices. If for example a new spectrum analyser or signal generator has been acquired, it is only needed to configure the corresponding device driver in the {{RadiMation}} software, and it will be able to use the new device. All the configurations of all the tests will remain the same.
This chapter only describes the device driver specific configuration. All the configuration of the device drivers that are managed by the {{RadiMation}} Core, are described in {{ManualLink|14|3. Device Driver configuration}}. Also the common device driver settings are described in {{ManualLink|14|Device Driver Settings}}.
== Virtual device drivers ==
For testing and demonstration purposes, virtual device drivers can be used and added as well.
Virtual device drivers are devices that act like normal devices but are emulated by software. These virtual devices do not physically exist, and thus do not control any real hardware. If a test site with only virtual devices is created, it can be used to perform completely virtual tests.
Table of contents:
==Active device drivers and passive device drivers==
{{RadiMation}} supports two types of device drivers: active and passive device drivers.
15.2 Configurable device drivers vs. non configurable device drivers 54
15.3 Configuration 54
15.3.1 New 54
15.3.2 Edit 54
15.3.3 Remove 54
15.3.4 RS 232 Setting 54
15.3.5 IEEE Setting 54
15.4 General Settings 54
15.4.1 Description 54
15.4.2 Brand 54
15.4.3 Type 54
15.4.4 ID 54
15.4.5 Serial number 54
15.4.6 Hardware version 54
15.4.7 Software version 54
15.4.8 Calibration Expires 54
15.4.9 Advanced 54
15.5 Device specific configuration 54
15.5.1 Signal generator 54
15.5.2 Amplifier 54
15.5.3 Antenna 54
15.5.4 Coupler 54
15.5.5 Powermeter 54
15.5.6 Field Sensor 54
15.5.7 AD converter 54
15.5.8 Injection device 54
15.5.9 Calibration Jigs 54
15.5.10 Current Sensor 54
15.5.11 Pre Amplifiers 54
15.5.12 Receivers / Spectrum analyser 54
15.5.13 LISN 54
15.5.14 Turn Table 54
15.5.15 Antenna Tower 54
15.5.16 Absorbing Clamps 54
15.5.17 Clamp positioner 54
15.5.18 Cables 54
15.5.19 Switch matrix 54
15.5.20 EUT Controller 54
15.6 Virtual devices 54
15.7 Configurable devices 54
15.7.1 Generic settings 54
15.7.2 Signal generator 54
15.7.3 Amplifier 54
15.7.4 Antenna 54
15.7.5 Powermeter 54
15.7.6 AD converter 54
15.7.7 Calibration Jigs 54
15.7.8 Current Sensor 54
15.7.9 Pre Amplifiers 54
15.7.10 Receivers / Spectrum analyser 54
15.7.11 LISN 54
15.7.12 Turn Table 54
15.7.13 Antenna Tower 54
15.7.14 Absorbing Clamps 54
15.7.15 Clamp positioner 54
15.7.16 Cables 54
15.7.17 Switch matrix 54
15.7.18 EUT Controller 54
15.8 Messages 54
15.8.1 This Device cannot be configured 54
15.8.2 Unknown Device Driver 54
15.8.3 GPIB: Device is not connected 54
15.8.4 Device not connected 54
15.9 How to make a buscapture 54
15.10 Ddlogger.dat 54
15.11 AD convertors 54
15.11.1 National Instruments 6023E 8 Analog inputs 54
15.11.2 Hewlett Packard 34401A 54
15.11.3 Hewlett Packard 54600 54
15.11.4 Hewlett Packard 3562A 54
15.11.5 Hewlett Packard 59313 54
15.11.6 Hewlett Packard 59313 54
15.11.7 Marconi 2305 54
15.11.8 Fluke 45 AC Current 54
15.11.9 Fluke 45 DC Current 54
15.11.10 Fluke 45 AC Voltage 54
15.11.11 Fluke 45 DC Voltage 54
15.11.12 Fluke 45 Frequency 54
15.11.13 Fluke 45 Resistance 54
15.11.14 LeCroy 9304AM Channel A,B,C,D 54
15.11.15 Fluke 8840A AC Current 54
15.11.16 Fluke 8840A DC Current 54
15.11.17 Fluke 8840A AC Voltage 54
15.11.18 Fluke 8840A DC Voltage 54
15.11.19 Hewlett Packard 3478A AC Current 54
15.11.20 Hewlett Packard 3478A DC Current 54
15.11.21 Hewlett Packard 3478A AC Voltage 54
15.11.22 Hewlett Packard 3478A DC Voltage 54
15.11.23 Tektronix TDS 400 Series 54
15.11.24 Tektronix TDS 500A Series 54
15.11.25 Tektronix TDS 600A Series 54
15.11.26 Tektronix TDS 3000 Series 54
15.11.27 D.A.R.E!! Development Radimate 2 and 3 54
15.11.28 EPI 575 54
15.11.29 Parallel Port Input 0x3BC and 0x378 54
Active devices are test and measurement equipment like [[powermeter]], [[signal generator]]s, [[spectrum analyser]]s, [[field sensor]]s, etc. For these devices the communication settings (like the IEEE address or the serial port) can be set which {{RadiMation}} has to use to be able to communicate with these devices.
15 Device drivers
15.1 Introduction
Passive devices are test and measurement equipment like [[coupler]]s, [[antenna]], [[cable]]s, [[calibration jig]]s, [[current sensor]]s, etc. Passive equipment do also need a device driver.
The reason for this is that {{RadiMation}} also has to know a number of details for these devices. Among others, the following information is relevant:
* Usable frequency range
* Maximum input power
* General information of the device, to include a list of used equipment during a test in the report
* Correction files that are relevant for the device
With this information {{RadiMation}} can prevent the test engineer from making large mistakes, like for example using a [[coupler]] in an unsupported frequency range.
Device drivers are the lowest layer in Radimation, they make it possible to communicate with thousands of different devices. They make the system flexible, while the Core stays generic. The Core does not even know which device it is controlling, only the type of device. So if you have bought a new spectrum analyser or signal generator you only have to tell the software to use this new device and all the rest stays the same.
==Device specific configuration==
===Signal generator===
Pressing advanced will open an IEEE configuration screen. Please view chapter IEEE setting for a complete description. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device.
Radimation has two types of device drivers, configurable and none configurable device drivers. Configurable devices for instance are powermeter, signal generators and spectrum analysers. For these devices you can set the IEEE address or the serial port which Radimation has to use to be able to communicate with these devices. None configurable devices are couplers and calibration jigs. You might think that it is useless to have non-configurable drivers, but the opposite has been proven many times. Non-configurable drivers still contain important information like the start and stop frequency of a device. With this information Radimation can prevent the test engineer from making large mistakes, like using the wrong coupler in a certain frequency range.
===Amplifier===
When an amplifier can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the amplifier can not be remotely controlled the message This Device cannot be configured will appear. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting and protecting this device.
15.3 Configuration
You can configure a device in the configuration screen. You can access the configuration screen by selecting configuration ==> configuration in the main menu. The window showed below will appear.
===Antenna===
When pressing advanced the message This Device cannot be configured will appear because this device can not be controlled remotely. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device.
You can select the type of device you want in the box that is situated top right. The big window will display all currently configured device drivers, like in the picture all configured AD Converters.
===Coupler===
15.3.1 New
If you press New a new window will appear, like the one below. Select the device you want and press New, if you do not whish to create a new device press Close.
===Powermeter===
[[Image:Power Meter Configuration Window.png]]
If you select New the window below will appear. You can now give the device a unique name, which will be used in the software.
{{ScreenElementDescription|Pre Wait time|    The minimum time that should be waited before the measurements starts.}}
{{ScreenElementDescription|Measure|            The minimum amount of measurements that {{RadiMation}} has to perform to determine if the <Max. Difference> condition is met.}}
{{ScreenElementDescription|Wait time|          The minimum time that {{RadiMation}} should be waiting between each measurement.}}
{{ScreenElementDescription|Max. Difference| The maximum difference that is allowed between the <Measure> measurement value(s); the highest and the lowest values are compared.}}
{{ScreenElementDescription|Max. Measure|    The maximum amount of measurements that ''could'' be performed by {{RadiMation}}, before there is a final measurement value available (which could meet these conditions, or not).}}
Now that all windows have been generally explained, the procedure {{RadiMation}} uses is the following. {{RadiMation}} takes the amount of measurements as defined in the Measure window. After that {{RadiMation}} determines the minimum, maximum and difference. When the difference is equal or smaller than defined in Max difference {{RadiMation}} determines the power. If the difference is greater than defined, {{RadiMation}} takes one new measurement. Replaces the oldest value with the
new one. Determines the minimum, maximum and difference again. This continues as long as the difference is larger then defined and the maximum amount of measurement has not yet been reached. If the maximum amount of measurement has been reached the last measurement is taken as the measured value.
15.3.2 Edit
To edit the configuration of a device, select the device in the list and press Edit.
Now two things can happen, either you seen an configuration screen of the device or an message box that tells you that you can not configure this device. The message box is explained in the chapter Messages and the configuration screen is explained in the chapter Device specific configuration
15.3.3 Remove
If you want to remove a device driver, select the device and press Remove. Please note that this driver will then not be available in the test sites, not even if you insert the exact same type of device with the exact the same name.
In the textbox you can type the COM port which Radimation has to use to able to communicate with the device. Baudrate, 7 or 8 bits and parity do not need to set because this is device specific.
15.3.5 IEEE Setting Basic
Primary Address
The primary address specifies the address of the device, so if the device you want to configure is on address is on 20 you enter 20. Please consult the manual of the device how to get the GPIB address of the device.
Show advanced configuration
Advanced configuration enables the user to setup the GPIB board. The configuration is discussed in “IEEE settings advanced” This option should only be used if the default settings are not sufficient enough.
15.3.6 IEEE Setting advanced
GPIB Board
With GPIB Board you can specify the GPIB board that is used. The default value is 0 and can only be different when multiple GPIB boards are present.
Primary Address
Secondary Address
The second address is default 0, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.
GPIB Delay
GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast pc’s (>2.5 GHz). This is most of the time noticeable when a driver is sometimes working and some times gives a GPIB(EABO or TIMO) error. These problems are mostly random. If this is your problem insert a GPIB delay time of 3000. Run the test again and see of the problem as disappeared. Is the problem has disappear then your problem was timing, if not please contact your reseller and report the problems you are having.
General Settings
General settings are settings that can be configured in every device driver. The picture below shows the setting of a signal generator.
15.3.7 Description
In the description window the description of the device can be changed/viewed. The description can be used in the report generator. Changing the description does not result in losing the device in the test sites. the description is automatically updated in test site. When generating a report with a test prior to the change the old description is inserted.
15.3.8 Brand
In the brand window the brand of the device can be changed/viewed. This brand can be used in the report generator.
15.3.9 Type
In the type window the brand of the device can be changed/viewed. This type can be used in the report generator.
{{Note|'''Speed up(1):''' Sometimes slower means faster. For instance, if you have a slower powermeter (resistor head) it may be quicker to have a longer waiting time. Because the powermeter has more time to determine the right value, {{RadiMation}} needs less measurements to determine the value. It is up to the engineer to find the right accuracy vs. time setting.
15.3.10 ID
In the ID window the internally used ID of the device can be changed/viewed. This ID can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction.
'''Speed up(2):''' When using a spectrum analyser a small amount of samples may be sufficient to determine the right amount of power. It is up to the engineer to find the right accuracy vs. time setting.
15.3.11 Serial number
'''Measuring:''' Powermeters which do not have a RMS detector, should not be used in Fast constant mode.
15.3.12 Hardware version
In the hardware version the brand of the device can be changed/viewed. This brand can be used in the report generator.
===Field Sensor===
15.3.13 Software version
==== Measurement setting panel ====
[[Image:Field Sensor Configuration Window.png]]
In the software version the brand of the device can be changed/viewed. This brand can be used in the report generator.
15.3.14 Calibration Expires
You can set the date that this device needs to be calibrated. Depending upon the setting in the configuration window Radimation will use this date. When you press the control a small window will appear, allowing you to select the new date.
Now that all windows and buttons have been generally explained, the procedure {{RadiMation}} uses is the following. {{RadiMation}} takes the amount of measurements as defined in the Measure window. After that {{RadiMation}} determines the minimum, maximum and difference. When the difference is equal or smaller than defined in Max difference {{RadiMation}} determines the field. If the difference is greater than defined, {{RadiMation}} takes one new measurement. Replaces the oldest value with the
15.3.15 Advanced
Pressing advanced opens the advanced menu of the device driver. Device specific configuration will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view chapter Configurable devices or Virtual devices for these devices.
15.4 Device specific configuration
15.4.1 Signal generator
Pressing advanced will open an IEEE configuration screen. Please view chapter IEEE setting for a complete description. See chapter 14.2 and 14.3 for correcting this device.
15.4.2 Amplifier
When an amplifier can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the amplifier can not be remotely controlled the message This Device cannot be configured will appear. See chapter 14.2 and 14.3 for correcting and protecting this device.
15.4.3 Antenna
When pressing advanced the message This Device cannot be configured will appear because this device can not be controlled remotely. See chapter 14.2 and 14.3 for correcting this device.
15.4.4 Coupler
15.4.5 Powermeter
Wait Time: Wait time is the time that Radimation may take for one measurement
Max Difference: The maximum allowed difference between the minimum and the maximum of measurements defined in the measurement window
Measure: Measure is the minimum amount of measurements Radimation has to make to determine the power.
Max Measure: Maximum measurement is the maximum amount of measurement Radimation may take to determine the power.
Now that all windows have been generally explained, the procedure Radimation uses is the following. Radimation takes the amount of measurements as defined in the Measure window. After that Radimation determines the minimum, maximum and difference. When the difference is equal or smaller then defined in Max difference Radimation determines the power. If the difference is greater then defined, Radimation takes one new measurement. Replaces the oldest value with the
{{Note|'''Speed up:''' Sometimes slower means faster. For instance, if you have a slower field sensor it may be quicker to have a longer waiting time. Because the field sensor has more time to determine the right value, {{RadiMation}} needs less measurements to determine the value. It is up to the engineer to find the right accuracy vs. time setting.}}
Some notes:
==== Axis setting panel ====
[[Image:Field sensor Advanced settings Axis Configuration.png‎]]
Speed up(1): Sometimes slower means faster. For instance, if you have a slower powermeter (resistor head) it may be quicker to have a longer waiting time. Because the powermeter has more time to determine the right value, Radimation needs less measurements to determine the value. It is up to the engineer to find the right accuracy vs. time setting.
The Axis configuration is a generic driver setting panel.
There are several (old) field sensors which don’t support the readout of an isotropic value: instead they provide the measurement values of all three axes separately.
Using the ‘Software isotropic’ setting in the device driver can then activate that the measurement data of all three axes are being interpreted and used to calculate the isotropic value and have this isotropic value being returned as the measured field strength.
When the field sensor already supports the retrieval of the isotropic measurement value, often this value is also returned when the software option is being selected in the device configuration: both isotropic as well as software isotropic will return the isotropic value of the field sensor. This is depending on the device driver and used equipment.
Measuring: Powermeter who doesn't have a RMS detector, should not be used in Fast constant mode.
15.4.6 Field Sensor
Measure: Measure is the minimum amount of measurements Radimation has to make to determine the field.
Max Measure: Maximum measurement is the maximum amount of measurement Radimation may take to determine the field.
Advanced: Opens the advanced menu, this can be RS 232, IEEE or device specific settings like Range mode.
Now that all windows and buttons have been generally explained, the procedure Radimation uses is the following. Radimation takes the amount of measurements as defined in the Measure window. After that Radimation determines the minimum, maximum and difference. When the difference is equal or smaller then defined in Max difference Radimation determines the field. If the difference is greater then defined, Radimation takes one new measurement. Replaces the oldest value with the
Speed up: Sometimes slower means faster. For instance, if you have a slower field sensor it may be quicker to have a longer waiting time. Because the field sensor has more time to determine the right value, Radimation needs less measurements to determine the value. It is up to the engineer to find the right accuracy vs. time setting.
15.4.7 AD converter
===AD converter===
15.4.8 Injection device
15.4.9 Calibration Jigs
===Injection device===
===Calibration Jigs===
15.4.10 Current Sensor
===Current Sensor===
15.4.11   Pre Amplifiers
===Pre Amplifiers===
15.4.12   Receivers / Spectrum analyser
===Receivers / Spectrum analyser===
15.4.13 LISN
When an LISN can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the LISN can not be remotely controlled the message This Device cannot be configured will appear. See chapter 14.2 and 14.3 for correcting this device.
15.4.14 Turn Table
===LISN===
When an LISN can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the LISN can not be remotely controlled the message This Device cannot be configured will appear. See {{ManualLink|14|Correction files}} and {{ManualLink|14|Correction file uses}} for correcting this device.
===Turn Table===
15.4.15 Antenna Tower
===Antenna Tower===
15.4.16 Absorbing Clamps
===Absorbing Clamps===
15.4.17 Clamp positioner
===Clamp positioner===
15.4.18 Cables
===Cables===
15.4.19 Switch matrix
===Switch matrix===
15.4.20 EUT Controller
===EUT Controller===
15.5 Virtual devices
===Auxiliary equipment===
When a Auxiliary equipment can be remotely controlled the specific window will appear. Please view chapter IEEE or RS232 setting for a complete description. When the Auxiliary equipment can not be remotely controlled the message This Device cannot be configured will appear.
Virtual devices are devices that act like normal devices but do not really exist. If you create a test site with only virtual devices, you can perform complete virtual tests. You might wonder why Radimation supports virtual device drivers. It is mainly used for debugging and solving software problems, but is has proven to be very useful when used as a temporarily workaround. For example a test site has an amplifier that is IEEE controlled and one day somebody accidentally destroys the communication between computer and amplifier. The test engineer has to perform some tests, he switches the driver for the amplifier for a virtual one. Sets the amplifier in operate by hand and was able to perform the tests that day. When the communication was repaired he changed the driver back again
== Communication interfaces and protocols ==
Different test and measurement instrumentation devices, can be controlled by different electronic communication buses and connectors, with on top of that different communication protocols. Which communication stream setting should be used to control a measurement instruments, cannot be easily based alone on the connector that is used to connect the measurement instrument. The following list shows the possible communication interfaces and protocols that are possible with a specific connector. This table can help to identify the actual communication protocol that is used. The manual of the test and measurement instrument also often lists the communication protocol that should be used, with similar terms, as is shown in the table. Based on the actual communication protocol the best communication stream to use is suggested.
15.6 Configurable devices
{|class="wikitable"
! Connector || Communication protocol || Suggested communication stream || Alternative communication stream
| [[GPIB]] / [[wikipedia:IEEE-488|IEEE-488]] || GPIB || [[#GPIB_Setting|GPIB]] || [[#VISA_Setting|VISA]] (with GPIB sub option)
|rowspan="5"| Ethernet / [[wikipedia:Rj45|RJ45]] || VXI-11 || [[#VISA_Setting|VISA]] (with LAN sub option) ||
| [[wikipedia:High Speed LAN Instrument Protocol|HiSLIP]] || [[#VISA_Setting|VISA]] (with LAN sub option) ||
| [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier)
| HTTP || [[#HTTP_Setting|HTTP]] ||
| UDP || [[#UDP_Setting|UDP]] ||
|rowspan="4"| [[wikipedia:USB|USB]]|| [[wikipedia:Virtual COM port|Virtual COM Port (VCP)]] (RS-232) || [[#RS-232_Setting|RS-232]] || [[#VISA_Setting|VISA]] (with RS-232 sub option)
| USBTMC (Test and Measurement Class) || [[#VISA_Setting|VISA]] (with Visa-ID sub option) ||
| Raditeq USB devices || [[#USB_Setting|USB]] || [[#RS-232_Setting|RS-232]]<BR/>[[#VISA_Setting|VISA]] (with RS-232 sub option)
| Device specific USB ||  || should be implemented device specifically
| [[wikipedia:D-subminiature|Sub-D9]] / [[wikipedia:RS-232|RS-232]] || RS-232 || [[#RS-232_Setting|RS-232]] || [[#VISA_Setting|VISA]] (with RS-232 sub option)
|rowspan="2"| other software || [[wikipedia:Network_socket#Types|Raw socket]] || [[#TCPIP_Setting|TCPIP]] || [[#VISA_Setting|VISA]] (with Visa-ID sub option with <code>::SOCKET</code> identifier)
The details for the configuration of each possible communication stream is described in the next section.
== Communication settings ==
=== GPIB Setting ===
[[Image:GPIB DeviceStream Configuration.png]]
2. Press the “New” button.
{{ScreenElementDescription|Primary Address|The primary address specifies the address of the device, so if the device you want to configure is on address 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device.}}
{{ScreenElementDescription|Use advanced configuration|Allows the end-user to enable the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient.}}
{{ScreenElementDescription|GPIB Board|With GPIB Board you can specify the GPIB board that is used. The default value is 0 and can only be different when multiple GPIB boards are present.}}
7. Press the “Edit” button.
{{ScreenElementDescription|Secondary Address|The second address is 0 by default, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.}}
8. A configuration screen for the device driver will appear. All required control commands for the device must be entered. Refer to the operating manual of the equipment for these codes.
{{ScreenElementDescription|GPIB Delay|GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PCs (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and sometimes gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see if the problem has disappeared. If the problem has disappeared then your problem was timing, if not please contact your reseller and report the problems you are having.}}
{{ScreenElementDescription|Clear device during initialisation|When checked, a low level GPIB command to reset the device will be sent to the device during initialisation.}}
{{ScreenElementDescription|Readdress device|When checked, a low level GPIB command to select the correct address will be transmitted every time a command is sent to the device.}}
=== HTTP Setting ===
The HTTP communication settings are used to specify the HTTP communication parameters, for the communication with a measurement device over HTTP. In this HTTP communication mode, the [[wikipedia:Hypertext Transfer Protocol|HyperText Transfer Protocol (HTTP)]] is used, and actual GET requests are performed.
[[File:HTTPDeviceStreamConfiguration.png]]
Remark:
Passive equipment such as antenna’s, current probes, cables etc. do also need a device driver. The reason for this is that RadiMation® has to know a number of parameters of these devices. Among others the following information is relevant:
{{ScreenElementDescription|Address|Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. By default the socket port number is 80, but this can be changed to another port if desired.}}
• Frequency range
If for example in a configurable device driver this HTTP communication stream is used, and the specified command is for example "SETA=1" (which set relay A of a Mini-Circuits relay to position 1), an actual "GET /SETA=1" HTTP request is transmitted to the specified hostname and socket port that is configured for the {{ScreenElement|Address}}.
• Maximum input power
15.6.1 Generic settings
=== Manual Setting ===
The manual communication stream can be used to show message boxes on the screen for the end-user, to manually perform some operation on a device. The commands that are specified in the device driver that is using the manual communication stream, are thus the messages that are shown to the end-user in a message box, for example with the instruction to change a setting on a test and measurement device.
Start and stop frequency
No further configuration in the manual communication stream is needed.
The most generic setting is the start and stop frequency. The default value for the start frequency is 1Hz and for the stop frequency is 40 GHz. The creator of the device is encouraged to set the right start and stop frequency. If the start and stop frequency are set correctly Radimation can warn the test engineer when he/she want to use the device out of its valid frequency range.
[[File:ManualDeviceStreamConfiguration.png]]
Reset
===RS-232 Setting===
The RS-232 communication settings specify the serial port parameters when communicating with a measurement device over RS-232. Although many modern devices connect via USB, they often appear as a virtual COM port (VCP). In that case you still configure the RS-232 settings as below.
[[Image:RS-232 DeviceStream Configuration.png]]
Init
{{ScreenElementDescription|COM Port|The COM port which {{RadiMation}} has to be used to communicate with the measurement device.}}
{{ScreenElementDescription|Baud rate|Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Data bits|Allows to configure the number of data bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Parity|Allows to configure the parity bit that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Stop bits|Allows to configure the number of stop bits that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Send termination|Allows to configure if and which terminator should be used during data transmissions to the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Receive termination|Allows to configure if and which terminator should be used during the receiving of data from the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
{{ScreenElementDescription|Receiving a NewLine ends read|Allows to configure if a read should be ended as soon as the receive terminator is received from the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.}}
In the init window you need to specify the string that the software needs to send when it want to Init  the device. For example *RST is commonly used reset string.
=== RSIB Setting ===
The RSIB communication settings are used to communicate with an ethernet connected measurement device through the [[Rohde & Schwarz]] RSISB32.DLL, which implements a VISA VXI-11 compatible layer.
Get ID
[[File:RSIBDeviceStreamConfiguration.png]]
In the Get ID window you need to specify the string that the software needs to send when it want to get the ID string of the device. For example *IDN? is commonly used Get ID string.
Returned ID
{{ScreenElementDescription|RSIB Address|Allows to specify RSIB identifier of the device that should be controlled.}}
=== TCPIP Setting ===
The TCPIP communication settings are used to specify that a (raw-)socket based [[wikipedia:TCP/IP|TCP/IP]] connection to a measurement device should be used.
[[Image:TCPIP DeviceStream Configuration.png]]
15.6.2 Signal generator
{{ScreenElementDescription|Address|Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. If the socket port number is a fixed port number, it is already shown as the default value, and it will be automatically added when no socket port number is specified.}}
Set Frequency
=== UDP Setting ===
The UPD communication settings are used to specify that an UDP based communication over ethernet should be used to control the measurement.
[[Image:UDPDeviceStreamConfiguration.png]]
Set Frequency is the string that needs to send to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz.
Example: “FRQ__freq__MHZ”.
=== USB Setting ===
The USB Communication settings are used to specify that a USB connection to a [[Raditeq]] or [[DARE!! Instruments]] measurement device is used. For devices that appear as a Virtual COM port (VCP) over USB, configure the [[Chapter_15#RS-232_Setting|RS-232 setting]] rather than the USB setting.
__freq__ will be replaced by the value that Radimation want to set the signal generator to.  The “__” of  “__freq__” are two “_”.
Set Carrier
{{ScreenElementDescription|Device Identifier|The device identifier (which is a unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.}}
{{ScreenElementDescription|Detect|Will automatically determine the correct {{ScreenElement|Device Identifier}} for the measurement device that is connected.}}
Set Carrier is the string that needs to send to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.
=== VISA Setting ===
The VISA Communication settings are used to specify the VISA communication parameters, for the communication with a measurement device. The VISA library is a higher level communication library that supports different kinds of communication methods. {{RadiMation}} doesn't provide a VISA library itself, and thus requires that a VISA library from another supplier like [[National Instruments]] or [[Keysight Technologies]] is installed. The VISA manager that is provided by that VISA library can be used to determine the correct VISA resource. The selected VISA configuration thus also has an influence if a measurement device is controlled by GPIB, RS-232, LAN, USBTMC, VXI-11, network-socket or another communication method .
[[Image:VISA DeviceStream Configuration.png]]
__carrier__ will be replaced by the value that Radimation want to set the signal generator to. The “__” of “__carrier__ ” are two “_”.
{{ScreenElementDescription|Alias|Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. [[National Instruments]] MAX or [[Keysight Technologies]] connection expert) that is installed on the PC.}}
{{ScreenElementDescription|GPIB|Allows to specify that GPIB should be used for the communication with the measurement device. The GPIB address of the measurement device should be specified.}}
{{ScreenElementDescription|LAN|Allows to specify that a VXI-11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.}}
{{ScreenElementDescription|RS-232|Allows to specify that a RS232 (ASRL in VISA terms) connection should be used for the communication with the measurement device. The COM port of the measurement device should be specified.}}
{{ScreenElementDescription|Visa-ID|Allows to specify a VISA Identifier that should be used for the communication with the measurement device. Any VISA Identifier that is supported by VISA is accepted. If one of the other communication methods is selected the corresponding VISA Identifier is also shown in this setting. Often used VISA identifiers are:
* <code>TCPIP[board]::<IP-address>::INSTR</code>: VXI-11 or LXI ethernet communication with the device with '<IP-address>'.
* <code>TCPIP[board]::<IP-address>::<port-number>::SOCKET</code>: Raw socket based ethernet communication on the port '<port-number>' with the the device '<IP-address>'
* <code>GPIB[board]::<primary address>::INSTR</code>: GPIB communication to the GPIB device with the address '<primary address>'
* <code>ASRL<port>::INSTR</code>: Serial communcation on COM-port '<port>'
Carrier on
==Configurable device drivers==
{{RadiMation}} allows the user to create their own device drivers for test and measurement equipment, which is (not yet) implemented in the standard device driver list.
Carrier on is the string that needs to send to set the signal generator carrier on.
Example: “OUTPUT ON”.
# In the device driver’s menu, select the required device driver type.
Carrier off
# Press the “New” button.
# Select the driver called “Configurable xx” (i.e. if you want to make a signal generator device driver you would select the “Configurable signal generator” device driver).
# Enter a description for the device driver (for example the type number of the generator) and press OK.
# The name of the device driver will be added in the available device drivers list.
# Select the new device driver from the available device drivers list.
# Press the “Edit” button.
# A configuration screen for the device driver will appear. For active controlled equipment all required control commands for the device must be entered. Refer to the operating manual of the equipment for these codes.
Carrier off is the string that needs to send to set the signal generator carrier off.
Example: “OUTPUT OFF”.
===Generic settings===
{{ScreenElementDescription|Start and stop frequency|The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, {{RadiMation}} can warn the test engineer when they want to use the device out of its valid frequency range.}}
{{ScreenElementDescription|Reset|In the reset window you need to specify the string that the software needs to send when it want to reset the device. For example *RST is commonly used reset string. If you don’t know the string then leave this window blank, and make sure that the device in a neutral state.}}
{{ScreenElementDescription|Init|In the init window you need to specify the string that the software needs to send when it want to Init  the device. For example *RST is commonly used reset string.}}
{{ScreenElementDescription|Get ID|In the Get ID window you need to specify the string that the software needs to send when it wants to get the ID string of the device. For example *IDN? is commonly used Get ID string.}}
{{ScreenElementDescription|Returned ID|In the Returned ID window you need to specify the string that the software will receive so that it knows that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.}}
AM on
=== Configurable Signal generator ===
[[Image:Configurable Signal Generator Window.png]]
AM on is the string that needs to send to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.
{{ScreenElementDescription|Set Frequency|Set Frequency is the string that needs to be sent to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz. }}
Example: “SOURCE2:FREQ  __freq__  KHZ;SOURCE:AM:DEPTH  __depth__
'''Example:''' “FRQ__freq__MHZ”.
                  PCT;SOURCE:AM:STATE ON”.
__freq__ will be replaced by the frequency and __depth__ by the modulation depth
'''Example:''' “FRQ__freqHz__HZ”.
“__freq__  ” are two “_”.
AM off
'''Example:''' “FRQ__freqkHz__KHZ”.
AM off is the string that needs to send to set the internal am modulation of the signal generator off.
'''Example:''' “FRQ__freqMHz__MHZ”.
'''Example:''' “FRQ__freqGHz__GHZ”.
External on
__freq__ will be replaced by the value that {{RadiMation}} want to set the signal generator to.  The “__” of  “__freq__” are two “_”.
External on is the string that needs to send to set the modulator to the external input.
{{ScreenElementDescription|Set Carrier|Set Carrier is the string that needs to be sent to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.}}
External off
'''Example:''' “:SOURCE:POWER __carrier__ DBM”.
External off is the string that needs to send to set the modulator to the internal input.
__carrier__ will be replaced by the value that {{RadiMation}} wants to set the signal generator to.  The “__” of  “__carrier__ ” are two “_”.
{{ScreenElementDescription|Carrier on|Carrier on is the string that needs to be sent to set the signal generator carrier on.}}
'''Example:''' “OUTPUT ON”.
PM on
{{ScreenElementDescription|Carrier off|Carrier off is the string that needs to be sent to set the signal generator carrier off.}}
'''Example:''' “OUTPUT OFF”.
PM on is the string that needs to send to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.
{{ScreenElementDescription|AM on|AM on is the string that needs to be sent to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.}}
Example: “SOURCE2:FREQ  __freq__  KHZ;SOURCE:PM:DUTY __duty__
'''Example:''' “SOURCE2:FREQ  __freq__  KHZ;SOURCE:AM:DEPTH __depth__ PCT;SOURCE:AM:STATE ON”.
                  PCT;SOURCE:PM:STATE ON”.
__freq__ will be replaced by the frequency and __duty__ by the duty cycle.
PM off
{{ScreenElementDescription|AM off|AM off is the string that needs to be sent to set the internal am modulation of the signal generator off.}}
'''Example:''' “SOURCE:AM:STATE OFF”.
PM off is the string that needs to send to set the internal PM modulation of the signal generator off.
{{ScreenElementDescription|External on|External on is the string that needs to be sent to set the modulator to the external input.}}
'''Example:''' “SOURCE:AM:SOURCE EXT”.
Sine Wave
Sine wave is the string that needs to be send when Radimation wants to set the wave form of the internal source to sine
{{ScreenElementDescription|External off|External off is the string that needs to be sent to set the modulator to the internal input.}}
Example: “SOURCE2:FUNC SIN”
'''Example:''' “SOURCE:AM:SOURCE INT”.
Square Wave
{{ScreenElementDescription|PM on|PM on is the string that needs to be sent to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.}}
Square wave is the string that needs to be send when Radimation wants to set the wave form of the internal source to square.
'''Example:''' “SOURCE2:FREQ  __freq__  KHZ;SOURCE:PM:DUTY  __duty__ PCT;SOURCE:PM:STATE ON”.
Example: “SOURCE2:FUNC SQU”
15.6.3 Amplifier
{{ScreenElementDescription|PM off|PM off is the string that needs to be sent to set the internal PM modulation of the signal generator off.}}
'''Example:''' “SOURCE:PM:STATE OFF”.
{{ScreenElementDescription|Sine Wave|Sine wave is the string that needs to be sent when {{RadiMation}} wants to set the wave form of the internal source to sine.}}
15.6.4 Antenna
'''Example:''' “SOURCE2:FUNC SIN”
{{ScreenElementDescription|Square Wave|Square wave is the string that needs to be sent when {{RadiMation}} wants to set the wave form of the internal source to square.}}
Coupler
'''Example:''' “SOURCE2:FUNC SQU”
15.6.5 Powermeter
=== Configurable Amplifier ===
{{:Configurable Amplifier}}
Set frequency
=== Configurable Antenna ===
{{:Configurable Antenna}}
Set Frequency is the string that needs to send to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.
=== Configurable Powermeter ===
[[Image:Configurable Power Meter Window.png]]
__freq__ will be replaced by the value that Radimation want to set the powermeter too. The “__” of  “__freq__” are two “_”.
{{ScreenElementDescription|Set frequency|Set Frequency is the string that needs to be sent to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.}}
Select Channel
Set Select Channel is the string that Radimation needs to send to set the powermeter channel. This is only necessary when the powermeter has more the one channel.
Example: “P1,U1”.
__freq__ will be replaced by the value that {{RadiMation}} want to set the powermeter too.  The “__” of  “__freq__” are two “_”.
Trigger
{{ScreenElementDescription|Select Channel|Set Select Channel is the string that {{RadiMation}} needs to send to set the powermeter channel. This is only necessary when the powermeter has more the one channel.}}
Trigger is the string that Radimation needs to send to trigger the powermeter.
'''Example:''' “P1,U1”.
Example: “X1”
{{ScreenElementDescription|Trigger|Trigger is the string that {{RadiMation}} needs to send to trigger the powermeter.}}
Result Format
'''Example:''' “X1”
Result Format is the string that Radimation needs to use to decode the value from the string send by the powermeter. The return value is interpreted in dBm
{{ScreenElementDescription|Result Format|Result Format is the string that {{RadiMation}} needs to use to decode the value from the string sent by the powermeter. The return value is interpreted in dBm.}}
Example: “__result__”
'''Example:''' “__result__”
__result__ will be replaced by the value that Radimation get from the powermeter.
__result__ will be replaced by the value that {{RadiMation}} get from the powermeter.
The “__” of  “__result__” are two “_”.
Start Zero
{{ScreenElementDescription|Start Zero|Start Zero is the string that {{RadiMation}} needs to send to start zeroing the powermeter.}}
Start Zero is the string that Radimation needs to send to start zeroing the powermeter.
Example: “O1”
Duration
Duration is the time that Radimation waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable result.
Field Sensor
'''Example:''' “O1”
15.6.6 AD converter
{{ScreenElementDescription|Duration|Duration is the time that {{RadiMation}} waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable results.}}
Currently unavailable.
=== Configurable AD converter ===
{{:Configurable_AD_Convertor}}
15.6.7 Calibration Jigs
=== Configurable Calibration Jig ===
{{:Configurable Calibration Jig}}
=== Configurable Current Sensor ===
15.6.8 Current Sensor
=== Configurable Pre Amplifier ===
15.6.9 Pre Amplifiers
=== Configurable Receivers / Spectrum analyser ===
=== Configurable LISN ===
{{:Configurable LISN}}
15.6.10 Receivers / Spectrum analyser
=== Configurable Turn Table===
{{:Configurable TurnTable}}
=== Configurable Antenna Tower===
{{:Configurable Antenna Tower}}
15.6.11 LISN
15.6.12 Turn Table
15.6.13 Antenna Tower
15.6.14 Absorbing Clamps
=== Configurable Absorbing Clamps===
15.6.15 Clamp positioner
=== Configurable Clamp positioner===
15.6.16 Cables
=== Configurable Cable===
15.6.17 Switch matrix
=== Configurable Switch matrix===
{{:Configurable Switch Matrix}}
15.6.18 EUT Controller
{{:Configurable EUT Controller}}
15.7 Messages
15.7.1 This Device cannot be configured
{{:Configurable equipment}}
==Messages ==
===This Device cannot be configured===
15.7.2 Unknown Device Driver
[[Image:Amplifier Device Driver Can Not Be Configured Window.png]]
This message box is displayed when Radimation is trying to locate a device driver and was unable to find it. If you see this message please contact your reseller and tell him which driver you are trying to use. The reseller will take action so that you will receive the right device driver.
===Unknown Device Driver===
15.7.3 GPIB: Device is not connected
[[Image:Unknown Device Driver Window.png]]
This message box is displayed when Radimation is unable to connect to a device when using GPIB. Please check the cable and the GPIB device driver address setting.
This message box is displayed when {{RadiMation}} is trying to locate a device driver and was unable to find it. If you see this message please contact your reseller and tell him which driver you are trying to use. The reseller will take action so that you will receive the right device driver.
15.7.4 Device not connected
===GPIB: Device is not connected===
This message box is displayed when Radimation is unable to connect to the device. Please check cables and device driver settings.
[[Image:GPIB Not Connected Window.png]]
15.8 How to make a buscapture
This message box is displayed when {{RadiMation}} is unable to connect to a device when using GPIB. Please check the cable and the GPIB device driver address setting.
===Device not connected===
Please verify your settings with the picture showed above and press OK.
Now start capturing by pressing the blue arrow pointing to the right. Start Radimation, and perform the test you need to capture. After the test is done, press the button with the red circle. Then select File and then save as to save to hard disk.
15.9 Ddlogger.dat
Ddlogger.dat is an output file for a debugging tool used by the software engineers at D.A.R.E!!. The tool is extremely useful for remote problem solving. The general idea is quite simple yet really helpful. Radimation inserts all the communication messages with devices in this file. So the software engineer can see what the software is sending and receiving when it communicates with the devices. With this information the engineer is more capable of reproducing/solving the problem, sometimes in combination with a buscapture. Do you need to install some additional software for this functionality? The answer is no, the tool is integrated in Radimation. But how do I turn it on or off? At startup Radimation looks for the ddlogger.dat file in the root of the drive. When it finds the file it will turn on the debugging tool, otherwise the debugging tool is turned off. What are the consequences for the performance? When the debugging tool is turned on, there is a small performance loss because there is additional writing to the harddisk. Therefore the default should be to run Radimation with the debugging tool turned off.
This message box is displayed when {{RadiMation}} is unable to connect to the device. Please check cables and device driver settings.
Step by step procedure
==How to Report an Error==
1. Make sure Radimation is not running.
When encountering a problem with the software you might like to report it to the RadiMation support.
2. Insert an empty text file in the root of the drive and rename it to “ddlogger.dat”. Make sure that the extension (“.txt”) is also changed.
The RadiMation error popups, contains a detailed section. This can be expanded with the details button on the error popup.
3. Start Radimation, during the start the following screen will appear.
In the expanded detailed error popup, a Report Error button is present. This allows Error Reporting to the RadiMation support within RadiMation.
More information about Error Reporting can be found here:
https://wiki.radimation.com/wiki/index.php/Error_Report
4. Select no.
[[File:ErrorPopupWindow.png]]
5. Perform the measurement you have problems with.
6. Close Radimation.
7. Send the ddlogger.dat file to D.A.R.E!!, preferably in a compressed file format (Zip)
8. Remove the ddlogger.dat file from your harddrive.
==AD convertors==
15.10 AD convertors
15.10.1 National Instruments 6023E 8 Analog inputs
===National Instruments 6023E 8 Analog inputs===
Type of communication: IEEE.
Channels 1:
Channels 1:
15.10.2 Hewlett Packard 34401A
===Hewlett Packard 34401A===
Line 621: Line 503:
Type of measurement: AC Volt
Minimum value: 0Volt
Maximum value: 1 kVolt.
Channel 2:
Channel 2:
Minimum value: 0 Amp.
Maximum value: 3 Amp.
Channel 3
Channel 3
Type of measurement: DC Volt
Channel 4:
Channel 4:
Maximum value: 3 Amp. (rms)
Line 645: Line 538:
Channel 5
Channel 5
Type of measurement: Resistance (Ohm)
Type of measurement: Resistance ({{Ohm}})
Minimum value: 0 Ohm
Maximum value: 100 M Ohm
Minimum value: 0 {{Ohm}}
Maximum value: 100 M{{Ohm}}
Channel 6
Channel 6
Minimum value: 0 Hz.
Maximum value: 300 kHz
Channel 7 and 8
Channel 7 and 8
Not used
Not used
15.10.3 Hewlett Packard 54600
===Hewlett Packard 54600===
Channel 1
Channel 1:
Type of measurement: V Max.
Maximum value: 1000 Volt.
Channel 2:
Channel 2:
Type of measurement: V Min
Minimum value: 0 Volt.
Channel 3
Channel 3
Channel 4:
Channel 4:
Channel 5
Channel 5
Minimum value: 0 Hz
Minimum value: 0 Hz
Maximum value: 100 kHz
Channel 6
Channel 6
Type of measurement: Period
Minimum value: 0 ms
Minimum value: 0 ms
Maximum value: 10000 ms
Channel 7
Channel 7
Minimum value: 0 ms
Minimum value: 0 ms
Channel 8
Channel 8
Minimum value: 0 ms
Minimum value: 0 ms
15.10.4 Hewlett Packard 3562A
===Hewlett Packard 3562A===
Type of communication: GPIB
Line 719: Line 642:
Minimum value: 0 dB
Minimum value: 0 dB
Maximum value: 100 dB
15.10.5 Hewlett Packard 59313
===Hewlett Packard 59313===
Line 730: Line 654:
Minimum value: 0
Minimum value: 0
Maximum value: 2048
Maximum value: 2048
Channel 2:
Channel 2:
Minimum value: 0
Minimum value: 0
Maximum value: 2048
Maximum value: 2048
Channel 3
Channel 3
Minimum value: 0
Minimum value: 0
Maximum value: 2048
Maximum value: 2048
Channel 4:
Channel 4:
Minimum value: 0
Minimum value: 0
Maximum value: 2048
Maximum value: 2048
Channel 5 to 8
Channel 5 to 8
Line 755: Line 691:
Not used.
Not used.
15.10.6 Hewlett Packard 59313
Line 768: Line 704:
Maximum value: 100
Maximum value: 100
15.10.7 Marconi 2305
===Marconi 2305===
Line 775: Line 711:
Minimum value: 0
Minimum value: 0
Maximum value: 1000 MHz
Channel 2:
Channel 2:
Minimum value: 0
Minimum value: 0
Maximum value: 1000
Maximum value: 1000
Channel 3 to 8
Channel 3 to 8
Not used.
Not used.
15.10.8 Fluke 45 AC Current
===Fluke 45 AC Current===
Line 796: Line 739:
Minimum value: 0 mA
Minimum value: 0 mA
Maximum value: 10.000 mA
15.10.9 Fluke 45 DC Current
===Fluke 45 DC Current===
Line 806: Line 751:
Minimum value: 0 mA
Minimum value: 0 mA
15.10.10 Fluke 45 AC Voltage
===Fluke 45 AC Voltage===
Line 816: Line 763:
Minimum value: 0 mV
Minimum value: 0 mV
Maximum value: 1.000.000 mV
15.10.11 Fluke 45 DC Voltage
===Fluke 45 DC Voltage===
Line 826: Line 775:
Minimum value: 0 mV
Minimum value: 0 mV
15.10.12 Fluke 45 Frequency
===Fluke 45 Frequency===
Line 836: Line 787:
Minimum value: 0 Hz
Minimum value: 0 Hz
Maximum value: 1.000.000 Hz
15.10.13 Fluke 45 Resistance
===Fluke 45 Resistance===
Line 845: Line 798:
Maximum value: 100.000.000 Ohm
15.10.14 LeCroy 9304AM Channel A,B,C,D
Maximum value: 100.000.000 {{Ohm}}
===LeCroy 9304AM Channel A,B,C,D===
Line 856: Line 811:
Maximum value: 353.55 Volt.
Channel 2:
Channel 2:
Channel 3
Channel 3
Channel 4:
Channel 4:
Maximum value: 707.1 Volt
Channel 5
Channel 5
Type of measurement: RMS
Maximum value: 250 Volt
Channel 6
Channel 6
Maximum value: 200 MHz
Channel 7 and 8
Channel 7 and 8
Line 893: Line 866:
Not used
Not used
15.10.15 Fluke 8840A AC Current
===Fluke 8840A AC Current===
Line 902: Line 875:
Minimum value: 0 mA
Minimum value: 0 mA
15.10.16 Fluke 8840A DC Current
===Fluke 8840A DC Current===
Line 912: Line 887:
Minimum value: 0 mA
Minimum value: 0 mA
15.10.17 Fluke 8840A AC Voltage
===Fluke 8840A AC Voltage===
Line 922: Line 899:
Minimum value: 0 mV
Minimum value: 0 mV
15.10.18 Fluke 8840A DC Voltage
===Fluke 8840A DC Voltage===
Line 932: Line 911:
Minimum value: 0 mV
Minimum value: 0 mV
15.10.19 Hewlett Packard 3478A AC Current
===Hewlett Packard 3478A AC Current===
Line 943: Line 923:
Minimum value: 0 mA
Minimum value: 0 mA
15.10.20 Hewlett Packard 3478A DC Current
===Hewlett Packard 3478A DC Current===
Line 953: Line 935:
Minimum value: 0 mA
Minimum value: 0 mA
15.10.21 Hewlett Packard 3478A AC Voltage
===Hewlett Packard 3478A AC Voltage===
Line 963: Line 947:
Minimum value: 0 mV
Minimum value: 0 mV
Maximum value: 300.000 mV
15.10.22 Hewlett Packard 3478A DC Voltage
===Hewlett Packard 3478A DC Voltage===
Line 973: Line 959:
Minimum value: 0 mV
Minimum value: 0 mV
15.10.23 Tektronix TDS 400 Series
===Tektronix TDS 400 Series===
Line 984: Line 971:
Minimum value: 0
Minimum value: 0
Maximum value: 100
Maximum value: 100
Channels 5 to 8
Channels 5 to 8
Not used
Not used
15.10.24 Tektronix TDS 500A Series
===Tektronix TDS 500A Series===
Line 998: Line 988:
Minimum value: 0
Minimum value: 0
Maximum value: 100
Maximum value: 100
Channels 5 to 8
Channels 5 to 8
Not used
Not used
15.10.25 Tektronix TDS 600A Series
===Tektronix TDS 600A Series===
Line 1,012: Line 1,005:
Minimum value: 0
Minimum value: 0
Maximum value: 100
Maximum value: 100
Channels 5 to 8
Channels 5 to 8
Line 1,018: Line 1,013:
Not used
Not used
15.10.26 Tektronix TDS 3000 Series
===Tektronix TDS 3000 Series===
Line 1,027: Line 1,022:
Minimum value: 0
Minimum value: 0
Maximum value: 100
Maximum value: 100
Channels 5 to 8
Channels 5 to 8
Line 1,033: Line 1,030:
Not used
Not used
15.10.27 D.A.R.E!! Development Radimate 2 and 3
===DARE!! Development Radimate 2 and 3===
Line 1,042: Line 1,039:
Minimum value: 0
Minimum value: 0
Maximum value: 16383
15.10.28 EPI 575
===EIP 575===
Line 1,051: Line 1,049:
Minimum value:  0 Hz
Minimum value:  0 Hz
Maximum value: 10 kHz
Channel 2:
Channel 2:
Channel 3
Channel 3
Minimum value: 0 Hz
Minimum value: 0 Hz
Maximum value: 1 MHz.
Channel 4:
Channel 4:
Maximum value: 10 MHz.
Channel 5
Channel 5
Minimum value: 0 Hz
Minimum value: 0 Hz
Maximum value: 100 MHz
Channel 6
Channel 6
Minimum value: 0 Hz
Minimum value: 0 Hz
Maximum value: 1 GHz
Channel 7
Channel 7
Minimum value: 0 Hz
Minimum value: 0 Hz
Maximum value: 10 GHz
Channel 8
Channel 8
Minimum value: 0 Hz
Minimum value: 0 Hz
Maximum value: 100 GHz
15.10.29 Parallel Port Input 0x3BC and 0x378
===Parallel Port Input 0x3BC and 0x378===
Channel 1 to 8
Channel 1 to 8
[[Category:Manual]]
Device drivers[edit]
Virtual device drivers[edit]
  • Usable frequency range
Signal generator[edit]
Amplifier[edit]
Antenna[edit]
Coupler[edit]
Powermeter[edit]
Measuring: Powermeters which do not have a RMS detector, should not be used in Fast constant mode.
Field Sensor[edit]
Axis setting panel[edit]
AD converter[edit]
Injection device[edit]
Calibration Jigs[edit]
Current Sensor[edit]
Pre Amplifiers[edit]
LISN[edit]
Turn Table[edit]
Antenna Tower[edit]
Absorbing Clamps[edit]
Clamp positioner[edit]
Cables[edit]
Switch matrix[edit]
EUT Controller[edit]
Auxiliary equipment[edit]
Connector Communication protocol Suggested communication stream Alternative communication stream
GPIB / IEEE-488 GPIB GPIB VISA (with GPIB sub option)
Ethernet / RJ45 VXI-11 VISA (with LAN sub option)
HiSLIP VISA (with LAN sub option)
Raw socket TCPIP VISA (with Visa-ID sub option with ::SOCKET identifier)
HTTP HTTP
UDP UDP
USB Virtual COM Port (VCP) (RS-232) RS-232 VISA (with RS-232 sub option)
USBTMC (Test and Measurement Class) VISA (with Visa-ID sub option)
Raditeq USB devices USB RS-232
Device specific USB should be implemented device specifically
Sub-D9 / RS-232 RS-232 RS-232 VISA (with RS-232 sub option)
other software Raw socket TCPIP VISA (with Visa-ID sub option with ::SOCKET identifier)
Communication settings[edit]
GPIB Setting[edit]
ScreenElementDescription.svg Primary Address The primary address specifies the address of the device, so if the device you want to configure is on address 20 you enter 20. Please consult the manual of the device how to determine the GPIB address of the device.
ScreenElementDescription.svg Use advanced configuration Allows the end-user to enable the more advanced configuration parameters. Those advanced option should only be used if the default settings are not sufficient.
ScreenElementDescription.svg Secondary Address The second address is 0 by default, and should only be changed when needed. For further information please look in the help of National Instruments 488.2.
ScreenElementDescription.svg GPIB Delay GPIB delay is the delay between GPIB reading and writing actions. Some older IEEE 488.2 machines have difficulty communicating with fast PCs (> 2.5 GHz). This is most of the time noticeable when a driver is sometimes working and sometimes gives a GPIB (EABO or TIMO) error. These errors are most of the time, generated randomly. Specifying a GPIB delay time of 3000 uSeconds can fix these random errors. Run the test again and see if the problem has disappeared. If the problem has disappeared then your problem was timing, if not please contact your reseller and report the problems you are having.
ScreenElementDescription.svg Clear device during initialisation When checked, a low level GPIB command to reset the device will be sent to the device during initialisation.
ScreenElementDescription.svg Readdress device When checked, a low level GPIB command to select the correct address will be transmitted every time a command is sent to the device.
HTTP Setting[edit]
ScreenElementDescription.svg Address Allows to specify the IP-address or the (FQDN) hostname and the socket port number of the measurement device. This is normally done with a string like: "tcpip://<address-or-name>:<port-number>", where '<address-or-name>' is the IP address or hostname, and '<port-number>' is the socket port number on which the connection should be initiated. By default the socket port number is 80, but this can be changed to another port if desired.
Manual Setting[edit]
RS-232 Setting[edit]
ScreenElementDescription.svg COM Port The COM port which RadiMation® has to be used to communicate with the measurement device.
ScreenElementDescription.svg Baud rate Allows to configure the baud rate that should be used to communicate with the measurement device. This setting can be a fixed device specific value, in which case it cannot be configured and then the setting will be disabled.
RSIB Setting[edit]
ScreenElementDescription.svg RSIB Address Allows to specify RSIB identifier of the device that should be controlled.
TCPIP Setting[edit]
UDP Setting[edit]
USB Setting[edit]
ScreenElementDescription.svg Device Identifier The device identifier (which is a unique identifier of 8 groups of digits) that identifies the measurement device that is connected over USB.
VISA Setting[edit]
ScreenElementDescription.svg Alias Allows to specify a VISA Alias that should be used for the communication with the measurement device. Any Alias that is supported by VISA is accepted. The correct Alias can be determined by using the VISA manager (eg. National Instruments MAX or Keysight Technologies connection expert) that is installed on the PC.
ScreenElementDescription.svg LAN Allows to specify that a VXI-11 or LXI connection should be used for the communication with the measurement device. It is possible to specify the IP-address or the (FQDN) hostname of the measurement device.
  • TCPIP[board]::<IP-address>::INSTR: VXI-11 or LXI ethernet communication with the device with '<IP-address>'.
  • TCPIP[board]::<IP-address>::<port-number>::SOCKET: Raw socket based ethernet communication on the port '<port-number>' with the the device '<IP-address>'
  5. The name of the device driver will be added in the available device drivers list.
  6. Select the new device driver from the available device drivers list.
Generic settings[edit]
ScreenElementDescription.svg Start and stop frequency The default value for the start frequency is 1 Hz and for the stop frequency is 120 GHz. The creator of the device is encouraged to set the correct start and stop frequency. If the start and stop frequency are set correctly, RadiMation® can warn the test engineer when they want to use the device out of its valid frequency range.
ScreenElementDescription.svg Get ID In the Get ID window you need to specify the string that the software needs to send when it wants to get the ID string of the device. For example *IDN? is commonly used Get ID string.
ScreenElementDescription.svg Returned ID In the Returned ID window you need to specify the string that the software will receive so that it knows that it has the right device. For example "Hewlett_Packard,8643A," can be used for the Hewlett Packard 8643A. If you leave this window empty then all returned strings are accepted.
ScreenElementDescription.svg Set Frequency Set Frequency is the string that needs to be sent to set the signal generator frequency. The unit is in MHZ, so the string should be made for MHz.
Example: “FRQ__freqHz__HZ”.
Example: “FRQ__freqkHz__KHZ”.
Example: “FRQ__freqMHz__MHZ”.
Example: “FRQ__freqGHz__GHZ”.
ScreenElementDescription.svg Set Carrier Set Carrier is the string that needs to be sent to set the signal generator carrierlevel. The unit is in dBm, so the string should be made for dBM.
ScreenElementDescription.svg Carrier on Carrier on is the string that needs to be sent to set the signal generator carrier on.
ScreenElementDescription.svg Carrier off Carrier off is the string that needs to be sent to set the signal generator carrier off.
ScreenElementDescription.svg AM on AM on is the string that needs to be sent to set the internal AM generator of the signal generator. The unit of frequency is KHz and the unit of Modulation Depth is %.
ScreenElementDescription.svg AM off AM off is the string that needs to be sent to set the internal am modulation of the signal generator off.
ScreenElementDescription.svg External on External on is the string that needs to be sent to set the modulator to the external input.
ScreenElementDescription.svg External off External off is the string that needs to be sent to set the modulator to the internal input.
ScreenElementDescription.svg PM on PM on is the string that needs to be sent to set the internal PM generator of the signal generator. The unit of frequency is KHz and the unit of duty cycle is %.
ScreenElementDescription.svg PM off PM off is the string that needs to be sent to set the internal PM modulation of the signal generator off.
ScreenElementDescription.svg Sine Wave Sine wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to sine.
ScreenElementDescription.svg Square Wave Square wave is the string that needs to be sent when RadiMation® wants to set the wave form of the internal source to square.
Configurable Amplifier[edit]
Main[edit]
Operation[edit]
Example[edit]
1 = RF ON
1 = BAND 2
1 = LINE ON
Configurable Antenna[edit]
Configurable Powermeter[edit]
ScreenElementDescription.svg Set frequency Set Frequency is the string that needs to be sent to set the powermeter frequency. The unit is in MHz, so the string should be made for MHz.
Example: “FRQ__freq__MHZ”.
Example: “P1,U1”.
Example: “X1”
ScreenElementDescription.svg Result Format Result Format is the string that RadiMation® needs to use to decode the value from the string sent by the powermeter. The return value is interpreted in dBm.
Example: “O1”
ScreenElementDescription.svg Duration Duration is the time that RadiMation® waits so that the powermeter can zero properly. Make sure that this time is big enough, an incorrect value may result in unpredictable results.
Channels[edit]
Read Back examples[edit]
Communication[edit]
Configurable LISN[edit]
Configurable Turn Table[edit]
Configurable commands[edit]
Check device[edit]
Device is initialised[edit]
Device is deinitialised[edit]
Some examples are:
Communication[edit]
Identification[edit]
Antenna Tower[edit]
Antenna Tower Advanced[edit]
Currently unavailable.
Configurable Cable[edit]
sending the next command.
keyword inserted value
__calculatedvoltage__ The calculated voltage expressed in V.
Messages[edit]
Unknown Device Driver[edit]
Device not connected[edit]
How to Report an Error[edit]
AD convertors[edit]
Type of communication: IEEE.
Channels 1:
Hewlett Packard 34401A[edit]
Channel 1
Type of measurement: AC Volt
Minimum value: 0Volt
Maximum value: 1 kVolt.
Channel 2:
Minimum value: 0 Amp.
Maximum value: 3 Amp.
Channel 3
Type of measurement: DC Volt
Channel 4:
Maximum value: 3 Amp. (rms)
Channel 5
Minimum value: 0 Ω
Maximum value: 100 MΩ
Channel 6
Minimum value: 0 Hz.
Maximum value: 300 kHz
Channel 7 and 8
Not used
Hewlett Packard 54600[edit]
Channel 1:
Type of measurement: V Max.
Maximum value: 1000 Volt.
Channel 2:
Type of measurement: V Min
Minimum value: 0 Volt.
Channel 3
Channel 4:
Channel 5
Minimum value: 0 Hz
Maximum value: 100 kHz
Channel 6
Type of measurement: Period
Minimum value: 0 ms
Maximum value: 10000 ms
Channel 7
Minimum value: 0 ms
Channel 8
Minimum value: 0 ms
Hewlett Packard 3562A[edit]
Type of communication: GPIB
Channels
Minimum value: 0 dB
Maximum value: 100 dB
Hewlett Packard 59313[edit]
Channel 1
Minimum value: 0
Maximum value: 2048
Channel 2:
Minimum value: 0
Maximum value: 2048
Channel 3
Minimum value: 0
Maximum value: 2048
Channel 4:
Minimum value: 0
Maximum value: 2048
Channel 5 to 8
Not used.
Channels
Marconi 2305[edit]
Channel 1
Minimum value: 0
Maximum value: 1000 MHz
Channel 2:
Minimum value: 0
Maximum value: 1000
Channel 3 to 8
Not used.
Fluke 45 AC Current[edit]
Channels
Minimum value: 0 mA
Maximum value: 10.000 mA
Fluke 45 DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 45 AC Voltage[edit]
Channels
Minimum value: 0 mV
Maximum value: 1.000.000 mV
Fluke 45 DC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 45 Frequency[edit]
Channels
Minimum value: 0 Hz
Maximum value: 1.000.000 Hz
Fluke 45 Resistance[edit]
Channels
Minimum value: 0 Ω
Maximum value: 100.000.000 Ω
Channel 1
Maximum value: 353.55 Volt.
Channel 2:
Channel 3
Channel 4:
Maximum value: 707.1 Volt
Channel 5
Type of measurement: RMS
Maximum value: 250 Volt
Channel 6
Maximum value: 200 MHz
Channel 7 and 8
Not used
Fluke 8840A AC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A DC Current[edit]
Channels
Minimum value: 0 mA
Fluke 8840A AC Voltage[edit]
Channels
Minimum value: 0 mV
Fluke 8840A DC Voltage[edit]
Channels
Minimum value: 0 mV
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mA
Channels
Minimum value: 0 mV
Maximum value: 300.000 mV
Channels
Minimum value: 0 mV
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 4
Minimum value: 0
Maximum value: 100
Channels 5 to 8
Not used
Channels 1 to 8
Minimum value: 0
Maximum value: 16383
EIP 575[edit]
Channel 1
Minimum value: 0 Hz
Maximum value: 10 kHz
Channel 2:
Channel 3
Minimum value: 0 Hz
Maximum value: 1 MHz.
Channel 4:
Maximum value: 10 MHz.
Channel 5
Minimum value: 0 Hz
Maximum value: 100 MHz
Channel 6
Minimum value: 0 Hz
Maximum value: 1 GHz
Channel 7
Minimum value: 0 Hz
Maximum value: 10 GHz
Channel 8
Minimum value: 0 Hz
Maximum value: 100 GHz
Channel 1 to 8