Line 430: Line 430:
[[Image:Calibration_jig_factor_9.5_dB_for_150_Ohm_system.png]]
[[Image:Calibration_jig_factor_150_Ohm_for_150_Ohm_system.png]]
[[Image:Calibration_jig_factor_0_dB_for_50_Ohm_system.png]]
[[Image:Calibration_jig_factor_50_Ohm_for_50_Ohm_system.png]]
===Current sensor impedance===
Revision as of 11:25, 16 March 2015
Software configuration[edit]
In the main menu, under the Configuration pull down menu, a number of standard configurations can be made. The selected and / or entered configurations will be used as default.
The next paragraphs will explain all possible configurations.
Warning: To prevent unauthorised persons to change system configurations, a password protection can be used. The password is by default activated with the default password: “Radimation”. This password protection can be switched off, changed and configured under:
   Menu.svg Configuration
      Menu.svg Password
Test engineers[edit]
      Menu.svg Test engineers
In this window, test engineer information can be changed.
  • The default client
  • The default manufacturer
  • The default test site
Data entered under “default address information”, will automatically be entered in the EUT window and must therefore only be entered once.
      Menu.svg Default address information
Default Units[edit]
  • Power : dBuW, dBm, dBmW, dBW, uW, mW and W
  • Field Strength : dBuV/m, dBmV/m, dBV/m, uV/m, mV/m and V/m
  • Voltage : dBuVrms, dBmVrms, dBVrms, uVrms, mVrms and Vrms
  • Current : dBuA, dBmA, dBA, uA, mA and A
  • Emission : dBuV, dBmv, dBV, V, W, dBuV/m, dBmV/m and V/m
  • And many more
The RadiMation® software can use different directories to store test data (*.EUT files). The following file types are used:
  • Sequence files (*.SEQ files)
Under "RadiMation® file types and locations" the different file types are described. For each data type the default path name can be entered through the configuration, directories pull down menu.
The path name can be either a local directory or a directory on a network drive. The second makes it possible to store EUT data central on a computer network.
To modify the path name select the:
         Menu.svg Directories
Device drivers are used to let your software know which equipment has to be used and controlled during EMC tests.
In RadiMation®, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information like the IEEE address or RS232 COM-port has to be entered.
You can configure a device in the configuration screen. You can access the configuration screen by selecting:
         Menu.svg Device Drivers
You can select the type of device you want in the box that is situated top right. The big window will display all currently configured device drivers, like in the picture all configured Signal Generators.
ScreenElementDescription.svg Device Driver Type Different types of devices are available in RadiMation®. Only the created device drivers of the type that is selected for Device Driver Type will be shown in the list
ScreenElementDescription.svg Available Device Drivers Shows a list of all the available device drivers for the selected Device Driver Type
ScreenElementDescription.svg New Creates a new RadiMation® device driver for a specific device. It is only possible to create a new device driver for the selected Device Driver Type.
ScreenElementDescription.svg Edit To edit the configuration of a device, select the device in the list and press Edit. This allows to change the generic set of device driver settings of the selected device driver from the list of Available Device Drivers. See Chapter 14: Device Driver settings for a detailed description of these settings.
ScreenElementDescription.svg Remove Removes the configuration of the selected device driver from the list of Available Device Drivers. Please note that this driver will then not be available anymore in the test sites. Even if a new device driver is created with the exact same name, it will not automatically reappear in the test sites.
The required device driver can be selected from this list. The new device driver will then be created when New button is pressed. At that moment it is possible to use the default name of the device driver, or to change it to another (eg. more specific) name.
If you do not whis to create a new device driver press Close.
Device Driver settings[edit]
All device drivers have a generic set of settings that can be configured for every device driver. The picture below shows an example.
ScreenElementDescription.svg Software version In the software version the brand of the device can be changed/viewed. This brand can be used in the report generator.
ScreenElementDescription.svg Description In the description window the description of the device can be changed/viewed. The description can be used in the report generator. Changing the description does not result in losing the device in the test sites. the description is automatically updated in test site. When generating a report with a test prior to the change the old description is inserted.
ScreenElementDescription.svg Brand In the brand window the brand of the device can be changed/viewed. This brand can be used in the report generator.
ScreenElementDescription.svg Hardware version In the hardware version the brand of the device can be changed/viewed. This brand can be used in the report generator.
ScreenElementDescription.svg Serial number In the brand window the brand of the device can be changed/viewed. This brand can be used in the report generator.
ScreenElementDescription.svg Type In the type window the brand of the device can be changed/viewed. This type can be used in the report generator.
ScreenElementDescription.svg Device driver DLL Version Shows the exact version of the DLL in which the device driver is contained. This is an unique identifier that specifies the version of the device driver. The contents of this cell cannot be changed
ScreenElementDescription.svg ID In the ID window the internally used ID of the device can be changed/viewed. This ID can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction.
ScreenElementDescription.svg Device drivers installation date Shows the date and time at which the last update of the device drivers has been installed. The contents of this cell cannot be changed
ScreenElementDescription.svg Device drivers versions Shows the version number of the last installed device drivers update. The contents of this cell cannot be changed
ScreenElementDescription.svg Add The Add button can be used to add additional information items to the general device driver settings. The additional items will be shown in the list. Additional information items can for example be used to note the storage location, the location of the user manual or something else.
ScreenElementDescription.svg Edit When the Edit button is pressed, when an information item is selected, it is possible to change the value of the information item. Information items that cannot be modified, are not allowed to be edited.
ScreenElementDescription.svg Remove When the Remove button is pressed, the selected information item will be removed from the list. Only custom information items can be removed. Fixed information items, like described above, cannot be removed.
ScreenElementDescription.svg Zero Interval When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be rezeroed. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation however generally takes several seconds to be completed, and it may be valid for a few hours. To prevent that a device is rezeroed during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zeroed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval, can be a large reduction of the time that is needed to initialize each test.
This setting is only available for powermeter and field strength sensor devices.
ScreenElementDescription.svg Calibration Expire Date This field can be set to the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window RadiMation® will use this date, to give warnings/notifications of the recalibration. When you press the control a small window will appear, allowing you to select the new date.
ScreenElementDescription.svg Advanced Pressing Advanced opens the advanced menu of the device driver. Device specific configuration will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view chapter Configurable devices or Virtual devices for these devices.
ScreenElementDescription.svg Check Pressing Check performs a check if the device is connected. This check is the same as the check that is performed during each test initialization. This check can be used to determine if the correct addressing is set, and if communication with the device is possible.
ScreenElementDescription.svg Knowledgebase Pressing the Knowledgebase button, opens a new web-browser window which will show the RadiWiki page that is specific for that device driver. That RadiWiki page can contain some additional information or a detailed explanation of all advanced settings.
Graphics configuration[edit]
  • Background colour
  • Grid colour
  • Text colour
  • Grid style (solid line or dashed)
  • Default linear or log scale (during each test the user is able to switch between linear log scales).
To configure the graphical interface please select:
Now select the “Graphics” TAB and make all required selections.
Database Configuration[edit]
Select
   Menu.svg Start
      Menu.svg Settings
         Menu.svg Control Panel
            Menu.svg ODBC Data Sources
Configuration of the characterset[edit]
Characterset value
UTF-8 (without bom) CharacterSet=65001
UTF-16 LE (without bom) CharacterSet=Unicode
MySQL ODBC database[edit]
MSSQL ODBC database[edit]
Language Selection[edit]
RadiMation® supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require.
         Menu.svg Language
The selected language is activated after the RadiMation® package is restarted.
Please note that not all languages are available. Please contact your local reseller for language availability.
Measurement Settings[edit]
The measurement settings are settings to assist the test engineer with the measurement. With Starting Level Signal Generatorhe engineer can determine the starting level of the signal generator. For example increase the starting level so the test is starting just above the noise level.
Warning: It is the responsibility of the test engineer that the Starting Level Signal Generator is set correctly. RadiMation® nor its producers can not be held responsible for setting this value wrong, and any damages / injuries resulting from setting this value wrong. The most optimal setting is just below the noise level of the power meters. Please contact your contact person when in doubt about the correct setting of this value.
If Perform Power delivery test after pause is checked, RadiMation® performs a power delivery test after leaving manual mode during immunity testing.
RadiMation® can warn or even stop a measurement when a used device is out of calibration. In the top combo box you can select if RadiMation® only warns you that a device is out of calibration or, if it should stop the measurement. With the second combo you can specify the amount days RadiMation® should warn you before the calibration expires. Please note that you type in the value yourself or select one of the predefined periods.
With the checkboxes in the Required environment settings you can define the environment settings that need to be set for every measurement. For sequences the engineer only has to set the selected values once.
The error message window will ask you to abort the test or to ignore the error message and continue the test with different test equipment. Of course, to achieve maximum accuracy, one should use the same test equipment during the calibration and during the substitution tests. However, in some circumstances, changing test equipment between calibration and test will not have any influence on test results. This is for example true when a RI substitution list is made using the forward power calibration file. Changing the amplifier won’t change test results as long as the new amplifier covers the required frequency range and can deliver the required output power.
To add, modify or delete test equipment lists, select the:
      Menu.svg Test equipment
Password protection[edit]
To activate, modify or de-activate passwords, select the:
      Menu.svg Engineers
         Menu.svg Edit
            Menu.svg Change Password
In the engineer edit window, the password can be enabled, disabled and changed (to disable or change the password, the current password is required). Furthermore, in this engineer window, for the following items, the protection can be enabled or disabled:
Test equipment list, prevents unauthorised changes to test equipment configuration Test engineers, prevents unauthorised changes to test engineers configuration Technical Setup Files (TSF files), prevents unauthorised changes tot TSF files.
Correction files[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction file
Entering data points[edit]
When a data point is incorrect, it can be removed by selecting the data point and pressing the “Remove row” button.
Saving correction files[edit]
   Menu.svg File
      Menu.svg Save correction file
Attaching a correction files[edit]
Most correction files are used for correcting a measured value of a device driver. The made correction file should therefor be 'attached' to the correct device driver.
      Menu.svg configuration
You can select the type of device you want in the box that is situated top right. The big window will display all currently configured device drivers, like in the picture all configured AD Converters. To edit the configuration of a device, select the device in the list and press Edit.
In the lower part of this window, the made correction files can be selected, for one of the specified uses. The RadiMation® software will use the specified correction files, every time when the device driver is used during a test. Depending on the type of the device driver, different uses of the correction files are available.
Correction file uses[edit]
Signal Generator Output Correction[edit]
  • Frequency
  • Correction
Amplifier Input Protection Level[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Power
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum power that will be inputed to the amplifier.
Amplifier Forward Protection Level[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Power
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum forward power that will be outputted by the amplifier.
Amplifier Reflected Protection Level[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Power
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum reflected power. During the regulation of the test-level is this maximum reflected power monitored.
Coupler Forward Power Attenuation Correction[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
Coupler Reflected Power Attenuation Correction[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
Powermeter correction[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
Cable loss correction[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
The values in the correction column should be expressed in dB. Cable device drivers can be used in different parts of a test-site setup. But they all have in common that an 'interesting' power-level is attenuated by the cable before it is measured by a measuring device (most of the time a powermeter, receiver or an analyzer). The values of the correction and attenuation column are therefore added to the measured powerlevel. The values in the gain column are inverted before they are added to the measured powerlevel.
Absorbing clamp loss correction[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
Antenna Factor[edit]
For converting the reading dBuV into dBuV/m the antenna factor is needed.
  • Frequency
  • Correction
Where the correction column is the antenna factor.
GPSHeight till AntennaHeight Correction[edit]
  • Height
AntennaGain versus Angle Correction[edit]
  • Angle
  • Attenuation
Calibration Jig Transfer correction[edit]
  • Frequency
  • Correction
If the jig is not "Configurable" there is a default correction factor of 9.5dB (theoretical value). How is it taken into account the presence of a correction factor? Is this value added or taken as is? We explain these questions with an example: At 150 kHz the real correction factor of the jig is 10dB. So, this value is saved in the correction file. What will be the behavior of RadiMation? Will it now take 10dB as the total transfer correction?
This implies that:
Current sensor impedance[edit]
  • Frequency
  • Correction
The values in the correction column should be expressed in Ohm (Ω). The value specified in the correction column is used to correctly calculate the current flowing through the current sensor.
Field sensor correction[edit]
  • Frequency
  • Correction
LISN Amplitude correction[edit]
  • Frequency
  • Correction
Preamplifier Gain[edit]
The gain of an used preamplifier can be corrected inside the software. A correction file with the frequency dependant gain of the preamplifier should be attached to the device driver of the preamplifier. The correction file should contain the columns:
  • Frequency
  • Correction
The values in the correction column should be expressed in dB. The values of the correction column are subtracted of the measured emission level. In this way the real emission level can be calculated.
Receiver loss correction[edit]
  • Frequency
  • Correction
Correction Factor (dBpT/uV) correction[edit]
During emission measurements when the measured emission values have to be shown in Ampere-units, the impedance correction factor should be configured in the software. A correction file with the impedance should be attached to the device driver of the receiver or spectrum analyser. The correction file should contain the columns:
  • Frequency
  • Correction
The value specified in the correction column should be expressed in Ohm (Ω). The values in the correction column is used in the law of Ohm to calculate the correct to be displayed emission values in an Ampere unit.
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the receiver or spectrum analyser. The correction file should contain the columns:
  • Frequency
  • Correction
The value specified in the correction column should be expressed in dB. The values in the correction column are added to the measured emission levels to calculate the correct to be displayed emission values in a Tesla unit.
Resistor Correction[edit]
  • Frequency
  • Correction
