(→Database Configuration) |
(→Device driver configuration) |
Line 258: | Line 258: | ||
Device drivers are used to let your software know which equipment has to be used and controlled during EMC tests. | Device drivers are used to let your software know which equipment has to be used and controlled during EMC tests. | ||
− | In {{RadiMation}}, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information like the IEEE address or | + | In {{RadiMation}}, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information like the IEEE address or RS232 COM-port has to be entered. |
− | + | You can configure a device in the configuration screen. You can access the configuration screen by selecting: | |
− | {{Menu|Configuration | + | {{Menu|Configuration|Configuration|Device Drivers}} |
[[Image:Configuration DeviceDrivers.png]] | [[Image:Configuration DeviceDrivers.png]] | ||
+ | You can select the type of device you want in the box that is situated top right. The big window will display all currently configured device drivers, like in the picture all configured [[Signal Generator]]s. | ||
{{ScreenElementDescription|Device Driver Type|Different [[Device Types|types]] of devices are available in {{RadiMation}}. Only the created device drivers of the type that is selected for {{ScreenElement|Device Driver Type}} will be shown in the list}} | {{ScreenElementDescription|Device Driver Type|Different [[Device Types|types]] of devices are available in {{RadiMation}}. Only the created device drivers of the type that is selected for {{ScreenElement|Device Driver Type}} will be shown in the list}} | ||
{{ScreenElementDescription|Available Device Drivers|Shows a list of all the available device drivers for the selected {{ScreenElement|Device Driver Type}}}} | {{ScreenElementDescription|Available Device Drivers|Shows a list of all the available device drivers for the selected {{ScreenElement|Device Driver Type}}}} | ||
{{ScreenElementDescription|New|Creates a new [[RadiMation|{{RadiMation}}]] device driver for a specific device. It is only possible to create a new device driver for the selected {{ScreenElement|Device Driver Type}}.}} | {{ScreenElementDescription|New|Creates a new [[RadiMation|{{RadiMation}}]] device driver for a specific device. It is only possible to create a new device driver for the selected {{ScreenElement|Device Driver Type}}.}} | ||
− | {{ScreenElementDescription|Edit| | + | {{ScreenElementDescription|Edit|To edit the configuration of a device, select the device in the list and press {{ScreenElement|Edit}}. This allows to change the configuration of the selected device driver from the list of {{ScreenElement|Available Device Drivers}}}} |
− | {{ScreenElementDescription|Remove|Removes the configuration of the selected device driver from the list of {{ScreenElement|Available Device Drivers}}}} | + | {{ScreenElementDescription|Remove|Removes the configuration of the selected device driver from the list of {{ScreenElement|Available Device Drivers}}. Please note that this driver will then not be available anymore in the test sites. Even if a new device driver is created with the exact same name, it will not automatically reappear in the test sites.}} |
When the {{ScreenElement|New}} button is pressed, a list of all supported device drivers will be shown. | When the {{ScreenElement|New}} button is pressed, a list of all supported device drivers will be shown. | ||
Line 278: | Line 279: | ||
[[Image:Configuration DeviceDrivers NewDeviceDriver.png]] | [[Image:Configuration DeviceDrivers NewDeviceDriver.png]] | ||
− | The required device driver can be selected from this list. The new device driver will then be created when {{ScreenElement|New}} button is pressed. At that moment it is possible to use the default name of the device driver, or to change it to another (eg. more specific) name. | + | The required device driver can be selected from this list. The new device driver will then be created when {{ScreenElement|New}} button is pressed. At that moment it is possible to use the default name of the device driver, or to change it to another (eg. more specific) name. |
+ | [[Image:Device Driver Configuration Window.png]] | ||
+ | If you do not whis to create a new device driver press {{ScreenElement|Close}}. | ||
==== Virtual device drivers ==== | ==== Virtual device drivers ==== |
Revision as of 14:17, 7 February 2011
Getting started
Software overview
RadiMation® consists of 7 separate software modules to perform 5 different types of EMC tests, calibrations, automatic report generation and sequence testing.
- Calibration
- Report generator
The combination of the different software modules in one software package has a number of advantages compared to other software packages:
- Easy to use software by implementation same "look & feel" over all modules.
- Seamless data exchange between the different software modules.
- Calibration functionality included in the package enables easy exchange of calibration files.
- Different types of tests can be performed, fully automatic after each other (Sequence testing).
The figure below represents an overview of the RadiMation® software:
- The Visual and Graph System
- Sequence Testing
- Calibration
- Data Logging
- Database Functionality
- The Device Driver System
Version 2023.2 of the RadiMation® package contains the following modules:
- Radiated immunity
- Pulsed immunity
- Radiated emission
- Report generator
RadiMation® has some requirements for the PC(-environment) that is used to perform the EMC tests on.
Hardware requirements
The RadiMation® software extensively makes use of data calculations, bus interactions (to control the test equipment) and screen interactions (for the use of graphs). For this reason, processor performance is quite important and a dedicated PC is required for RadiMation®. Running other programs on the same computer can cause compatibility problems and is not supported. Programs which are used by RadiMation®(like Word, Excel and MySQL) are supported to be used with RadiMation®.
To use the software to its full extent, we require the following hardware configuration:
- Disk drives: CDROM drive
- Monitor: SXGA (1280 x 1024) or larger. (at least 1280 pixels wide and at least 900 pixels high)
Memory
RadiMation® needs a considerable amount of memory to perform at its best. As a guiding line there should be at least 512 MB (Mega Bytes) of free memory. Free memory is defined the amount of free ram without having to use the swap file. Not the total amount of memory that put in to the machine. How do you determine the amount of free memory?
- Start you machine from power off, so no hibernate or restart.
- Determine the total amount of memory put into the PC.
- Calculate the amount of free memory by subtracting the amount of memory from point 3 from the amount of memory from point 2.
If the amount of free memory is less then 512 MB there are several options to create more free memory, depending on the configuration:
- Insert more memory modules into the machine
- Remove programs with large memory usage and that do not involve measuring, for example VNC viewers.
- Prevent that programs are started automatically during the booting of the PC
GPIB card
The GPIB hardware and software needs to be installed correctly before the installation of RadiMation®. If the card is not installed before the installation, D.A.R.E!! is not responsible for any additional costs or delays that may occurs if the card is not installed. The GPIB Card has to be a card that is supported by the National Instruments NI-488.2 software library.
Software requirements
The RadiMation® software will operate (and is tested) with the Microsoft Windows XP and Microsoft Windows 7 operating systems. RadiMation® is using Microsoft Office Word, to generate and print the printout's. It is therefore required that Microsoft Office Word is installed.
Microsoft office
The components of the Microsoft office suite must be fully installed and working, before they can be used for report generation. The formats in which the RadiMation® report generator can create reports in are:
- Microsoft Word
- Microsoft Excel
Installation
During the installation an network/PC administrator from the IT department must to be available, and in the building. This person must have access in granting rights, for one day or more. Also this person must have knowledge of the network situation. D.A.R.E!! is not responsible for any additional costs or delays that may occurs is this support can not be provided.
The first time RadiMation® needs to be installed with administrator rights. Upgrades have no need for full rights. For an update the following rights needs to be set:
- Read rights for the 'C:\' drive
- Full rights on the 'C:\Program Files\Dare Development' folder. (Please note that the path may be different depending on the language of the OS.)
Conventions
RadiMation® makes use of the Windows structure of pull down menu’s. In this manual, the pull down menu’s are described in the following way:
The above example shows that first the File item from the menu bar should be selected, and after that the Open item, and the EUT item to show the file open selection window.
Software installation
To install the RadiMation® software please insert the RadiMation® CD in the CD ROM drive of your computer. The setup program of RadiMation® should now start automatically. If the program doesn't start automatically, run E:\setup.exe (where E:\ is the CD-ROM drive) from the CD. The Installation Wizard will guide you through the installation process. After installation, the computer must be restarted before running RadiMation®. Please see User rights and installation for rights needed for first time installation.
In a network environment multiple computers are connected with each other with a shared communication medium. This had advantages because data can be shared between the connected computers. The RadiMation® software is designed to optimally use this advantage. All data that will be used or generated by RadiMation® can be shared on the network.
To optimally use this sharing of data the configuration of RadiMation® should be slightly changed. Use the following steps to do this.
- Create a shared network directory on one of the Network Servers. Your Network Administrator can help you with this. This directory will be called <NetworkDir> from now on.
- Create a sub-directory called Setup in the <NetworkDir>
- Place the RadiMation® setup files from the RadiMation® CD (SETUP.EXE and all SETUP.W?? files) in the Setup sub-directory.
- Install RadiMation® from the Setup directory on one of the connected PC's. This can be done by starting the SETUP.EXE program. Use the installation procedure as described in the previous section.
- EUT_File
- CAL_File
- COR_File
- CONFDVDR
- LLF_File
- RGT_File
- RGO_File
- Printout
- SEQ_File
- TSF_File
- Start the installed RadiMation® program, and open the Configuration window. On the Directories tab, change the location of the Device Driver Configuration directory to the CONFDVDR directory in the <NetworkDir>.
- Restart the RadiMation® program
- Open the Configuration window again and on the {{ScreenElement|Directories} tab, change the location of all the directories to the corresponding directories in the <NetworkDir>.
- From now on the RadiMation® program will use the <NetworkDir> as its main data directory.
When the RadiMation® software should be installed on another PC in the Network, the following steps can be used.
- Install the RadiMation® software from the Setup sub-directory in the <NetworkDir>. Use the installation procedure as described in the previous section.
Installing software updates
Software updates are distributed on a CD-ROM. To install the new RadiMation® version, please insert the CD-ROM in the CD-ROM drive of the computer. Run the installation program (i.e. E:\setup.exe, where E:\ is the CD-ROM drive). The setup program will detect that a version of RadiMation® is already installed on your PC. The setup program will install the new RadiMation® version at the same location as the already installed version. A back up of the previously installed version of RadiMation® will be copied to the backup directory automatically. All the settings of the previous version will also be used for the new installed RadiMation® version. When problems do occur with the newly installed RadiMation® version, the old version of RadiMation® can be recovered.
If the already installed RadiMation® version was installed from a network directory, the new setup files can be copied from the CD-ROM to the Setup sub-directory of the <NetworkDir>. On each PC the new RadiMation® version can then be installed from the Setup sub-directory
This chapter describes what rights a user must have in order to perform an installation or an upgrade of the RadiMation® software.
Upgrade
For an upgrade the user must have the rights to read and write in the RadiMation® folder in the program files directory. And must have rights to add and check items to the "HKEY_LOCAL_MACHINE\Software\Classes\" tree in the registry as well as the "HKEY_LOCAL_MACHINE\Software\DARE Development\" tree. Please note that upgrading like described here can only be performed with RadiMation® version greater then version 4.3.20. And upgrading from a version earlier or equal to version 4.3.20 can only be performed by an administrator.
New installation
Software configuration
Password
To prevent unauthorised persons to change system configurations, a password protection can be used. The password is standard turned on with the default password Radimation. This password protection can be switched off, changed and configured under:
Test engineers
Shows the currently configured list of engineer's. |
The Add button can be used to add another engineer's name to the list of available engineers. After the Add button is pressed, the following window will be shown, and another engineer's name can be specified. |
The Edit button can be used to change the name of the selected engineer. |
The Delete button can be used to remove the name of the selected engineer from the list of available engineers. |
The Ok button will close the dialog. |
- The default client
- The default manufacturer
- The default test site
Data entered under default address information, will automatically be entered in the EUT window and therefore only needs to be entered once.
Default Units
- Power: dBuW, dBm, dBmW, dBW, uW, mW, W
- Field Strength: dBuV/m, dBmV/m, dBV/m, uV/m, mV/m, V/m
- Voltage: dBuVrms, dBmVrms, dBVrms, uVrms, mVrms, Vrms
- Current: dBuA, dBmA, dBA, uA, mA, A
- Emission: dBuV, dBmv, dBV, V, W, dBuV/m, dBmV/m, V/m
Will show the Unit configuration dialog, for the selected usage of the unit, so more specific details of the used unit, and the formatting of the values can be configured. |
The path name can be either a local directory or a directory on a network drive. If a network drive is selected, it makes it possible to store for example EUT data central on a computer network.
To modify the path name select:
And then select the Directories tab.
A directory location can be changed by selecting the directory from the list and then pressing the Change button.
Configuring the Report Generator
The Report generator must be configured for the type of word-processor you are using. To configure the report generator please select:
Now select the Report Generator tab and select the correct word-processor that should be used to the generate the reports in.
In RadiMation®, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information like the IEEE address or RS232 COM-port has to be entered.
You can select the type of device you want in the box that is situated top right. The big window will display all currently configured device drivers, like in the picture all configured Signal Generators.
Different types of devices are available in RadiMation®. Only the created device drivers of the type that is selected for Device Driver Type will be shown in the list |
Shows a list of all the available device drivers for the selected Device Driver Type |
Creates a new RadiMation® device driver for a specific device. It is only possible to create a new device driver for the selected Device Driver Type. |
The required device driver can be selected from this list. The new device driver will then be created when New button is pressed. At that moment it is possible to use the default name of the device driver, or to change it to another (eg. more specific) name.
If you do not whis to create a new device driver press Close.
Virtual device drivers do not have limitations like normal device drivers (frequency Band, max power, etc.)
Graphics configuration
To configure the looks of the shown graphs, please select:
Then select the Graphs tab.
An example graph is shown that is created with the currently selected graph settings. When the Customize button is pressed, the Graph Customization Configuration window is shown, which allows to change the look of all graphs.
Shows the currently selected color for the back(ground) of the graph. The color can be changed by clicking on the color itself. |
Shows the currently selected color for the grid of the graph. The color can be changed by clicking on the color itself. |
Shows the currently selected color for the text around the graph. The color can be changed by clicking on the color itself. |
Shows the currently selected type of style that is used for the grid of the graph. Possible line styles are:
Determines if the initial x-axis scaling of the graph should be shown linear or logarithmic |
The selected settings in this Graph Customization Configuration window are used as the RadiMation® default settings. It is still possible to overrule these settings in each test to more specific settings.
Introduction
Language Selection
RadiMation® supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require.
Then select the Language tab:
The currently selected language is marked with the text ' (current) ' appended. If another language should be used in RadiMation®, one of the other languages can just easily be selected. The selected language will only be active after the RadiMation® package is restarted.
Product Standard
Introduction
- Washing machine
- Hair dryer
- Car radio
- Television
Adding a Product Standard
And then select the Product standards tab:
After pressing Add the end-user can insert all necessary information.
Please note that the Official name can only be inserted the first time and has to be unique. All fields speak for themselves, except for the File location. In the File location field, it is possible to associate a file to this standard. For example a PDF, DOC or HTML file that describes the standard. RadiMation® will open the file, with the default reader for this file type, when the engineer selects the magnifying glass.
Add(ing) Basic Standard(s)
Adding a basic standard to the product standard can be done simply by pressing the Add button. The following window will appear:
It is possible to select multiple items by selecting the top item, then hold the 'shift' key and selecting the lowest items. Holding the 'Ctrl' key allows you to select or deselect one or more items. Double clicking on an item will display the item details in a new window.
Basic Standard
The error message window will ask you to abort the test or to ignore the error message and continue the test with different test equipment. Of course, to achieve maximum accuracy, one should use the same test equipment during the calibration and during the substitution tests. However, in some circumstances, changing test equipment between calibration and test will not have any influence on test results. This is for example true when a RI substitution list is made using FWD power calibration file. Changing the amplifier won’t change test results as long as the new amplifier covers the required frequency range and can deliver the required output power.
Password protection
To activate, modify or de-activate passwords, select:
In the Change Password window, the password can be enabled, disabled and changed (to disable or change the password, the current password is required). Furthermore, in the password window, for the following items, the protection can be enabled or disabled:
prevents unauthorised changes to test equipment configuration. |
prevents unauthorised changes to test engineers configuration. |
prevents unauthorised changes to Test Setup Files (TSF) files. |
prevents unauthorised changes to limit line files. |
prevents unauthorised changes to correction files. |
User configurable device drivers
To make your own device driver, open the configuration window:
and select the Device drivers tab.
- Press the New button.
- Select the driver called Configurable <XXX> (i.e. if you want to make a signal generator device driver you would select the Configurable signal generator device driver).
- Enter a description for the device driver (for example the type number of the generator) and press Ok.
- Press the Edit button.
EUT files
Opening an existing EUT file
The first column displays the testnumber. The second column displays the description of the test. The third column shows the notes you inserted. The three columns can be sorted one at a time. This can be done by pressing the desired column header. If you wish to sort on testnumbers, press the first column header. If you wish to sort by note press the third column header. Pressing a column header twice will reverse the sorting order. For example the first time you press the third column header the notes will be sorted ascending, the second time you press the header the notes will be sorted descending. This pertains to all the columns.
Cable information of the EUT
RadiMation® is capable of storing a lot of EUT related information. One of these items is the cables connected to the EUT. You can specify them by selecting the EUT Information tab in the EUT window. On the bottom of the window you will find a control looking like the one below.
You can specify every cable of the EUT, pressing Add or Edit will popup the following window:
Limit line files
Creating limit lines
The window below will be displayed:
The user can enter a file description to the limit line in the Description field.
Entering data points
To enter data point, press the Insert button. A new line will appear and you can enter frequency and level. When entering data points, the limit line will be drawn directly in the graph. This makes it easy to check if the limit line is entered correctly.
When a data point in incorrect, it can be removed by selecting the data point and pressing the Remove button.
Clicking on the coloured box behind the Color selection box will change the line color. The selected line colour will be used during emission tests. |
In between the specified frequency points, the correct value of the limit line will be calculated by using a logarithmic interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. |
In between the specified frequency points, the correct value of the limit line will be calculated by using a linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. |
During emission tests, RadiMation® will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. In the Detectors menu the operator can select for which detector(s) the limit line is valid. |
Saving limit line files
RadiMation® will prompt you for a filename for the new limit line file. This can be the same or a different filename as the limit line Description field.
Correction files
Creating correction files
To enter data points, press the Add Row button. A new line will appear and you can enter frequency and level. |
When a data point is incorrect, it can be removed by selecting the data point and pressing the Remove Row button. |
Saving correction files
Using Test Setup Files (T.S.F.)
RadiMation® is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes Radiated immunity tests [R.I.], Conducted immunity tests [C.I.] as well as Radiated emission tests [R.E.] and Conducted emission tests [C.E.]. (See also "software overview").
For each kind of test, the test configuration data (start/stop frequency, test level, dwell time etc.) can be saved in a Test Configuration File (*.TSF). Separate from the test configuration data, the TSF file also stores the calibration file and test site information. The description of the TSF file can be 40 characters max. When the same kind of test must be performed, one can simply load the TSF file and start the test.
In order to organise the different TSF files RadiMation® enables the user to create a directory tree thus enabling the user to store the TSF files in a structured manner.
A more sophisticated use of TSF files can be found during sequence testing. In a sequence, different TSF files (even from different types of tests) can be placed after each other, and started without the intervention of a test engineer (See also “Sequence testing”)
Viewing test data
RadiMation® stores all test results for the corresponding EUT in the EUT file. To view test data, open the corresponding EUT file through the File, Open, EUT pull down menu.
The EUT window will show a list of all performed tests for the selected EUT. To view the extended data from a single test, select the test from the list and press the view button. The following data will be displayed:
- Time and date of the test.
- Equipment used during the test (Including serial numbers).
- Test engineer.
- Test configuration data.
- All graphs of AD channels used Field graphs (If used).
- Current graph (If used).
- Notes.
- Signal generator level graph (Used for fault diagnoses only).
- Forward power level graph (If applicable).
- Reflected power level graph (If applicable).
- Net power level graph (If applicable).
- Ambient information:
- Temperature graph (If used).
- Pressure graph (If used).
- Relative humidity graph (If used).
- Calibration file used during the test (If applicable).
Printing test data
Test data can be printed in several different ways:
- When only a part of the test data needs to be printed, select the desired test, then select view and select the graph you want to print. The graph will be displayed on the screen. Above the graph you will find a print button. This print button will only print the current selected graph.
- From the list of performed tests, a single test can be selected. When all data of this test must be printed, simply press the print button at the right side of the list.
- When you want to print the data of more than one test, select the tests to print while holding down the shift button. All selected tests will be printed when the print button is pressed.
Report generation
RadiMation® can also use the information of the performed tests, to automatically generate a report, or export all the data to Microsoft Excel. The Report Generator and Excel Exporter can be used to perfrom these tasks. See Chapter 13 for a detailed description.
Ambient information