Line 60: Line 60:
* Monitor: SXGA (1280 x 1024) or larger. (at least 1280 pixels wide and at least 900 pixels high)
* IEEE card: Any GPIB Card that is supported by the [[National Instruments]] NI-488.2 Software Library
==== Memory ====
==== Memory ====
Line 73: Line 74:
* Remove/uninstall programs with large memory usage and that do not involve measuring, for example VNC viewers.
* Prevent that programs are started automatically during the booting of the computer.
==== GPIB card ====
==== GPIB card ====
The GPIB hardware and software needs to be installed correctly before the installation of {{RadiMation}}. If the card is not installed before the installation, DARE!! is not responsible for any additional costs or delays that may occurs if the card is not installed.
The GPIB Card has to be a card that is supported by the [[National Instruments]] [[NI-488.2]] software library.
=== Software requirements ===
The {{RadiMation}} software will operate (and is tested) with the [[Microsoft]] Windows 7 (32-bit and 64-bit) and [[Microsoft]] Windows 8 (32-bit and 64-bit) operating systems. {{RadiMation}} is using Microsoft Office Word, to generate and print the printout's. It is therefore required that Microsoft Office Word is installed.
==== Microsoft office ====
Line 86: Line 90:
* Microsoft Word
* Microsoft Word
* Microsoft Excel
* Microsoft Excel
==== Installation ====
Revision as of 08:16, 1 September 2015
Getting started[edit]
Software overview[edit]
  • The Visual and Graph System
  • Sequence Testing
  • Calibration
  • Data Logging
  • Database Functionality
  • The Device Driver System
  1. Radiated immunity testing (RI)
  2. Conducted immunity testing (CI)
  3. Pulsed immunity (ESD, EFT, Surge)
  4. Radiated emission testing (RE)
  5. Conducted emission testing (CE)
  6. Report generator
  7. Excel exporter
Version 2025.1 of the RadiMation® package contains the following modules:
  3. Pulsed immunity
  6. Report generator
RadiMation® has some requirements for the PC(-environment) that is used to perform the EMC tests on.
Hardware requirements[edit]
The RadiMation® software extensively makes use of data calculations, bus interactions (to control the test equipment) and screen interactions (for the use of graphs). For this reason, processor performance is quite important and a dedicated PC is required for RadiMation®. Running other programs on the same computer can cause compatibility problems and is not supported. Programs which are used by RadiMation®(like Word, Excel and MySQL) are supported to be used with RadiMation®.
To use the software to its full extent, we require the following hardware configuration:
  • Processor type: 2 GHz minimum, > 3 GHz recommended
  • Free RAM memory: 512 MB minimal, > 1 GB recommended
  • Monitor: SXGA (1280 x 1024) or larger. (at least 1280 pixels wide and at least 900 pixels high)
Memory[edit]
RadiMation® needs a considerable amount of memory to perform at its best. As a guiding line there should be at least 512 MB (Mega Bytes) of free memory. Free memory is defined as the amount of free RAM without having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. So, how do you determine the amount of free memory?
  1. Cold boot your machine from power off, so no hibernate or restart.
  4. Calculate the amount of free memory by subtracting the amount of memory from point three from the amount of memory from point two.
If the amount of free memory is less then 512 MB there are several options to obtain more free memory, depending on the configuration and situation:
  • Remove/uninstall programs with large memory usage and that do not involve measuring, for example VNC viewers.
  • Prevent that programs are started automatically during the booting of the computer.
GPIB card[edit]
The GPIB hardware and software needs to be installed correctly before the installation of RadiMation®. If the card is not installed before the installation, DARE!! is not responsible for any additional costs or delays that may occurs if the card is not installed. The GPIB Card has to be a card that is supported by the National Instruments NI-488.2 software library.
Software requirements[edit]
The RadiMation® software will operate (and is tested) with the Microsoft Windows 7 (32-bit and 64-bit) and Microsoft Windows 8 (32-bit and 64-bit) operating systems. RadiMation® is using Microsoft Office Word, to generate and print the printout's. It is therefore required that Microsoft Office Word is installed.
Microsoft office[edit]
The components of the Microsoft office suite must be fully installed and working, before they can be used for report generation. The formats in which the RadiMation® report generator can create reports in are:
  • Microsoft Word
  • Microsoft Excel
Installation[edit]
During the installation a network/PC administrator from the IT department must be available. This person must have access in granting rights, for one day or more. Also this person must have knowledge of the network situation. DARE!! is not responsible for any additional costs or delays that may occurs if this support can not be provided.
The first time RadiMation® is installed, it needs to be installed with administrator rights. Upgrades have no need for full rights (although rights to perform installations are required). For an update the following rights needs to be set:
  • Read rights for the 'C:\' drive
  • Full rights for the installation directory (e.g. 'C:\Program Files\Dare Development' or 'C:\Program Files\Dare Instruments', please note that the path may be different depending on the used OS and language.)
Conventions[edit]
RadiMation® makes use of the Windows structure of pull down menu’s. In this manual, the pull down menu’s are described in the following way:
   Menu.svg File
      Menu.svg Open
         Menu.svg EUT
The above example shows that first the File item from the menu bar should be selected, and after that the Open item and then the EUT item to show the file open selection window.
Software installation[edit]
To install the RadiMation® software please insert the RadiMation® CD in the CD-ROM drive of your computer. The setup program of RadiMation® should now start automatically. If the program doesn't start automatically, run E:\setup.exe (where E:\ is the letter of your CD-ROM drive) from the CD. The Installation Wizard will guide you through the installation process. After installation, the computer must be restarted before running RadiMation®. Please see User rights and installation for the needed user rights and information for a first time installation.
In a network environment multiple computers are connected with each other with a shared communication medium. Normally this has multiple advantages over a local installation. One of the benefits is that data can be shared between the connected computers. The RadiMation® software is designed to optimally use this advantage. Therefore, all data that will be used or generated by RadiMation® can also be shared on the network. Most of the time it is advised to use a network configuration over a local configuration.
Note: When RadiMation® is making use of a network configuration the program will still be executed from the local computer. Therefore, it is neither cached nor running directly from the network.
To fully use this functionality of sharing these data, the configuration of RadiMation® should slightly be changed. In order to achieve this following these steps below.
  1. Create a shared network directory with the appropriate rights on one of the Network Servers. Your Network Administrator can help you with this. This directory will be called <NetworkDir> from now on.
  2. Create a drive assignment for this specific <NetworkDir>. For instance, this will avoid the use of excessive path depths. Your Network Administrator can also help you with this step.
  4. Copy the RadiMation® setup.exe file from the CD-ROM (or directly from local disk if you downloaded RadiMation®) to this <Setup> directory in the <NetworkDir>.
    • CAL_File
    • CONFDVDR
    • COR_File
    • EUT_File
    • LLF_File
    • Printout
    • RGO_File
    • RGT_File
    • SEQ_File
    • TSF_File
  7. Start the installed RadiMation® program, and open the Configuration window. On the Directories tab, change the location of the Device Driver Configuration directory to the CONFDVDR directory in the <NetworkDir>. (See Default directory configuration. Note: please wait changing the other directories.)
  9. Open the Configuration window again and on the Directories tab, now change the location of all the other directories to the corresponding directories in the <NetworkDir>.
  11. From now on the RadiMation® program will use the <NetworkDir> as its main data directory.
When the RadiMation® software should be installed on another computer in the network, the following steps can be used.
  2. Start the installed RadiMation® program, and open the Configuration window. On the Directories tab, change the location of the Device Driver Configuration directory to the CONFDVDR directory in the <NetworkDir>.
  3. Restart the RadiMation® program
Software updates are distributed on CD-ROM and can be retrieved from our webshop. To install the new RadiMation® version, please insert the CD-ROM in the CD-ROM drive of the computer. Run the installation program (i.e. E:\setup.exe, where E:\ is the CD-ROM drive letter). The setup program will detect that a previous version of RadiMation® is already installed on your PC. The setup program will install the new RadiMation® version at the same location as the already installed version. A back up of the previously installed version of RadiMation® will be copied to the backup directory automatically. All the settings of the previous version will also be used for the new installed RadiMation® version. When problems do occur with the newly installed RadiMation® version, the old version of RadiMation® can be recovered.
Note: Before installing a RadiMation® update it is strongly advised that you completely backup all your files and directories e.g. TSF/Sequence/Correction/Calibration and EUT.
Warning: Do NOT uninstall RadiMation®!! This will cause removal of all user settings and TSF, sequence, calibration and correction files. This is unrecoverable.
If the already installed RadiMation® version was installed from a network directory, the new setup file can be copied to the <Setup> sub-directory of the <NetworkDir>. On each computer the new RadiMation® version can then easily be installed from this directory.
This chapter describes the required rights a user must have in order to perform an installation or an upgrade of the RadiMation® software.
Upgrade[edit]
For an upgrade the user must have the rights to read and write in the RadiMation® folder, which is normally located in the program files directory. In addition to these rights it is also needed to add and check items to the "HKEY_LOCAL_MACHINE\Software\Classes\" tree in the registry as well as the "HKEY_LOCAL_MACHINE\Software\DARE Development\" tree. Please note that upgrading like described here can only be performed with RadiMation® version greater then version 4.3.20. Upgrading from a version earlier or equal to version 4.3.20 can only be performed by an administrator.
New installation[edit]
Software configuration[edit]
Password protection[edit]
RadiMation® allows you to protect a number of configurations through a password protection. To activate, modify or de-activate passwords follow the same procedure by adding or changing an engineer. To prevent unauthorised persons to change system configurations, a password protection can be used. The password is standard turned on for the default used test engineer 'Administrator'. The default password is Radimation. This password protection can be switched off, changed and configured under:
   Menu.svg Configuration
      Menu.svg Engineers
Then select the Administrator engineer and click 'Edit'. In this window, of the Administrator Engineer, the password can be enabled, disabled and changed (to disable or change the password, the current password is required). Furthermore, for the following items, the protection can be enabled or disabled.
ScreenElementDescription.svg Equipment prevents unauthorised changes to test equipment configuration.
ScreenElementDescription.svg Engineer prevents unauthorised changes to test engineers configuration.
ScreenElementDescription.svg TSF Files prevents unauthorised changes to Test Setup Files (TSF) files.
ScreenElementDescription.svg Limit Line prevents unauthorised changes to Limit Line Files (LLF) files.
ScreenElementDescription.svg Correction File prevents unauthorised changes to correction files.
ScreenElementDescription.svg General Configuration prevents unauthorised changes to the general configuration.
Click 'Change Password' to get the Change Password window. After typing the 'Old Password' and now leaving the 'Password' and 'Retype Password' fields empty, the password protection will be switched off. This can be verified by the message marked in red displayed on the bottom of the window.
Test engineers[edit]
      Menu.svg Engineers
ScreenElementDescription.svg Engineers Shows the currently configured list of engineers.
ScreenElementDescription.svg Add The Add button can be used to add another engineers name to the list of available engineers. After the Add button is pressed, the following window will be shown, and another engineers name can be specified.
ScreenElementDescription.svg Edit The Edit button can be used to change the name of the selected engineers.
ScreenElementDescription.svg Delete The Delete button can be used to remove the name of the selected engineers from the list of available engineers.
ScreenElementDescription.svg Ok The Ok button will close the dialog.
  • The default client
  • The default manufacturer
  • The default test site
Data entered under default address information, will automatically be entered in the EUT window and therefore only needs to be entered once.
Default Units[edit]
To configure these settings go to:
      Menu.svg Units
It is possible to define different units depending on the test selected from the drop down menu. It is possible to use a different unit setting for:
  • Radiated Immunity
  • Conducted Immunity
  • Radiated Emission
  • Conducted Emission
  • Pulsed Immunity
  • Antenna Diagram
  • RadiCal
For instance the units for the following parameters can be defined:
  • Power: dBuW, dBm, dBmW, dBW, uW, mW, W
  • Field Strength: dBuV/m, dBmV/m, dBV/m, uV/m, mV/m, V/m
  • Voltage: dBuVrms, dBmVrms, dBVrms, uVrms, mVrms, Vrms
  • Current: dBuA, dBmA, dBA, uA, mA, A
  • Emission: dBuV, dBmv, dBV, V, W, dBuV/m, dBmV/m, V/m
ScreenElementDescription.svg Change Will show the Unit configuration dialog, for the selected usage of the unit, so more specific details of the used unit, and the formatting of the values can be configured.
The path name can be either a local directory or a directory on a network drive. If a network drive is selected, it makes it possible to store for example EUT data central on a computer network.
To modify the path name select:
And then select the Directories tab.
A directory location can be changed by selecting the directory from the list and then pressing the Change button.
Device drivers are used to let your software know which equipment has to be used and controlled during EMC tests.
In RadiMation®, the device drivers must be configured to customer requirements once. During the configuration of the device driver, information like the IEEE address or RS232 COM-port has to be entered.
You can configure a device in the configuration screen. You can access the configuration screen in two different ways. By selecting:
         Menu.svg Device Drivers
Or by selecting:
   Menu.svg Devices
      Menu.svg Configure...
You can select the type of device you want by the drop-down menu that is situated at the top. The big window will display all currently configured device drivers, like in the picture all configured Signal Generators.
ScreenElementDescription.svg Device Driver Type Different types of devices are available in RadiMation®. Only the created device drivers of the type that is selected for Device Driver Type will be shown in the list
ScreenElementDescription.svg Available Device Drivers Shows a list of all the available device drivers for the selected Device Driver Type
ScreenElementDescription.svg New Creates a new RadiMation® device driver for a specific device. It is only possible to create a new device driver for the selected Device Driver Type.
ScreenElementDescription.svg Edit To edit the configuration of a device, select the device in the list and press Edit. This allows to change the configuration of the selected device driver from the list of Available Device Drivers
ScreenElementDescription.svg Remove Removes the configuration of the selected device driver from the list of Available Device Drivers. Please note that this driver will then not be available anymore in the test sites. Even if a new device driver is created with the exact same name, it will not automatically reappear in the test sites.
The required device driver can be searched in or directly selected from this list. The search will start instantly when typing in this 'Search' field. Therefore, the length of the list results will probably get shorter during typing, depending on the keyword(s). The new device driver will then be created when New button is pressed. At that moment it is possible to use the default name of the device driver, or to change it to another (eg. more specific) name.
If you do not whish to create a new device driver press Cancel.
Virtual device drivers[edit]
Virtual device drivers do not have the same limitations like normal device drivers (frequency Band, max power, etc.).
Graphics configuration[edit]
To configure the looks of the shown graphs, please select:
Then select the Graphs tab.
An example graph is shown that is created with the currently selected graph settings. When the Customize button is pressed, the Graph Customization Configuration window is shown, which allows to change the look of all graphs.
ScreenElementDescription.svg Back Color Shows the currently selected color for the back(ground) of the graph. The color can be changed by clicking on the color itself.
ScreenElementDescription.svg Grid Color Shows the currently selected color for the grid of the graph. The color can be changed by clicking on the color itself.
ScreenElementDescription.svg Text Color Shows the currently selected color for the text around the graph. The color can be changed by clicking on the color itself.
ScreenElementDescription.svg Line Style Shows the currently selected type of style that is used for the grid of the graph. Possible line styles are:
  • Dash
  • Dot
  • Thick
  • Thin
ScreenElementDescription.svg Scaling Determines if the initial x-axis scaling of the graph should be shown linear or logarithmic
The selected settings in this Graph Customization Configuration window are used as the RadiMation® default settings. It is still possible to overrule these settings in each test to more specific settings.
Database Configuration[edit]
Language Selection[edit]
RadiMation® supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require.
Then select the Language tab:
The currently selected language is marked with the text ' (current) ' appended. If another language should be used in RadiMation®, one of the other languages can just easily be selected. The selected language will only be active after the RadiMation® package is restarted.
Product Standard[edit]
  • Washing machine
  • Hair dryer
  • Car radio
  • Television
And then select the Product standards tab:
After pressing Add the end-user can insert all necessary information.
Please note that the Official name can only be inserted the first time and has to be unique. All fields speak for themselves, except for the File location. In the File location field, it is possible to associate a file to this standard. For example a PDF, DOC or HTML file that describes the standard. RadiMation® will open the file, with the default reader for this file type, when the engineer selects the magnifying glass.
Adding a basic standard to the product standard can be done simply by pressing the Add button. The following window will appear:
It is possible to select multiple items by selecting the top item, then hold the 'shift' key and selecting the lowest items. Holding the 'Ctrl' key allows you to select or deselect one or more items. Double clicking on an item will display the item details in a new window.
Note: When on the EUT window, a product standard is selected, or added, the data of the standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that if the configuration of a standard in the Configuration window of RadiMation® is changed, the data in the EUT file will still have the old data of the standard!!! This behaviour is correct, because it is the information of the standard that is used at the time that the EUT was tested.
Basic Standard[edit]
The error message window will ask you to abort the test or to ignore the error message and continue the test with different test equipment. Of course, to achieve maximum accuracy, one should use the same test equipment during the calibration and during the substitution tests. However, in some circumstances, changing test equipment between calibration and test will not have any influence on test results. This is for example true when a RI substitution list is made using FWD power calibration file. Changing the amplifier won’t change test results as long as the new amplifier covers the required frequency range and can deliver the required output power.
      Menu.svg Test Equipment
To make your own device driver, open the configuration window:
and select the Device drivers tab.
  2. Press the New button.
  3. Select the driver called Configurable <XXX> (i.e. if you want to make a signal generator device driver you would select the Configurable signal generator device driver).
  4. Enter a description for the device driver (for example the type number of the generator) and press Ok.
  7. Press the Edit button.
  8. A configuration screen for the device driver will appear. All required control commands for the device must be entered. Refer to the operating manual of the equipment for these codes.
Note: Passive equipment such as antenna’s, current probes, cables etc. do also need a device driver. The reason for this is that RadiMation® has to know a number of parameters of these devices. Among others the following information is relevant:
  • Frequency range
  • Maximum input power
EUT files[edit]
Creating a new EUT file[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg EUT
   Menu.svg File
      Menu.svg Open
         Menu.svg EUT
The first column displays the testnumber. The second column displays the description of the test. The third column shows the notes you inserted. The three columns can be sorted one at a time. This can be done by pressing the desired column header. If you wish to sort on testnumbers, press the first column header. If you wish to sort by note press the third column header. Pressing a column header twice will reverse the sorting order. For example the first time you press the third column header the notes will be sorted ascending, the second time you press the header the notes will be sorted descending. This pertains to all the columns.
Cable information of the EUT[edit]
RadiMation® is capable of storing a lot of EUT related information. One of these items is the cables connected to the EUT. You can specify them by selecting the EUT Information tab in the EUT window. On the bottom of the window you will find a control looking like the one below.
You can specify every cable of the EUT, pressing Add or Edit will popup the following window:
Limit line files[edit]
Creating limit lines[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Limit line
The window below will be displayed:
The user can enter a file description to the limit line in the Description field.
Entering data points[edit]
To enter data point, press the Insert button. A new line will appear and you can enter frequency and level. When entering data points, the limit line will be drawn directly in the graph. This makes it easy to check if the limit line is entered correctly.
When a data point in incorrect, it can be removed by selecting the data point and pressing the Remove button.
ScreenElementDescription.svg Color Clicking on the coloured box behind the Color selection box will change the line color. The selected line colour will be used during emission tests.
ScreenElementDescription.svg Logarithmic Interpolation In between the specified frequency points, the correct value of the limit line will be calculated by using a logarithmic interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.
ScreenElementDescription.svg Linear Interpolation In between the specified frequency points, the correct value of the limit line will be calculated by using a linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.
ScreenElementDescription.svg Detectors During emission tests, RadiMation® will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. In the Detectors menu the operator can select for which detector(s) the limit line is valid.
Saving limit line files[edit]
   Menu.svg File
      Menu.svg Save Limit line
RadiMation® will prompt you for a filename for the new limit line file. This can be the same or a different filename as the limit line Description field.
Correction files[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
ScreenElementDescription.svg Entering data points To enter data points, press the Add Row button. A new line will appear and you can enter frequency and level.
ScreenElementDescription.svg Remove Row When a data point is incorrect, it can be removed by selecting the data point and pressing the Remove Row button.
Saving correction files[edit]
   Menu.svg File
      Menu.svg Save correction
RadiMation® is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes Radiated immunity tests [R.I.], Conducted immunity tests [C.I.] as well as Radiated emission tests [R.E.] and Conducted emission tests [C.E.]. (See also "software overview").
For each kind of test, the test configuration data (start/stop frequency, test level, dwell time etc.) can be saved in a Test Configuration File (*.TSF). Separate from the test configuration data, the TSF file also stores the calibration file and test site information. The description of the TSF file can be 40 characters max. When the same kind of test must be performed, one can simply load the TSF file and start the test.
In order to organise the different TSF files RadiMation® enables the user to create a directory tree thus enabling the user to store the TSF files in a structured manner.
A more sophisticated use of TSF files can be found during sequence testing. In a sequence, different TSF files (even from different types of tests) can be placed after each other, and started without the intervention of a test engineer (See also “Sequence testing”)
Viewing test data[edit]
RadiMation® stores all test results for the corresponding EUT in the EUT file. To view test data, open the corresponding EUT file through the File, Open, EUT pull down menu.
The EUT window will show a list of all performed tests for the selected EUT. To view the extended data from a single test, select the test from the list and press the view button. The following data will be displayed:
  • Time and date of the test.
  • Equipment used during the test (Including serial numbers).
  • Test engineer.
  • Test configuration data.
  • All graphs of AD channels used Field graphs (If used).
  • Current graph (If used).
  • Notes.
  • Signal generator level graph (Used for fault diagnoses only).
  • Forward power level graph (If applicable).
  • Reflected power level graph (If applicable).
  • Net power level graph (If applicable).
  • Ambient information:
  • Temperature graph (If used).
  • Pressure graph (If used).
  • Relative humidity graph (If used).
  • Calibration file used during the test (If applicable).
Printing test data[edit]
Test data can be printed in several different ways:
  • When only a part of the test data needs to be printed, select the desired test, then select view and select the graph you want to print. The graph will be displayed on the screen. Above the graph you will find a print button. This print button will only print the current selected graph.
  • From the list of performed tests, a single test can be selected. When all data of this test must be printed, simply press the print button at the right side of the list.
  • When you want to print the data of more than one test, select the tests to print while holding down the shift button. All selected tests will be printed when the print button is pressed.
Warning: All print jobs from RadiMation® will be printed on the default Windows printer. To use another printer, please change the default printer in your Windows environment.
Report generation[edit]
RadiMation® can also use the information of the performed tests, to automatically generate a report, or export all the data to Microsoft Excel. The Report Generator and Excel Exporter can be used to perfrom these tasks. See Chapter 13 for a detailed description.
Ambient information[edit]