Line 378: Line 378:
You can select the type of device you want by the drop-down menu that is situated at the top. The big window will display all currently configured device drivers, like in the picture all configured [[Signal Generator]]s.
To configure a device driver, go to the Device Drivers tab in the Configuration window. Select the desired Device Driver [[Device Types|type]] from the drop-down menu (e.g. Signal Generator). The available drivers for that type will be shown.
{{ScreenElementDescription|Device Driver Type|Different [[Device Types|types]] of devices are available in {{RadiMation}}. Only the created device drivers of the type that is selected for {{ScreenElement|Device Driver Type}} will be shown in the list}}
{{ScreenElementDescription|Available Device Drivers|Shows a list of all the available device drivers for the selected {{ScreenElement|Device Driver Type}}}}
{{ScreenElementDescription|New|Creates a new [[RadiMation|{{RadiMation}}]] device driver for a specific device. It is only possible to create a new device driver for the selected {{ScreenElement|Device Driver Type}}.}}
{{ScreenElementDescription|Edit|To edit the configuration of a device, select the device in the list and press {{ScreenElement|Edit}}. This allows to change the configuration of the selected device driver from the list of {{ScreenElement|Available Device Drivers}}}}
{{ScreenElementDescription|Remove|Removes the configuration of the selected device driver from the list of {{ScreenElement|Available Device Drivers}}. Please note that this driver will then not be available anymore in the test sites. Even if a new device driver is created with the exact same name, it will not automatically reappear in the test sites.}}
When the {{ScreenElement|New}} button is pressed, a list of all supported device drivers will be shown.
[[Image:Configuration DeviceDrivers NewDeviceDriver.png]]
To create a new {{RadiMation}} device driver, select the desired device driver type (e.g. Signal Generator) and click ‘Add’.
The required device driver can be searched in or directly selected from this list. The search will start instantly when typing in this 'Search' field. Therefore, the length of the list results will probably get shorter during typing, depending on the keyword(s). The new device driver will then be created when {{ScreenElement|New}} button is pressed. At that moment it is possible to use the default name of the device driver, or to change it to another (eg. more specific) name.
[[Image:Device driver add.png]]
[[Image:Device Driver Configuration Window.png]]
If you do not whish to create a new device driver press {{ScreenElement|Cancel}}.
Select a new device driver (in the scroll menu or through the search function) and click ‘New’. To cancel your selection and return to the Configuration window, click ‘Close’.
[[Image:New signal generators.png]]
==== Virtual device drivers ====
Virtual device drivers are added for testing and demonstration purposes. These virtual device drivers act as normal devices but do not control actual hardware.
Virtual device drivers do not have the same limitations as normal device drivers (frequency Band, max power, etc.).
The name/description of the chosen device driver can be changed in the Device Driver Configuration window. To confirm your changes (or the provided description), click ‘OK’, the Device Driver Settings window will appear. To cancel your changes, click ‘Cancel’. You will be returned to the previous window (e.g. New Signal generators).
[[Image:Device driver configuration.png]]
The Device Driver Settings window will appear with information on the selected device driver. In this window information can be added, edited and removed. To confirm the settings, click ‘OK’, to cancel your changes, click ‘Cancel’. You will be returned to the Configuration window. More detailed Device Driver Settings are described in [[Device Driver settings|Chapter 14 "Device Driver settings"]].
[[Image:Device driver settings.png]]
To edit an existing device driver, select the desired (device driver) type and device driver, then click ‘Edit’. (The Device Driver Settings window will appear.)
[[Image:Device driver edit.png]]
To remove an existing device driver, select the desired (device driver) type and device driver, then click ‘Remove’.
[[Image:Device driver remove.png]]
A confirmation window will appear. To remove the selected device driver, click ‘Yes’, to cancel, click ‘No’. You will be returned to the Configuration window.
[[Image:Device driver remove2.png]]
Click ‘Close’ to leave the Configuration window.
=== Graphics configuration ===
Revision as of 09:44, 16 September 2015
Getting started[edit]
Software overview[edit]
  • The Visual and Graph System
  • Sequence Testing
  • Calibration
  • Data Logging
  • Database Functionality
  • The Device Driver System
  1. Radiated immunity testing (RI)
  2. Conducted immunity testing (CI)
  3. Pulsed immunity (ESD, EFT, Surge)
  4. Radiated emission testing (RE)
  5. Conducted emission testing (CE)
  6. Report generator
  7. Excel Exporter
Hardware requirements[edit]
  • Processor type: 2 GHz minimum, > 3 GHz recommended
  • Free RAM memory: 512 MB minimal, > 1 GB recommended
  • Monitor: SXGA (1280 x 1024) or larger (at least 1280 pixels wide and at least 900 pixels high)
Memory[edit]
RadiMation® needs a considerable amount of memory to perform at its best. We recommend that you have at least 512 MB (Mega Bytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM without having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps to help you determine the amount of free memory in your PC:
GPIB card[edit]
The GPIB hardware and software needs to be installed correctly before RadiMation® is installed. DARE!! is not responsible for any additional costs or delays that may occur if this card is not installed prior to the installation of RadiMation®. The GPIB Card has to be supported by the National Instruments NI-488.2 software library.
Software requirements[edit]
The RadiMation® software will operate (and is tested) with Microsoft Windows 7 (32-bit and 64-bit) and Microsoft Windows 8 (32-bit and 64-bit) operating systems. It is required to have Microsoft Office Word installed on the PC because RadiMation® uses Microsoft Office Word to generate and print the printout's.
Microsoft office[edit]
  • Microsoft Word
  • Microsoft Excel
Installation[edit]
During the installation of RadiMation® a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. DARE!! is not responsible for any additional costs or delays that may occurs if this support is not provided.
  • Full rights to the installation directory (e.g. 'C:\Program Files\Dare Development' or 'C:\Program Files\Dare Instruments', please note that the path may be different depending on the used OS and language)
Conventions[edit]
   Menu.svg File
      Menu.svg Open
         Menu.svg EUT
The example above shows that, to show the file open selection window, the File item from the menu bar should be selected first, followed by the Open item and then the EUT item.
Software installation[edit]
To install the RadiMation® software, please insert the RadiMation® CD in the CD-ROM drive of your computer. The setup program of RadiMation® should start automatically. If the program does not start automatically, run E:\setup.exe from the CD (where E:\ is the letter of your CD-ROM drive). The Installation Wizard will guide you through the installation process. After installation the computer must be restarted before running RadiMation®. Please read User rights and installation about the user rights and information that is required for first time installation.
In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages compaired to a local installation. One of these benefits is that data can be shared between the connected computers. The RadiMation® software is designed to optimally use this advantage. All the data that is used or generated by RadiMation® can also be shared on the network. In most cases we recommend the use a of network configuration over a local configuration.
    • CAL_File
    • CONFDVDR
    • COR_File
    • EUT_File
    • LLF_File
    • Printout
    • RGO_File
    • RGT_File
    • SEQ_File
    • TSF_File
Software updates can be distributed on CD or downloaded from our website at http://www.dare.eu/instruments/emc-test-software/download.
To install the downloaded RadiMation® update, please execute the startup program.
To install the RadiMation® update from a CD, please insert the CD in the CD-ROM drive of your computer. Run the installation program (i.e. E:\setup.exe, where E:\ is the CD-ROM drive letter). The setup program will detect that a previous version of RadiMation® has already been installed on your PC. The setup program will install the new RadiMation® version at the same location as the previously installed version. A back up of this older version will automatically be copied to the backup directory. All the settings of the older version will also be used for the newly installed RadiMation®. If problems should occur with the newly installed RadiMation®, the old version of RadiMation® can be recovered.
If the already installed RadiMation® version was installed from a network directory, the new setup file can be copied to the <Setup> sub-directory of the <NetworkDir>. The new RadiMation® version can then be installed on each computer from this directory.
This chapter describes the rights a user must have to install or upgrade RadiMation® software.
New installation[edit]
Upgrade[edit]
To upgrade the RadiMation® software, the user must have the rights to read and write in the RadiMation® folder, which is normally located in the C:\Program Files (x86)\DARE Instruments\ directory.
Software configuration[edit]
Password protection[edit]
RadiMation® allows you to protect a number of configurations with a password. The password is standard turned on for the default used test engineer 'Administrator'. The default password is Radimation.
Select the Administrator Engineer and click ‘Edit’.
In the (Administrator) Test Engineer window the password can be enabled, disabled or changed. The current password setting for the selected engineer is displayed at the bottom of the window. This will indicate whether a password has been ‘set’ or ‘not set’.
To enable, disable or change the password, click 'Change Password'.
To change the password, enter the old password and the new password (twice), then click ‘OK’. You will be returned to the Test Engineer window.
To disable the password, enter the old password and leave the other two fields empty, then click ‘OK’. You will be returned to the Test Engineer window. (The password indication will change to: “Warning: Password is not set for this engineer.”)
To enable the password, enter the new password (twice), then click ‘OK’. You will be returned to the Test Engineer window. (The password indication will change to: “Password is set for this engineer.”)
To leave the Engineers window, click ‘OK’.
Test engineers[edit]
To add, delete or edit (the names of) test engineers, go to the Engineers window in the Configuration tab.
The Engineers window will show the current configured list of engineers. To leave the Engineers window, click ‘OK’.
To add a new test engineer, click ‘Add’.
Enter the name and set the rights of the new engineer, then click ‘OK’. You will be returned to the Engineer window.
The new engineer will not have a protection password (yet), this can be enabled by clicking 'Change Password'. Please read Password protection for more information.
To delete a test engineer, select the engineer and click ‘Delete’.
To confirm that you want to delete the engineer from the configuration list, click ‘Yes’. If you do not want to delete this engineer from the list, click ‘No’ or ‘Cancel’. You will be returned to the Engineer window.
To edit the name, rights and/or password of an engineer, select the engineer and click ‘Edit’.
Make the desired changes in the Test Engineer window and click ‘OK’. If you want to undo or cancel your changes, click ‘Cancel’. You will be returned to the Engineer window.
When you make a new EUT file the name and address information of the producer, test house and client can be entered separately. In many situations one or more of this information will be the same for different EUT files (i.e.: when a test house owns the software, the test house data will always be the same).
  • The default client
  • The default manufacturer
  • The default test site
Data entered under Default Address Information, will automatically be entered in the EUT window and therefore only needs to be entered once.
To add, modify or delete the default address information, go to the Default Address Information window in the Configuration tab.
Enter the desired information as shown in the example below. To confirm the data input, click ‘OK’, to cancel the data input, click ‘Cancel’. The Default Address Information window will close.
Default Units[edit]
To configure the unit settings, go to the Configuration window in the Configuration tab.
The default settings for the different units can be set in the Units tab. It is possible to define different units for different tests. The following can be selected:
  • Radiated Immunity
  • Conducted Immunity
  • Radiated Emission
  • Conducted Emission
  • Pulsed Immunity
  • Antenna Diagram
  • RadiCal
To change the settings of a unit, choose the desired test (e.g. Radiated Immunity) from the drop down menu in the Units tab.
Select the measurement (e.g. Power) and click the ‘Change’ button.
The Unit configuration window will appear, allowing you to change the parameters of the unit. To confirm your changes, click ‘OK’, to cancel your changes, click ‘Cancel’. You will be returned to the Configuration window.
For each data type the default location can be entered through the configuration. The path name can link to a directory on a local or network drive. Selecting a network drive will allow you to store data (e.g. EUT data) on a computer network, making it more accessible.
To modify a path name, go to the Configuration window in the Configuration tab.
Go to the Directories tab, select the directory that you wish to change and click the ‘Change’ button.
The Browse For Folder window will appear, allowing you to select (or make) a new folder in your directory. To confirm your change, click ‘OK’, to cancel your change, click ‘Cancel’. You will be returned to the Configuration window.
In RadiMation®, the device drivers only need to be configured to the customer requirements once. During the configuration of the device driver, information such as the IEEE address or RS232 COM-port has to be entered.
You can configure the device drivers in the configuration screen. This screen can be accessed in two ways, either through the Configuration tab or through the Devices tab.
To configure a device driver, go to the Device Drivers tab in the Configuration window. Select the desired Device Driver type from the drop-down menu (e.g. Signal Generator). The available drivers for that type will be shown.
To create a new RadiMation® device driver, select the desired device driver type (e.g. Signal Generator) and click ‘Add’.
The Device Driver Settings window will appear with information on the selected device driver. In this window information can be added, edited and removed. To confirm the settings, click ‘OK’, to cancel your changes, click ‘Cancel’. You will be returned to the Configuration window. More detailed Device Driver Settings are described in Chapter 14 "Device Driver settings".
Graphics configuration[edit]
To configure the style of the shown graphs, please select:
   Menu.svg Configuration
Then select the Graphs tab.
An example graph is shown that is created with the currently selected graph settings. When the Customize button is pressed, the Graph Customization Configuration window is shown, which allows to change the look of all graphs.
ScreenElementDescription.svg Back Color Shows the currently selected color for the back(ground) of the graph. The color can be changed by clicking on the color itself.
ScreenElementDescription.svg Grid Color Shows the currently selected color for the grid of the graph. The color can be changed by clicking on the color itself.
ScreenElementDescription.svg Text Color Shows the currently selected color for the text around the graph. The color can be changed by clicking on the color itself.
ScreenElementDescription.svg Line Style Shows the currently selected type of style that is used for the grid of the graph. Possible line styles are:
  • Dash
  • Dot
  • Thick
  • Thin
ScreenElementDescription.svg Scaling Determines if the initial x-axis scaling of the graph should be shown linear or logarithmic
The selected settings in this Graph Customization Configuration window are used as the RadiMation® default settings. It is still possible to overrule these settings in each test to more specific settings.
Database Configuration[edit]
Language Selection[edit]
RadiMation® supports multi-language operation. This means that the software package can be operated in any language. Further more, the multi- language support allows easy translation from English to virtually every other language. Please contact your local reseller for the availability of the language driver you require.
Then select the Language tab:
The currently selected language is marked with the text ' (current) ' appended. If another language should be used in RadiMation®, one of the other languages can just easily be selected. The selected language will only be active after the RadiMation® package is restarted.
Product Standard[edit]
  • Washing machine
  • Hair dryer
  • Car radio
  • Television
And then select the Product standards tab:
After pressing Add the end-user can insert all necessary information.
Please note that the Official name can only be inserted the first time and has to be unique. All fields speak for themselves, except for the File location. In the File location field, it is possible to associate a file to this standard. For example a PDF, DOC or HTML file that describes the standard. RadiMation® will open the file, with the default reader for this file type, when the engineer selects the magnifying glass.
Adding a basic standard to the product standard can be done simply by pressing the Add button. The following window will appear:
It is possible to select multiple items by selecting the top item, then hold the 'shift' key and selecting the lowest items. Holding the 'Ctrl' key allows you to select or deselect one or more items. Double clicking on an item will display the item details in a new window.
Note: When on the EUT window, a product standard is selected, or added, the data of the standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that if the configuration of a standard in the Configuration window of RadiMation® is changed, the data in the EUT file will still have the old data of the standard!!! This behaviour is correct, because it is the information of the standard that is used at the time that the EUT was tested.
Basic Standard[edit]
The error message window will ask you to abort the test or to ignore the error message and continue the test with different test equipment. Of course, to achieve maximum accuracy, one should use the same test equipment during the calibration and during the substitution tests. However, in some circumstances, changing test equipment between calibration and test will not have any influence on test results. This is for example true when a RI substitution list is made using FWD power calibration file. Changing the amplifier won’t change test results as long as the new amplifier covers the required frequency range and can deliver the required output power.
      Menu.svg Test Equipment
To make your own device driver, open the configuration window:
and select the Device drivers tab.
  2. Press the New button.
  3. Select the driver called Configurable <XXX> (i.e. if you want to make a signal generator device driver you would select the Configurable signal generator device driver).
  4. Enter a description for the device driver (for example the type number of the generator) and press Ok.
  7. Press the Edit button.
  8. A configuration screen for the device driver will appear. All required control commands for the device must be entered. Refer to the operating manual of the equipment for these codes.
Note: Passive equipment such as antenna’s, current probes, cables etc. do also need a device driver. The reason for this is that RadiMation® has to know a number of parameters of these devices. Among others the following information is relevant:
  • Frequency range
  • Maximum input power
EUT files[edit]
Creating a new EUT file[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg EUT
   Menu.svg File
      Menu.svg Open
         Menu.svg EUT
The first column displays the testnumber. The second column displays the description of the test. The third column shows the notes you inserted. The three columns can be sorted one at a time. This can be done by pressing the desired column header. If you wish to sort on testnumbers, press the first column header. If you wish to sort by note press the third column header. Pressing a column header twice will reverse the sorting order. For example the first time you press the third column header the notes will be sorted ascending, the second time you press the header the notes will be sorted descending. This pertains to all the columns.
Cable information of the EUT[edit]
RadiMation® is capable of storing a lot of EUT related information. One of these items is the cables connected to the EUT. You can specify them by selecting the EUT Information tab in the EUT window. On the bottom of the window you will find a control looking like the one below.
You can specify every cable of the EUT, pressing Add or Edit will popup the following window:
Limit line files[edit]
Creating limit lines[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Limit line
The window below will be displayed:
The user can enter a file description to the limit line in the Description field.
Entering data points[edit]
To enter data point, press the Insert button. A new line will appear and you can enter frequency and level. When entering data points, the limit line will be drawn directly in the graph. This makes it easy to check if the limit line is entered correctly.
When a data point in incorrect, it can be removed by selecting the data point and pressing the Remove button.
ScreenElementDescription.svg Color Clicking on the coloured box behind the Color selection box will change the line color. The selected line colour will be used during emission tests.
ScreenElementDescription.svg Logarithmic Interpolation In between the specified frequency points, the correct value of the limit line will be calculated by using a logarithmic interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.
ScreenElementDescription.svg Linear Interpolation In between the specified frequency points, the correct value of the limit line will be calculated by using a linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.
ScreenElementDescription.svg Detectors During emission tests, RadiMation® will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. In the Detectors menu the operator can select for which detector(s) the limit line is valid.
Saving limit line files[edit]
   Menu.svg File
      Menu.svg Save Limit line
RadiMation® will prompt you for a filename for the new limit line file. This can be the same or a different filename as the limit line Description field.
Correction files[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction
ScreenElementDescription.svg Entering data points To enter data points, press the Add Row button. A new line will appear and you can enter frequency and level.
ScreenElementDescription.svg Remove Row When a data point is incorrect, it can be removed by selecting the data point and pressing the Remove Row button.
Saving correction files[edit]
   Menu.svg File
      Menu.svg Save correction
RadiMation® is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes Radiated immunity tests [R.I.], Conducted immunity tests [C.I.] as well as Radiated emission tests [R.E.] and Conducted emission tests [C.E.]. (See also "software overview").
For each kind of test, the test configuration data (start/stop frequency, test level, dwell time etc.) can be saved in a Test Configuration File (*.TSF). Separate from the test configuration data, the TSF file also stores the calibration file and test site information. The description of the TSF file can be 40 characters max. When the same kind of test must be performed, one can simply load the TSF file and start the test.
In order to organise the different TSF files RadiMation® enables the user to create a directory tree thus enabling the user to store the TSF files in a structured manner.
A more sophisticated use of TSF files can be found during sequence testing. In a sequence, different TSF files (even from different types of tests) can be placed after each other, and started without the intervention of a test engineer (See also “Sequence testing”)
Viewing test data[edit]
RadiMation® stores all test results for the corresponding EUT in the EUT file. To view test data, open the corresponding EUT file through the File, Open, EUT pull down menu.
The EUT window will show a list of all performed tests for the selected EUT. To view the extended data from a single test, select the test from the list and press the view button. The following data will be displayed:
  • Time and date of the test.
  • Equipment used during the test (Including serial numbers).
  • Test engineer.
  • Test configuration data.
  • All graphs of AD channels used Field graphs (If used).
  • Current graph (If used).
  • Notes.
  • Signal generator level graph (Used for fault diagnoses only).
  • Forward power level graph (If applicable).
  • Reflected power level graph (If applicable).
  • Net power level graph (If applicable).
  • Ambient information:
  • Temperature graph (If used).
  • Pressure graph (If used).
  • Relative humidity graph (If used).
  • Calibration file used during the test (If applicable).
Printing test data[edit]
Test data can be printed in several different ways:
  • When only a part of the test data needs to be printed, select the desired test, then select view and select the graph you want to print. The graph will be displayed on the screen. Above the graph you will find a print button. This print button will only print the current selected graph.
  • From the list of performed tests, a single test can be selected. When all data of this test must be printed, simply press the print button at the right side of the list.
  • When you want to print the data of more than one test, select the tests to print while holding down the shift button. All selected tests will be printed when the print button is pressed.
Warning: All print jobs from RadiMation® will be printed on the default Windows printer. To use another printer, please change the default printer in your Windows environment.
Report generation[edit]
RadiMation® can also use the information of the performed tests, to automatically generate a report, or export all the data to Microsoft Excel. The Report Generator and Excel Exporter can be used to perfrom these tasks. See Chapter 13 for a detailed description.
Ambient information[edit]