| Line 392: | Line 392: | ||
Device drivers let your software know which equipment has to be used and controlled during EMC tests. | Device drivers let your software know which equipment has to be used and controlled during EMC tests. | ||
In {{RadiMation}}, the device drivers only need to be configured to the customer requirements once | In {{RadiMation}}, the device drivers only need to be configured to the customer requirements once. | ||
During the configuration of the device driver, information such as the IEEE address or RS232 COM-port has to be entered. | |||
You can configure the device drivers in the ''' 'Configuration' window'''. This window can be accessed in one of two ways, either through the ''' 'Configuration' tab''' or through the ''' 'Devices' tab'''. | |||
You can configure the device drivers in the ''' 'Configuration' window'''. | |||
[[Image:Config configuration.png]] | [[Image:Config configuration.png]] | ||
| Line 404: | Line 409: | ||
To configure a device driver, go to the ''' 'Device Drivers' tab''' in the ''' 'Configuration' window'''. Select the desired ''' '[[Device Types|Device Driver Type]]' ''' from the drop-down menu (e.g. Signal Generator). The available drivers for that type will be shown. | To configure a device driver, go to the ''' 'Device Drivers' tab''' in the ''' 'Configuration' window'''. | ||
Select the desired ''' '[[Device Types|Device Driver Type]]' ''' from the drop-down menu (e.g. Signal Generator). The available drivers for that type will be shown. | |||
[[Image:Device driver type.png]] | [[Image:Device driver type.png]] | ||
| Line 414: | Line 421: | ||
Select a new device driver (in the scroll menu or through the search function) and click ''' 'New' '''. To cancel your selection and return to the ''' 'Configuration' window''', click ''' 'Close' '''. | Select a new device driver (in the scroll menu or through the search function) and click ''' 'New' '''. | ||
To cancel your selection and return to the ''' 'Configuration' window''', click ''' 'Close' '''. | |||
[[Image:New signal generators.png]] | [[Image:New signal generators.png]] | ||
The name/description of the chosen device driver can be changed in the ''' 'Device Driver Configuration' window'''. To confirm your changes (or the provided description), click ''' 'OK' ''', the ''' 'Device Driver Settings' window''' will appear. To cancel your changes, click ''' 'Cancel' '''. You will be returned to the previous window (e.g. New Signal generators). | The name/description of the chosen device driver can be changed in the ''' 'Device Driver Configuration' window'''. | ||
To confirm your changes (or the provided description), click ''' 'OK' ''', the ''' 'Device Driver Settings' window''' will appear. | |||
To cancel your changes, click ''' 'Cancel' '''. You will be returned to the previous window (e.g. New Signal generators). | |||
[[Image:Device driver configuration.png]] | [[Image:Device driver configuration.png]] | ||
The ''' 'Device Driver Settings' window''' will appear with information on the selected device driver. In this window information can be added, edited and removed. To confirm the settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. More detailed device driver settings are described in [[#Device Driver settings|Chapter 14 "Device Driver settings"]]. | The ''' 'Device Driver Settings' window''' will appear with information on the selected device driver. In this window information can be added, edited and removed. | ||
To confirm the settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | |||
More detailed device driver settings are described in [[#Device Driver settings|Chapter 14 "Device Driver settings"]]. | |||
[[Image:Device driver settings.png]] | [[Image:Device driver settings.png]] | ||
Revision as of 12:59, 23 October 2015
Getting started[edit]
Software overview[edit]
- The Visual and Graph System
- Sequence Testing
- Calibration
- Data Logging
- Database Functionality
- The Device Driver System
- Report generator
- Excel Exporter
Hardware requirements[edit]
- Monitor: SXGA (1280 x 1024) or larger (at least 1280 pixels wide and at least 900 pixels high)
Memory[edit]
RadiMation® needs a considerable amount of memory to perform at its best. We recommend that you have at least 512 MB (Mega Bytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM without having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps to help you determine the amount of free memory in your PC:
GPIB card[edit]
The GPIB hardware and software needs to be installed correctly before RadiMation® is installed. DARE!! is not responsible for any additional costs or delays that may occur if this card is not installed prior to the installation of RadiMation®. The GPIB Card has to be supported by the National Instruments NI-488.2 software library.
Software requirements[edit]
The RadiMation® software will operate (and is tested) with Microsoft Windows 7 (32-bit and 64-bit) and Microsoft Windows 8 (32-bit and 64-bit) operating systems. It is required to have Microsoft Office Word installed on the PC because RadiMation® uses Microsoft Office Word to generate and print the printout's.
Microsoft office[edit]
- Microsoft Word
- Microsoft Excel
Installation[edit]
During the installation of RadiMation® a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. DARE!! is not responsible for any additional costs or delays that may occurs if this support is not provided.
- Full rights to the installation directory (e.g. 'C:\Program Files\Dare Development' or 'C:\Program Files\Dare Instruments', please note that the path may be different depending on the used OS and language)
Conventions[edit]
File
Open
EUT
Software installation[edit]
To install the RadiMation® software, please insert the RadiMation® CD in the CD-ROM drive of your computer. The setup program of RadiMation® should start automatically. If the program does not start automatically, run E:\setup.exe from the CD (where E:\ is the letter of your CD-ROM drive). The Installation Wizard will guide you through the installation process. After installation the computer must be restarted before running RadiMation®. Please read User rights and installation about the user rights and information that is required for first time installation.
In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages compaired to a local installation. One of these benefits is that data can be shared between the connected computers. The RadiMation® software is designed to optimally use this advantage. All the data that is used or generated by RadiMation® can also be shared on the network. In most cases we recommend the use a of network configuration over a local configuration.
- CAL_File
- CONFDVDR
- COR_File
- EUT_File
- LLF_File
- Printout
- RGO_File
- RGT_File
- SEQ_File
- TSF_File
Software updates can be distributed on CD or downloaded from our website at http://www.dare.eu/instruments/emc-test-software/download.
To install the downloaded RadiMation® update, please execute the startup program.
To install the RadiMation® update from a CD, please insert the CD in the CD-ROM drive of your computer. Run the installation program (i.e. E:\setup.exe, where E:\ is the CD-ROM drive letter). The setup program will detect that a previous version of RadiMation® has already been installed on your PC. The setup program will install the new RadiMation® version at the same location as the previously installed version. A back up of this older version will automatically be copied to the backup directory. All the settings of the older version will also be used for the newly installed RadiMation®. If problems should occur with the newly installed RadiMation®, the old version of RadiMation® can be recovered.
| Note: | We strongly advice that a you make a complete backup of all your files and directories (e.g. TSF/Sequence/Correction/Calibration and EUT) before installing a RadiMation® update. |
| Warning: | Do NOT uninstall RadiMation®!! This will cause removal of all user settings, TSF, sequence, calibration and correction files. This is unrecoverable. |
If the already installed RadiMation® version was installed from a network directory, the new setup file can be copied to the <Setup> sub-directory of the <NetworkDir>. The new RadiMation® version can then be installed on each computer from this directory.
This chapter describes the rights a user must have to install or upgrade RadiMation® software.
New installation[edit]
Upgrade[edit]
To upgrade the RadiMation® software, the user must have the rights to read and write in the RadiMation® folder, which is normally located in the C:\Program Files (x86)\DARE Instruments\ directory.
Software configuration[edit]
Password protection[edit]
RadiMation® allows you to protect a number of configurations with a password.
The password is standard turned on for the default used test engineer 'Administrator'.
The default password is Radimation.
To change the password protection settings, go to the 'Engineers' window in the 'Configuration' tab.
Select the Administrator Engineer and click 'Edit' .
In the (Administrator) 'Test Engineer' window the password can be enabled, disabled or changed.
To enable, disable or change the password, click 'Change Password' .
To leave the 'Engineers' window, click 'OK' .
Test engineers[edit]
- The default client
- The default manufacturer
- The default test site
Default Units[edit]
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
More detailed device driver settings are described in Chapter 14 "Device Driver settings".
Graphics configuration[edit]
To configure the style of the shown graphs, go to the Configuration window in the Configuration tab.
An example graph is shown in the Graphs tab that is created with the current settings. To change these graph settings, click ‘Customize’.
The Graph Customization Configuration window contains three tabs; Chart, Graph and Markings.
The Chart tab contains the settings related to the overall graph and can be used to determine the colors, layout and titles.
The Graph tab contains the settings related to the graph lines shown in the graph and can be used to determine the colors, scaling and layout of each and every shown graph line.
The Markers tab contains the settings related to the markers on the graph and can be used to determine the colors, style and scaling of each and every shown marker.
Click ‘Update’ to update the graph settings without closing the Graph Customization Configuration window. This will allow you to see the effect of the new settings and make additional changes.
Click ‘Close’ to update (and save) the graph settings and close the Graph Customization Configuration window. You will be returned to the Configuration window.
The selected settings in the Graph Customization Configuration window are used as the RadiMation® default settings. It is still possible to overrule these settings for individual tests to more specific settings.
Click ‘Restore factory default’ to return to the factory settings.
Database Configuration[edit]
Language Selection[edit]
RadiMation® supports multi-language use. This means that the software package can be operated in several language. In addition, the multi- language support allows for easy translation from English to most other languages. Please contact your local reseller for the availability of your required language driver.
The default language of RadiMation® is English. To change the language settings, go to the Configuration window in the Configuration tab.
Go to the Language tab. The current language setting is marked with the text (current).
To change the language setting, select one of the other languages, click ‘Close’ (to leave the Configuration window) and restart RadiMation®. The selected language will only be active after RadiMation® is restarted.
| Note: | Not all languages are currently available. Please contact your local reseller for language availability. |
Product Standard[edit]
Product Standards describe the requirements (e.g. dimensional, technical, etc.) and the methods of testing (as well as grading, marking, etc.) of a product. The objective of these standards is to define and impose ‘standards’ that will ensure that the product in question will not represent harm or hazard to consumers.
There are many Product Standards for individual products and product types. These, in turn, often refer to multiple Basic Standards. Where a Product Standard describes all the tests that should be performed on a EUT, a Basic Standard only describes one single test.
For example: The Product Standard for product X describes four different requirements/tests that the product needs to comply with. These four tests are described in four Basic Standards, where each basic standard describes the requirements for one test.
Basic Standards can also be added to RadiMation®. In addition, one single Technical Setup File (TSF) can be generated for each Basic Standard, configured to perform the test as described in the standard.
To add a Product Standard to RadiMation®, go to the Configuration window in the Configuration tab.
Select the Product standards tab and click ‘Add’.
(To edit a Product Standard, select the standard and click ‘Edit’. To remove a Product Standard, select the standard and click ‘Remove’.)
Insert the required information in the Standard details window (as visible in the example below).
Please note that the Official name can only be inserted the first time and has to be unique. In addition, it is possible to link a file to the standard in the File location field. For example a PDF, DOC or HTML file that describes the standard. RadiMation® will open the file (with the default reader for this file type) when the engineer selects the magnifying glass button next to that field.
To confirm your data input, click ‘Ok’, to cancel your input, click ‘Cancel’. You will be returned to the Configuration window.
To add a Basic Standard to the product standard, click ‘Add’. The Select Standard window will appear. Then, select one or more Basic Standards (hold the ‘Ctrl’ key to select multiple standards) and click ‘Select’. To cancel your selection, click ‘Cancel’. You will be returned to the Standard details window.
| Note: | When a product standard is selected or added in the EUT window, the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the Configuration window of RadiMation® is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested. |
Basic Standard[edit]
To add a Basic Standard to RadiMation®, go to the Configuration window in the Configuration tab.
Select the Basic standards tab and click ‘Add’.
(To edit a Basic Standard, select the standard and click ‘Edit’. To remove a Basic Standard, select the standard and click ‘Remove’. Please note that removing a Basic Standard from the list will also remove any references of this standard entered in the Product Standards.)
The test equipment configuration contains a list of all the test equipment used during a test. The software uses this list to determine which device drivers to use to control the test equipment. RadiMation® allows you to configure specific test equipment lists for each type of test as well as different test equipment lists for the same type of tests. (This can be useful if you have, for example, more than one signal generator to perform conducted immunity tests with).
In the Test Equipment window a set of test equipment will be selected. When a calibration file is generated, RadiMation® will record this test equipment configuration together with the calibration data. When a substitution test is started, the test equipment selected for the substitution test will be compared with the test equipment used (as recorded) during the calibration test. If the equipment does not match, an error message will be displayed.
To add, modify or delete test equipment lists, go to the Test Equipment window in the Configuration tab.
If no test equipment has been added to RadiMation®, click the ‘Add’ button in the top of the window.
Enter a name for the new test equipment in the New Test Equipment window (e.g. Example Test Equipment). Click ‘Ok’ to confirm the name of the new test equipment, or click ‘Cancel’ to cancel the new name. You will be returned to the Test Equipment window.
To delete a test equipment from RadiMation®, select the test equipment from the pull down menu and click the ‘Delete’ button in the top of the window. A confirmation window will appear. Click ‘Yes’ to remove the device from the list, click ‘Cancel’ to cancel. You will be returned to the Test Equipment window.
To configure the equipment of a test, select the desired ‘Test Equipment’ from the pull down menu (e.g. Example Test Equipment). Specify the equipment for the test by adding, editing and/or removing devices.
To add a device, click ‘Add’ (at the bottom left of the window), a menu will open below the button to show the available devices for each Device Type.
Another way to add a device is to click the desired Device Type in the equipment overview. A menu will open (at your mouse point) with the available devices for that Device Type. Select and click on the desired device. The device will then appear in the test equipment overview.
To edit a device, select the device and click ‘Edit’. The Device Driver Settings window will appear. Edit the settings as desired, to confirm your changes, click ‘OK’, to cancel your changes, click ‘Cancel’. You will be returned to the Test Equipment window.
To remove a device, select the device and click ‘Remove’. A confirmation window will appear. Click ‘Yes’ to remove the device from the list, click ‘Cancel’ to cancel. You will be returned to the Test Equipment window.
To save the changes made in the test equipment list, click ‘SAVE’ (you will remain in the Test Equipment window), to save the test equipment list under a new name, click ‘SAVE AS’.
Enter a new name for the test equipment in the New Test Equipment window (e.g. Example Test Equipment 2). Click ‘Ok’ to confirm the name of the new test equipment, or click ‘Cancel’ to cancel the new name. You will be returned to the Test Equipment window.
To close the Test Equipment window, click ‘Close’. If you have not saved the changes to the test equipment list, a dedicated window will appear asking you to save these changes. To save the changes, click ‘Yes’, to not save the changes, click ‘NO’. The Test Equipment window will close.
To make your own device driver, go to the Configuration window in the Configuration tab.
Go to the Device Drivers tab and select the required device driver type from the pull down menu (e.g. Signal generators), then click ‘Add’.
Select the Configurable driver (e.g. the Configurable Signal Generator) and click ‘New’.
Enter a description for the device driver (e.g. the type number of the generator) and click ‘OK’, the Device Driver Settings window will appear. To cancel your changes, click ‘Cancel’. You will be returned to the previous window (e.g. New Signal generators).
Edit he information about your ‘user configurable device driver’ in the Device Driver Settings window. Click ‘Advanced’ and enter the required control commands for the device in the tabs of the dedicated window.
After the commands are entered, the device driver is ready for use. Please refer to the operating manual of the equipment for the required commands.
To confirm the advanced settings, click ‘OK’, to cancel your changes, click ‘Cancel’. You will be returned to the Device Driver Settings window.
To confirm your (final) settings, click ‘OK’, to cancel your changes, click ‘Cancel’. You will be returned to the Configuration window.
Your new ‘user configurable device driver’ is now visible on the Available Device Drivers list in the Configuration window and can be used as any other device driver (by selecting it in the Equipment list).
| Note: | Passive equipment, such as antennas, current probes, cables etc., also need a device driver. This is required because RadiMation® has to know a number of parameters for these devices. Among others the following information is relevant:
EUT files[edit]
Creating a new EUT file[edit]
To perform test(s) on a new device, first create a new EUT file. Select ‘File’, ‘New’, ‘EUT’ or press ‘Ctrl+E’.
RadiMation® will ask you to specify and save a file name for this new EUT file. Name the EUT file and click ‘Save’. (All the test data for this EUT will be stored under this EUT file name.) The EUT window will open.
To perform test(s) on a previously tested device, open the existing EUT file for that device. Select ‘File’, ‘Open’, ‘EUT’.
RadiMation® will ask you to specify and open an existing EUT file. Select the desired EUT file and click ‘Open’. (The new test data will be added to the already existing EUT file.)
EUT window overview[edit]
After performing a number of test the EUT window will look something like the example shown below.
The EUT window contains five tabs; ‘Main’, ‘EUT Information’, ‘Monitoring input channels’, ‘Standards’ and ‘Reports’.
- The Main tab consists of six areas where Client, Manufacturer, EUT, Order Number, Test Site and Test information can be found and specified.
- The EUT information tab consists of two areas; EUT and Cables. Here (more detailed) EUT information can be specified. This includes the cables that are connected to the EUT.
- The Monitoring input channels tab consists of two areas that can be used to monitor the input channels and setup short cut keys (that can be used during measurements).
- The Standards tab consists of two areas; Product Standards and Basic Standards. Standards related to the EUT can be added in these fields. (Documentation for standards can be added to RadiMation® in the ‘Product standards’ and ‘Basic standards’ tabs of the Configuration window.)
- The Reports tab can be used to generate reports with the configured report generator, export reports to a spread sheet program and print detailed EUT information.
More information about the content of these tabs can be found in Chapter 3, EUT Information.
Limit line files[edit]
Creating limit lines[edit]
To create a Limit Line File, select ‘File’, ‘New’, ‘Limit Line’. The Limit Line window will open.
The default Limit Line window will appear. This window contains six fields; Description, Data, Options, Detectors, NB/BB and the graph area.
| Enter a file description for the limit line in the Description field (e.g. Limit Line Example). |
| Enter frequency-value pairs for the limit line in the Data field (see ‘Enter, edit & remove data pairs’ below). |
Setup the Options field:
| Setup the Detectors field by selecting one or more detectors. During emission tests, RadiMation® will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. |
| Setup the NB/BB field by selecting the option(s) Narrowband and/or Broadband. |
To enter a frequency-value pair, click the ‘Add Row’ or ‘Insert Row’ button, a new row will appear in the data field. Select the desired row and enter the frequency and value data, the limit line will be drawn directly in the graph.
To edit a data pair, select the desired value and/or unit of the data that you wish to edit and insert the new value and/or unit.
To remove a frequency-value pair, select the desired data pair and click ‘Remove Row’.
Saving limit line files[edit]
Before a limit line can be saved, make sure the required information has been entered into the ‘Description’, ‘Data’ and ‘Detectors’ fields. Otherwise a RadiMation Warning will appear when you try to save the limit line.
To save a Limit Line File, select ‘File’, ‘Save Limit line’ (or ‘Save Limit line As…’). The Save Limit line As window will open. (The Limit Line window will remain open in the background.)
RadiMation® will ask you to specify and save a file name for the (new) Limit Line File. This can be the same or a different filename as used in the limit line ‘Description’ field of the Limit Line window. Name the Limit Line File and click ‘Save’. (The Limit Line window will remain open in the background.)
To open an existing limit line file, select ‘File’, ‘Open’, ‘Limit Line’. The Limit Line window will open.
RadiMation® will ask you to specify and open an existing Limit Line File. Select the desired Limit Line File and click ‘Open’. The specified Limit Line window will appear.
Correction files[edit]
Correction files can be used to correct the frequency response of cables, generators, clamps, antenna’s, couplers, etc. When the frequency response of a device is known, the operator can create a new correction file and assign it to the device.
To create a Correction File, select ‘File’, ‘New’, ‘Correction’ or press ‘Ctrl+O’. The Correction File window will open.
The default Correction File window will appear. This window contains three fields; Data, Options and the graph area.
| Enter data pairs for the correction value in the Data field (see ‘Enter, edit & remove data pairs’ below). |
| Select either logarithmic or linear interpolation. In between the specified frequency points, the correct value of the correction value will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. |
The ‘Data columns’ need to be defined before data pairs can be added.
To select the desired columns (based on the use of the correction file), click ‘Columns/Units’. The Correction Columns selection window will appear where columns can be added and/or removed. (E.g. an amplifier correction file will need two columns; frequency and power. A radiated emission test requires three columns; frequency, angle and height.) Click ‘Close’ to return to the Correction File window.
To enter a data pair, click the ‘Add Row’ or ‘Insert Row’ button, a new row will appear in the data field. Select the desired row and enter the information for the data pair, the correction value will be drawn directly in the graph.
To remove a data pair, select the desired data point row and click ‘Remove Row’.
Saving correction files[edit]
To save a Correction File, select ‘File’, ‘Save Correction’ (or ‘Save Correction As…’). The Save Correction As window will open. (The Correction File window will remain open.)
RadiMation® will ask you to specify and save a file name for the (new) Correction File. Name the Correction File and click ‘Save’. (The Correction File window will remain open.)
To open an existing limit line file, select ‘File’, ‘Open’, ‘Correction’. The Open Correction window will open.
RadiMation® will ask you to specify and open an existing Correction File. Select the desired Correction File and click ‘Open’. The specified Correction File window will appear.
A more advanced use of the TSF files takes place during sequence testing. In a sequence, different TSF files (even from different types of tests) can be lined up and started without the intervention of a test engineer (See Chapter 11 "Sequence Testing" for more information.)
Viewing test data[edit]
RadiMation® stores all the test data in the corresponding EUT files. To view the test data, open the desired EUT file through the 'File, 'Open', 'EUT' pull down menu .
The EUT window will show a list of all the performed tests for the selected EUT (in the Main tab). To view the extended data from a single test, select the desired test from the list and click the ‘Info’ button on the right side of the window.
- Time and date of the test
- Test engineer
- Notes
- Ambient information
- Test graphs
- Test configuration data
Printing test data[edit]
Test data can be printed in a few different ways (from the EUT window):
- To print part of the data from one test; select the desired test and click ‘Info’. In the 'test data window', select the graph that you want to print. The graph will be displayed on the screen. Click the ‘print graph’ button above the graph (containing a graph and printer icon). The currently selected graph will be printed.
- To print all the data from one test; select the desired test and click the ‘Print’ button at the right side of the list (in the EUT window). The data from the selected test will be printed.
- To print the data of more than one test; select the desired tests (in the EUT window) while holding the ‘Shift’ button (on your keyboard). Click the ‘Print’ button at the right side of the list, the selected tests will be printed.
| Warning: | Printing from RadiMation® will be done by your default Windows printer (with its default settings). To change the printer/settings used by RadiMation®, please change the default printer settings in your Windows environment. |
Report generation[edit]
RadiMation® can use the information of the performed tests to automatically generate a report as well as export all the data to Microsoft Excel. The Report Generator and Excel Exporter can be used to perform these tasks. See Chapter 13 for a detailed description.
Ambient information[edit]