(→Configuration) |
(→Antenna: Antenna Efficiency) |
(65 intermediate revisions by 4 users not shown) | |||
Line 15: | Line 15: | ||
== Engineers == | == Engineers == | ||
− | An unlimited number of | + | On a new installation, only a single 'Administrator' engineer is created by default. An unlimited number of engineers can be added, each with an unique name. For each engineer a login password and different permissions can be assigned which give them the right to modify specific parts of the {{RadiMation}} configuration. There is no relation between the engineer in {{RadiMation}} and the Windows Logon account. Also the password that can be configured for the engineer in {{RadiMation}} is not linked and also not synchronized with the Windows User account password. |
+ | Automatic login of the engineer can be disabled or enabled. If it is enabled, it can only be selected for one engineer, and that engineer will be automatically logged in the next time that {{RadiMation}} starts. If the {{ScreenElement|Automatically login with this engineer}} checkbox is selected for an engineer, the automatic login of all the other engineers will not be performed anymore. For changing setting of the the automatically logged in engineer, it is still necessary to close the login dialog by specifying the corresponding {{ScreenElement|Password}} and using the {{ScreenElement|Ok}} button. | ||
+ | Switching between the active engineers while {{RadiMation}} is running can be done through the menu: | ||
{{Menu|File|Change Engineer}} | {{Menu|File|Change Engineer}} | ||
− | To add, delete or edit the names and permissions of the | + | [[Image:File Change Engineer.png]] |
+ | To add, delete or edit the names and permissions of the engineers, go to the {{ScreenElement|Engineers}} window, which can be accessed in the menu: | ||
{{Menu|Configuration|Engineers}} | {{Menu|Configuration|Engineers}} | ||
Line 212: | Line 218: | ||
[[Image:Test equipment close.png]] | [[Image:Test equipment close.png]] | ||
=== Before Action === | === Before Action === | ||
Line 266: | Line 279: | ||
== Default address information == | == Default address information == | ||
− | + | In many situations one or more of those address information will be the same for different [[EUT]] files (for example, when a test site use {{RadiMation}}, the test site address information will always be the same). | |
− | + | To prevent users from having to enter repetitive data input, {{RadiMation}} provides the possibility to set default address information for: | |
− | + | Address information entered on the {{WindowTitle|Default Address Information}} window, will be automatically included on the {{WindowTitle|EUT}} window and therefore only needs to be entered once. | |
− | + | To add, modify or delete the default address information, use the menu: | |
− | + | {{Menu|Configuration|Default Address Information}} | |
− | + | In the {{WindowTitle|Default Address Information}} window, the default address information can be configured. Specifying this information is not required, however it has the benefit that address information that is always the same doesn't need to be entered over and over again in every new [[EUT]] file. | |
+ | [[Image:Default Address Information.png]] | ||
− | + | {{ScreenElementDescriptionStart}} | |
− | + | {{ScreenElementDescription|Client Company|The default company name for the client.}} | |
− | + | {{ScreenElementDescription|Client Contact Person|The default contact person for the client.}} | |
− | + | {{ScreenElementDescription|Client Address|The default address for the client.}} | |
− | + | {{ScreenElementDescription|Client Zip Code|The default zip code for the client.}} | |
− | + | {{ScreenElementDescription|Client City|The default city for the client.}} | |
− | + | The default address information is stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured in the global {{RadiMation}} configuration. | |
== Advanced options == | == Advanced options == | ||
Line 356: | Line 399: | ||
=== 2. Directories configuration === | === 2. Directories configuration === | ||
− | The {{RadiMation}} software can use different directories to store different | + | The {{RadiMation}} software can use different directories to store different files. |
− | The location of the | + | The location of the Device Driver Configuration Files can be specified: |
+ | * Device Driver Configuration Files (CONFDVDR), | ||
+ | The location of the following data file types can be specified: | ||
* Correction files (*.COR files) | * Correction files (*.COR files) | ||
Line 370: | Line 415: | ||
− | For each data type the default location can be entered through the configuration. The path name can link to a directory on a local or network drive. | + | For each data type the default location can be entered through the configuration. The path name can link to a directory on a local or on a shared network drive. Selecting a network location will allow you to store data (e.g. EUT data) on a computer network, making it more accessible. This will allow you to configure {{RadiMation}}, to make use of a networked configuration with multiple test PC's, sharing the same configuration. |
− | + | ==== Open the 'Directories' configuration screen ==== | |
+ | Open the directory configuration by selecting from the menu: | ||
+ | {{Menu|Configuration|Configuration}} | ||
− | + | Go to the {{ScreenElement|Directories}} tab. | |
− | + | [[Image:Configuration Directories.png]] | |
− | + | ==== Modifying the 'Device Driver Configuration Files' directory ==== | |
+ | With the {{ScreenElement|Use standalone configuration}} option, it is possible to create an version specific isolated configuration directory with it's own content only used for this version of RadiMation. This can for example be used to verify and validate another {{RadiMation}} version without influencing another {{RadiMation}} version. | ||
− | + | The location of the {{ScreenElement|Device Driver Configuration Files}} can by changed by clicking on the {{ScreenElement|'Open Folder' icon}}. A ''' 'Browse For Folder' ''' window will appear, allowing to select an existing device driver configuration files directory. If an empty or an invalid device driver configuration files directory is selected, an error will be shown, and the original selected directory will remain selected. | |
− | + | ==== Creating a new 'Device Driver Configuration Files' folder ==== | |
+ | ==== Duplicating a 'Device Driver Configuration Files' folder ==== | ||
+ | It is possible to duplicate an existing configuration folder, by clicking on {{ScreenElement|Edit}} and either choose {{ScreenElement|Duplicate and set from current global folder}} or {{ScreenElement|Duplicate and set from current standalone folder}}. A directory chooser will be prompted to create and / or select a new empty folder to copy the content of the source configuration to. After the duplication of the folder, the new {{ScreenElement|Device Driver Configuration Files}} directory will be selected. | ||
− | + | ==== Change the path for a file type ==== | |
+ | Select the directory that you wish to change and click on the {{ScreenElement|Open Folder}} icon to select a different directory. | ||
− | + | The {{ScreenElement|Browse For Folder}} window will appear, allowing you to select (or make) a new folder in your directory. | |
+ | To confirm your change, click {{ScreenElement|Select folder}}, or to cancel your change, click {{ScreenElement|Cancel}}. You will be returned to the {{ScreenElement|Configuration}} window. | ||
− | + | [[Image:Browse Calibration File Directory.png]] | |
+ | Click {{ScreenElement|Close}} to leave the {{ScreenElement|Configuration}} window. | ||
=== 3. Device Driver configuration === | === 3. Device Driver configuration === | ||
Line 533: | Line 589: | ||
This is a very helpful tool because it is a fast method to check whether the advanced settings (especially the ([[GPIB]]) address) of the device are set correctly. If this 'Check' is successful, the 'device check' during the test initialization (which performs the same check) will also be successful. | This is a very helpful tool because it is a fast method to check whether the advanced settings (especially the ([[GPIB]]) address) of the device are set correctly. If this 'Check' is successful, the 'device check' during the test initialization (which performs the same check) will also be successful. | ||
− | The check that is performed depends on the implemented device driver. For example, some devices (like antenna's, couplers, LISN's, injection clamps, etc.) cannot be checked for the correct connection to the PC (done at step 4), because there is no connection at all. In addition, some device drivers assume that the | + | The check that is performed depends on the implemented device driver. For example, some devices (like antenna's, couplers, LISN's, injection clamps, etc.) cannot be checked for the correct connection to the PC (done at step 4), because there is no connection at all. In addition, some device drivers assume that the laser of the device is activated before the check is successful. Despite these deviations, the 'Check' functionality will still work and show the message displayed above, as long as the device driver passes the (appropriate) check. |
==== Storing the driver version in the test file ==== | ==== Storing the driver version in the test file ==== | ||
Line 936: | Line 989: | ||
*Power | *Power | ||
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum allowed reflected power. During the regulation of the test-level the reflected power level is measured and the regulation is ensuring that the reflected power will not become higher than the maximum reflected power as specified in the correction file. | The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum allowed reflected power. During the regulation of the test-level the reflected power level is measured and the regulation is ensuring that the reflected power will not become higher than the maximum reflected power as specified in the correction file. | ||
+ | ===Antenna: Ant Gain vs Angle=== | ||
+ | This correction is only available when the Antenna Diagram module is activated. The antenna gain is depend on the measuring angle during antenna diagram measurements. A correction file can be used to specify the AntennaGain depending on the receiving angle. This correction file should be attached to the device driver of the antenna. The correction file should contain the columns: | ||
+ | *Angle | ||
+ | *Attenuation | ||
+ | The Antenna Efficiency is expressed as a unit less number and the value should be between -1 and 1. | ||
+ | *Frequency | ||
+ | *Correction | ||
+ | Where the values in the correction column are used as a correction which is added on top of the the Antenna Efficiency that is specified in the antenna device driver itself. | ||
===Antenna: Antenna Factor=== | ===Antenna: Antenna Factor=== | ||
− | For converting the reading dBuV into dBuV/m the antenna factor is needed. | + | For converting the reading dBuV into dBuV/m the antenna factor is needed. This is expressed as dB/m. |
The correction file should contain the columns: | The correction file should contain the columns: | ||
Line 944: | Line 1,013: | ||
− | Where the correction column is the antenna factor. | + | Where the correction column is the antenna factor. The values are added to the calculation to account for the non-linearity in the antenna. |
+ | *Frequency | ||
+ | *Antenna gain | ||
+ | *Height | ||
===Antenna: Magnetic Flux Density (pT/A) correction=== | ===Antenna: Magnetic Flux Density (pT/A) correction=== | ||
− | During immunity measurements when a | + | During immunity measurements when a magnetic field has to be generated, this correction file is used to determine the current that should be generated during the magnetic field generation loop antenna. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns: |
*Frequency | *Frequency | ||
The value specified in the correction column should be expressed in pT/A. The values in the correction column are multiplied with the generated current to calculate the generated magnetic field in picoTesla (pT). | The value specified in the correction column should be expressed in pT/A. The values in the correction column are multiplied with the generated current to calculate the generated magnetic field in picoTesla (pT). | ||
− | This correction is primarily used during the MIL-STD | + | This correction is primarily used during the [[MIL-STD-461]] RS101 magnetic field tests. In the MIL-STD 461, this correction value is specified as 9.5x10^7 pT/A. |
===Antenna: Magnetic Field Factor (dBpT/uV) correction=== | ===Antenna: Magnetic Field Factor (dBpT/uV) correction=== | ||
Line 958: | Line 1,042: | ||
− | The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels to calculate the correct to be displayed emission values in a Tesla unit. | + | The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla unit. |
+ | This same correction is also used during the [[MIL-STD-461]] RS101 Magnetic Field calibration for the 'calibration antenna'. | ||
− | + | ===Antenna: Reflection coefficient=== | |
− | + | *Frequency | |
− | * | + | *Correction |
===Cable: Cable loss=== | ===Cable: Cable loss=== | ||
Line 1,030: | Line 1,113: | ||
[[Image:Calibration_jig_factor_50_Ohm_for_50_Ohm_system.png]] | [[Image:Calibration_jig_factor_50_Ohm_for_50_Ohm_system.png]] | ||
+ | == Using a transfer correction file provided by the supplier == | ||
+ | Sometimes a file with values for the transfer correction is delivered by the manufacturer. It is unusual and does not happen that often. If you would like to use the transfer file provided by the manufacturer you have to set the correction factor described above to zero dB. The correction file is then attached directly to the jig. However you should then use a congifurable driver for the jig where you manually set the transfer factor to 0dB. If you use the original driver then the correction factor is added with the programmed factor for that driver. | ||
+ | Normally it is not needed to add this transfer file and using the correct driver for the jig should have the correct factor already added nominally. It is also unusual for manufacturers to add the transfer factor but sometimes this is delivered. | ||
===Coupler: Forward Correction=== | ===Coupler: Forward Correction=== | ||
The forward power attenuation of a coupler can be corrected by a correction file. A correction file with the correction should be attached to the device driver of the coupler. The correction file should contain the columns: | The forward power attenuation of a coupler can be corrected by a correction file. A correction file with the correction should be attached to the device driver of the coupler. The correction file should contain the columns: | ||
− | *Correction | + | * Correction. The values in the 'Correction' column should be expressed in dB, and these values are added to the typical attenuation of the coupler. |
Or | Or | ||
Or | Or | ||
− | The values | + | |
===Coupler: Reflected Correction=== | ===Coupler: Reflected Correction=== | ||
Line 1,058: | Line 1,149: | ||
The impedance of a current sensor is very important to calculate the correct measured current. The impedance is current sensor and frequency dependant. Before a current sensor can be used, an impedance correction file should be attached to the device driver of the current sensor. The value specified in the correction file is used to correctly calculate the current flowing through the current sensor. The correction file should contain the columns: | The impedance of a current sensor is very important to calculate the correct measured current. The impedance is current sensor and frequency dependant. Before a current sensor can be used, an impedance correction file should be attached to the device driver of the current sensor. The value specified in the correction file is used to correctly calculate the current flowing through the current sensor. The correction file should contain the columns: | ||
− | * Correction | + | * Correction. The values in the 'Correction' column should be expressed in Ohm ({{Ohm}}). |
Or | Or | ||
− | * Resistance | + | * Resistance. The values in the 'Resistance' column will use the unit that is configured for the 'Resistance' column. |
− | The values in the | ||
===Field sensor: Field offset=== | ===Field sensor: Field offset=== | ||
The measured value of a field sensor, can be corrected by the software. The measured value can be corrected with an offset, or it can be corrected by a multiply factor. Both corrections can also be used together but the multiply correction will be performed first, and after that the offset correction will be added. The correction files with the correction should be attached to the device driver of the field sensor. The correction file for the field offset should contain the columns: | The measured value of a field sensor, can be corrected by the software. The measured value can be corrected with an offset, or it can be corrected by a multiply factor. Both corrections can also be used together but the multiply correction will be performed first, and after that the offset correction will be added. The correction files with the correction should be attached to the device driver of the field sensor. The correction file for the field offset should contain the columns: | ||
− | *Correction | + | * Correction. The values in the column used for the offset correction should be expressed in V/m. This correction is then applied on the isotropic value. |
− | The values in the correction column | + | or |
+ | * Frequency | ||
+ | When the correction is specified for the individual axis, those axis are first corrected, after which the isotropic value is calculated based on the individually corrected axes. | ||
− | To summarize this in an example, let’s | + | To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field offset isotropic correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m). |
===Field sensor: Field Multiply=== | ===Field sensor: Field Multiply=== | ||
The measured value of a field sensor, can be corrected by the software. The measured value can be corrected with an offset, or it can be corrected by a multiply factor. Both corrections can also be used together but the multiply correction will be performed first, and after that the offset correction will be added. The correction files with the correction should be attached to the device driver of the field sensor. The correction file for the multiply should contain the columns: | The measured value of a field sensor, can be corrected by the software. The measured value can be corrected with an offset, or it can be corrected by a multiply factor. Both corrections can also be used together but the multiply correction will be performed first, and after that the offset correction will be added. The correction files with the correction should be attached to the device driver of the field sensor. The correction file for the multiply should contain the columns: | ||
− | * Correction | + | * Correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is then applied on the isotropic value. |
− | To summarize this in an example, let’s | + | To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply isotropic correction for that same frequency of 1.1x is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1). |
===LISN: Amplitude correction=== | ===LISN: Amplitude correction=== | ||
Line 1,088: | Line 1,190: | ||
===Powermeter: Power correction=== | ===Powermeter: Power correction=== | ||
The measured power level of a power meter can be corrected with a certain value. A correction file with the correction should be attached to the device driver of the powermeter. The correction file should contain the columns: | The measured power level of a power meter can be corrected with a certain value. A correction file with the correction should be attached to the device driver of the powermeter. The correction file should contain the columns: | ||
− | + | or | |
− | *Attenuation | + | * Attenuation |
− | + | or | |
− | *Gain | + | * Gain |
The values in the correction column should be expressed in dB. The value specified in the attenuation column is subtracted from the measured power level. The values in the correction or gain column is added to the measured power level. For example: A power level of -14.58 dBm is measured with a power-meter. A correction file with a frequency and attenuation column is connected to the device driver. The value specified in the attenuation column for 81.7 MHz. is 0.27 dB. Then the software will use a value of -14.85 dBm as the value that was measured with the power-meter. | The values in the correction column should be expressed in dB. The value specified in the attenuation column is subtracted from the measured power level. The values in the correction or gain column is added to the measured power level. For example: A power level of -14.58 dBm is measured with a power-meter. A correction file with a frequency and attenuation column is connected to the device driver. The value specified in the attenuation column for 81.7 MHz. is 0.27 dB. Then the software will use a value of -14.85 dBm as the value that was measured with the power-meter. | ||
===Pre amplifier: Amplifier Gain=== | ===Pre amplifier: Amplifier Gain=== | ||
− | The gain of an used preamplifier can be corrected inside the software. A correction file with the frequency | + | The gain of an used preamplifier can be corrected inside the software. A correction file with the frequency dependent gain of the preamplifier should be attached to the device driver of the preamplifier. The values in the correction column should be expressed in dB. |
− | + | The actually used calculation is depending on the configuration of the correction file that is selected for the preamplifier. The correction file can contain the columns: | |
− | * | + | * Frequency |
− | * | + | * Correction: The values in the Correction column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be subtracted from the values that are measured with the receiver/analyser. |
− | * | + | or |
+ | or | ||
===Receiver: Amplitude correction=== | ===Receiver: Amplitude correction=== |
Engineers
Test engineer settings
Adding an engineer
Deleting an engineer
Editing engineer settings
Password protection
Rights
Change password
Test Equipment
Adding test equipment
Deleting test equipment
Editing test equipment
Devices 1 & 2
Adding devices
Editing devices
Removing devices
Saving test equipment
Save As
Before Action
Example
After Action
The default company name for the client. |
The default contact person for the client. |
The default address for the client. |
The default zip code for the client. |
The default city for the client. |
The default state for the client. |
The default country for the client. |
The default telephone number for the client. |
The default fax numberfor the client. |
The default e-mail address for the client. |
The default company name for the manufacturer. |
The default contact person for the manufacturer. |
The default address for the manufacturer. |
The default zip code for the manufacturer. |
The default city for the manufacturer. |
The default state for the manufacturer. |
The default country for the manufacturer. |
The default telephone number for the manufacturer. |
The default fax number for the manufacturer. |
The default e-mail address for the manufacturer. |
The default company name for the test site. |
The default contact person for the test site. |
The default address for the test site. |
The default zip code for the test site. |
The default city for the test site. |
The default state for the test site. |
The default country for the test site. |
The default telephone number for the test site. |
The default fax number for the test site. |
The default e-mail address for the test site. |
Stores the modified information and closes the window. |
Discards the modified information and closes the window. |
Advanced options
- Units
- Directories
- Device Drivers
- Graphs
- Database
- Language
- Measurement settings
- Product standards
- Basic standards
- Enhanced Status Window
1. Units configuration
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Changing unit settings
Go to the Directories tab.
2. Through the 'Devices' tab.
Select a device driver type
Adding a new device driver
Editing a device driver
Removing a device driver
Confirm settings
Fixed information
Shows the brand of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the description of the device. This information can be changed/viewed in this cell and can be used in the report generator. Changing the description does not result in losing the device in the test sites. The description is automatically updated in the test site information. If a report is generated from a test done prior to the description change the old description will be inserted. |
Shows the exact version of the DLL in which the device driver is contained. This is a unique identifier that specifies the version of the device driver. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Shows the date and time of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Shows the version number of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Shows the hardware version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the internally used ID of the device. This information can be changed/viewed in this cell and can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction. |
Shows the serial number of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the software version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the type of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Additional options
The ‘’’ ‘Add’ button’’’ can be used to add additional information items to the general device driver settings. The additional items will be shown in the list. Additional information items can be used to note the storage location, the location of the user manual, etc. |
The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited. |
The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed. |
When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can be a significantly reduce the time that is needed to initialize each test.
This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, RadiMation® will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date. |
The ‘’’ ‘Advanced’ button’’’ can be used to open the advanced menu of the device driver. In chapter 15, the paragraph ‘Device specific configuration’ will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view the paragraphs about ‘Configurable devices’ or ‘Virtual devices’ (also in chapter 15) for these devices. |
The ‘’’ ‘Check’ button’’’ can be used to check if the device is connected. This check is the same as the check that is performed during each test initialization. This check can be used to determine if the correct addressing is set, and whether communication with the device is possible. (This option is explained further below in the paragraph ‘Checking the device connection’.) |
The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the RadiWiki page about that device driver. That RadiWiki page can contain additional information or a detailed explanation of all the advanced settings. |
Now we can conclude that:
4. Graphs configuration
5. Database configuration
Introduction
Select
This just a name for the setup you are making, name this Radimation, or another descriptive name for the used database |
Allows to select the directory in which the text based database files are stored |
Overrules the selected directory to use the current directory. It is best to turn this checkbox off, and select a specific directory by using the Select Directory... button |
Shows a list of filename extensions which will be used to determine the correct text files. Remove *.asc and *.tab from the extentions list |
Will show the Define Text Format dialog, which can be used to define the format of the database as it is maintained in the text file |
MySQL ODBC database
An unique name that uniquely identifies the Data Source from the other available Data Sources. |
The name (or IP address) of the MySQL server, to which a connection should be made |
The port-number to which the which the connection should be made. By default this is port-number 3306 |
The name of the MySQL user that should be used to connect to the MySQL Server |
The password of the MySQL user that should be used to connect to the MySQL Server |
The name of the MySQL database that should be connected to |
Will perform a connection test to the specified MySQL server and Database |
Will show additional configuration details for the MySQL connection. Your MySQL database administrator can help you with setting this details. |
MSSQL ODBC database
Enable the option, and specify or select the name of the database on the MSSQL Server that should be used. |
Will perform a connection test, to determine if a connection with the MSSQL server can be established, using the configured parameters. A confirmation or error dialog with some additional information will be shown. |
Stores the configured ODBC connection. |
Closes the dialogs without creating an ODBC connection. |
6. Language Selection
7. Measurement Settings
Calibration
Required environment settings
8. Product Standard
Introduction
Adding a Product Standard
Add(ing) Basic Standard(s)
9. Basic Standard
Correction files
Creating correction files
Adding and removing columns
Entering data points
9 29.69
10 29.68
etc.
9 29.69 10 29.68
Saving correction files
Attaching correction files
Correction file uses
Absorbing clamp: Clamp loss
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Amplifier: Maximum Input
- Frequency
- Correction
Or
- Frequency
- Power
Amplifier: Maximum Forward
- Frequency
- Correction
Or
- Frequency
- Power
Amplifier: Maximum Reflected
- Frequency
- Correction
Or
- Frequency
- Power
Antenna: Ant Gain vs Angle
- Angle
- Attenuation
Antenna: Antenna Efficiency
- Frequency
- Correction
Antenna: Antenna Factor
- Frequency
- Correction
Antenna: Antenna Gain
- Frequency
- Antenna gain
Antenna: GpsHt - AntHeight
- Height
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction
Cable: Cable loss
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
This implies that:
Coupler: Forward Correction
- Frequency
Or
- Frequency
Or
- Frequency
Coupler: Reflected Correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Current sensor: impedance
- Frequency
Or
- Frequency
Field sensor: Field offset
- Frequency
or
- Frequency
Field sensor: Field Multiply
- Frequency
or
- Frequency
LISN: Amplitude correction
- Frequency
- Correction
Powermeter: Power correction
- Frequency
- Correction
or
- Frequency
- Attenuation
or
- Frequency
- Gain
Pre amplifier: Amplifier Gain
- Frequency
or
- Frequency
or
- Frequency
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction