(91 intermediate revisions by 5 users not shown)
__NOTOC__
__NOTOC__
= Configuration =
= Configuration =
The global configuration of {{RadiMation}} is available in the menu
{{Menu|Configuration}}
In the main menu, under the configuration pull down menu, a number of standard configurations can be made with regards to the engineers, test equipment, etc. The selected and/or entered configurations will be used as the default settings for future use.
The configured global {{RadiMation}} configuration will be used as the configuration for future use. This configuration is independent of the actual Windows User Account that is logged on to the Microsoft Windows PC. The configured global {{RadiMation}} configuration is shared and accessible for all Windows User Accounts that have access to the PC.
{{Warning|To prevent unauthorised individuals from changing the system configurations, a password protection can be used. The password is activated by default with the password: “{{RadiMation}}”. This password protection can be switched off, changed and configured under:
{{Warning|To prevent unauthorised individuals from changing the global {{RadiMation}} configurations, a password protection can be used for specific engineers. The password is activated by default for the "'''Administrator'''" engineer with the password "'''Radimation'''". This password protection can be switched off, changed and configured in the menu
{{Menu|Configuration|Engineers|(Edit)}}
{{Menu|Configuration|Engineers|Edit}}
== Engineers ==
On a new installation, only a single 'Administrator' engineer is created by default. An unlimited number of engineers can be added, each with an unique name. For each engineer a login password and different permissions can be assigned which give them the right to modify specific parts of the {{RadiMation}} configuration. There is no relation between the engineer in {{RadiMation}} and the Windows Logon account. Also the password that can be configured for the engineer in {{RadiMation}} is not linked and also not synchronized with the Windows User account password.
== Engineers ==
During the startup of {{RadiMation}} the actual engineer that is going to start tests can be selected in the {{WindowTitle|Login}} dialog. If a password is configured for that engineer, it should also be specified at that moment.
For each test, the name and password settings of the test engineer performing the current test can be entered.
[[Image:Login.png]]
To add, delete or edit (the names and settings of) the test engineers, go to the ''' 'Engineers' window''' in the ''' 'Configuration' tab'''.
Automatic login of the engineer can be disabled or enabled. If it is enabled, it can only be selected for one engineer, and that engineer will be automatically logged in the next time that {{RadiMation}} starts. If the {{ScreenElement|Automatically login with this engineer}} checkbox is selected for an engineer, the automatic login of all the other engineers will not be performed anymore. For changing setting of the the automatically logged in engineer, it is still necessary to close the login dialog by specifying the corresponding {{ScreenElement|Password}} and using the {{ScreenElement|Ok}} button.
Switching between the active engineers while {{RadiMation}} is running can be done through the menu:
{{Menu|File|Change Engineer}}
[[Image:File Change Engineer.png]]
The Engineers window will open and show the current configured list of engineers. To leave the 'Engineers' window, click ''' 'OK' '''.
To add, delete or edit the names and permissions of the engineers, go to the {{ScreenElement|Engineers}} window, which can be accessed in the menu:
{{Menu|Configuration|Engineers}}
=== Test engineer settings ===
To leave the 'Engineers' window, click {{ScreenElement|OK}}.
==== Adding an engineer ====
To add a new test engineer to the configuration, click {{ScreenElement|Add}}.
To add a new test engineer to the configuration, click ''' 'Add' '''(in the 'Engineers' window).
[[Image:Engineer add engineer.png]]
Enter the name and set the rights of the new engineer, then click {{ScreenElement|OK}}. You will be returned to the 'Engineers' window.
Enter the name and set the rights of the new engineer, then click ''' 'OK' '''. You will be returned to the 'Engineers' window.
The new engineer will not have a protection password (yet), this can be enabled by clicking {{ScreenElement|[[Change Password]]}}. Please read the paragraph 'Password protection' (located further down) for more information.
The new engineer will not have a protection password (yet), this can be enabled by clicking ''' '[[Change Password|Change Password]]' '''. Please read the paragraph 'Password protection' (located further down) for more information.
[[Image:Test engineer add engineer.png]]
==== Deleting an engineer ====
To delete a test engineer from the configuration, select the engineer and click {{ScreenElement|Delete}}.
To delete a test engineer from the configuration, select the engineer and click ''' 'Delete' '''(in the 'Engineers' window).
[[Image:Engineer delete engineer.png]]
To confirm that you want to delete the engineer from the configuration list, click {{ScreenElement|Yes}}.
To confirm that you want to delete the engineer from the configuration list, click ''' 'Yes' '''.
If you do not want to delete this engineer from the list, click {{ScreenElement|No}} or {{ScreenElement|Cancel}}. You will be returned to the 'Engineers' window.
If you do not want to delete this engineer from the list, click ''' 'No' ''' or ''' 'Cancel' '''. You will be returned to the 'Engineers' window.
[[Image:Engineer delete engineer confirm.png]]
==== Editing engineer settings ====
To edit the name, rights and/or password of an engineer in the configuration, select the engineer and click {{ScreenElement|Edit}}.
To edit the name, rights and/or password of an engineer in the configuration, select the engineer and click ''' 'Edit' '''(in the 'Engineers' window).
[[Image:Engineer edit engineer.png]]
Make the desired changes in the 'Test Engineer' window and click {{ScreenElement|OK}}.
Make the desired changes in the 'Test Engineer' window and click ''' 'OK' '''.
If you want to undo or cancel your changes, click {{ScreenElement|Cancel}}. You will be returned to the 'Engineers' window.
If you want to undo or cancel your changes, click ''' 'Cancel' '''. You will be returned to the 'Engineers' window.
=== Password protection ===
{{RadiMation}} allows you to protect a number of configurations with a password. This password can be enabled, disabled or changed in the ''' 'Test Engineer' window'''.
* The password is enabled by default for the default test engineer 'Administrator'.
Line 79: Line 79:
[[Image:Test engineer.png]]
==== Rights ====
==== Rights ====
Line 95: Line 94:
==== Change password ====
To enable, disable or change the password, click ''' '[[Change Password|Change Password]]' '''(in the 'Test Engineer' window).
To change the password, enter the old password and the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the 'Test Engineer' window.
To '''change the password''', enter the old password and the new password (twice), then click ''' 'OK' '''. You will be returned to the 'Test Engineer' window.
[[Image:Change password.png]]
To disable the password, enter the old password and leave the two new password fields empty, then click {{ScreenElement|OK}}. You will be returned to the 'Test Engineer' window, and the password indication will change to: “Warning: Password is not set for this engineer.”
To '''disable the password''', enter the old password and leave the other two fields empty, then click ''' 'OK' '''. You will be returned to the 'Test Engineer' window.
To enable the password, enter the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the 'Test Engineer' window, and the password indication will change to: “Password is set for this engineer.”
To '''enable the password''', enter the new password (twice), then click ''' 'OK' '''. You will be returned to the 'Test Engineer' window.
[[Image:Enable password.png]]
== Test Equipment ==
Line 228: Line 218:
[[Image:Test equipment close.png]]
== Before Actions & After Actions ==
The before and after actions of a testsite are designed to be triggered when the testsite is activated (before actions) or deactivated (after actions).
These before and after actions can be seen as like switching relays to desired positions when a testsite is activated, and switching the relays to a default position again when the testsite is deactivated.
When a testsite is used in two consecutive bands, the after and before actions are not executed, because there is no change in the activated testsite.
When a test with 3 bands is configured, and the first and third band are using the same testsite, while another testsite is selected for the second band, then the after and before actions will be executed when the test is switching from band 1 to band 2, and again when the test is switching from band 2 to band 3.
=== Before Action ===
Line 282: Line 279:
== Default address information ==
When a new [[EUT]] file is created, the name and address information of the client, manufacturer and test site can be entered.
In many situations one or more of those address information will be the same for different [[EUT]] files (for example, when a test site use {{RadiMation}}, the test site address information will always be the same).
To prevent users from having to enter repetitive data input, {{RadiMation}} provides the possibility to set default address information for:
* the client that has ordered that tests should be performed on the EUT
* the manufacturer that has produced the EUT
* the test site that is performing the tests on the EUT
Address information entered on the {{WindowTitle|Default Address Information}} window, will be automatically included on the {{WindowTitle|EUT}} window and therefore only needs to be entered once.
* The default client
* The default manufacturer
{{Menu|Configuration|Default Address Information}}
* The default test site
Data entered under ''' 'Default Address Information' ''', will automatically be entered in the [[EUT]] window and therefore only needs to be entered once.
In the {{WindowTitle|Default Address Information}} window, the default address information can be configured. Specifying this information is not required, however it has the benefit that address information that is always the same doesn't need to be entered over and over again in every new [[EUT]] file.
To add, modify or delete the default address information, go to the ''' 'Default Address Information' window''' in the ''' 'Configuration' tab'''.
{{ScreenElementDescription|Client Company|The default company name for the client.}}
{{ScreenElementDescription|Client Contact Person|The default contact person for the client.}}
{{ScreenElementDescription|Client Address|The default address for the client.}}
{{ScreenElementDescription|Client Zip Code|The default zip code for the client.}}
{{ScreenElementDescription|Client City|The default city for the client.}}
{{ScreenElementDescription|Client State|The default state for the client.}}
To confirm the data input, click ''' 'OK' ''', to cancel the data input, click ''' 'Cancel' '''. The 'Default Address Information' window will close.
{{ScreenElementDescription|Client Country|The default country for the client.}}
{{ScreenElementDescription|Client Telephone Number|The default telephone number for the client.}}
{{ScreenElementDescription|Client Fax Number|The default fax numberfor the client.}}
{{ScreenElementDescription|Client Email Address|The default e-mail address for the client.}}
{{ScreenElementDescription|Manufacturer Company|The default company name for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Contact Person|The default contact person for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Address|The default address for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Zip Code|The default zip code for the manufacturer.}}
{{ScreenElementDescription|Manufacturer City|The default city for the manufacturer.}}
{{ScreenElementDescription|Manufacturer State|The default state for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Country|The default country for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Telephone Number|The default telephone number for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Fax Number|The default fax number for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Email Address|The default e-mail address for the manufacturer.}}
{{ScreenElementDescription|Test Site Company|The default company name for the test site.}}
{{ScreenElementDescription|Test Site Contact Person|The default contact person for the test site.}}
{{ScreenElementDescription|Test Site Address|The default address for the test site.}}
{{ScreenElementDescription|Test Site Zip Code|The default zip code for the test site.}}
{{ScreenElementDescription|Test Site City|The default city for the test site.}}
{{ScreenElementDescription|Test Site State|The default state for the test site.}}
{{ScreenElementDescription|Test Site Country|The default country for the test site.}}
{{ScreenElementDescription|Test Site Telephone Number|The default telephone number for the test site.}}
{{ScreenElementDescription|Test Site Fax Number|The default fax number for the test site.}}
{{ScreenElementDescription|Test Site Email Address|The default e-mail address for the test site.}}
{{ScreenElementDescription|Ok|Stores the modified information and closes the window.}}
{{ScreenElementDescription|Cancel|Discards the modified information and closes the window.}}
The default address information is stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured in the global {{RadiMation}} configuration.
== Advanced options ==
Line 311: Line 338:
There are a several [[advanced options]] in {{RadiMation}}, that allow to tune or optimize some specific settings. Making changes to one or more of those settings, can have a major influence in the correct functioning of the software. Changes to one or more of these advanced options are only needed to circumvent a specific situation. Before changes can be made, a specific password should be specified. The required password is not related to any of the passwords of existing engineers. If it is necessary to make a change to one or more of the advanced options, the password will be provided to the end-user by the [[reseller]] or  {{CompanyName}} If the password is specified correctly, a configuration window is shown that includes a table to review and edit the list of advanced options.
There are a several [[advanced options]] in {{RadiMation}}, that allow to tune or optimize some specific settings. Making changes to one or more of those settings, can have a major influence on the correct functioning of the software. Changes to one or more of these advanced options are only needed to circumvent a specific situation. Before changes can be made, a specific password should be specified. The required password is not related to any of the passwords of existing engineers. If it is necessary to make a change to one or more of the advanced options, the password will be provided to the end-user by the [[reseller]] or  {{CompanyName}}. If the password is specified correctly, a configuration window is shown that includes a table to review and edit the list of advanced options.
== Configuration ==
== Configuration ==
Line 372: Line 399:
=== 2. Directories configuration ===
The {{RadiMation}} software can use different directories to store different data file types.
The {{RadiMation}} software can use different directories to store different files.
* Device Driver Configuration Files (CONFDVDR),
* Equipment Under Test Data files (*.EUT files),
* Correction files (*.COR files)
Line 386: Line 415:
The different file types are described in [[Chapter 18|Chapter 18 “{{RadiMation}} file types and locations”]].
For each data type the default location can be entered through the configuration. The path name can link to a directory on a local or on a shared network drive. Selecting a network location will allow you to store data (e.g. EUT data) on a computer network, making it more accessible. This will allow you to configure {{RadiMation}}, to make use of a networked configuration with multiple test PC's, sharing the same configuration.
==== Open the 'Directories' configuration screen ====
{{Menu|Configuration|Configuration}}
==== Modifying a path name ====
Go to the {{ScreenElement|Directories}} tab.
To modify a path name, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''.
[[Image:Configuration Directories.png]]
==== Modifying the 'Device Driver Configuration Files' directory ====
With the {{ScreenElement|Use standalone configuration}} option, it is possible to create a version specific isolated configuration directory with its own content only used for this version of RadiMation. This can for example be used to verify and validate another {{RadiMation}} version without influencing another {{RadiMation}} version.
Go to the ''' 'Directories' tab''', select the directory that you wish to change and click ''' 'Change' '''.
The location of the {{ScreenElement|Device Driver Configuration Files}} can by changed by clicking on the {{ScreenElement|'Open Folder' icon}}. A ''' 'Browse For Folder' ''' window will appear, allowing the selection of an existing device driver configuration files directory. If an empty or an invalid device driver configuration files directory is selected, an error will be shown, and the original selected directory will remain selected.
When it's needed to create a new {{ScreenElement|Device Driver Configuration Files}} folder, click {{ScreenElement|Edit}} and select {{ScreenElement|Create and set new folder}}. A directory chooser will be prompted to create and / or select a new empty folder. The selected folder should be completely empty otherwise it cannot be initialized to become a {{ScreenElement|Device Driver Configuration Files}} directory.
==== Duplicating a 'Device Driver Configuration Files' folder ====
It is possible to duplicate an existing configuration folder, by clicking on {{ScreenElement|Edit}} and either choose {{ScreenElement|Duplicate and set from current global folder}} or {{ScreenElement|Duplicate and set from current standalone folder}}. A directory chooser will be prompted to create and / or select a new empty folder to copy the content of the source configuration to. After the duplication of the folder, the new {{ScreenElement|Device Driver Configuration Files}} directory will be selected.
The ''' 'Browse For Folder' window''' will appear, allowing you to select (or make) a new folder in your directory.
==== Change the path for a file type ====
Select the directory that you wish to change and click on the {{ScreenElement|Open Folder}} icon to select a different directory.
To confirm your change, click ''' 'OK' ''', to cancel your change, click ''' 'Cancel' '''. You will be returned to the 'Configuration' window.
[[Image:Browse Calibration File Directory.png]]
The {{ScreenElement|Browse For Folder}} window will appear, allowing you to select (or make) a new folder in your directory.
To confirm your change, click {{ScreenElement|Select folder}}, or to cancel your change, click {{ScreenElement|Cancel}}. You will be returned to the {{ScreenElement|Configuration}} window.
Click {{ScreenElement|Close}} to leave the {{ScreenElement|Configuration}} window.
=== 3. Device Driver configuration ===
Line 433: Line 473:
The first step to create or configure a device driver is to select the required ''' '[[Device Types|Device Driver Type]]' ''' (e.g. Signal Generator). This can be done from the drop-down menu in the ‘Configuration’ window. One’s a type has been selected, the available drivers for that type will be shown.
The first step to create or configure a device driver is to select the required ''' '[[Device Types|Device Driver Type]]' ''' (e.g. Signal Generator). This can be done from the drop-down menu in the ‘Configuration’ window. Once a type has been selected, the available drivers for that type will be shown.
Line 517: Line 556:
{{ScreenElementDescription|Zero Interval|When the {{ScreenElement|Zero interval}} is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that [[Power Meter]]s and [[Field Sensor]]s are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can be a significantly reduce the time that is needed to initialize each test.
{{ScreenElementDescription|Zero Interval|When the {{ScreenElement|Zero interval}} is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that [[Power Meter]]s and [[Field Sensor]]s are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can significantly reduce the time that is needed to initialize each test.
Line 524: Line 563:
Line 549: Line 587:
This is a very helpful tool because it is a fast method to check whether the advanced settings (especially the ([[GPIB]]) address) of the device are set correctly. If this 'Check' is successful, the 'device check' during the test initialization (which performs the same check) will also be successful.
==== Storing the driver version in the test file ====
Line 624: Line 659:
Text, CSV, Excel, [[MSSQL]] and [[MySQL]] dabase files need some additional setup before they can be used. The reason for this is that {{RadiMation}} uses [[ODBC]] to connect and read out thoes databases. This additional setup of an [[ODBC]] connection must be performed before the database driver can be configured in {{RadimationLink}}. {{:ODBC Configuration}}
Text, CSV, Excel, [[MSSQL]] and [[MySQL]] dabase files need some additional setup before they can be used. The reason for this is that {{RadiMation}} uses [[ODBC]] to connect and read out those databases. This additional setup of an [[ODBC]] connection must be performed before the database driver can be configured in {{RadimationLink}}. {{:ODBC Configuration}}
Line 643: Line 678:
=== 6. Language Selection ===
{{RadiMation}} supports multi-language use. This means that the software package can be operated in several languages.
Line 673: Line 707:
With measurement settings you can change the behavior of the software at certain points. Be very careful what you do with these setting, because they are used through out {{RadiMation}}.
With measurement settings you can change the behavior of the software at certain points. Be very careful what you do with these setting, because they are used throughout {{RadiMation}}.
Line 708: Line 742:
=== 8. Product Standard ===
Line 829: Line 862:
[[Image:ColumnsUnits.png]]
When you have added a column, you will set it on the left-hand side in the list Columns. You can define the unit being used by the column.
Line 854: Line 887:
Line 952: Line 985:
*Power
*Power
*Angle
*Attenuation
The Antenna Efficiency is expressed as a unitless number and the value should be between -1 and 1.
*Frequency
*Correction
===Antenna: Antenna Factor===
For converting the reading dBuV into dBuV/m the antenna factor is needed.
Line 960: Line 1,009:
*Correction
*Correction
Where the correction column is the antenna factor.
===Antenna: Antenna Gain===
The Antenna Gain correction is used to correct the calculated logarithmic gain for horn antennas based on their horn dimensions.
This value is used by the calculations applicable to the calculated electrical field test levels.
*Frequency
*Antenna gain
===Antenna: GpsHt - AntHeight===
*Height
===Antenna: Magnetic Flux Density (pT/A) correction===
*Frequency
*Frequency
*Correction
*Correction
This correction is primarily used during the [[MIL-STD-461]] RS101 magnetic field tests. In the MIL-STD 461, this correction value is specified as 9.5x10^7 pT/A.
===Antenna: Magnetic Field Factor (dBpT/uV) correction===
Line 974: Line 1,038:
*Frequency
*Frequency
*Correction
*Correction
This same correction is also used during the [[MIL-STD-461]] RS101 Magnetic Field calibration for the 'calibration antenna'.
The Reflection coefficient is used to correct the calculated required net power in Watts for horn antennas to a calculated required forward power level in Watts.
The Reflection coefficient is expressed as a unitless number.
*Frequency
*Height
*Correction
*Angle
*Attenuation
===Cable: Cable loss===
Line 1,008: Line 1,071:
Line 1,046: Line 1,109:
[[Image:Calibration_jig_factor_50_Ohm_for_50_Ohm_system.png]]
== Using a transfer correction file provided by the supplier ==
===Coupler: Forward Correction===
*Frequency
* Frequency
*Correction
* Correction. The values in the 'Correction' column should be expressed in dB, and these values are added to the typical attenuation of the coupler.
Or
Or
*Frequency
* Frequency
*Attenuation
* Attenuation. The values in the 'Attenuation' column are added to the typical attenuation of the coupler.
Or
Or
*Frequency
* Frequency
*Gain
* Gain. The values specified in the 'Gain' column are subtracted from the typical attenuation of the coupler.
===Coupler: Reflected Correction===
Line 1,069: Line 1,140:
*Frequency
*Frequency
*Gain
*Gain
* Frequency
* Frequency
* Correction
* Correction. The values in the 'Correction' column should be expressed in Ohm ({{Ohm}}).
Or
Or
* Frequency
* Frequency
* Resistance
* Resistance. The values in the 'Resistance' column will use the unit that is configured for the 'Resistance' column.
The values in the correction column should be expressed in Ohm ({{Ohm}}).
===Field sensor: Field offset===
*Frequency
* Frequency
*Correction
* Correction. The values in the column used for the offset correction should be expressed in V/m. This correction is then applied on the isotropic value.
The values in the correction column of the offset correction should be expressed in V/m.
or
* Frequency
* X-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the x-axis reading only.
* Y-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the y-axis reading only.
To summarize this in an example, let’s say that the RAW field probe measurement at 90 MHz is 10.2 V/m. When a Field offset correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m).
===Field sensor: Field Multiply===
* Frequency
* Frequency
* Correction
* Correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is then applied on the isotropic value.
The values in the correction column of the multiply factor should be expressed as x (multiplier).
or
* Frequency
* X-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the x-axis reading only.
* Y-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the y-axis reading only.
To summarize this in an example, let’s say that the RAW field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply correction for that same frequency of 1.1 is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1).
===LISN: Amplitude correction===
Line 1,104: Line 1,186:
===Powermeter: Power correction===
*Frequency
* Frequency
*Correction
* Correction
Or
or
*Frequency
* Frequency
*Attenuation
* Attenuation
Or
or
*Frequency
* Frequency
*Gain
* Gain
* 'Frequency' and 'Correction' column: The values in the 'Correction' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
* Frequency
* 'Frequency' and 'Gain' column: The values in the 'Gain' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
* 'Frequency' and 'Attenuation' column: The values in the 'Attenuation' column are interpreted as 'attenuation' values. A positive value is interpreted as a loss caused by the preamplifier. A negative value is interpreted as a positive amplification of the signal, caused by the preamplifier. These 'attenuation' values will be added to the values that are measured with the receiver/analyser.
or
* Frequency
* Gain: The values in the Gain column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be subtracted from the values that are measured with the receiver/analyser.
or
* Frequency
* Attenuation: The values in the Attenuation column are interpreted as 'attenuation' values. A positive value is interpreted as a loss caused by the preamplifier. A negative value is interpreted as a positive amplification of the signal, caused by the preamplifier. These 'attenuation' values will be added to the values that are measured with the receiver/analyser.
===Receiver: Amplitude correction===
Line 1,136: Line 1,224:
*Frequency
* Frequency
*Correction
* Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3.
or
* Frequency
* Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2.
or
* Frequency
* Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4.
[[Category:Manual]]
[[Category:Manual]]
Revision as of 08:45, 2 April 2025
   Menu.svg Configuration
      Menu.svg Engineers
         Menu.svg Edit
Engineers[edit]
   Menu.svg File
      Menu.svg Change Engineer
      Menu.svg Engineers
Test engineer settings[edit]
Adding an engineer[edit]
Deleting an engineer[edit]
Password protection[edit]
  • The password is enabled by default for the default test engineer 'Administrator'.
Rights[edit]
Change password[edit]
Test Equipment[edit]
Adding test equipment[edit]
Deleting test equipment[edit]
Editing test equipment[edit]
Devices 1 & 2[edit]
Adding devices[edit]
Editing devices[edit]
Removing devices[edit]
Saving test equipment[edit]
Save As[edit]
Before Action[edit]
Example[edit]
After Action[edit]
Advanced options[edit]
  1. Units
  2. Directories
  3. Device Drivers
  4. Graphs
  5. Database
  6. Language
  7. Measurement settings
  8. Product standards
  9. Basic standards
  10. Enhanced Status Window
1. Units configuration[edit]
  • Radiated Immunity
  • Conducted Immunity
  • Radiated Emission
  • Conducted Emission
  • Pulsed Immunity
  • Antenna Diagram
  • RadiCal
Changing unit settings[edit]
Go to the Directories tab.
2. Through the 'Devices' tab.
Editing a device driver[edit]
Device Driver Settings[edit]
Confirm settings[edit]
Fixed information[edit]
Additional options[edit]
ScreenElementDescription.svg Zero Interval When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can significantly reduce the time that is needed to initialize each test.
Now we can conclude that:
4. Graphs configuration[edit]
Select
   Menu.svg Start
      Menu.svg Settings
         Menu.svg Control Panel
            Menu.svg ODBC Data Sources
Configuration of the characterset[edit]
Characterset value
UTF-8 (without bom) CharacterSet=65001
UTF-16 LE (without bom) CharacterSet=Unicode
MySQL ODBC database[edit]
MSSQL ODBC database[edit]
6. Language Selection[edit]
7. Measurement Settings[edit]
8. Product Standard[edit]
9. Basic Standard[edit]
Correction files[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction file
Entering data points[edit]
9 29.69 10 29.68 etc.
9 29.69 10 29.68
Saving correction files[edit]
   Menu.svg File
      Menu.svg Save correction file
      Menu.svg configuration
Correction file uses[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
  • Frequency
  • Correction
Or
  • Frequency
  • Power
  • Frequency
  • Correction
Or
  • Frequency
  • Power
  • Frequency
  • Correction
Or
  • Frequency
  • Power
  • Angle
  • Attenuation
  • Frequency
  • Correction
Antenna: Antenna Factor[edit]
  • Frequency
  • Correction
Antenna: Antenna Gain[edit]
  • Frequency
  • Antenna gain
  • Height
  • Frequency
  • Correction
  • Frequency
  • Correction
  • Frequency
  • Correction
Cable: Cable loss[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
  • Frequency
  • Correction
This implies that:
  • Frequency
Or
  • Frequency
Or
  • Frequency
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
  • Frequency
Or
  • Frequency
  • Frequency
or
  • Frequency
  • Frequency
or
  • Frequency
  • Frequency
  • Correction
  • Frequency
  • Correction
or
  • Frequency
  • Attenuation
or
  • Frequency
  • Gain
  • Frequency
or
  • Frequency
or
  • Frequency
  • Frequency
  • Correction
  • Frequency
  • Correction
  • Frequency
  • Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3.
or
  • Frequency
  • Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2.
or
  • Frequency
  • Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4.
