(52 intermediate revisions by 4 users not shown)
Line 15: Line 15:
== Engineers ==
== Engineers ==
An unlimited number of Engineers can be added, each with an unique name. For each engineer different permissions can be assigned which give them the right to modify specific parts of the {{RadiMation}} configuration.
On a new installation, only a single 'Administrator' engineer is created by default. An unlimited number of engineers can be added, each with an unique name. For each engineer a login password and different permissions can be assigned which give them the right to modify specific parts of the {{RadiMation}} configuration. There is no relation between the engineer in {{RadiMation}} and the Windows Logon account. Also the password that can be configured for the engineer in {{RadiMation}} is not linked and also not synchronized with the Windows User account password.
There is no relation between the Engineer in {{RadiMation}} and the Windows Logon account. Also the password that can be configured for the engineer in {{RadiMation}} is not linked and also not synchronized with the Windows User account password.
During the startup of {{RadiMation}} the actual engineer that is going to start tests can be selected in the {{WindowTitle|Login}} dialog. If a password is configured for that engineer, it should also be specified at that moment.
For each test, the name and password settings of the test engineer performing the current test can be selected. Also during the startup of {{RadiMation}} the actual Engineer that is going to start tests can be selected. Switching between the active engineers while {{RadiMation}} is running can be done through the menu:
[[Image:Login.png]]
Automatic login of the engineer can be disabled or enabled. If it is enabled, it can only be selected for one engineer, and that engineer will be automatically logged in the next time that {{RadiMation}} starts. If the {{ScreenElement|Automatically login with this engineer}} checkbox is selected for an engineer, the automatic login of all the other engineers will not be performed anymore. For changing setting of the the automatically logged in engineer, it is still necessary to close the login dialog by specifying the corresponding {{ScreenElement|Password}} and using the {{ScreenElement|Ok}} button.
{{Menu|File|Change Engineer}}
To add, delete or edit the names and permissions of the test engineers, go to the {{ScreenElement|Engineers window}}, which can be accessed in the menu:
[[Image:File Change Engineer.png]]
To add, delete or edit the names and permissions of the engineers, go to the {{ScreenElement|Engineers}} window, which can be accessed in the menu:
{{Menu|Configuration|Engineers}}
Line 67: Line 73:
* The password is enabled by default for the default test engineer 'Administrator'.
Line 273: Line 279:
== Default address information ==
When a new [[EUT]] file is created, the name and address information of the client, manufacturer and test site can be entered.
In many situations one or more of those address information will be the same for different [[EUT]] files (for example, when a test site use {{RadiMation}}, the test site address information will always be the same).
To prevent users from having to enter repetitive data input, {{RadiMation}} provides the possibility to set default address information for:
* the client that has ordered that tests should be performed on the EUT
* the manufacturer that has produced the EUT
* the test site that is performing the tests on the EUT
Address information entered on the {{WindowTitle|Default Address Information}} window, will be automatically included on the {{WindowTitle|EUT}} window and therefore only needs to be entered once.
* The default client
* The default manufacturer
{{Menu|Configuration|Default Address Information}}
* The default test site
Data entered under ''' 'Default Address Information' ''', will automatically be entered in the [[EUT]] window and therefore only needs to be entered once.
In the {{WindowTitle|Default Address Information}} window, the default address information can be configured. Specifying this information is not required, however it has the benefit that address information that is always the same doesn't need to be entered over and over again in every new [[EUT]] file.
To add, modify or delete the default address information, go to the ''' 'Default Address Information' window''' in the ''' 'Configuration' tab'''.
{{ScreenElementDescription|Client Company|The default company name for the client.}}
{{ScreenElementDescription|Client Contact Person|The default contact person for the client.}}
{{ScreenElementDescription|Client Address|The default address for the client.}}
{{ScreenElementDescription|Client Zip Code|The default zip code for the client.}}
{{ScreenElementDescription|Client City|The default city for the client.}}
{{ScreenElementDescription|Client State|The default state for the client.}}
{{ScreenElementDescription|Client Country|The default country for the client.}}
{{ScreenElementDescription|Client Telephone Number|The default telephone number for the client.}}
{{ScreenElementDescription|Client Fax Number|The default fax numberfor the client.}}
{{ScreenElementDescription|Client Email Address|The default e-mail address for the client.}}
{{ScreenElementDescription|Manufacturer Company|The default company name for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Contact Person|The default contact person for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Address|The default address for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Zip Code|The default zip code for the manufacturer.}}
{{ScreenElementDescription|Manufacturer City|The default city for the manufacturer.}}
{{ScreenElementDescription|Manufacturer State|The default state for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Country|The default country for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Telephone Number|The default telephone number for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Fax Number|The default fax number for the manufacturer.}}
{{ScreenElementDescription|Manufacturer Email Address|The default e-mail address for the manufacturer.}}
{{ScreenElementDescription|Test Site Company|The default company name for the test site.}}
{{ScreenElementDescription|Test Site Contact Person|The default contact person for the test site.}}
{{ScreenElementDescription|Test Site Address|The default address for the test site.}}
{{ScreenElementDescription|Test Site Zip Code|The default zip code for the test site.}}
{{ScreenElementDescription|Test Site City|The default city for the test site.}}
{{ScreenElementDescription|Test Site State|The default state for the test site.}}
{{ScreenElementDescription|Test Site Country|The default country for the test site.}}
{{ScreenElementDescription|Test Site Telephone Number|The default telephone number for the test site.}}
{{ScreenElementDescription|Test Site Fax Number|The default fax number for the test site.}}
{{ScreenElementDescription|Test Site Email Address|The default e-mail address for the test site.}}
{{ScreenElementDescription|Ok|Stores the modified information and closes the window.}}
{{ScreenElementDescription|Cancel|Discards the modified information and closes the window.}}
The default address information is stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured in the global {{RadiMation}} configuration.
To confirm the data input, click ''' 'OK' ''', to cancel the data input, click ''' 'Cancel' '''. The 'Default Address Information' window will close.
== Advanced options ==
Line 302: Line 338:
There are a several [[advanced options]] in {{RadiMation}}, that allow to tune or optimize some specific settings. Making changes to one or more of those settings, can have a major influence in the correct functioning of the software. Changes to one or more of these advanced options are only needed to circumvent a specific situation. Before changes can be made, a specific password should be specified. The required password is not related to any of the passwords of existing engineers. If it is necessary to make a change to one or more of the advanced options, the password will be provided to the end-user by the [[reseller]] or  {{CompanyName}} If the password is specified correctly, a configuration window is shown that includes a table to review and edit the list of advanced options.
There are a several [[advanced options]] in {{RadiMation}}, that allow to tune or optimize some specific settings. Making changes to one or more of those settings, can have a major influence on the correct functioning of the software. Changes to one or more of these advanced options are only needed to circumvent a specific situation. Before changes can be made, a specific password should be specified. The required password is not related to any of the passwords of existing engineers. If it is necessary to make a change to one or more of the advanced options, the password will be provided to the end-user by the [[reseller]] or  {{CompanyName}}. If the password is specified correctly, a configuration window is shown that includes a table to review and edit the list of advanced options.
== Configuration ==
== Configuration ==
Line 394: Line 430:
With the {{ScreenElement|Use standalone configuration}} option, it is possible to create an version specific isolated configuration directory with it's own content only used for this version of RadiMation. This can for example be used to verify and validate another {{RadiMation}} version without influencing another {{RadiMation}} version.
With the {{ScreenElement|Use standalone configuration}} option, it is possible to create a version specific isolated configuration directory with its own content only used for this version of RadiMation. This can for example be used to verify and validate another {{RadiMation}} version without influencing another {{RadiMation}} version.
The location of the {{ScreenElement|Device Driver Configuration Files}} can by changed by clicking on the {{ScreenElement|'Open Folder' icon}}. A ''' 'Browse For Folder' ''' window will appear, allowing to select an existing device driver configuration files directory. If an empty or an invalid device driver configuration files directory is selected, an error will be shown, and the original selected directory will remain selected.
The location of the {{ScreenElement|Device Driver Configuration Files}} can by changed by clicking on the {{ScreenElement|'Open Folder' icon}}. A ''' 'Browse For Folder' ''' window will appear, allowing the selection of an existing device driver configuration files directory. If an empty or an invalid device driver configuration files directory is selected, an error will be shown, and the original selected directory will remain selected.
Line 437: Line 473:
The first step to create or configure a device driver is to select the required ''' '[[Device Types|Device Driver Type]]' ''' (e.g. Signal Generator). This can be done from the drop-down menu in the ‘Configuration’ window. One’s a type has been selected, the available drivers for that type will be shown.
The first step to create or configure a device driver is to select the required ''' '[[Device Types|Device Driver Type]]' ''' (e.g. Signal Generator). This can be done from the drop-down menu in the ‘Configuration’ window. Once a type has been selected, the available drivers for that type will be shown.
Line 521: Line 556:
{{ScreenElementDescription|Zero Interval|When the {{ScreenElement|Zero interval}} is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that [[Power Meter]]s and [[Field Sensor]]s are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can be a significantly reduce the time that is needed to initialize each test.
{{ScreenElementDescription|Zero Interval|When the {{ScreenElement|Zero interval}} is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that [[Power Meter]]s and [[Field Sensor]]s are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can significantly reduce the time that is needed to initialize each test.
Line 528: Line 563:
Line 553: Line 587:
This is a very helpful tool because it is a fast method to check whether the advanced settings (especially the ([[GPIB]]) address) of the device are set correctly. If this 'Check' is successful, the 'device check' during the test initialization (which performs the same check) will also be successful.
==== Storing the driver version in the test file ====
Line 628: Line 659:
Text, CSV, Excel, [[MSSQL]] and [[MySQL]] dabase files need some additional setup before they can be used. The reason for this is that {{RadiMation}} uses [[ODBC]] to connect and read out thoes databases. This additional setup of an [[ODBC]] connection must be performed before the database driver can be configured in {{RadimationLink}}. {{:ODBC Configuration}}
Text, CSV, Excel, [[MSSQL]] and [[MySQL]] dabase files need some additional setup before they can be used. The reason for this is that {{RadiMation}} uses [[ODBC]] to connect and read out those databases. This additional setup of an [[ODBC]] connection must be performed before the database driver can be configured in {{RadimationLink}}. {{:ODBC Configuration}}
Line 647: Line 678:
=== 6. Language Selection ===
{{RadiMation}} supports multi-language use. This means that the software package can be operated in several languages.
Line 677: Line 707:
With measurement settings you can change the behavior of the software at certain points. Be very careful what you do with these setting, because they are used through out {{RadiMation}}.
With measurement settings you can change the behavior of the software at certain points. Be very careful what you do with these setting, because they are used throughout {{RadiMation}}.
Line 712: Line 742:
=== 8. Product Standard ===
Line 833: Line 862:
[[Image:ColumnsUnits.png]]
When you have added a column, you will set it on the left-hand side in the list Columns. You can define the unit being used by the column.
Line 858: Line 887:
Line 956: Line 985:
*Power
*Power
*Angle
*Attenuation
The Antenna Efficiency is expressed as a unitless number and the value should be between -1 and 1.
*Frequency
*Correction
===Antenna: Antenna Factor===
* Frequency
* Correction
Where the correction column is the antenna factor expressed in dB/m. The values are added to the calculation to account for the non-linearity in the antenna.
As {{RadiMation}} only has a correction file for the antenna factor that contains the frequency and actual antenna factor value, it is not known for {{RadiMation}} at which distance this antenna factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct antenna factor is specified, corresponding to the antenna distance that is actually used during the measurement.
===Antenna: Antenna Gain===
The Antenna Gain correction is used to correct the calculated logarithmic gain for horn antennas based on their horn dimensions.
This value is used by the calculations applicable to the calculated electrical field test levels.
*Frequency
*Frequency
*Correction
*Antenna gain
===Antenna: GpsHt - AntHeight===
*Height
===Antenna: Magnetic Flux Density (pT/A) correction===
Line 975: Line 1,037:
===Antenna: Magnetic Field Factor (dBpT/uV) correction===
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla or A/m) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns:
*Frequency
* Frequency
*Correction
* Correction
The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla or A/m unit.
As {{RadiMation}} only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for {{RadiMation}} at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement.
This same correction is also used during the [[MIL-STD-461]] RS101 Magnetic Field calibration for the 'calibration antenna'.
===Antenna: Reflection coëfficiënt ===
The Reflection coëfficiëntis used to correct the calculated required netto power in Watts for horn antennas to a calculated required forward power level in Watts.
The Reflection coefficient is used to correct the calculated required net power in Watts for horn antennas to a calculated required forward power level in Watts.
The Reflection coëfficiëntis expressed as a unit less number.
The Reflection coefficient is expressed as a unitless number.
Line 990: Line 1,054:
*Correction
*Correction
Where the correction column is the Reflection coëfficiënt.
*Height
*Angle
*Attenuation
===Cable: Cable loss===
Line 1,022: Line 1,075:
Line 1,060: Line 1,113:
[[Image:Calibration_jig_factor_50_Ohm_for_50_Ohm_system.png]]
== Using a transfer correction file provided by the supplier ==
Line 1,092: Line 1,144:
*Frequency
*Frequency
*Gain
*Gain
Line 1,105: Line 1,157:
* Frequency
* Frequency
* Correction. The values in the correction column of the offset correction should be expressed in V/m.
* Correction. The values in the column used for the offset correction should be expressed in V/m. This correction is then applied on the isotropic value.
or
* Frequency
* X-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the x-axis reading only.
* Y-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the y-axis reading only.
To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field offset correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m).
===Field sensor: Field Multiply===
* Frequency
* Frequency
* Correction. The values in the 'Correction' column should be expressed as x (multiplier).
* Correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is then applied on the isotropic value.
or
* Frequency
* X-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the x-axis reading only.
* Y-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the y-axis reading only.
To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply correction for that same frequency of 1.1x is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1).
===LISN: Amplitude correction===
Line 1,124: Line 1,190:
===Powermeter: Power correction===
*Frequency
* Frequency
*Correction
* Correction
Or
or
*Frequency
* Frequency
*Attenuation
* Attenuation
Or
or
*Frequency
* Frequency
*Gain
* Gain
* 'Frequency' and 'Correction' column: The values in the 'Correction' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
* Frequency
* 'Frequency' and 'Gain' column: The values in the 'Gain' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
* 'Frequency' and 'Attenuation' column: The values in the 'Attenuation' column are interpreted as 'attenuation' values. A positive value is interpreted as a loss caused by the preamplifier. A negative value is interpreted as a positive amplification of the signal, caused by the preamplifier. These 'attenuation' values will be added to the values that are measured with the receiver/analyser.
or
* Frequency
* Gain: The values in the Gain column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be subtracted from the values that are measured with the receiver/analyser.
or
* Frequency
* Attenuation: The values in the Attenuation column are interpreted as 'attenuation' values. A positive value is interpreted as a loss caused by the preamplifier. A negative value is interpreted as a positive amplification of the signal, caused by the preamplifier. These 'attenuation' values will be added to the values that are measured with the receiver/analyser.
===Receiver: Amplitude correction===
Line 1,156: Line 1,228:
*Frequency
* Frequency
*Correction
* Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3.
or
* Frequency
* Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2.
or
* Frequency
* Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4.
[[Category:Manual]]
[[Category:Manual]]
Latest revision as of 09:26, 15 April 2025
   Menu.svg Configuration
      Menu.svg Engineers
         Menu.svg Edit
Engineers[edit]
   Menu.svg File
      Menu.svg Change Engineer
      Menu.svg Engineers
Test engineer settings[edit]
Adding an engineer[edit]
Deleting an engineer[edit]
Password protection[edit]
  • The password is enabled by default for the default test engineer 'Administrator'.
Rights[edit]
Change password[edit]
Test Equipment[edit]
Adding test equipment[edit]
Deleting test equipment[edit]
Editing test equipment[edit]
Devices 1 & 2[edit]
Adding devices[edit]
Editing devices[edit]
Removing devices[edit]
Saving test equipment[edit]
Save As[edit]
Before Action[edit]
Example[edit]
After Action[edit]
Advanced options[edit]
  1. Units
  2. Directories
  3. Device Drivers
  4. Graphs
  5. Database
  6. Language
  7. Measurement settings
  8. Product standards
  9. Basic standards
  10. Enhanced Status Window
1. Units configuration[edit]
  • Radiated Immunity
  • Conducted Immunity
  • Radiated Emission
  • Conducted Emission
  • Pulsed Immunity
  • Antenna Diagram
  • RadiCal
Changing unit settings[edit]
Go to the Directories tab.
2. Through the 'Devices' tab.
Editing a device driver[edit]
Device Driver Settings[edit]
Confirm settings[edit]
Fixed information[edit]
Additional options[edit]
ScreenElementDescription.svg Zero Interval When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can significantly reduce the time that is needed to initialize each test.
Now we can conclude that:
4. Graphs configuration[edit]
Select
   Menu.svg Start
      Menu.svg Settings
         Menu.svg Control Panel
            Menu.svg ODBC Data Sources
MySQL ODBC database[edit]
MSSQL ODBC database[edit]
6. Language Selection[edit]
7. Measurement Settings[edit]
8. Product Standard[edit]
9. Basic Standard[edit]
Correction files[edit]
   Menu.svg File
      Menu.svg New
         Menu.svg Correction file
Entering data points[edit]
9 29.69 10 29.68 etc.
9 29.69 10 29.68
Saving correction files[edit]
   Menu.svg File
      Menu.svg Save correction file
      Menu.svg configuration
Correction file uses[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
  • Frequency
  • Correction
Or
  • Frequency
  • Power
  • Frequency
  • Correction
Or
  • Frequency
  • Power
  • Frequency
  • Correction
Or
  • Frequency
  • Power
  • Angle
  • Attenuation
  • Frequency
  • Correction
Antenna: Antenna Factor[edit]
  • Frequency
  • Correction
Where the correction column is the antenna factor expressed in dB/m. The values are added to the calculation to account for the non-linearity in the antenna.
As RadiMation® only has a correction file for the antenna factor that contains the frequency and actual antenna factor value, it is not known for RadiMation® at which distance this antenna factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct antenna factor is specified, corresponding to the antenna distance that is actually used during the measurement.
Antenna: Antenna Gain[edit]
  • Frequency
  • Antenna gain
  • Height
  • Frequency
  • Correction
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla or A/m) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns:
  • Frequency
  • Correction
The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla or A/m unit.
As RadiMation® only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for RadiMation® at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement.
  • Frequency
  • Correction
Cable: Cable loss[edit]
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
  • Frequency
  • Correction
This implies that:
  • Frequency
Or
  • Frequency
Or
  • Frequency
  • Frequency
  • Correction
Or
  • Frequency
  • Attenuation
Or
  • Frequency
  • Gain
  • Frequency
Or
  • Frequency
  • Frequency
or
  • Frequency
  • Frequency
or
  • Frequency
  • Frequency
  • Correction
  • Frequency
  • Correction
or
  • Frequency
  • Attenuation
or
  • Frequency
  • Gain
  • Frequency
or
  • Frequency
or
  • Frequency
  • Frequency
  • Correction
  • Frequency
  • Correction
  • Frequency
  • Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3.
or
  • Frequency
  • Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2.
or
  • Frequency
  • Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4.
