(→Antenna: Magnetic Flux Density (pT/A) correction) |
(→Signal Generator Output Correction) |
Line 867: | Line 867: | ||
Correction files can be used in many situations. The type and number of columns that should be used in the correction file is depending on the situation. | Correction files can be used in many situations. The type and number of columns that should be used in the correction file is depending on the situation. | ||
− | ===Signal Generator Output Correction=== | + | ===Signal Generator: Output Correction=== |
The output level of a signal generator can be corrected. A correction file with the correction should be attached to the device driver of the signal generator. The correction file should contain the columns: | The output level of a signal generator can be corrected. A correction file with the correction should be attached to the device driver of the signal generator. The correction file should contain the columns: | ||
*Frequency | *Frequency |
Revision as of 13:08, 31 May 2017
In the main menu, under the configuration pull down menu, a number of standard configurations can be made with regards to the engineers, test equipment, etc. The selected and/or entered configurations will be used as the default settings for future use.
Engineers
For each test, the name and password settings of the test engineer performing the current test can be entered.
To add, delete or edit (the names and settings of) the test engineers, go to the 'Engineers' window in the 'Configuration' tab.
The Engineers window will open and show the current configured list of engineers. To leave the 'Engineers' window, click 'OK' .
Test engineer settings
Adding an engineer
To add a new test engineer to the configuration, click 'Add' (in the 'Engineers' window).
The new engineer will not have a protection password (yet), this can be enabled by clicking 'Change Password' . Please read the paragraph 'Password protection' (located further down) for more information.
Deleting an engineer
To delete a test engineer from the configuration, select the engineer and click 'Delete' (in the 'Engineers' window).
Editing engineer settings
To edit the name, rights and/or password of an engineer in the configuration, select the engineer and click 'Edit' (in the 'Engineers' window).
Password protection
Rights
Change password
To enable, disable or change the password, click 'Change Password' (in the 'Test Engineer' window).
Test Equipment
Adding test equipment
Deleting test equipment
Editing test equipment
Devices 1 & 2
Adding devices
Editing devices
Removing devices
Saving test equipment
Save As
Before Action
Example
After Action
Default address information
- The default client
- The default manufacturer
- The default test site
Advanced options
There are a several advanced options in RadiMation®, that allow to tune or optimize some specific settings. Making changes to one or more of those settings, can have a major influence in the correct functioning of the software. Changes to one or more of these advanced options are only needed to circumvent a specific situation. Before changes can be made, a specific password should be specified. The required password is not related to any of the passwords of existing engineers. If it is necessary to make a change to one or more of the advanced options, the password will be provided to the end-user by the reseller or DARE!! Instruments. If the password is specified correctly, a configuration window is shown that includes a table to review and edit the list of advanced options.
- Units
- Directories
- Device Drivers
- Graphs
- Database
- Language
- Measurement settings
- Product standards
- Basic standards
- Enhanced Status Window
To configure these settings, go to the 'Configuration' window in the 'Configuration' tab.
1. Unit configuration
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Changing unit settings
2. Directory configuration
Modifying a path name
3. Device Driver configuration
2. Through the 'Devices' tab.
Select a device driver type
Adding a new device driver
Editing a device driver
Removing a device driver
Confirm settings
Fixed information
Brand | Shows the brand of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Description | Shows the description of the device. This information can be changed/viewed in this cell and can be used in the report generator. Changing the description does not result in losing the device in the test sites. The description is automatically updated in the test site information. If a report is generated from a test done prior to the description change the old description will be inserted. |
Device driver DLL Version | Shows the exact version of the DLL in which the device driver is contained. This is a unique identifier that specifies the version of the device driver. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Device drivers installation date | Shows the date and time of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Device drivers versions | Shows the version number of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Hardware version | Shows the hardware version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
ID | Shows the internally used ID of the device. This information can be changed/viewed in this cell and can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction. |
Serial number | Shows the serial number of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Software version | Shows the software version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Type | Shows the type of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Additional options
Add | The ‘’’ ‘Add’ button’’’ can be used to add additional information items to the general device driver settings. The additional items will be shown in the list. Additional information items can be used to note the storage location, the location of the user manual, etc. |
Edit | The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited. |
Remove | The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed. |
Zero Interval | When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can be a significantly reduce the time that is needed to initialize each test.
Calibration Expire Date | This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, RadiMation® will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date. |
Advanced | The ‘’’ ‘Advanced’ button’’’ can be used to open the advanced menu of the device driver. In chapter 15, the paragraph ‘Device specific configuration’ will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view the paragraphs about ‘Configurable devices’ or ‘Virtual devices’ (also in chapter 15) for these devices. |
Check | The ‘’’ ‘Check’ button’’’ can be used to check if the device is connected. This check is the same as the check that is performed during each test initialization. This check can be used to determine if the correct addressing is set, and whether communication with the device is possible. (This option is explained further below in the paragraph ‘Checking the device connection’.) |
Knowledgebase | The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the RadiWiki page about that device driver. That RadiWiki page can contain additional information or a detailed explanation of all the advanced settings. |
Now we can conclude that:
4. Graphics configuration
5. Database configuration
Introduction
Select
Data Source Name | This just a name for the setup you are making, name this Radimation, or another descriptive name for the used database |
Select Directory... | Allows to select the directory in which the text based database files are stored |
Use Current Directory | Overrules the selected directory to use the current directory. It is best to turn this checkbox off, and select a specific directory by using the Select Directory... button |
Extensions List | Shows a list of filename extensions which will be used to determine the correct text files. Remove *.asc and *.tab from the extentions list |
Define Format... | Will show the Define Text Format dialog, which can be used to define the format of the database as it is maintained in the text file |
MySQL ODBC database
Data Source Name | An unique name that uniquely identifies the Data Source from the other available Data Sources. |
Description | Any additional information regarding the Data Source |
TCP/IP Server | The name (or IP address) of the MySQL server, to which a connection should be made |
Port | The port-number to which the which the connection should be made. By default this is port-number 3306 |
User | The name of the MySQL user that should be used to connect to the MySQL Server |
Password | The password of the MySQL user that should be used to connect to the MySQL Server |
Database | The name of the MySQL database that should be connected to |
Test | Will perform a connection test to the specified MySQL server and Database |
Details >> | Will show additional configuration details for the MySQL connection. Your MySQL database administrator can help you with setting this details. |
MSSQL ODBC database
Server | The name (or IP address) of the MSSQL server, to which a connection should be made |
Windows NT Authentication | Select this option to use the credential of the logged in user to connect to the MSSQL server. |
SQL Server Authentication | Select this option to use a SQL Server specific username and password to connect to the MSSQL server. |
Client Configuration... | Will show additional configuration details for the MSSQL connection. Your MSSQL database administrator can help you with setting this details. |
Login ID | The name of the MSSQL user that should be used to connect to the MSSQL Server |
Password | The password of the MSSQL user that should be used to connect to the MSSQL Server |
Change default database | Enable the option, and specify or select the name of the database on the MSSQL Server that should be used. |
Test Data Source... | Will perform a connection test, to determine if a connection with the MSSQL server can be established, using the configured parameters. A confirmation or error dialog with some additional information will be shown. |
Ok | Stores the configured ODBC connection. |
Cancel | Closes the dialogs without creating an ODBC connection. |
6. Language Selection
7. Measurement Settings
Calibration
Required environment settings
8. Product Standard
Introduction
Adding a Product Standard
Add(ing) Basic Standard(s)
9. Basic Standard
Correction files
Creating correction files
Entering data points
When a data point is incorrect, it can be removed by selecting the data point and pressing the “Remove row” button.
Saving correction files
Attaching a correction files
Most correction files are used for correcting a measured value of a device driver. The made correction file should therefor be 'attached' to the correct device driver.
You can select the type of device you want in the box that is situated top right. The big window will display all currently configured device drivers, like in the picture all configured AD Converters. To edit the configuration of a device, select the device in the list and press Edit.
In the lower part of this window, the made correction files can be selected, for one of the specified uses. The RadiMation® software will use the specified correction files, every time when the device driver is used during a test. Depending on the type of the device driver, different uses of the correction files are available.
Correction file uses
- Frequency
- Correction
Amplifier Input Protection Level
- Frequency
- Correction
Or
- Frequency
- Power
Amplifier Forward Protection Level
- Frequency
- Correction
Or
- Frequency
- Power
Amplifier Reflected Protection Level
- Frequency
- Correction
Or
- Frequency
- Power
Coupler Forward Power Attenuation Correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Coupler Reflected Power Attenuation Correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Powermeter correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Cable loss correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
The values in the correction column should be expressed in dB. Cable device drivers can be used in different parts of a test-site setup. But they all have in common that an 'interesting' power-level is attenuated by the cable before it is measured by a measuring device (most of the time a powermeter, receiver or an analyzer). The values of the correction and attenuation column are therefore added to the measured powerlevel. The values in the gain column are inverted before they are added to the measured powerlevel.
Absorbing clamp loss correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Antenna Factor
For converting the reading dBuV into dBuV/m the antenna factor is needed.
- Frequency
- Correction
Where the correction column is the antenna factor.
GPSHeight till AntennaHeight Correction
- Height
AntennaGain versus Angle Correction
- Angle
- Attenuation
Calibration Jig Transfer correction
- Frequency
- Correction
This implies that:
Current sensor impedance
- Frequency
- Correction
The values in the correction column should be expressed in Ohm (Ω). The value specified in the correction column is used to correctly calculate the current flowing through the current sensor.
Field sensor correction
- Frequency
- Correction
LISN Amplitude correction
- Frequency
- Correction
Preamplifier Gain
- 'Frequency' and 'Correction' column: The values in the 'Correction' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
- 'Frequency' and 'Gain' column: The values in the 'Gain' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
- 'Frequency' and 'Attenuation' column: The values in the 'Attenuation' column are interpreted as 'attenuation' values. A positive value is interpreted as a loss caused by the preamplifier. A negative value is interpreted as a positive amplification of the signal, caused by the preamplifier. These 'attenuation' values will be added to the values that are measured with the receiver/analyser.
Receiver loss correction
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction
Resistor Correction
- Frequency
- Correction