The correction file should contain the columns: | The correction file should contain the columns: | ||
*Frequency | * Frequency | ||
*Correction | * Correction | ||
Where the correction column is the antenna factor expressed in dB/m. The values are added to the calculation to account for the non-linearity in the antenna. | |||
As {{RadiMation}} only has a correction file for the antenna factor that contains the frequency and actual antenna factor value, it is not known for {{RadiMation}} at which distance this antenna factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct antenna factor is specified, corresponding to the antenna distance that is actually used during the measurement. | |||
===Antenna: Antenna Gain=== | ===Antenna: Antenna Gain=== | ||
| Line 1,035: | Line 1,037: | ||
===Antenna: Magnetic Field Factor (dBpT/uV) correction=== | ===Antenna: Magnetic Field Factor (dBpT/uV) correction=== | ||
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns: | During emission measurements when the measured emission values have to be shown in magnetic field (Tesla or A/m) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns: | ||
The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla unit. | The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla or A/m unit. | ||
As {{RadiMation}} only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for {{RadiMation}} at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement. | |||
This same correction is also used during the [[MIL-STD-461]] RS101 Magnetic Field calibration for the 'calibration antenna'. | This same correction is also used during the [[MIL-STD-461]] RS101 Magnetic Field calibration for the 'calibration antenna'. | ||
| Line 1,160: | Line 1,164: | ||
* Z-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the z-axis reading only. | * Z-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the z-axis reading only. | ||
When the correction is specified for the individual | When the correction is specified for the individual axes, those axes are first corrected, after which the isotropic value is calculated based on the individually corrected axes. | ||
To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field offset isotropic correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m). | To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field offset isotropic correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m). | ||
| Line 1,174: | Line 1,178: | ||
* Z-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the z-axis reading only. | * Z-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the z-axis reading only. | ||
To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply isotropic correction for that same frequency of 1.1x is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1). | To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply isotropic correction for that same frequency of 1.1x is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1). | ||
| Line 1,198: | Line 1,202: | ||
===Pre amplifier: Amplifier Gain=== | ===Pre amplifier: Amplifier Gain=== | ||
The gain of | The gain of a used preamplifier can be corrected inside the software. A correction file with the frequency dependent gain of the preamplifier should be attached to the device driver of the preamplifier. The values in the correction column should be expressed in dB. | ||
The actually used calculation | The actually used calculation depends on the configuration of the correction file that is selected for the preamplifier. The correction file can contain the columns: | ||
* Correction: The values in the Correction column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be subtracted from the values that are measured with the receiver/analyser. | * Correction: The values in the Correction column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be subtracted from the values that are measured with the receiver/analyser. | ||
| Line 1,224: | Line 1,228: | ||
===Signal Generator: Output Correction=== | ===Signal Generator: Output Correction=== | ||
The output level of a signal generator can be corrected. A correction file with the correction should be attached to the device driver of the signal generator. The correction file should contain the columns: | The output level of a signal generator can be corrected. A correction file with the correction should be attached to the device driver of the signal generator. The correction file should contain the columns: | ||
*Correction | * Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3. | ||
The values in the correction column should be expressed in | or | ||
* Frequency | |||
* Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2. | |||
or | |||
* Frequency | |||
* Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4. | |||
Latest revision as of 09:26, 15 April 2025
| Warning: | To prevent unauthorised individuals from changing the global RadiMation® configurations, a password protection can be used for specific engineers. The password is activated by default for the "Administrator" engineer with the password "Radimation". This password protection can be switched off, changed and configured in the menu
Engineers[edit]
File
Change Engineer
Engineers
Test engineer settings[edit]
Adding an engineer[edit]
Deleting an engineer[edit]
Password protection[edit]
- The password is enabled by default for the default test engineer 'Administrator'.
Rights[edit]
Change password[edit]
Test Equipment[edit]
Adding test equipment[edit]
Deleting test equipment[edit]
Editing test equipment[edit]
Devices 1 & 2[edit]
Adding devices[edit]
Editing devices[edit]
Removing devices[edit]
Saving test equipment[edit]
Save As[edit]
Before Action[edit]
Example[edit]
After Action[edit]
| The default company name for the client. |
| The default contact person for the client. |
| The default address for the client. |
| The default zip code for the client. |
| The default city for the client. |
| The default state for the client. |
| The default country for the client. |
| The default telephone number for the client. |
| The default fax numberfor the client. |
| The default e-mail address for the client. |
| The default company name for the manufacturer. |
| The default contact person for the manufacturer. |
| The default address for the manufacturer. |
| The default zip code for the manufacturer. |
| The default city for the manufacturer. |
| The default state for the manufacturer. |
| The default country for the manufacturer. |
| The default telephone number for the manufacturer. |
| The default fax number for the manufacturer. |
| The default e-mail address for the manufacturer. |
| The default company name for the test site. |
| The default contact person for the test site. |
| The default address for the test site. |
| The default zip code for the test site. |
| The default city for the test site. |
| The default state for the test site. |
| The default country for the test site. |
| The default telephone number for the test site. |
| The default fax number for the test site. |
| The default e-mail address for the test site. |
| Stores the modified information and closes the window. |
| Discards the modified information and closes the window. |
Advanced options[edit]
- Units
- Directories
- Device Drivers
- Graphs
- Database
- Language
- Measurement settings
- Product standards
- Basic standards
- Enhanced Status Window
1. Units configuration[edit]
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Changing unit settings[edit]
Go to the Directories tab.
2. Through the 'Devices' tab.
Editing a device driver[edit]
Device Driver Settings[edit]
Confirm settings[edit]
Fixed information[edit]
| Shows the brand of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the description of the device. This information can be changed/viewed in this cell and can be used in the report generator. Changing the description does not result in losing the device in the test sites. The description is automatically updated in the test site information. If a report is generated from a test done prior to the description change the old description will be inserted. |
| Shows the exact version of the DLL in which the device driver is contained. This is a unique identifier that specifies the version of the device driver. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the date and time of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the version number of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the hardware version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the internally used ID of the device. This information can be changed/viewed in this cell and can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction. |
| Shows the serial number of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the software version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the type of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Additional options[edit]
| The ‘’’ ‘Add’ button’’’ can be used to add additional information items to the general device driver settings. The additional items will be shown in the list. Additional information items can be used to note the storage location, the location of the user manual, etc. |
| The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited. |
| The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed. |
| When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can significantly reduce the time that is needed to initialize each test.
| This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, RadiMation® will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date. |
| The ‘’’ ‘Advanced’ button’’’ can be used to open the advanced menu of the device driver. In chapter 15, the paragraph ‘Device specific configuration’ will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view the paragraphs about ‘Configurable devices’ or ‘Virtual devices’ (also in chapter 15) for these devices. |
| The ‘’’ ‘Check’ button’’’ can be used to check if the device is connected. This check is the same as the check that is performed during each test initialization. This check can be used to determine if the correct addressing is set, and whether communication with the device is possible. (This option is explained further below in the paragraph ‘Checking the device connection’.) |
| The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the RadiWiki page about that device driver. That RadiWiki page can contain additional information or a detailed explanation of all the advanced settings. |
Now we can conclude that:
4. Graphs configuration[edit]
Select
Start
Settings
Control Panel
ODBC Data Sources
| This is just a name for the setup you are making, name this Radimation, or another descriptive name for the used database |
| Allows to select the directory in which the text based database files are stored |
| Overrules the selected directory to use the current directory. It is best to turn this checkbox off, and select a specific directory by using the Select Directory... button |
| Shows a list of filename extensions which will be used to determine the correct text files. Remove *.asc and *.tab from the extentions list |
| Will show the Define Text Format dialog, which can be used to define the format of the database as it is maintained in the text file |
MySQL ODBC database[edit]
| A unique name that uniquely identifies the Data Source from the other available Data Sources. |
| Any additional information regarding the Data Source |
| The name (or IP address) of the MySQL server, to which a connection should be made |
| The port-number to which the which the connection should be made. By default this is port-number 3306 |
| The name of the MySQL user that should be used to connect to the MySQL Server |
| The password of the MySQL user that should be used to connect to the MySQL Server |
| The name of the MySQL database that should be connected to |
| Will perform a connection test to the specified MySQL server and Database |
| Will show additional configuration details for the MySQL connection. Your MySQL database administrator can help you with setting this details. |
MSSQL ODBC database[edit]
| The name (or IP address) of the MSSQL server, to which a connection should be made |
| Select this option to use the credential of the logged in user to connect to the MSSQL server. |
| Select this option to use a SQL Server specific username and password to connect to the MSSQL server. |
| Will show additional configuration details for the MSSQL connection. Your MSSQL database administrator can help you with setting this details. |
| The name of the MSSQL user that should be used to connect to the MSSQL Server |
| The password of the MSSQL user that should be used to connect to the MSSQL Server |
| Enable the option, and specify or select the name of the database on the MSSQL Server that should be used. |
| Will perform a connection test, to determine if a connection with the MSSQL server can be established, using the configured parameters. A confirmation or error dialog with some additional information will be shown. |
| Stores the configured ODBC connection. |
| Closes the dialogs without creating an ODBC connection. |
6. Language Selection[edit]
7. Measurement Settings[edit]
| Warning: | Disabling the 'zoom-in in emission tests' in combination with analysers will very likely result in an incorrect measurement. |
8. Product Standard[edit]
| Note: | When a product standard is selected or added in the 'EUT' window, the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the 'Configuration' window of RadiMation® is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested. |
9. Basic Standard[edit]
Correction files[edit]
File
New
Correction file
Entering data points[edit]
9 29.69
10 29.68
etc.
9 29.69 10 29.68
Saving correction files[edit]
File
Save correction file
configuration
Correction file uses[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
Or
- Frequency
- Power
- Frequency
- Correction
Or
- Frequency
- Power
- Frequency
- Correction
Or
- Frequency
- Power
- Angle
- Attenuation
- Frequency
- Correction
Antenna: Antenna Factor[edit]
- Frequency
- Correction
Where the correction column is the antenna factor expressed in dB/m. The values are added to the calculation to account for the non-linearity in the antenna.
As RadiMation® only has a correction file for the antenna factor that contains the frequency and actual antenna factor value, it is not known for RadiMation® at which distance this antenna factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct antenna factor is specified, corresponding to the antenna distance that is actually used during the measurement.
Antenna: Antenna Gain[edit]
- Frequency
- Antenna gain
- Height
- Frequency
- Correction
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla or A/m) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns:
- Frequency
- Correction
The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla or A/m unit.
As RadiMation® only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for RadiMation® at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement.
- Frequency
- Correction
Cable: Cable loss[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
This implies that:
- Frequency
Or
- Frequency
Or
- Frequency
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
Or
- Frequency
- Frequency
or
- Frequency
- Frequency
or
- Frequency
- Frequency
- Correction
- Frequency
- Correction
or
- Frequency
- Attenuation
or
- Frequency
- Gain
- Frequency
or
- Frequency
or
- Frequency
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3.
or
- Frequency
- Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2.
or
- Frequency
- Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4.