| (34 intermediate revisions by 5 users not shown) | |||
| Line 15: | Line 15: | ||
== Engineers == | == Engineers == | ||
An unlimited number of engineers can be added, each with an unique name. For each engineer a login password and different permissions can be assigned which give them the right to modify specific parts of the {{RadiMation}} configuration. There is no relation between the engineer in {{RadiMation}} and the Windows Logon account. Also the password that can be configured for the engineer in {{RadiMation}} is not linked and also not synchronized with the Windows User account password. | On a new installation, only a single 'Administrator' engineer is created by default. An unlimited number of engineers can be added, each with an unique name. For each engineer a login password and different permissions can be assigned which give them the right to modify specific parts of the {{RadiMation}} configuration. There is no relation between the engineer in {{RadiMation}} and the Windows Logon account. Also the password that can be configured for the engineer in {{RadiMation}} is not linked and also not synchronized with the Windows User account password. | ||
During the startup of {{RadiMation}} the actual engineer that is going to start tests can be selected in the {{WindowTitle|Login}} dialog. If a password is configured for that engineer, it should also be specified at that moment. | During the startup of {{RadiMation}} the actual engineer that is going to start tests can be selected in the {{WindowTitle|Login}} dialog. If a password is configured for that engineer, it should also be specified at that moment. | ||
| Line 73: | Line 73: | ||
{{RadiMation}} allows you to protect a number of configurations with a password. This password can be enabled, disabled or changed in the {{ScreenElement|'Test Engineer' window}}. | {{RadiMation}} allows you to protect a number of configurations with a password. This password can be enabled, disabled or changed in the {{ScreenElement|'Test Engineer' window}}. | ||
* The password is | * The password is enabled by default for the default test engineer 'Administrator'. | ||
| Line 130: | Line 130: | ||
=== Adding test equipment === | === Adding test equipment === | ||
To add a (set of) test equipment to the configuration, click the ''' 'Add' ''' button in the top of the window. The 'New Test Equipment' window will appear. | To add a new (set of) test equipment to the configuration, click the ''' 'Add' ''' button in the top of the window. The 'New Test Equipment' window will appear. | ||
Enter a name for the new test equipment in the ''' 'New Test Equipment' ''' window (e.g. Example Test Equipment). | Enter a name for the new test equipment in the ''' 'New Test Equipment' ''' window (e.g. Example Test Equipment). | ||
| Line 150: | Line 150: | ||
=== Editing test equipment === | === Editing test equipment === | ||
To configure | To configure a test equipment set, select the desired configuration from the pull-down menu (e.g. ''Example Test Equipment''). | ||
Specify the | Specify the required devices by adding, editing and/or removing equipment under the corresponding tabs, as described further below. | ||
=== Test Equipment configuration === | |||
Each tab in the Test Equipment window represents a category of devices. | |||
All tabs support adding, editing, and removing devices using the buttons at the bottom of the window. | |||
The available tabs are: | |||
* '''Devices 1''' | |||
* '''Devices 2''' | |||
* '''Field Probes''' | |||
* '''Pulsed''' | |||
* '''Cables''' | |||
* '''Data Logging''' | |||
* '''Auxiliary''' | |||
* '''Monitoring''' | |||
* '''Before Action''' | |||
* '''After Action''' | |||
==== Adding devices ==== | ==== Adding devices ==== | ||
| Line 194: | Line 208: | ||
[[Image:Test equipment remove selected devices.png]] | [[Image:Test equipment remove selected devices.png]] | ||
=== Test Equipment tabs overview === | |||
==== Devices 1 & Devices 2 ==== | |||
These two tabs contain general-purpose equipment lists. | |||
==== Field Probes ==== | |||
=== | This tab allows configuring up to eight field probe slots ('''Field Probe 1''' through '''Field Probe 8'''). | ||
==== Pulsed ==== | |||
This tab contains equipment used for pulsed immunity testing, including: | |||
* ESD generators | |||
* EFT/Burst generators | |||
* Surge generators | |||
* Voltage dips & interruptions equipment | |||
==== Cables ==== | |||
The Cables tab lists all available cable device entries that describe the physical signal path between equipment components. | |||
==== Data Logging ==== | |||
This tab contains devices used for extended data logging. | |||
==== Auxiliary ==== | |||
Auxiliary devices include all supporting equipment that does not fall under other categories. | |||
This may include passive objects (fixtures, tables, loads) or configurable auxiliary devices. | |||
[[Image:Test Equipment Auxiliary.png]] | |||
==== Monitoring ==== | |||
The Monitoring tab is used to configure '''Monitoring input channels'''. | |||
== Before Actions & After Actions == | ==== Before Actions & After Actions ==== | ||
The before and after actions of a testsite are designed to be triggered when the testsite is activated (before actions) or deactivated (after actions). | The before and after actions of a testsite are designed to be triggered when the testsite is activated (before actions) or deactivated (after actions). | ||
| Line 226: | Line 254: | ||
When a test with 3 bands is configured, and the first and third band are using the same testsite, while another testsite is selected for the second band, then the after and before actions will be executed when the test is switching from band 1 to band 2, and again when the test is switching from band 2 to band 3. | When a test with 3 bands is configured, and the first and third band are using the same testsite, while another testsite is selected for the second band, then the after and before actions will be executed when the test is switching from band 1 to band 2, and again when the test is switching from band 2 to band 3. | ||
=== Before Action === | ===== Before Action ===== | ||
Under the tab 'Before Action' any and all actions that the end-user wants to have carried out before the equipment of the selected 'test equipment' is initialized can be specified. | Under the tab 'Before Action' any and all actions that the end-user wants to have carried out before the equipment of the selected 'test equipment' is initialized can be specified. | ||
| Line 240: | Line 268: | ||
The easiest way to explain this function is with an example: | The easiest way to explain this function is with an example: | ||
==== Example ==== | ====== Example ====== | ||
{{RadiMation}} needs to be configured for people who do not perform EMC measurements every day. To make sure the engineers do not forget to use the right equipment and procedure, the manager has decided to incorporate a reminder in the test setup. | {{RadiMation}} needs to be configured for people who do not perform EMC measurements every day. To make sure the engineers do not forget to use the right equipment and procedure, the manager has decided to incorporate a reminder in the test setup. | ||
| Line 267: | Line 295: | ||
[[IMAGE:AddACtion_PopUp.PNG]] | [[IMAGE:AddACtion_PopUp.PNG]] | ||
=== After Action === | ===== After Action ===== | ||
Under the tab 'After Action' any and all actions that the end-user wants to have carried out after the equipment of the selected 'test equipment' is disabled can be specified. | Under the tab 'After Action' any and all actions that the end-user wants to have carried out after the equipment of the selected 'test equipment' is disabled can be specified. | ||
| Line 277: | Line 305: | ||
(It is also possible to specify actions that should be carried out before the equipment of the selected Test Equipment is initialized. This functionality is called 'Before Action'.) | (It is also possible to specify actions that should be carried out before the equipment of the selected Test Equipment is initialized. This functionality is called 'Before Action'.) | ||
=== Saving test equipment === | |||
To save the changes made in the test equipment list, click ''' 'Save' ''' (you will remain in the 'Test Equipment' window). | |||
==== Save As ==== | |||
To save the test equipment list under a new name, click ''' 'Save As' '''. | |||
Enter a new name for the test equipment in the ''' 'New Test Equipment' window''' (e.g. Example Test Equipment 2). | |||
Click ''' 'OK' ''' to confirm the name of the new test equipment, or click ''' 'Cancel' ''' to cancel the new name. You will be returned to the 'Test Equipment' window. | |||
[[Image:Test equipment save as.png]] | |||
To close the 'Test Equipment' window, click ''' 'Close' '''. | |||
If you have not saved the changes to the test equipment list, a dedicated window will appear asking you to save these changes. | |||
To save the changes, click ''' 'Yes' ''', to not save the changes, click ''' 'No' '''. The 'Test Equipment' window will close. | |||
[[Image:Test equipment close.png]] | |||
== Default address information == | == Default address information == | ||
| Line 293: | Line 345: | ||
{{Menu|Configuration|Default Address Information}} | {{Menu|Configuration|Default Address Information}} | ||
In the {{WindowTitle|Default Address Information}} window, the default address information can be configured. | In the {{WindowTitle|Default Address Information}} window, the default address information can be configured. Specifying this information is not required, however it has the benefit that address information that is always the same doesn't need to be entered over and over again in every new [[EUT]] file. | ||
| Line 338: | Line 390: | ||
[[Image:Configuration pull down menu.png]] | [[Image:Configuration pull down menu.png]] | ||
There are a several [[advanced options]] in {{RadiMation}}, that allow to tune or optimize some specific settings. Making changes to one or more of those settings, can have a major influence | There are a several [[advanced options]] in {{RadiMation}}, that allow to tune or optimize some specific settings. Making changes to one or more of those settings, can have a major influence on the correct functioning of the software. Changes to one or more of these advanced options are only needed to circumvent a specific situation. Before changes can be made, a specific password should be specified. The required password is not related to any of the passwords of existing engineers. If it is necessary to make a change to one or more of the advanced options, the password will be provided to the end-user by the [[reseller]] or {{CompanyName}}. If the password is specified correctly, a configuration window is shown that includes a table to review and edit the list of advanced options. | ||
| Line 430: | Line 482: | ||
With the {{ScreenElement|Use Global Configuration}} option, it is possible to share the configuration with other versions of RadiMation on the same computer. This is the default setting. | With the {{ScreenElement|Use Global Configuration}} option, it is possible to share the configuration with other versions of RadiMation on the same computer. This is the default setting. | ||
With the {{ScreenElement|Use standalone configuration}} option, it is possible to create | With the {{ScreenElement|Use standalone configuration}} option, it is possible to create a version specific isolated configuration directory with its own content only used for this version of RadiMation. This can for example be used to verify and validate another {{RadiMation}} version without influencing another {{RadiMation}} version. | ||
The location of the {{ScreenElement|Device Driver Configuration Files}} can by changed by clicking on the {{ScreenElement|'Open Folder' icon}}. A ''' 'Browse For Folder' ''' window will appear, allowing | The location of the {{ScreenElement|Device Driver Configuration Files}} can by changed by clicking on the {{ScreenElement|'Open Folder' icon}}. A ''' 'Browse For Folder' ''' window will appear, allowing the selection of an existing device driver configuration files directory. If an empty or an invalid device driver configuration files directory is selected, an error will be shown, and the original selected directory will remain selected. | ||
==== Creating a new 'Device Driver Configuration Files' folder ==== | ==== Creating a new 'Device Driver Configuration Files' folder ==== | ||
| Line 473: | Line 525: | ||
==== Select a device driver type ==== | ==== Select a device driver type ==== | ||
The first step to create or configure a device driver is to select the required ''' '[[Device Types|Device Driver Type]]' ''' (e.g. Signal Generator). This can be done from the drop-down menu in the ‘Configuration’ window. | The first step to create or configure a device driver is to select the required ''' '[[Device Types|Device Driver Type]]' ''' (e.g. Signal Generator). This can be done from the drop-down menu in the ‘Configuration’ window. Once a type has been selected, the available drivers for that type will be shown. | ||
[[Image:Device driver type.png]] | [[Image:Device driver type.png]] | ||
==== Adding a new device driver ==== | ==== Adding a new device driver ==== | ||
| Line 557: | Line 608: | ||
{{ScreenElementDescription|Edit|The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited.}} | {{ScreenElementDescription|Edit|The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited.}} | ||
{{ScreenElementDescription|Remove|The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed.}} | {{ScreenElementDescription|Remove|The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed.}} | ||
{{ScreenElementDescription|Zero Interval|When the {{ScreenElement|Zero interval}} is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that [[Power Meter]]s and [[Field Sensor]]s are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can | {{ScreenElementDescription|Zero Interval|When the {{ScreenElement|Zero interval}} is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that [[Power Meter]]s and [[Field Sensor]]s are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can significantly reduce the time that is needed to initialize each test. | ||
This setting is only available for power meter and field strength sensor devices.}} | This setting is only available for power meter and field strength sensor devices.}} | ||
{{ScreenElementDescription|Calibration Expire Date|This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, {{Radimation}} will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date.}} | {{ScreenElementDescription|Calibration Expire Date|This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, {{Radimation}} will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date.}} | ||
| Line 564: | Line 615: | ||
{{ScreenElementDescription|Knowledgebase|The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the [[RadiWiki]] page about that device driver. That [[RadiWiki]] page can contain additional information or a detailed explanation of all the advanced settings.}} | {{ScreenElementDescription|Knowledgebase|The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the [[RadiWiki]] page about that device driver. That [[RadiWiki]] page can contain additional information or a detailed explanation of all the advanced settings.}} | ||
==== Checking the device connection ==== | ==== Checking the device connection ==== | ||
| Line 661: | Line 711: | ||
===Additional Setup for ODBC Connected databases === | ===Additional Setup for ODBC Connected databases === | ||
Text, CSV, Excel, [[MSSQL]] and [[MySQL]] dabase files need some additional setup before they can be used. The reason for this is that {{RadiMation}} uses [[ODBC]] to connect and read out | Text, CSV, Excel, [[MSSQL]] and [[MySQL]] dabase files need some additional setup before they can be used. The reason for this is that {{RadiMation}} uses [[ODBC]] to connect and read out those databases. This additional setup of an [[ODBC]] connection must be performed before the database driver can be configured in {{RadimationLink}}. {{:ODBC Configuration}} | ||
| Line 680: | Line 730: | ||
{{:SQL Server ODBC Driver Configuration}} | {{:SQL Server ODBC Driver Configuration}} | ||
=== 6. Language Selection === | === 6. Language Selection === | ||
{{RadiMation}} supports multi-language use. This means that the software package can be operated in several | {{RadiMation}} supports multi-language use. This means that the software package can be operated in several languages. | ||
In addition, the multi- language support allows for easy translation from English to most other languages. | In addition, the multi- language support allows for easy translation from English to most other languages. | ||
| Line 710: | Line 759: | ||
=== 7. Measurement Settings=== | === 7. Measurement Settings=== | ||
With measurement settings you can change the behavior of the software at certain points. Be very careful what you do with these setting, because they are used | With measurement settings you can change the behavior of the software at certain points. Be very careful what you do with these setting, because they are used throughout {{RadiMation}}. | ||
[[Image:Measurement Settings.PNG]] | [[Image:Measurement Settings.PNG]] | ||
| Line 745: | Line 794: | ||
With the required environment settings you can force the test engineer to fill in specific [[Environmental Data]] settings. The first test done with a new EUT document requires the engineer to fill in the requested information. The second time a button will appear that will allow you to automatically fill in the same information as used in the previous test. | With the required environment settings you can force the test engineer to fill in specific [[Environmental Data]] settings. The first test done with a new EUT document requires the engineer to fill in the requested information. The second time a button will appear that will allow you to automatically fill in the same information as used in the previous test. | ||
=== 8. Product Standard === | === 8. Product Standard === | ||
| Line 866: | Line 914: | ||
[[Image:ColumnsUnits.png]] | [[Image:ColumnsUnits.png]] | ||
When you have added a column, you will | When you have added a column, you will set it on the left-hand side in the list Columns. You can define the unit being used by the column. | ||
This is done by selecting the column and click units. | This is done by selecting the column and click units. | ||
| Line 891: | Line 939: | ||
To facilitate this you have two buttons to import and export from the clipboard. This also makes it easy to transfer the data to excel and use calculations if needed. | To facilitate this you have two buttons to import and export from the clipboard. This also makes it easy to transfer the data to excel and use calculations if needed. | ||
This can be done from excel files, text files, pdf files etc. In the example above you need two rows: Frequency and Correction. To be able to use the file that has been given by the | This can be done from excel files, text files, pdf files etc. In the example above you need two rows: Frequency and Correction. To be able to use the file that has been given by the manufacturer to insert the correction data most of the time you have to edit it a bit. In this example this means that you must make sure that the list you are copying to the clipboard looks as follows: | ||
| Line 989: | Line 1,037: | ||
*Power | *Power | ||
The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum allowed reflected power. During the regulation of the test-level the reflected power level is measured and the regulation is ensuring that the reflected power will not become higher than the maximum reflected power as specified in the correction file. | The values in the correction column should be expressed in dBm. The value specified in the correction or power column is the maximum allowed reflected power. During the regulation of the test-level the reflected power level is measured and the regulation is ensuring that the reflected power will not become higher than the maximum reflected power as specified in the correction file. | ||
===Antenna: Ant Gain vs Angle=== | |||
This correction is only available when the Antenna Diagram module is activated. The antenna gain depends on the measuring angle during antenna diagram measurements. A correction file can be used to specify the AntennaGain depending on the receiving angle. This correction file should be attached to the device driver of the antenna. The correction file should contain the columns: | |||
*Angle | |||
*Attenuation | |||
The value of the attenuation column is added to the measured radiated power. | |||
===Antenna: Antenna Efficiency === | |||
The Antenna Efficiency is used to indicate the Efficiency of the antenna. | |||
The Antenna Efficiency is expressed as a unitless number and the value should be between -1 and 1. | |||
*Frequency | |||
*Correction | |||
Where the values in the correction column are used as a correction which is added on top of the the Antenna Efficiency that is specified in the antenna device driver itself. | |||
===Antenna: Antenna Factor=== | ===Antenna: Antenna Factor=== | ||
For converting the reading | For converting the reading dBμV into dBμV/m the antenna factor is needed. This is expressed as dB/m. | ||
* Frequency | |||
* Correction | |||
Where the correction column is the antenna factor expressed in dB/m. The values are added to the calculation to account for the non-linearity in the antenna. | |||
As {{RadiMation}} only has a correction file for the antenna factor that contains the frequency and actual antenna factor value, it is not known for {{RadiMation}} at which distance this antenna factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct antenna factor is specified, corresponding to the antenna distance that is actually used during the measurement. | |||
===Antenna: Antenna Gain=== | |||
The Antenna Gain correction is used to correct the calculated logarithmic gain for horn antennas based on their horn dimensions. | |||
This value is used by the calculations applicable to the calculated electrical field test levels. | |||
The Antenna Gain is expressed in dB(i) | |||
The correction file should contain the columns: | The correction file should contain the columns: | ||
*Frequency | *Frequency | ||
* | *Antenna gain | ||
===Antenna: GpsHt - AntHeight=== | |||
This correction is only available when the Antenna Diagram module is activated. The height distance between the GPS receiver and the Antenna can be corrected by a correction file. A correction file with the height distance should be attached to the device driver of the antenna. The correction file should contain the column: | |||
*Height | |||
Only one row should be present in the correction file. If multiple rows are available, only the height value of the first row will be used. The value of the height column, is subtracted from the height received by the GPS receiver. The result is the exact height of the receiving antenna for the antenna diagram measurements. | |||
===Antenna: Magnetic Flux Density (pT/A) correction=== | ===Antenna: Magnetic Flux Density (pT/A) correction=== | ||
| Line 1,008: | Line 1,089: | ||
===Antenna: Magnetic Field Factor (dBpT/uV) correction=== | ===Antenna: Magnetic Field Factor (dBpT/uV) correction=== | ||
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns: | During emission measurements when the measured emission values have to be shown in magnetic field (Tesla or A/m) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns: | ||
*Frequency | * Frequency | ||
*Correction | * Correction | ||
The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla unit. | The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla or A/m unit. | ||
As {{RadiMation}} only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for {{RadiMation}} at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement. | |||
This same correction is also used during the [[MIL-STD-461]] RS101 Magnetic Field calibration for the 'calibration antenna'. | This same correction is also used during the [[MIL-STD-461]] RS101 Magnetic Field calibration for the 'calibration antenna'. | ||
===Antenna: Reflection coefficient=== | ===Antenna: Reflection coefficient=== | ||
The Reflection coefficient is used to correct the calculated required | The Reflection coefficient is used to correct the calculated required net power in Watts for horn antennas to a calculated required forward power level in Watts. | ||
The Reflection coefficient is expressed as a | The Reflection coefficient is expressed as a unitless number. | ||
| Line 1,024: | Line 1,107: | ||
Where the correction column is the Reflection coefficient. | Where the correction column is the Reflection coefficient. | ||
===Cable: Cable loss=== | ===Cable: Cable loss=== | ||
| Line 1,065: | Line 1,127: | ||
If the jig is not "Configurable" there is a default correction factor of 9.5 dB (theoretical value). We will explain how the presence of a correction factor is | If the jig is not "Configurable" there is a default correction factor of 9.5 dB (theoretical value). We will explain how the presence of a correction factor is taken into account by RadiMation and how this value is used by an example. For instance we have the following situation: At 150 kHz the real total correction factor of the jig is 10 dB. When this value is saved in the correction file, what will be the behavior of RadiMation? Will it now take the 10 dB as the total transfer correction? The answer is that the correction value will be added, but RadiMation will use this specified 10 dB correction on top of the internal correction value. In this case it is not correct and a correction of app. 0,5 should have been used. | ||
There are two categories of Injection devices: | There are two categories of Injection devices: | ||
| Line 1,103: | Line 1,165: | ||
[[Image:Calibration_jig_factor_50_Ohm_for_50_Ohm_system.png]] | [[Image:Calibration_jig_factor_50_Ohm_for_50_Ohm_system.png]] | ||
== Using a transfer correction file provided by the supplier == | == Using a transfer correction file provided by the supplier == | ||
| Line 1,135: | Line 1,196: | ||
*Gain | *Gain | ||
The value specified in the gain column is | The value specified in the gain column is subtracted from the typical attenuation of the coupler. The values in the correction or attenuation column is added to the typical attenuation of the coupler. For example: A coupler is used with a typical reflected attenuation of 40 dB. A correction file with a frequency and attenuation column is connected to the device driver. The value specified in the attenuation column for 130 kHz. is 0.13 dB. Then the software will use a value of 40.13 dB for the total attenuation of the coupler. | ||
===Current sensor: impedance=== | ===Current sensor: impedance=== | ||
| Line 1,155: | Line 1,216: | ||
* Z-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the z-axis reading only. | * Z-axis correction. The values in the column used for the offset correction should be expressed in V/m. This correction is applied on the z-axis reading only. | ||
When the correction is specified for the individual | When the correction is specified for the individual axes, those axes are first corrected, after which the isotropic value is calculated based on the individually corrected axes. | ||
To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field offset isotropic correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m). | To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field offset isotropic correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m). | ||
| Line 1,169: | Line 1,230: | ||
* Z-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the z-axis reading only. | * Z-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the z-axis reading only. | ||
To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply isotropic correction for that same frequency of 1.1x is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1). | To summarize this in an example, let’s assume that the uncorrected field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply isotropic correction for that same frequency of 1.1x is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1). | ||
| Line 1,193: | Line 1,254: | ||
===Pre amplifier: Amplifier Gain=== | ===Pre amplifier: Amplifier Gain=== | ||
The gain of | The gain of a used preamplifier can be corrected inside the software. A correction file with the frequency dependent gain of the preamplifier should be attached to the device driver of the preamplifier. The values in the correction column should be expressed in dB. | ||
The actually used calculation | The actually used calculation depends on the configuration of the correction file that is selected for the preamplifier. The correction file can contain the columns: | ||
* Correction: The values in the Correction column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be subtracted from the values that are measured with the receiver/analyser. | * Correction: The values in the Correction column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be subtracted from the values that are measured with the receiver/analyser. | ||
| Line 1,219: | Line 1,280: | ||
===Signal Generator: Output Correction=== | ===Signal Generator: Output Correction=== | ||
The output level of a signal generator can be corrected. A correction file with the correction should be attached to the device driver of the signal generator. The correction file should contain the columns: | The output level of a signal generator can be corrected. A correction file with the correction should be attached to the device driver of the signal generator. The correction file should contain the columns: | ||
*Correction | * Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3. | ||
The values in the correction column should be expressed in | or | ||
* Frequency | |||
* Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2. | |||
or | |||
* Frequency | |||
* Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4. | |||
Latest revision as of 11:05, 11 December 2025
| Warning: | To prevent unauthorised individuals from changing the global RadiMation® configurations, a password protection can be used for specific engineers. The password is activated by default for the "Administrator" engineer with the password "Radimation". This password protection can be switched off, changed and configured in the menu
Engineers[edit]
File
Change Engineer
Engineers
Test engineer settings[edit]
Adding an engineer[edit]
Deleting an engineer[edit]
Password protection[edit]
- The password is enabled by default for the default test engineer 'Administrator'.
Rights[edit]
Change password[edit]
Test Equipment[edit]
Adding test equipment[edit]
To add a new (set of) test equipment to the configuration, click the 'Add' button in the top of the window. The 'New Test Equipment' window will appear.
Deleting test equipment[edit]
Editing test equipment[edit]
To configure a test equipment set, select the desired configuration from the pull-down menu (e.g. Example Test Equipment).
Test Equipment configuration[edit]
Each tab in the Test Equipment window represents a category of devices. All tabs support adding, editing, and removing devices using the buttons at the bottom of the window.
- Devices 1
- Devices 2
- Field Probes
- Pulsed
- Cables
- Data Logging
- Auxiliary
- Monitoring
- Before Action
- After Action
Adding devices[edit]
Editing devices[edit]
Removing devices[edit]
Test Equipment tabs overview[edit]
Devices 1 & Devices 2[edit]
Field Probes[edit]
This tab allows configuring up to eight field probe slots (Field Probe 1 through Field Probe 8).
Pulsed[edit]
- ESD generators
- EFT/Burst generators
- Surge generators
- Voltage dips & interruptions equipment
Cables[edit]
Data Logging[edit]
Auxiliary[edit]
Auxiliary devices include all supporting equipment that does not fall under other categories. This may include passive objects (fixtures, tables, loads) or configurable auxiliary devices.
Monitoring[edit]
The Monitoring tab is used to configure Monitoring input channels.
Before Action[edit]
Example[edit]
After Action[edit]
Saving test equipment[edit]
Save As[edit]
| The default company name for the client. |
| The default contact person for the client. |
| The default address for the client. |
| The default zip code for the client. |
| The default city for the client. |
| The default state for the client. |
| The default country for the client. |
| The default telephone number for the client. |
| The default fax numberfor the client. |
| The default e-mail address for the client. |
| The default company name for the manufacturer. |
| The default contact person for the manufacturer. |
| The default address for the manufacturer. |
| The default zip code for the manufacturer. |
| The default city for the manufacturer. |
| The default state for the manufacturer. |
| The default country for the manufacturer. |
| The default telephone number for the manufacturer. |
| The default fax number for the manufacturer. |
| The default e-mail address for the manufacturer. |
| The default company name for the test site. |
| The default contact person for the test site. |
| The default address for the test site. |
| The default zip code for the test site. |
| The default city for the test site. |
| The default state for the test site. |
| The default country for the test site. |
| The default telephone number for the test site. |
| The default fax number for the test site. |
| The default e-mail address for the test site. |
| Stores the modified information and closes the window. |
| Discards the modified information and closes the window. |
Advanced options[edit]
- Units
- Directories
- Device Drivers
- Graphs
- Database
- Language
- Measurement settings
- Product standards
- Basic standards
- Enhanced Status Window
1. Units configuration[edit]
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Changing unit settings[edit]
Go to the Directories tab.
2. Through the 'Devices' tab.
Editing a device driver[edit]
Device Driver Settings[edit]
Confirm settings[edit]
Fixed information[edit]
| Shows the brand of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the description of the device. This information can be changed/viewed in this cell and can be used in the report generator. Changing the description does not result in losing the device in the test sites. The description is automatically updated in the test site information. If a report is generated from a test done prior to the description change the old description will be inserted. |
| Shows the exact version of the DLL in which the device driver is contained. This is a unique identifier that specifies the version of the device driver. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the date and time of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the version number of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the hardware version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the internally used ID of the device. This information can be changed/viewed in this cell and can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction. |
| Shows the serial number of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the software version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the type of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Additional options[edit]
| The ‘’’ ‘Add’ button’’’ can be used to add additional information items to the general device driver settings. The additional items will be shown in the list. Additional information items can be used to note the storage location, the location of the user manual, etc. |
| The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited. |
| The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed. |
| When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can significantly reduce the time that is needed to initialize each test.
| This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, RadiMation® will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date. |
| The ‘’’ ‘Advanced’ button’’’ can be used to open the advanced menu of the device driver. In chapter 15, the paragraph ‘Device specific configuration’ will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view the paragraphs about ‘Configurable devices’ or ‘Virtual devices’ (also in chapter 15) for these devices. |
| The ‘’’ ‘Check’ button’’’ can be used to check if the device is connected. This check is the same as the check that is performed during each test initialization. This check can be used to determine if the correct addressing is set, and whether communication with the device is possible. (This option is explained further below in the paragraph ‘Checking the device connection’.) |
| The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the RadiWiki page about that device driver. That RadiWiki page can contain additional information or a detailed explanation of all the advanced settings. |
Now we can conclude that:
4. Graphs configuration[edit]
Select
Start
Settings
Control Panel
ODBC Data Sources
| This is just a name for the setup you are making, name this Radimation, or another descriptive name for the used database |
| Allows to select the directory in which the text based database files are stored |
| Overrules the selected directory to use the current directory. It is best to turn this checkbox off, and select a specific directory by using the Select Directory... button |
| Shows a list of filename extensions which will be used to determine the correct text files. Remove *.asc and *.tab from the extentions list |
| Will show the Define Text Format dialog, which can be used to define the format of the database as it is maintained in the text file |
Configuration of the characterset[edit]
| Characterset | value |
| UTF-8 (without bom) | CharacterSet=65001 |
| UTF-16 LE (without bom) | CharacterSet=Unicode |
MySQL ODBC database[edit]
| A unique name that uniquely identifies the Data Source from the other available Data Sources. |
| Any additional information regarding the Data Source |
| The name (or IP address) of the MySQL server, to which a connection should be made |
| The port-number to which the which the connection should be made. By default this is port-number 3306 |
| The name of the MySQL user that should be used to connect to the MySQL Server |
| The password of the MySQL user that should be used to connect to the MySQL Server |
| The name of the MySQL database that should be connected to |
| Will perform a connection test to the specified MySQL server and Database |
| Will show additional configuration details for the MySQL connection. Your MySQL database administrator can help you with setting this details. |
MSSQL ODBC database[edit]
| The name (or IP address) of the MSSQL server, to which a connection should be made |
| Select this option to use the credential of the logged in user to connect to the MSSQL server. |
| Select this option to use a SQL Server specific username and password to connect to the MSSQL server. |
| Will show additional configuration details for the MSSQL connection. Your MSSQL database administrator can help you with setting this details. |
| The name of the MSSQL user that should be used to connect to the MSSQL Server |
| The password of the MSSQL user that should be used to connect to the MSSQL Server |
| Enable the option, and specify or select the name of the database on the MSSQL Server that should be used. |
| Will perform a connection test, to determine if a connection with the MSSQL server can be established, using the configured parameters. A confirmation or error dialog with some additional information will be shown. |
| Stores the configured ODBC connection. |
| Closes the dialogs without creating an ODBC connection. |
6. Language Selection[edit]
7. Measurement Settings[edit]
| Warning: | Disabling the 'zoom-in in emission tests' in combination with analysers will very likely result in an incorrect measurement. |
8. Product Standard[edit]
| Note: | When a product standard is selected or added in the 'EUT' window, the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the 'Configuration' window of RadiMation® is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested. |
9. Basic Standard[edit]
Correction files[edit]
File
New
Correction file
Entering data points[edit]
9 29.69
10 29.68
etc.
9 29.69 10 29.68
Saving correction files[edit]
File
Save correction file
configuration
Correction file uses[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
Or
- Frequency
- Power
- Frequency
- Correction
Or
- Frequency
- Power
- Frequency
- Correction
Or
- Frequency
- Power
- Angle
- Attenuation
- Frequency
- Correction
Antenna: Antenna Factor[edit]
- Frequency
- Correction
Where the correction column is the antenna factor expressed in dB/m. The values are added to the calculation to account for the non-linearity in the antenna.
As RadiMation® only has a correction file for the antenna factor that contains the frequency and actual antenna factor value, it is not known for RadiMation® at which distance this antenna factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct antenna factor is specified, corresponding to the antenna distance that is actually used during the measurement.
Antenna: Antenna Gain[edit]
- Frequency
- Antenna gain
- Height
- Frequency
- Correction
During emission measurements when the measured emission values have to be shown in magnetic field (Tesla or A/m) units, the correction factor should be configured in the software. A correction file with the correction factor should be attached to the device driver of the antenna. The correction file should contain the columns:
- Frequency
- Correction
The value specified in the correction column should be expressed in dBpT/uV. The values in the correction column are added to the measured emission levels from the measurement receiver to calculate the correct to be displayed emission values in a Tesla or A/m unit.
As RadiMation® only has a correction file for the magnetic field factor that contains the frequency and actual factor value, it is not known for RadiMation® at which distance this magnetic field factor has been calibrated. Recalculating it to another distance is thus not possible as the original calibration distance is not known. It thus is important that the correct magnetic field factor is specified, corresponding to the antenna distance that is actually used during the measurement.
- Frequency
- Correction
Cable: Cable loss[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
This implies that:
- Frequency
Or
- Frequency
Or
- Frequency
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
Or
- Frequency
- Frequency
or
- Frequency
- Frequency
or
- Frequency
- Frequency
- Correction
- Frequency
- Correction
or
- Frequency
- Attenuation
or
- Frequency
- Gain
- Frequency
or
- Frequency
or
- Frequency
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction: The values in the correction column should be expressed in dB. The value specified in the correction column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.3 dB more than is specified on its display, can be corrected with a value of: -0.3.
or
- Frequency
- Gain: The values in the gain column should be expressed in dB. The value specified in the gain column is subtracted from the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.2 dB more than is specified on its display, can be corrected with a value of: 0.2.
or
- Frequency
- Attenuation: The values in the attenuation column should be expressed in dB. The value specified in the attenuation column is added to the desired carrier level and then send to the signal generator. For example: a signal generator that outputs 0.4 dB less than is specified on its display, can be corrected with a value of: 0.4.