(→Duplicate a configuration folder) |
(→Modifying a path for a filetype) |
Line 404: | Line 404: | ||
It is possible to duplicate an existing configuration folder, click '''Edit''' and either choose '''Duplicate and set from current global folder''' or '''Duplicate and set from current standalone folder'''. After selecting which folder to duplicate a directory chooser will be prompted to create and / or select a new empty folder to copy the configuration to. | It is possible to duplicate an existing configuration folder, click '''Edit''' and either choose '''Duplicate and set from current global folder''' or '''Duplicate and set from current standalone folder'''. After selecting which folder to duplicate a directory chooser will be prompted to create and / or select a new empty folder to copy the configuration to. | ||
− | ==== | + | ==== Change the path for a file type ==== |
Select the directory that you wish to change and click on the ''' 'Folder' icon''' to select a different directory. | Select the directory that you wish to change and click on the ''' 'Folder' icon''' to select a different directory. |
Revision as of 09:36, 13 May 2020
Engineers
An unlimited number of Engineers can be added, each with an unique name. For each engineer different permissions can be assigned which give them the right to modify specific parts of the RadiMation® configuration.
There is no relation between the Engineer in RadiMation® and the Windows Logon account. Also the password that can be configured for the engineer in RadiMation® is not linked and also not synchronized with the Windows User account password.
For each test, the name and password settings of the test engineer performing the current test can be selected. Also during the startup of RadiMation® the actual Engineer that is going to start tests can be selected. Switching between the active engineers while RadiMation® is running can be done through the menu:
To add, delete or edit the names and permissions of the test engineers, go to the Engineers window, which can be accessed in the menu:
Test engineer settings
Adding an engineer
Deleting an engineer
Editing engineer settings
Password protection
Rights
Change password
Test Equipment
Adding test equipment
Deleting test equipment
Editing test equipment
Devices 1 & 2
Adding devices
Editing devices
Removing devices
Saving test equipment
Save As
Before Action
Example
After Action
Default address information
- The default client
- The default manufacturer
- The default test site
Advanced options
- Units
- Directories
- Device Drivers
- Graphs
- Database
- Language
- Measurement settings
- Product standards
- Basic standards
- Enhanced Status Window
1. Units configuration
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Changing unit settings
2. Directories configuration
Open the 'Directories' configuration screen
Go to the 'Configuration' window in the 'Configuration' tab.
Go to the 'Directories' tab
Modifying the 'Device Driver Configuration Files' directory
Use global configuration and Use standalone configuration
When using the Use standalone configuration option, it is possible to create an isolated configuration directory with it's own content only used for this version of RadiMation.
Create a new 'Device Driver Configuration Files' folder
When it's needed to create a new Device Driver Configuration Files folder, click Edit and select Create and set new folder. A directory chooser will be prompted to create and / or select a new empty folder.
Duplicate a 'Device Driver Configuration Files' folder
It is possible to duplicate an existing configuration folder, click Edit and either choose Duplicate and set from current global folder or Duplicate and set from current standalone folder. After selecting which folder to duplicate a directory chooser will be prompted to create and / or select a new empty folder to copy the configuration to.
Select the directory that you wish to change and click on the 'Folder' icon to select a different directory.
To confirm your change, click 'Select folder' , or to cancel your change, click 'Cancel' . You will be returned to the 'Configuration' window.
3. Device Driver configuration
2. Through the 'Devices' tab.
Select a device driver type
Adding a new device driver
Editing a device driver
Removing a device driver
Confirm settings
Fixed information
Shows the brand of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the description of the device. This information can be changed/viewed in this cell and can be used in the report generator. Changing the description does not result in losing the device in the test sites. The description is automatically updated in the test site information. If a report is generated from a test done prior to the description change the old description will be inserted. |
Shows the exact version of the DLL in which the device driver is contained. This is a unique identifier that specifies the version of the device driver. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Shows the date and time of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Shows the version number of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
Shows the hardware version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the internally used ID of the device. This information can be changed/viewed in this cell and can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction. |
Shows the serial number of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the software version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Shows the type of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Additional options
The ‘’’ ‘Add’ button’’’ can be used to add additional information items to the general device driver settings. The additional items will be shown in the list. Additional information items can be used to note the storage location, the location of the user manual, etc. |
The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited. |
The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed. |
When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can be a significantly reduce the time that is needed to initialize each test.
This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, RadiMation® will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date. |
The ‘’’ ‘Advanced’ button’’’ can be used to open the advanced menu of the device driver. In chapter 15, the paragraph ‘Device specific configuration’ will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view the paragraphs about ‘Configurable devices’ or ‘Virtual devices’ (also in chapter 15) for these devices. |
The ‘’’ ‘Check’ button’’’ can be used to check if the device is connected. This check is the same as the check that is performed during each test initialization. This check can be used to determine if the correct addressing is set, and whether communication with the device is possible. (This option is explained further below in the paragraph ‘Checking the device connection’.) |
The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the RadiWiki page about that device driver. That RadiWiki page can contain additional information or a detailed explanation of all the advanced settings. |
Now we can conclude that:
4. Graphs configuration
5. Database configuration
Introduction
Select
This just a name for the setup you are making, name this Radimation, or another descriptive name for the used database |
Allows to select the directory in which the text based database files are stored |
Overrules the selected directory to use the current directory. It is best to turn this checkbox off, and select a specific directory by using the Select Directory... button |
Shows a list of filename extensions which will be used to determine the correct text files. Remove *.asc and *.tab from the extentions list |
Will show the Define Text Format dialog, which can be used to define the format of the database as it is maintained in the text file |
MySQL ODBC database
An unique name that uniquely identifies the Data Source from the other available Data Sources. |
The name (or IP address) of the MySQL server, to which a connection should be made |
The port-number to which the which the connection should be made. By default this is port-number 3306 |
The name of the MySQL user that should be used to connect to the MySQL Server |
The password of the MySQL user that should be used to connect to the MySQL Server |
The name of the MySQL database that should be connected to |
Will perform a connection test to the specified MySQL server and Database |
Will show additional configuration details for the MySQL connection. Your MySQL database administrator can help you with setting this details. |
MSSQL ODBC database
Enable the option, and specify or select the name of the database on the MSSQL Server that should be used. |
Will perform a connection test, to determine if a connection with the MSSQL server can be established, using the configured parameters. A confirmation or error dialog with some additional information will be shown. |
Stores the configured ODBC connection. |
Closes the dialogs without creating an ODBC connection. |
6. Language Selection
7. Measurement Settings
Calibration
Required environment settings
8. Product Standard
Introduction
Adding a Product Standard
Add(ing) Basic Standard(s)
9. Basic Standard
Correction files
Creating correction files
Adding and removing columns
Entering data points
9 29.69
10 29.68
etc.
9 29.69 10 29.68
Saving correction files
Attaching correction files
Correction file uses
Absorbing clamp: Clamp loss
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Amplifier: Maximum Input
- Frequency
- Correction
Or
- Frequency
- Power
Amplifier: Maximum Forward
- Frequency
- Correction
Or
- Frequency
- Power
Amplifier: Maximum Reflected
- Frequency
- Correction
Or
- Frequency
- Power
Antenna: Antenna Factor
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction
Antenna: GpsHt - AntHeight
- Height
Antenna: Ant Gain vs Angle
- Angle
- Attenuation
Cable: Cable loss
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
This implies that:
Coupler: Forward Correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Coupler: Reflected Correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Current sensor: impedance
- Frequency
- Correction
Or
- Frequency
- Resistance
The values in the correction column should be expressed in Ohm (Ω).
Field sensor: Field offset
- Frequency
- Correction
The values in the correction column of the offset correction should be expressed in V/m.
To summarize this in an example, let’s say that the RAW field probe measurement at 90 MHz is 10.2 V/m. When a Field offset correction for that same frequency of 2.0 V/m is specified, RadiMation will use a corrected field strength of 12.2 V/m (10.2 V/m + 2.0 V/m).
Field sensor: Field Multiply
- Frequency
- Correction
The values in the correction column of the multiply factor should be expressed as x (multiplier).
To summarize this in an example, let’s say that the RAW field probe measurement at 90 MHz is 10.2 V/m. When a Field multiply correction for that same frequency of 1.1 is specified, RadiMation will use a corrected field strength of 11.22 V/m (10.2 V/m * 1.1).
LISN: Amplitude correction
- Frequency
- Correction
Powermeter: Power correction
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
Pre amplifier: Amplifier Gain
- 'Frequency' and 'Correction' column: The values in the 'Correction' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
- 'Frequency' and 'Gain' column: The values in the 'Gain' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
- 'Frequency' and 'Attenuation' column: The values in the 'Attenuation' column are interpreted as 'attenuation' values. A positive value is interpreted as a loss caused by the preamplifier. A negative value is interpreted as a positive amplification of the signal, caused by the preamplifier. These 'attenuation' values will be added to the values that are measured with the receiver/analyser.
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction
