| Line 1,130: | Line 1,130: | ||
* X-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the x-axis reading only. | * X-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the x-axis reading only. | ||
* Y-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the | * Y-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the y-axis reading only. | ||
* Z-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the | * Z-axis correction. The values in the column used for the multiply correction should be expressed as x (multiplier). This correction is applied on the z-axis reading only. | ||
When the correction is specified for the individual axis, those axis are first corrected, after which the isotropic value is calculated based on the individually corrected axes. | When the correction is specified for the individual axis, those axis are first corrected, after which the isotropic value is calculated based on the individually corrected axes. | ||
Revision as of 08:15, 24 August 2023
| Warning: | To prevent unauthorised individuals from changing the global RadiMation® configurations, a password protection can be used for specific engineers. The password is activated by default for the "Administrator" engineer with the password "Radimation". This password protection can be switched off, changed and configured in the menu
Engineers[edit]
An unlimited number of Engineers can be added, each with an unique name. For each engineer different permissions can be assigned which give them the right to modify specific parts of the RadiMation® configuration.
There is no relation between the Engineer in RadiMation® and the Windows Logon account. Also the password that can be configured for the engineer in RadiMation® is not linked and also not synchronized with the Windows User account password.
For each test, the name and password settings of the test engineer performing the current test can be selected. Also during the startup of RadiMation® the actual Engineer that is going to start tests can be selected. Switching between the active engineers while RadiMation® is running can be done through the menu:
File
Change Engineer
To add, delete or edit the names and permissions of the test engineers, go to the Engineers window, which can be accessed in the menu:
Engineers
Test engineer settings[edit]
Adding an engineer[edit]
Deleting an engineer[edit]
Password protection[edit]
Rights[edit]
Change password[edit]
Test Equipment[edit]
Adding test equipment[edit]
Deleting test equipment[edit]
Editing test equipment[edit]
Devices 1 & 2[edit]
Adding devices[edit]
Editing devices[edit]
Removing devices[edit]
Saving test equipment[edit]
Save As[edit]
Before Action[edit]
Example[edit]
After Action[edit]
- The default client
- The default manufacturer
- The default test site
Advanced options[edit]
- Units
- Directories
- Device Drivers
- Graphs
- Database
- Language
- Measurement settings
- Product standards
- Basic standards
- Enhanced Status Window
1. Units configuration[edit]
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Changing unit settings[edit]
Go to the Directories tab.
2. Through the 'Devices' tab.
Editing a device driver[edit]
Device Driver Settings[edit]
Confirm settings[edit]
Fixed information[edit]
| Shows the brand of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the description of the device. This information can be changed/viewed in this cell and can be used in the report generator. Changing the description does not result in losing the device in the test sites. The description is automatically updated in the test site information. If a report is generated from a test done prior to the description change the old description will be inserted. |
| Shows the exact version of the DLL in which the device driver is contained. This is a unique identifier that specifies the version of the device driver. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the date and time of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the version number of the last update/installation of the device drivers. The contents of this cell cannot be changed. (More information on this data cell is explained further below in the paragraph ‘Storing the driver version in the test file’.) |
| Shows the hardware version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the internally used ID of the device. This information can be changed/viewed in this cell and can be used in the report generator. The main function of this window is to link your measurement to your devices for easier reproduction. |
| Shows the serial number of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the software version of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
| Shows the type of the device. This information can be changed/viewed in this cell and can be used in the report generator. |
Additional options[edit]
| The ‘’’ ‘Add’ button’’’ can be used to add additional information items to the general device driver settings. The additional items will be shown in the list. Additional information items can be used to note the storage location, the location of the user manual, etc. |
| The ‘’’ ‘Edit’ button’’’ can be used to edit the information of a selected item. Information items that cannot be modified cannot be edited. |
| The ‘’’ ‘Remove’ button’’’ can be used to remove a selected information item from the list. Only additional information items can be removed. Fixed information items, as described previously, cannot be removed. |
| When the Zero interval is activated, it is possible to specify a time period in minutes, in which a device does not need to be ‘re-zero-ed’. For accurate measurements it is necessary that Power Meters and Field Sensors are zero-ed. A zero operation generally takes several seconds to be completed and it may be valid for a few hours. To prevent that a device is ‘re-zero-ed’ during each test initialization, it is possible to set a zero interval in minutes. When a test is initialized and the device is already zero-ed within the specified zero interval, no zero will be performed at that time. Specifying and activating the zero interval can be a significantly reduce the time that is needed to initialize each test.
| This field can be used to set the date on which the device needs to be re-calibrated. Depending upon the setting in the configuration window, RadiMation® will use this date to give warnings/notifications on the recalibration. When you press the ‘calendar icon’ to the right of this field a small window will appear, allowing you to select the new date. |
| The ‘’’ ‘Advanced’ button’’’ can be used to open the advanced menu of the device driver. In chapter 15, the paragraph ‘Device specific configuration’ will explain this menu for every device type. Please note that the configuration does not apply on the configurable or virtual drivers. Please view the paragraphs about ‘Configurable devices’ or ‘Virtual devices’ (also in chapter 15) for these devices. |
| The ‘’’ ‘Check’ button’’’ can be used to check if the device is connected. This check is the same as the check that is performed during each test initialization. This check can be used to determine if the correct addressing is set, and whether communication with the device is possible. (This option is explained further below in the paragraph ‘Checking the device connection’.) |
| The ‘’’ ‘Knowledgebase’ button’’’ can be used to open a new web-browser window which will show the RadiWiki page about that device driver. That RadiWiki page can contain additional information or a detailed explanation of all the advanced settings. |
Now we can conclude that:
4. Graphs configuration[edit]
Select
Start
Settings
Control Panel
ODBC Data Sources
| This is just a name for the setup you are making, name this Radimation, or another descriptive name for the used database |
| Allows to select the directory in which the text based database files are stored |
| Overrules the selected directory to use the current directory. It is best to turn this checkbox off, and select a specific directory by using the Select Directory... button |
| Shows a list of filename extensions which will be used to determine the correct text files. Remove *.asc and *.tab from the extentions list |
| Will show the Define Text Format dialog, which can be used to define the format of the database as it is maintained in the text file |
Configuration of the characterset[edit]
| Characterset | value |
| UTF-8 (without bom) | CharacterSet=65001 |
| UTF-16 LE (without bom) | CharacterSet=Unicode |
MySQL ODBC database[edit]
| A unique name that uniquely identifies the Data Source from the other available Data Sources. |
| Any additional information regarding the Data Source |
| The name (or IP address) of the MySQL server, to which a connection should be made |
| The port-number to which the which the connection should be made. By default this is port-number 3306 |
| The name of the MySQL user that should be used to connect to the MySQL Server |
| The password of the MySQL user that should be used to connect to the MySQL Server |
| The name of the MySQL database that should be connected to |
| Will perform a connection test to the specified MySQL server and Database |
| Will show additional configuration details for the MySQL connection. Your MySQL database administrator can help you with setting this details. |
MSSQL ODBC database[edit]
| The name (or IP address) of the MSSQL server, to which a connection should be made |
| Select this option to use the credential of the logged in user to connect to the MSSQL server. |
| Select this option to use a SQL Server specific username and password to connect to the MSSQL server. |
| Will show additional configuration details for the MSSQL connection. Your MSSQL database administrator can help you with setting this details. |
| The name of the MSSQL user that should be used to connect to the MSSQL Server |
| The password of the MSSQL user that should be used to connect to the MSSQL Server |
| Enable the option, and specify or select the name of the database on the MSSQL Server that should be used. |
| Will perform a connection test, to determine if a connection with the MSSQL server can be established, using the configured parameters. A confirmation or error dialog with some additional information will be shown. |
| Stores the configured ODBC connection. |
| Closes the dialogs without creating an ODBC connection. |
6. Language Selection[edit]
7. Measurement Settings[edit]
| Warning: | Disabling the 'zoom-in in emission tests' in combination with analysers will very likely result in an incorrect measurement. |
8. Product Standard[edit]
| Note: | When a product standard is selected or added in the 'EUT' window, the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the 'Configuration' window of RadiMation® is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested. |
9. Basic Standard[edit]
Correction files[edit]
File
New
Correction file
Entering data points[edit]
9 29.69
10 29.68
etc.
9 29.69 10 29.68
Saving correction files[edit]
File
Save correction file
configuration
Correction file uses[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
Or
- Frequency
- Power
- Frequency
- Correction
Or
- Frequency
- Power
- Frequency
- Correction
Or
- Frequency
- Power
Antenna: Antenna Factor[edit]
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction
Antenna: Antenna Gain[edit]
- Frequency
- Antenna gain
- Height
- Angle
- Attenuation
Cable: Cable loss[edit]
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
- Correction
This implies that:
- Frequency
Or
- Frequency
Or
- Frequency
- Frequency
- Correction
Or
- Frequency
- Attenuation
Or
- Frequency
- Gain
- Frequency
Or
- Frequency
- Frequency
or
- Frequency
- Frequency
or
- Frequency
- Frequency
- Correction
- Frequency
- Correction
or
- Frequency
- Attenuation
or
- Frequency
- Gain
- 'Frequency' and 'Correction' column: The values in the 'Correction' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
- 'Frequency' and 'Gain' column: The values in the 'Gain' column are interpreted as 'gain' values. A positive value is interpreted as a positive amplification of the preamplifier. These 'gain' values will be substracted from the values that are measured with the receiver/analyser.
- 'Frequency' and 'Attenuation' column: The values in the 'Attenuation' column are interpreted as 'attenuation' values. A positive value is interpreted as a loss caused by the preamplifier. A negative value is interpreted as a positive amplification of the signal, caused by the preamplifier. These 'attenuation' values will be added to the values that are measured with the receiver/analyser.
- Frequency
- Correction
- Frequency
- Correction
- Frequency
- Correction