Chapter 2 “Getting Started” provides information relevant to first time use of {{RadiMation}}. This includes (but is not limited to) a product overview and installation guide, as well as the configuration of some basic settings (such as passwords, users, languages and test parameters). | |||
For more (detailed) information, follow the links and references in the text, or go straight to the chapter related to your topic of interest. | |||
Line 21: | Line 25: | ||
* [[Radiated immunity]] testing (RI) | |||
* [[Conducted immunity]] testing (CI) | |||
* [[Pulsed immunity]] (ESD, EFT, Surge) | |||
* [[Radiated emission]] testing (RE) | |||
* [[Conducted emission]] testing (CE) | |||
* [[Report generator]] | |||
* [[Excel Exporter]] | |||
Line 35: | Line 39: | ||
* Calibration functionality enables an easy exchange of the calibration files. | * Calibration functionality enables an easy exchange of the calibration files. | ||
* Different types of tests can be performed, fully automatic, after each other (Sequence testing). | * Different types of tests can be performed, fully automatic, after each other (Sequence testing). | ||
== Conventions == | |||
{{RadiMation}} uses the Windows structure of pull down menu’s. In this manual the pull down menu’s are visualized as followed: | |||
{{Menu|File|Open|EUT}} | |||
The example above shows that, to show the file open selection window, the ''' 'File' ''' item from the menu bar should be selected first, followed by the ''' 'Open' ''' item and then the ''' 'EUT' ''' item. | |||
A screenshot of these menu items is visible below: | |||
[[Image:MenuFileOpenEUT.png]] | |||
Line 54: | Line 59: | ||
* Processor architecture: x64 (ARM64 is not supported) | |||
* Processor | * Processor clock speed: 2 GHz minimum, > 3 GHz recommended | ||
* Free RAM memory: | * Free RAM memory: 2 GB minimal, > 4 GB recommended | ||
* Monitor: | * Monitor resolution: Full HD (1920 x 1080) or larger | ||
{{ | {{RadiMation}} needs a considerable amount of memory to perform at its best. We recommend that you have at least 512 MB (Mega Bytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM ''without'' having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps are to help you determine the amount of free memory in your PC: | ||
Line 75: | Line 80: | ||
If a GPIB connection to one of the measurement equipments is used, the GPIB hardware and software needs to be installed correctly before {{RadiMation}} is installed. {{CompanyName}} is not responsible for any additional costs or delays that may occur if this card is not installed prior to the installation of {{RadiMation}}. | |||
The GPIB | The used GPIB card has to be supported by the [[National Instruments]] NI-488.2 software library. At least version 15.0 or higher of the NI-488.2 software should be used. {{RadiMation}} is able to use any version of [[National Instruments]] NI-488.2 which is 15.0 or higher, so it is not needed to update the NI-488.2 library as long as the installed version is supported by National Instruments and {{RadiMation}}. | ||
=== Software requirements === | |||
The {{RadiMation}} software is supported on all the Microsoft Windows, Windows Enterprise and Windows Server 2019 (or higher) operating systems that are still supported by [[Microsoft]]. The IoT variant of these operating systems are not supported. If a specific version of Microsoft Windows is still supported, can be checked on [https://docs.microsoft.com/en-us/lifecycle/products/?products=windows Microsoft Lifecycle Policy pages]. Both the 32-bit and 64-bit versions of these operating systems are supported. In order to run RadiMation it is required that the used operating system is up to date. It can result in severe delays during an installation in the case that Windows Updates are not completely updated and installed. | |||
=== | ==== Microsoft .NET framework ==== | ||
The | The [[Microsoft]] [https://dotnet.microsoft.com/en-us/download/dotnet-framework/net48 .NET Framework 4.8] is also required, on Windows 10 1903 and higher this is installed by default. For Windows 10 October 2018 Update (Version 1809) the .NET framework 4.8 needs to be installed manually before installing {{RadiMation}} | ||
It is required to have a supported, licensed and working version of Microsoft Office Word installed on the PC. This should be at least Microsoft Word 2016 or higher, as Microsoft Word 2013 and lower has no mainstream support anymore from Microsoft. {{RadiMation}} uses Microsoft Office Word to generate and print the printouts, besides creating the reports. | |||
The {{RadiMation}} report | The {{RadiMation}} [[Report Generator|report generater]] can create reports in the following formats: | ||
* Microsoft Word | * [[Microsoft Word]] | ||
* Microsoft Excel | * [[Microsoft Excel]] | ||
During the installation of {{RadiMation}} a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. | During the installation of {{RadiMation}} a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. {{CompanyName}} is not responsible for any additional costs or delays that may occurs if this support is not provided. | ||
* Full rights to the installation directory (e.g. 'C:\Program Files\ | * Full rights to the installation directory (e.g. 'C:\Program Files\Raditeq', please note that the path may be different depending on the used OS and language) | ||
== Software installation == | |||
To install the {{RadiMation}} software, please | To install the {{RadiMation}} software, please download the {{RadiMation}} setup program from our download website at: https://www.raditeq.com/radimation-download/ . Once the download has been finished, the setup program can be started. The Installation Wizard will guide you through the installation process. After installation the computer must be restarted before running {{RadiMation}}. Please read [[#User rights and installation|User rights and installation]] about the user rights and information that is required for first time installation. | ||
In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages | In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages compared to a local installation. One of these benefits is that data can be shared between the connected computers. The {{RadiMation}} software is designed to optimally use this advantage. All the data that is used or generated by {{RadiMation}} can also be shared on the network. In most cases we recommend the use a of network configuration over a local configuration. | ||
Line 136: | Line 130: | ||
# Start the installed {{RadiMation}} program and open the Configuration window. On the {{ScreenElement|Directories}} tab, change the location of the {{ScreenElement|Device Driver Configuration}} directory to the CONFDVDR directory in the <NetworkDir>. ( | # Start the installed {{RadiMation}} program and open the Configuration window. On the {{ScreenElement|Directories}} tab, change the location of the {{ScreenElement|Device Driver Configuration}} directory to the CONFDVDR directory in the <NetworkDir>. (Note: do not change the other directories yet.) | ||
# Restart the {{RadiMation}} program. | # Restart the {{RadiMation}} program. | ||
# Open the Configuration window again and on the {{ScreenElement|Directories}} tab, change the location of all the other directories to the corresponding directories in the <NetworkDir>. | # Open the Configuration window again and on the {{ScreenElement|Directories}} tab, change the location of all the other directories to the corresponding directories in the <NetworkDir>. | ||
Line 142: | Line 136: | ||
Line 152: | Line 145: | ||
=== Automated / Silent installation === | |||
{{RadiMation}} is not provided as a .MSI file that can be used by network administrators to perform an automated installation on a lot of PC's. However {{RadiMation}} is provided as an .EXE file that does accept additional command line arguments that can be used to perform an automated silent installation. | |||
The following arguments can be specified to the {{RadiMation}} setup .EXE program: | |||
{|class="wikitable" | |||
!Argument || Purpose and explanation | |||
|/SP- || Disables the 'This will install ... Do you wish to continue?' prompt at the beginning of the setup | |||
|/SILENT || Instructs the setup program to be silent | |||
|/VERYSILENT || Instructs the setup program to be very silent | |||
|/SUPPRESSMSGBOXES || Instructs the setup to suppress message boxes | |||
|/LOG || Causes setup to create a log file in the user's TEMP directory | |||
|/LOG="filename" || Same as /LOG, except it allows to specify a fixed path/filename to use for the log file. | |||
|/NOCANCEL || Prevents the user from cancelling during the installation process. | |||
|/NORESTART || Prevents Setup from restarting the system following a successful installation, or after a Preparing to Install failure that requests a restart. | |||
|/DIR="x:\dirname" || Overrides the default directory name | |||
|/HELP || Shows a detailed explanation of all the command line parameters | |||
The {{RadiMation}} SETUP.EXE also contains the [[DRIVERS.EXE]] that was actual at the moment that the SETUP.EXE was build. That embedded [[DRIVERS.EXE]] will automatically be installed during the installation of the {{RadiMation}} version. If in the same directory in which the SETUP.EXE is located, also an 'AUTOINSTALLDRIVERS.EXE' is present, then that 'AUTOINSTALLDRIVERS.EXE' will be installed instead of the embedded [[DRIVERS.EXE]]. The 'AUTOINSTALLDRIVERS.EXE' can be created by just renaming the desired [[DRIVERS.EXE]] to 'AUTOINSTALLDRIVERS.EXE'. | |||
Software updates | Software updates can be downloaded from the [https://www.raditeq.com/radimation-download/ {{RadiMation}} Download] page. These updates are supplied as {{RadiMation}} versions with a higher version number. | ||
Once the download of the new version has been finished, the setup program can be started. The setup program will install the new {{RadiMation}} version at the same location as the previous older, but in a separate installation folder. This will allow you to use the previously installed {{RadiMation}} version, as well as the new one. | |||
==== Installing from a network directory ==== | |||
If the already installed {{RadiMation}} version was installed from a network directory, the new setup file can be copied to the <Setup> sub-directory of the <NetworkDir>. | |||
{{ | The new {{RadiMation}} version can then be installed on each computer from that <Setup> sub-directory. This installation still has to be started manually, {{RadiMation}} will not automatically detect that a newer version is available on the network directory, and it will not automatically start the installation of the newer version. | ||
==== Updating device drivers ==== | |||
After the new {{RadiMation}} version is installed, we strongly recommend that you install the latest [[device drivers]] as well. These device drivers can be installed by starting the ‘[[DRIVERS.EXE]]’ program that can be downloaded from our [https://wiki.dare.nl/wiki/index.php/Device_drivers RadiWiki website]. | |||
The [[DRIVERS.EXE]] program will detect the different {{RadiMation}} versions installed on your PC. A dialog will start, asking you to select the {{RadiMation}} version that needs updated device drivers. Select the latest version of {{RadiMation}} that was just installed. | |||
You have successfully obtained the latest released version of {{RadiMation}} when both the software and device drivers have been updated. | |||
{{Note|We strongly advice that a you make a complete backup of all your files and directories (e.g. TSF/Sequence/Correction/Calibration and EUT) before installing a {{RadiMation}} update.}} | |||
{{Warning| '''Do ''NOT'' uninstall {{RadiMation}}!!''' This will cause removal of all user settings, TSF, sequence, calibration and correction files. This is unrecoverable.}} | |||
For a new installation or an update, different user rights can be setup. | |||
=== Update === | |||
To update the {{RadiMation}} software, the user must have the rights to read and write in the {{RadiMation}} folder, which is normally located in the C:\Program Files (x86)\Raditeq\ directory. | |||
=== Daily usage === | |||
For daily usage of the {{RadiMation}} software, no specific rights are needed. Every standard end-user should be able to start and use the already installed {{RadiMation}} software. | |||
== | == Software protection key == | ||
[[Image:Dongle-keyhanger.jpg|thumb|Software protection key]] | |||
The {{RadiMation}} software is copy protected with an USB [[Software protection key]]. The USB [[Software protection key]] should be connected to the PC while a {{RadiMation}} test is running, or when the report generator or Excel Exporter is used. The USB Software protection key is not needed to open and review the test results. | |||
A network license or a floating license of the USB [[Software protection key]] is not available, and each {{RadiMation}} test PC should thus have it's own USB [[Software protection key]] physically connected to the PC. | |||
Line 177: | Line 217: | ||
The next paragraphs will explain the required configurations. | The next paragraphs will explain the required configurations. | ||
{{RadiMation}} allows | {{RadiMation}} allows to protect a number of configurations by assigning applicable rights to an engineer, where also a login password can be configured for those engineers. | ||
* The password protection is by default turned on for the default engineer with the name: 'Administrator'. | |||
* The default password is '''Radimation'''. | |||
* Configuring and using additional engineers with different passwords is optional. | |||
* The configured engineers are stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured using the menu: | |||
{{Menu|Configuration|Configuration|Directories}} | |||
To change the password protection settings, go to the Engineers window in the Configuration tab. | To change the password protection settings, go to the {{ScreenElement|Engineers}} window in the {{ScreenElement|Configuration}} tab. | ||
Select the {{ScreenElement|Administrator}} engineer and click {{ScreenElement|Edit}}. | |||
[[Image:Engineers admin.png]] | |||
In the {{ScreenElement|Test Engineer}} window for the 'Administrator' user, the password can be enabled, disabled or modified. | |||
The current password setting for the selected engineer is displayed at the bottom of the window. This will indicate whether a password has been ‘set’ or ‘not set’. | |||
In addition, the protection can be enabled or disabled for the following items | In addition, the protection can be enabled or disabled for the following items: | ||
Line 206: | Line 251: | ||
To enable, disable or change the password, click | To enable, disable or change the password, click {{ScreenElement|[[Change Password]]}}. | ||
To change the password, enter the old password and the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | |||
To change the password, enter the old password and the new password (twice), then click | |||
To disable the password, enter the old password and leave the other two fields empty, then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | |||
(The password indication will change to: "Warning: Password is not set for this engineer.") | |||
To enable the password, enter the new password (twice), then click | To enable the password, enter the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | ||
(The password indication will change to: "Password is set for this engineer.") | |||
To leave the {{ScreenElement|Engineers}} window, click {{ScreenElement|Ok}}. | |||
To leave the Engineers window, click | |||
Line 234: | Line 279: | ||
To add, delete or edit (the names of) test engineers, go to the Engineers window in the Configuration tab. | To add, delete or edit (the names of) test engineers, go to the ''' 'Engineers' window''' in the ''' 'Configuration' tab'''. | ||
The Engineers window will show the current configured list of engineers. To leave the Engineers window, click | The Engineers window will show the current configured list of engineers. To leave the ''' 'Engineers' window''', click ''' 'OK' '''. | ||
To add a new test engineer, click | To add a new test engineer, click ''' 'Add' '''. | ||
[[Image:Engineer add engineer.png]] | [[Image:Engineer add engineer.png]] | ||
Enter the name and set the rights of the new engineer, then click | Enter the name and set the rights of the new engineer, then click ''' 'OK' '''. You will be returned to the ''' 'Engineers' window'''. | ||
The new engineer will not have a protection password (yet), this can be enabled by clicking '[[Change Password|Change Password]]'. Please read [[#Password protection|Password protection]] for more information. | The new engineer will not have a protection password (yet), this can be enabled by clicking ''' '[[Change Password|Change Password]]' '''. Please read [[#Password protection|Password protection]] for more information. | ||
[[Image:Test engineer add engineer.png]] | [[Image:Test engineer add engineer.png]] | ||
To delete a test engineer, select the engineer and click | To delete a test engineer, select the engineer and click ''' 'Delete' '''. | ||
[[Image:Engineer delete engineer.png]] | [[Image:Engineer delete engineer.png]] | ||
To confirm that you want to delete the engineer from the configuration list, click | To confirm that you want to delete the engineer from the configuration list, click ''' 'Yes' '''. | ||
If you do not want to delete this engineer from the list, click ''' 'No' ''' or ''' 'Cancel' '''. You will be returned to the ''' 'Engineers' window'''. | |||
[[Image:Engineer delete engineer confirm.png]] | [[Image:Engineer delete engineer confirm.png]] | ||
To edit the name, rights and/or password of an engineer, select the engineer and click | To edit the name, rights and/or password of an engineer, select the engineer and click ''' 'Edit' '''. | ||
[[Image:Engineer edit engineer.png]] | [[Image:Engineer edit engineer.png]] | ||
Make the desired changes in the Test Engineer window and click | Make the desired changes in the ''' 'Test Engineer' window''' and click ''' 'OK' '''. | ||
If you want to undo or cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Engineers' window'''. | |||
[[Image:Test engineer edit engineer.png]] | [[Image:Test engineer edit engineer.png]] | ||
Line 274: | Line 323: | ||
When a new [[EUT]] file is created, the name and address information of the manufacturer, test house and client can be entered separately. | |||
The default address information is stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured using the menu: {{Menu|Configuration|Configuration|Directories}} | |||
Data entered under {{ScreenElement|Default Address Information}}, will automatically be copied and stored in the [[EUT]] file. Specifying this information is not required, however it has the benefit that address information that is always the same doesn't need to be entered over and over again in every new [[EUT]] file. | |||
To add, modify or delete the default address information, go to the {{ScreenElement|Default Address Information}} window in the {{ScreenElement|Configuration}} tab. | |||
[[Image:Config default address information.png]] | |||
Enter the desired information as shown in the example below. | |||
To confirm the data input, click {{ScreenElement|OK}}, to cancel the data input, click {{ScreenElement|Cancel}}. The {{ScreenElement|Default Address Information}} window will close. | |||
Line 302: | Line 351: | ||
To configure the unit settings, go to the Configuration window in the Configuration tab. | To configure the unit settings, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | ||
[[Image:Config configuration.png]] | [[Image:Config configuration.png]] | ||
The default settings for the different units can be set in the Units tab. It is possible to define different units for different tests. The following can be selected: | The default settings for the different units can be set in the ''' 'Units' tab'''. | ||
It is possible to define different units for different tests. The following can be selected: | |||
Line 318: | Line 369: | ||
To change the settings of a unit, choose the desired test (e.g. Radiated Immunity) from the drop down menu in the Units tab. | To change the settings of a unit, choose the desired test (e.g. Radiated Immunity) from the drop down menu in the ''' 'Units' tab'''. | ||
[[Image:Configuration units test selection.png]] | [[Image:Configuration units test selection.png]] | ||
Select the measurement (e.g. Power) and click | Select the measurement (e.g. Power) and click ''' 'Change'. | ||
[[Image:Configuration Units.png]] | [[Image:Configuration Units.png]] | ||
The {{WindowTitle|Unit configuration}} window will appear, allowing you to change the parameters of the unit. To confirm your changes, click | The ''' '{{WindowTitle|Unit configuration}}' window''' will appear, allowing you to change the parameters of the unit. | ||
To confirm your changes, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | |||
[[Image:Configuration units change2.png]] | [[Image:Configuration units change2.png]] | ||
Click | Click ''' 'Close' ''' to leave the ''' 'Configuration' window'''. | ||
The {{RadiMation}} software can use different directories to store different data file types. The location of the following file types can be specified: | |||
The {{RadiMation}} software can use different directories to store different data file types. | |||
Line 346: | Line 403: | ||
The different file types are described in [[Chapter 18|Chapter 18 “{{RadiMation}} file types and locations”]]. | |||
For each data type the default | |||
For each data type the default location can be entered through the configuration. The path name can link to a directory on a local or network drive. | |||
Selecting a network drive will allow you to store data (e.g. EUT data) on a computer network, making it more accessible. | |||
To modify | To modify a path name, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | ||
[[Image:Config configuration.png]] | |||
Go to the ''' 'Directories' tab''', select the directory that you wish to change and click ''' 'Change' '''. | |||
The ''' 'Browse For Folder' window''' will appear, allowing you to select (or make) a new folder in your directory. | |||
To confirm your change, click ''' 'OK' ''', to cancel your change, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | |||
Click ''' 'Close' ''' to leave the ''' 'Configuration' window'''. | |||
In {{RadiMation}}, the device drivers | Device drivers let your software know which equipment has to be used and controlled during EMC tests. | ||
In {{RadiMation}}, the device drivers only need to be configured to the customer requirements once. | |||
During the configuration of the device driver, information such as the IEEE address or RS232 COM-port has to be entered. | |||
You can configure the device drivers in the ''' 'Configuration' window'''. | |||
This window can be accessed in one of two ways, either through the ''' 'Configuration' tab''' or through the ''' 'Devices' tab'''. | |||
[[Image:Device configure.png]] | |||
To configure a device driver, go to the ''' 'Device Drivers' tab''' in the ''' 'Configuration' window'''. | |||
Select the desired ''' '[[Device Types|Device Driver Type]]' ''' from the drop-down menu (e.g. Signal Generator). The available drivers for that type will be shown. | |||
[[Image:Device driver type.png]] | |||
To create a new {{RadiMation}} device driver, select the desired ''' 'Device Driver Type' ''' (e.g. Signal Generator) and click ''' 'Add' '''. | |||
[[Image:Device driver add.png]] | |||
Select a new device driver (in the scroll menu or through the search function) and click ''' 'New' '''. | |||
To cancel your selection and return to the ''' 'Configuration' window''', click ''' 'Close' '''. | |||
[[Image:New signal generators.png]] | |||
The name/description of the chosen device driver can be changed in the ''' 'Device Driver Configuration' window'''. | |||
To confirm your changes (or the provided description), click ''' 'OK' ''', the ''' 'Device Driver Settings' window''' will appear. | |||
To cancel your changes, click ''' 'Cancel' '''. You will be returned to the previous window (e.g. New Signal generators). | |||
[[Image:Device driver configuration.png]] | |||
The ''' 'Device Driver Settings' window''' will appear with information on the selected device driver. In this window information can be added, edited and removed. | |||
To confirm the settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | |||
More detailed device driver settings are described in the paragraph "Device Driver settings" in [[Chapter 14]]. | |||
[[Image: | [[Image:Device driver settings.png]] | ||
To edit an existing device driver, select the desired (device driver) type and device driver, then click ''' 'Edit' '''. (The ''' 'Device Driver Settings' window''' will appear.) | |||
[[Image:Device driver edit.png]] | |||
To remove an existing device driver, select the desired (device driver) type and device driver, then click ''' 'Remove' '''. | |||
[[Image:Device | [[Image:Device driver remove.png]] | ||
A confirmation window will appear. To remove the selected device driver, click ''' 'Yes' ''', to cancel, click ''' 'No' '''. You will be returned to the ''' 'Configuration' window'''. | |||
[[Image:Device driver remove2.png]] | |||
{{RadiMation}} allows you to | {{RadiMation}} allows you to customize the way graphics are displayed on your screen. | ||
To configure the style of the shown graphs, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | |||
An example graph is shown in the ''' 'Graphs' tab''' that is created with the current settings. To change these graph settings, click ''' 'Customize' '''. | |||
The '{{WindowTitle|Graph Customization Configuration}}' '''window''' contains three tabs; ''' 'Chart' ''', ''' 'Graph' ''' and ''' 'Markings' '''. | |||
The ''' 'Chart' tab''' contains the settings related to the overall graph and can be used to determine the colours, layout and titles. | |||
The ''' 'Graph' tab''' contains the settings related to the graph lines shown in the graph and can be used to determine the colours, scaling and layout of each and every shown graph line. | |||
The ''' 'Markers' tab''' contains the settings related to the markers on the graph and can be used to determine the colours, style and scaling of each and every shown marker. | |||
To change the graph settings, select a graph (component) in the left section of the window, then make the desired changes in the right section of the window. | |||
{{ | Click ''' 'Update' ''' to update the graph settings without closing the '{{WindowTitle|Graph Customization Configuration}}' '''window'''. This will allow you to see the effect of the new settings and make additional changes. | ||
Click ''' 'Close' ''' to update (and save) the graph settings and close the '{{WindowTitle|Graph Customization Configuration}}' '''window'''. You will be returned to the ''' 'Configuration' window'''. | |||
The selected settings in the '{{WindowTitle|Graph Customization Configuration}}' '''window''' are used as the {{RadiMation}} default settings. It is still possible to overrule these settings for individual tests to more specific settings. | |||
Click ''' 'Restore factory default' ''' to return to the factory settings. | |||
Line 440: | Line 543: | ||
The default language of {{RadiMation}} is English. | {{RadiMation}} supports multi-language use. This means that the software package can be operated in several language. | ||
In addition, the multi- language support allows for easy translation from English to most other languages. | |||
Please contact your local reseller for the availability of your required language driver. | |||
The default language of {{RadiMation}} is English. | |||
To change the language settings, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | |||
Go to the ''' 'Language' tab'''. The current language setting is marked with the text '''(current)'''. | |||
To change the language setting, select one of the other languages, click ''' 'Close' ''' (to leave the ''' 'Configuration' window''') and restart {{RadiMation}}. | |||
The selected language will only be active after {{RadiMation}} is restarted. | |||
[[Image:Configuration language2.png]] | |||
Line 455: | Line 574: | ||
Product Standards describe | Product Standards describe the requirements (e.g. dimensional, technical, etc.) and the methods of testing (as well as grading, marking, etc.) of a product. | ||
The objective of these standards is to define and impose 'standards' that will ensure that the product in question will not represent harm or hazard to consumers. | |||
Because the standards with regards to product testing can be relevant to {{RadiMation}} users, {{RadiMation}} offers the possibility to add this documentation to your configuration. | |||
There are many Product Standards for individual products and product types. These, in turn, often refer to multiple Basic Standards. | |||
Where a Product Standard describes all the tests that should be performed on a EUT, a Basic Standard only describes one single test. | |||
For example: | |||
The Product Standard for product X describes four different requirements/tests that the product needs to comply with. | |||
These four tests are described in four Basic Standards, where each basic standard describes the requirements for one test. | |||
Basic Standards can also be added to {{RadiMation}}. | |||
In addition, one single Technical Setup File (TSF) can be generated for each Basic Standard, configured to perform the test as described in the standard. | |||
To add a Product Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | |||
Select the ''' 'Product standards' tab''' and click ''' 'Add' '''. | |||
(To edit a Product Standard, select the standard and click ''' 'Edit' '''. To remove a Product Standard, select the standard and click ''' 'Remove' '''.) | |||
Insert the required information in the ''' 'Standard details' window''' (as visible in the example below). | |||
Please note that the '''Official name''' can only be inserted the first time and has to be unique. | |||
In addition, it is possible to link a file to the standard in the '''File location''' field. For example a PDF, DOC or HTML file that describes the standard. | |||
{{RadiMation}} will open the file (with the default reader for this file type) when the engineer selects the magnifying glass button next to that field. | |||
To confirm your data input, click ''' 'OK' ''', to cancel your input, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | |||
To add a Basic Standard to the product standard, click ''' 'Add' '''. The ''' 'Select Standard' window''' will appear. | |||
Select one or more Basic Standards (hold the ‘Ctrl’ key to select multiple standards) and click ''' 'Select' '''. | |||
To cancel your selection, click ''' 'Cancel' '''. You will be returned to the ''' 'Standard details' window'''. | |||
To see the details of a Basic Standard, double click the standard or click ‘Details’. | |||
{{Note|When | {{Note|When a product standard is selected or added in the ''' 'EUT' window''', the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the ''' 'Configuration' window''' of {{RadiMation}} is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested.}} | ||
To add a Basic Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | |||
Select the ''' 'Basic standards' tab''' and click ''' 'Add' '''. | |||
(To edit a Basic Standard, select the standard and click ''' 'Edit' '''. To remove a Basic Standard, select the standard and click ''' 'Remove' '''. | |||
Please note that removing a Basic Standard from the list will also remove any references of this standard entered in the Product Standards.) | |||
=== Test equipment list configuration === | |||
The test equipment configuration contains a list of all the test equipment used during a test. The software uses this list to determine which device drivers should be used to control the test equipment. {{RadiMation}} allows you to configure specific test equipment lists for each type of test as well as different test equipment lists for the same type of tests. (This can be useful if you have, for example, more than one signal generator to perform conducted immunity tests with). | |||
In the ''' 'Test Equipment' window''' a set of test equipment will be selected. When a calibration file is generated, {{RadiMation}} will record this test equipment configuration together with the calibration data. When a substitution test is started, the test equipment selected for the substitution test will be compared with the test equipment used (as recorded) during the calibration test. If the equipment does not match, an error message will be displayed. | |||
The error message window will ask you to abort the test or to ignore the error message and continue the test with different test equipment. To achieve maximum accuracy, the same test equipment should be used (whenever possible) during both the calibration and the substitution tests. | |||
To add, modify or delete test equipment lists, go to the ''' 'Test Equipment' window''' in the ''' 'Configuration' tab'''. | |||
[[Image:Config test equipment.png]] | |||
If no test equipment has been added to {{RadiMation}}, click the ''' 'Add' ''' button in the top of the window. | |||
[[Image:Test equipment.png]] | |||
Enter a name for the new test equipment in the ''' 'New Test Equipment' window''' (e.g. Example Test Equipment). | |||
Click ''' 'OK' ''' to confirm the name of the new test equipment, or click ''' 'Cancel' ''' to cancel the new name. You will be returned to the ''' 'Test Equipment' window'''. | |||
[[Image:New test equipment.png]] | |||
To delete a test equipment from {{RadiMation}}, select the '''Test Equipment''' from the pull down menu and click the ''' 'Delete' button''' in the top of the window. A confirmation window will appear. | |||
Click ''' 'Yes' ''' to remove the device from the list, click ''' 'Cancel' ''' to cancel. You will be returned to the ''' 'Test Equipment' window'''. | |||
[[Image:Delete test equipment.png]] | |||
To configure the equipment of a test, select the desired '''Test Equipment''' from the pull down menu (e.g. Example Test Equipment). | |||
Specify the equipment for the test by adding, editing and/or removing devices. | |||
To add a device, click ''' 'Add' ''' (at the bottom left of the window), a menu will open below the button to show the available devices for each Device Type. | |||
[[Image:Test equipment add device 1.png]] | |||
Another way to add a device is to click the desired '''Device Type''' in the ''' 'Test Equipment' overview'''. | |||
A menu will open (at your mouse point) with the available devices for that Device Type. | |||
Select and click on the desired device. The device will then appear in the equipment overview. | |||
[[Image:Test equipment add device 2.png]] | |||
To edit a device, select the device and click ''' 'Edit' '''. The '''Device Driver Settings''' window will appear. | |||
Edit the settings as desired. To confirm your changes, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Test Equipment' window'''. | |||
[[Image:Test equipment device driver settings.png]] | |||
To remove a device, select the device and click ''' 'Remove' '''. A confirmation window will appear. | |||
[[Image:Test equipment remove selected devices.png]] | |||
To save the changes made in the test equipment list, click ''' 'Save' ''' (you will remain in the ''' 'Test Equipment' window'''). | |||
To save the test equipment list under a new name, click ''' 'Save As' '''. | |||
Enter a new name for the test equipment in the ''' 'New Test Equipment' window''' (e.g. Example Test Equipment 2). | |||
[[Image:Test equipment save as.png]] | |||
To close the ''' 'Test Equipment' window''', click ''' 'Close' '''. | |||
If you have not saved the changes to the test equipment list, a dedicated window will appear asking you to save these changes. | |||
To | To save the changes, click ''' 'Yes' ''', to not save the changes, click ''' 'No' '''. The ''' 'Test Equipment' window''' will close. | ||
[[Image:Test | [[Image:Test equipment close.png]] | ||
{{RadiMation}} allows | {{RadiMation}} allows users to create their own device drivers for test equipment (in addition to the standard device drivers). The so called ‘user configurable device drivers’. | ||
The option to set up a user configurable device driver is available for nearly all types of test equipment. Please note that this option is not available for spectrum analyzers and measurement receivers. The complexity (and differences between suppliers) of these devices is too high. | |||
To make your own device driver, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | |||
After | [[Image:Config configuration.png]] | ||
Go to the ''' 'Device Drivers' tab''' and select the required '''Device Driver Type''' from the pull down menu (e.g. Signal generators), then click ''' 'Add' '''. | |||
Select the '''Configurable driver''' (e.g. the Configurable Signal Generator) and click ''' 'New' '''. | |||
[[Image:Configurable signal generator.png]] | |||
Enter a description for the device driver (e.g. the type number of the generator) and click ''' 'OK' ''', the ''' 'Device Driver Settings' window''' will appear. | |||
[[Image:Device driver configuration2.png]] | |||
Edit he information about your ‘user configurable device driver’ in the ''' 'Device Driver Settings' window'''. | |||
Click ''' 'Advanced' ''' and enter the required control commands for the device in the tabs of the dedicated window. | |||
After the commands are entered the device driver is ready for use. Please refer to the operating manual of the equipment for the required commands. | |||
[[Image:Configurable signal generator edit.png]] | |||
{{Note|Passive equipment such as | To confirm the advanced settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Device Driver Settings' window'''. | ||
To confirm your (final) settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | |||
Your new ‘user configurable device driver’ is now visible on the '''Available Device Drivers''' list in the ''' 'Configuration' window''' and can be used as any other device driver (by selecting it in the Equipment list). | |||
{{Note|Passive equipment, such as antennas, current probes, cables etc., also need a device driver. This is required because {{RadiMation}} has to know a number of parameters for these devices. Among others the following information is relevant: | |||
Line 552: | Line 812: | ||
During the development of {{RadiMation}} it became clear to us that, not the EMC tests, but the Equipment Under Test (EUT) played the central role in EMC automation software. | |||
Because of this {{RadiMation}} requires that the EUT must be defined prior to any test. Defining a EUT consists of; creating a new EUT file and specifying the EUT, client and producer information. | |||
After a EUT has been defined, a test window can be opened (from the ‘Test’ menu) and a test can be started. Different types of tests can be performed from the same EUT window. All the test data for the EUT will be stored in the same EUT file. | |||
When, at a later time, tests are to be performed on the same EUT, the old EUT file can be opened and the new test results can be stored under this (old) EUT window. | |||
This allows test data from different tests, using different sets of test equipment, performed at different times and even at different locations, (regarding one EUT), to be stored in one clear EUT file. | |||
Make sure that a network is used to store all the test data when performing tests at different locations. | |||
To perform test(s) on a new device, start by creating a new EUT file. Select ''' 'File' , 'New' , 'EUT' ''' or press ''' 'Ctrl+E' '''. | |||
[[Image:File new eut.png]] | |||
{{RadiMation}} will ask you to specify and save a file name for this new EUT file. Name the EUT file and click ''' 'Save' '''. | |||
All test data for this EUT will be stored under this EUT file name. | (All the test data for this EUT will be stored under this EUT file name.) The ''' 'EUT' window''' will open. | ||
[[Image:Save eut as.png]] | |||
The new test data will be | To perform test(s) on a previously tested device, open the existing EUT file for that device. Select ''' 'File' , 'Open' , 'EUT' '''. | ||
[[Image:File open eut.png]] | |||
{{RadiMation}} will ask you to specify and open an existing EUT file. Select the desired EUT file and click ''' 'Open' '''. | |||
(The new test data will be added to the already existing EUT file.) | |||
[[Image:Open eut.png]] | |||
=== EUT window overview === | |||
After performing a number of test the ''' 'EUT' window''' will look something like the example shown below. | |||
[[Image:Equipment under test example.png]] | |||
The ''' 'EUT' window''' contains five tabs; 'Main', 'EUT Information', 'Monitoring input channels', 'Standards' and 'Reports'. | |||
The | *The ''' 'Main' tab''' consists of six areas where Client, Manufacturer, EUT, Order Number, Test Site and Test information can be found and specified. | ||
*The ''' 'EUT information' tab''' consists of two areas; EUT and Cables. Here (more detailed) EUT information can be specified. This includes the cables that are connected to the EUT. | |||
*The ''' 'Monitoring input channels' tab''' consists of two areas that can be used to monitor the input channels and setup short cut keys (that can be used during measurements). | |||
*The ''' 'Standards' tab''' consists of two areas; Product Standards and Basic Standards. Standards related to the EUT can be added in these fields. (Documentation for standards can be added to {{RadiMation}} in the 'Product standards' and 'Basic standards' tabs of the 'Configuration' window.) | |||
*The ''' 'Reports' tab''' can be used to generate reports with the configured report generator, export reports to a spread sheet program and print detailed EUT information. | |||
More information about the content of these tabs can be found in [[Chapter 3|Chapter 3 "EUT Information"]]. | |||
Line 609: | Line 882: | ||
To create a | To create a Limit Line File, select ''' 'File' , 'New' , 'Limit Line' '''. The ''' 'Limit Line' window''' will open. | ||
[[Image:File New Limit Line.png]] | |||
The | The default ''' 'Limit Line' window''' will appear. This window contains six fields; Description, Data, Options, Detectors, NB/BB and the graph area. | ||
[[Image:Limit Line Default.png]] | |||
==== Enter limit line information ==== | |||
To set up a limit line, enter the following information: | |||
{{ScreenElementDescription| | {{ScreenElementDescription|Description|Enter a file description for the limit line in the Description field (e.g. Limit Line Example).}} | ||
{{ScreenElementDescription| | {{ScreenElementDescription|Data|Enter frequency-value pairs for the limit line in the Data field (see ‘Enter, edit & remove data pairs’ below).}} | ||
{{ScreenElementDescription|Options|Setup the Options field: | |||
{{ScreenElementDescription|Detectors|During emission tests, {{RadiMation}} will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. | * Select a line style and colour for the graph line. | ||
* Select either logarithmic or linear interpolation. In between the specified frequency points, the correct value of the limit line will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. | |||
* Select the Master Unit (or Dimension) for the y-axis, (e.g. Electrical Field). | |||
* Select the Unit for the y-axis, (e.g. dBµV/m). | |||
* Mark the line as a Limit Line or Reference Line. The value of a limit line is compared to detected peaks. The value of a reference line is only usable as a graphical comparison with graphs.}} | |||
{{ScreenElementDescription|Detectors|Setup the Detectors field by selecting one or more detectors. During emission tests, {{RadiMation}} will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid.}} | |||
{{ScreenElementDescription|NB/BB|Setup the NB/BB field by selecting the option(s) Narrowband and/or Broadband.}} | |||
[[Image:Limitline file.png]] | |||
==== Enter, edit & remove data pairs==== | |||
* To enter a frequency-value pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field. Select the desired row and enter the frequency and value data, the limit line will be drawn directly in the graph. | |||
* To edit a data pair, select the desired value and/or unit of the data that you wish to edit and insert the new value and/or unit. | |||
* To remove a frequency-value pair, select the desired data pair and click ''' 'Remove Row' '''. | |||
Before a limit line can be saved, make sure the required information has been entered into the '''Description''', '''Data''' and '''Detectors''' fields. | |||
{{ | Otherwise a ''' 'RadiMation Warning' window''' will appear when you try to save the limit line. | ||
To save a Limit Line File, select ''' 'File' , 'Save Limit line' ''' (or ''' 'Save Limit line As…' '''). | |||
The ''' 'Save Limit line As' window''' will open. (The 'Limit Line' window will remain open in the background.) | |||
[[Image:File Save Limit Line.png]] | |||
{{RadiMation}} will ask you to specify and save a file name for the (new) Limit Line File. | |||
This can be the same or a different filename as used in the limit line '''Description''' field of the ''' 'Limit Line' window'''. | |||
Name the Limit Line File and click ''' 'Save' '''. (The 'Limit Line' window will remain open in the background.) | |||
[[Image:Save Limit line As.png]] | |||
=== Opening limit line files === | |||
To open an existing limit line file, select ''' 'File' , 'Open' , 'Limit Line' '''. The ''' 'Limit Line' window''' will open. | |||
[[Image:File Open Limit Line.png]] | |||
{{RadiMation}} will | {{RadiMation}} will ask you to specify and open an existing Limit Line File. | ||
Select the desired Limit Line File and click ''' 'Open' '''. The specified ''' 'Limit Line' window''' will appear. | |||
[[Image:Open Limit Line.png]] | |||
Correction files can be used to correct the frequency response of cables, generators, clamps, antenna’s, couplers etc. When the frequency response of a device is known, the operator can create a new correction file and assign it to the device. | Correction files can be used to for a lot of different purposes, for example to correct the frequency response of cables, generators, clamps, antenna’s, couplers, etc. When the frequency response of a device is known, the operator can create a new correction file and assign it to the device. From then on, every time that the device is used during a test, the values in selected correction are used to correct the measurement results. | ||
To create a | To create a Correction File, select ''' 'File' , 'New' , 'Correction' ''' or press ''' 'Ctrl+O' '''. | ||
The window | [[Image:File new correction file.png]] | ||
The default ''' 'Correction File' window''' will appear. | |||
This window contains three areas: Data, Options and the graph area. | |||
==== Enter correction file information ==== | |||
To set up a correction file, enter the following information: | |||
{{ScreenElementDescription| | {{ScreenElementDescription|Data|Enter data pairs for the correction value in the Data field (see ‘Enter, edit & remove data pairs’ below).}} | ||
{{ScreenElementDescription|Options|Select either logarithmic or linear interpolation. | |||
{{ScreenElementDescription| | In between the specified frequency points, the correct value of the correction value will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This depends on the requirements of the applicable standard.}} | ||
==== Enter, edit & remove data pairs ==== | |||
The '''Data columns''' need to be defined before data pairs can be added. | |||
To select the desired columns (based on the use of the correction file), click ''' 'Columns/Units' '''. The ''' 'Correction Columns' selection window''' will appear where columns can be added and/or removed. Depending on the usage of the correction file, one or more columns can be selected. E.g. an amplifier limitation correction file will need two columns: frequency and power. A radiated emission test frequency list requires three columns: frequency, angle and height. A simple frequency list for an immunity test only contains a single column: frequency. Also the unit of each of the columns can be modified to represent the desired unit. | |||
Click ''' 'Close' ''' to return to the ''' 'Correction File' window'''. | |||
[[Image:Correction file correction columns.png]] | |||
* To enter a data pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field. Select the desired row and enter the information for the data pair, the correction value will be drawn directly in the graph. | |||
* To remove a data pair, select the desired data point row and click ''' 'Remove Row' '''. | |||
==== Defining the units for the column ==== | |||
As shown in the paragraph above you can use the button ''' 'Columns/Units' ''' to add columns for specific types of information. | |||
[[Image:ColumnsCorrectionFile.png]] | |||
Correction files are being used in {{RadiMation}} to correct the data in a lot of different situations. An example of this would be when you add a new current sensor device to your setup, which should be corrected with the correct resistance. The correction factor you need to use is usually supplied by the manufacturer, but different manufacturers use different units to express the correction factor. To accommodate for this you can select a specific unit (like Ohm or dBOhm) for the resistance column. | |||
See the picture below on how to change the unit. | |||
[[Image:ChoosingUnitsForCorrectionColumn.png]] | |||
{{RadiMation}} will | To save a Correction File, select ''' 'File' , 'Save Correction' (or 'Save Correction As…'). | ||
(The 'Save Correction As' window will open. The 'Correction File' window will remain open in the background.) | |||
[[Image:File save correction.png]] | |||
{{RadiMation}} will ask you to specify and save a file name for the (new) Correction File. | |||
Name the Correction File and click ''' 'Save' '''. (The 'Correction File' window will remain open in the background.) | |||
[[Image:Save correction as.png]] | |||
=== Opening correction files === | |||
To open an existing limit line file, select ''' 'File' , 'Open', 'Correction' '''. (The 'Open Correction' window will open.) | |||
[[Image:File open correction.png]] | |||
{{RadiMation}} will ask you to specify and open an existing Correction File. | |||
Select the desired Correction File and click ''' 'Open' '''. (The specified 'Correction File' window will appear.) | |||
[[Image:Open correction.png]] | |||
{{RadiMation}} is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes [[Radiated immunity]] tests [R.I.], [[Conducted immunity]] tests [C.I.] | {{RadiMation}} is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes [[Radiated immunity]] tests [R.I.], [[Conducted immunity]] tests [C.I.], [[Radiated emission]] tests [R.E.] and [[Conducted emission]] tests [C.E.]. (For more information about the modular test package, consult the [[#Software overview|software overview]].) | ||
The test configuration data for each type of test (such as start/stop frequency, test level, dwell time, etc.) can be saved in a Test Setup File (*.TSF). This file also stores the calibration file and test site information. Save the TSF file with a clear name/description for future reference. A maximum of 40 characters can be used for the description of the TSF file. When the same type of test must be performed, one can simply load the matching TSF file and start the test. | |||
{{RadiMation}} enables the user to create a directory tree to organize and store the different TSF files in a structured way. | |||
A more | |||
A more advanced use of the TSF files takes place during sequence testing. In a sequence, different TSF files (even from different types of tests) can be lined up and started without the intervention of a test engineer. | |||
(See [[Chapter 11|Chapter 11 "Sequence Testing"]] for more information.) | |||
{{RadiMation}} stores all test | {{RadiMation}} stores all the test data in the corresponding EUT files. To view the test data, open the desired EUT file. (Select 'File , 'Open' , 'EUT' to do so.) | ||
The ''' 'EUT' window''' will show a list of all the performed tests for the selected EUT (in the 'Main' tab). | |||
To view the extended data from a single test, select the desired test from the list and click the ''' 'Info' button ''' on the right side of the window. | |||
[[Image:EUT main info.png]] | |||
A ''' 'test data window' ''' will open, displaying all the data for that test. This includes general information, such as: | |||
* Time and date of the test | |||
* Equipment used during the test (including serial numbers) | |||
* Test engineer | |||
* Notes | |||
* Ambient information | |||
And test specific information, such as: | |||
* Test graphs | |||
* Test configuration data | |||
* Calibration file used during the test (if applicable) | |||
[[Image:EUT test data example.png]] | |||
Test data can be printed in a few different ways. | |||
{{Warning| | Open the ''' 'EUT' window''' ('File' , 'Open' , 'EUT') that contains the desired test data and select the most appropriate print option for your situation: | ||
# To print part of the data from one test; | |||
#* Select the desired test (in the 'EUT' window) and click ''' 'Info' ''' at the right side of the list. | |||
#* Select the graph that you want to print (in the 'test data window'). The graph will be displayed on the screen. | |||
#* Click the ''' 'print graph' button ''' above the graph (containing a graph and printer icon). The selected graph will be printed. | |||
# To print all the data from one test; | |||
#* Select the desired test (in the 'EUT' window) and click ''' 'Print' ''' at the right side of the list. The data from the selected test will be printed. | |||
# To print the data of more than one test; | |||
#* Select the desired tests (in the 'EUT' window) while holding the ''' 'Shift' ''' button. | |||
#* Click ''' 'Print' ''' at the right side of the list, the selected tests will be printed. | |||
{{Warning|Printing from {{RadiMation}} will be done by your default Windows printer (with its default settings). To change the printer/settings used by {{RadiMation}}, please change the default printer settings in your Windows environment.}} | |||
{{RadiMation}} can | {{RadiMation}} can use the information of the performed tests to automatically generate a report as well as export all the data to Microsoft Excel. | ||
This can be done with the [[Report Generator]] and [[Excel Exporter]]. Please read [[Chapter 13]] for detailed descriptions of these functions. | |||
During EMC tests, ambient temperature, air pressure and relative humidity can influence test accuracy. | During EMC tests, ambient temperature, air pressure and relative humidity can influence test accuracy. Because of this most EMC standards define (at the very least) the temperature range in which EMC tests must be performed. | ||
Each test configuration menu allows you to enter a value for the temperature, pressure and relative humidity at the start of the test. This data can be entered, and viewed, in the ''' 'Environmental Data' window''', which can be opened from any ''' 'test data window' '''. | |||
To open the ''' 'Environmental Data' window''', click the ''' 'Environment' button''' in the ''' 'test data window' '''. | |||
[[Image:EUT new test environmental data.png]] | |||
== The first test == | |||
Now that we successfully have installed RadiMation and know how to configure the general settings we can launch RadiMation with a (default or created) test engineer account. | |||
Most importantly, we are able to create and configure the device drivers needed for our test setup and add them to our created test equipment. | |||
Also, when we need Correction files and/or Limit Line Files we know how to handle them. | |||
Creating or opening an EUT file and the general EUT overview has been explained. | |||
To start performing tests it is necessary to have an EUT opened and to create your first TSF (Test Setup File). | |||
Proceed to the next Chapters of the manual to read on how a test is configured. For some tests it is required to run a calibration first in order to run. | |||
Latest revision as of 18:50, 6 October 2024
Getting started[edit]
Software overview[edit]
- The Visual and Graph System
- Sequence Testing
- Calibration
- Data Logging
- Database Functionality
- The Device Driver System
- Report generator
- Excel Exporter
Conventions[edit]
- File
- Open
- EUT
- Open
- File
Hardware requirements[edit]
Memory[edit]
GPIB card[edit]
Software requirements[edit]
Microsoft office[edit]
Installation[edit]
Software installation[edit]
Note: | When RadiMation® is using a network configuration the program itself will still be executed from the local computer. It is therefore neither cached nor running directly from the network. |
- CAL_File
- CONFDVDR
- COR_File
- EUT_File
- LLF_File
- Printout
- RGO_File
- RGT_File
- SEQ_File
- TSF_File
Updating device drivers[edit]
Note: | We strongly advice that a you make a complete backup of all your files and directories (e.g. TSF/Sequence/Correction/Calibration and EUT) before installing a RadiMation® update. |
Warning: | Do NOT uninstall RadiMation®!! This will cause removal of all user settings, TSF, sequence, calibration and correction files. This is unrecoverable. |
New installation[edit]
Update[edit]
Daily usage[edit]
Software protection key[edit]
Software configuration[edit]
Password protection[edit]
- Configuration
- Configuration
- Directories
- Configuration
- Configuration
Test engineers[edit]
- The default client
- The default manufacturer
- The default test site
- Configuration
- Configuration
- Directories
- Configuration
- Configuration
Default Units[edit]
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Graphics configuration[edit]
Database Configuration[edit]
Language Selection[edit]
Product Standard[edit]
Basic Standard[edit]
Note: | Passive equipment, such as antennas, current probes, cables etc., also need a device driver. This is required because RadiMation® has to know a number of parameters for these devices. Among others the following information is relevant:
EUT files[edit]
Creating a new EUT file[edit]
EUT window overview[edit]
Limit line files[edit]
Creating limit lines[edit]
Saving limit line files[edit]
Correction files[edit]
Saving correction files[edit]
Viewing test data[edit]
- Time and date of the test
- Test engineer
- Notes
- Ambient information
- Test graphs
- Test configuration data
Printing test data[edit]
Warning: | Printing from RadiMation® will be done by your default Windows printer (with its default settings). To change the printer/settings used by RadiMation®, please change the default printer settings in your Windows environment. |
Report generation[edit]
Ambient information[edit]
The first test[edit]