== Software overview ==
Line 42: Line 41:
* Fast automatic report generation, made possible by one data repository for all test data.
== Conventions ==
{{RadiMation}} uses the Windows structure of pull down menu’s. In this manual the pull down menu’s are visualized as followed:
{{Menu|File|Open|EUT}}
The example above shows that, to show the file open selection window, the ''' 'File' ''' item from the menu bar should be selected first, followed by the ''' 'Open' ''' item and then the ''' 'EUT' ''' item.
[[Image:MenuFileOpenEUT.png]]
== Hardware and software requirements ==
Line 50: Line 59:
* Processor architecture: x64  (ARM64 is not supported)
* Processor type: 2 GHz minimum, > 3 GHz recommended
* Processor clock speed: 2 GHz minimum, > 3 GHz recommended
* Free RAM memory: 512 MB minimal, > 1 GB recommended
* Free RAM memory: 2 GB minimal, > 4 GB recommended
* Free hard disk space: 1 GB minimal
* Monitor: SXGA (1280 x 1024) or larger (at least 1280 pixels wide and at least 900 pixels high)
* Monitor resolution: Full HD (1920 x 1080) or larger
* IEEE card: Any GPIB Card that is supported by the [[National Instruments]] NI-488.2 Software Library
==== Memory ====
==== Memory ====
{{Radimation}} needs a considerable amount of memory to perform at its best. We recommend that you have at least 512 MB (Mega Bytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM ''without'' having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps to help you determine the amount of free memory in your PC:
{{RadiMation}} needs a considerable amount of memory to perform at its best. We recommend that you have at least 512 MB (Mega Bytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM ''without'' having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps are to help you determine the amount of free memory in your PC:
# Cold boot your machine from power off, not from hibernate or restart.
Line 71: Line 80:
==== GPIB card ====
==== GPIB card ====
The GPIB hardware and software needs to be installed correctly before {{RadiMation}} is installed. DARE!! is not responsible for any additional costs or delays that may occur if this card is not installed prior to the installation of {{RadiMation}}.
If a GPIB connection to one of the measurement equipments is used, the GPIB hardware and software needs to be installed correctly before {{RadiMation}} is installed. {{CompanyName}} is not responsible for any additional costs or delays that may occur if this card is not installed prior to the installation of {{RadiMation}}.
The GPIB Card has to be supported by the [[National Instruments]] NI-488.2 software library.
The used GPIB card has to be supported by the [[National Instruments]] NI-488.2 software library. At least version 15.0 or higher of the NI-488.2 software should be used. {{RadiMation}} is able to use any version of [[National Instruments]] NI-488.2 which is 15.0 or higher, so it is not needed to update the NI-488.2 library as long as the installed version is supported by National Instruments and {{RadiMation}}.
=== Software requirements ===
The {{RadiMation}} software is supported on all the Microsoft Windows, Windows Enterprise and Windows Server 2019 (or higher) operating systems that are still supported by [[Microsoft]]. The IoT variant of these operating systems are not supported. If a specific version of Microsoft Windows is still supported, can be checked on [https://docs.microsoft.com/en-us/lifecycle/products/?products=windows  Microsoft Lifecycle Policy pages]. Both the 32-bit and 64-bit versions of these operating systems are supported. In order to run RadiMation it is required that the used operating system is up to date. It can result in severe delays during an installation in the case that Windows Updates are not completely updated and installed.
==== Microsoft .NET framework ====
The {{RadiMation}} software will operate (and is tested) with [[Microsoft]] Windows 7 (32-bit and 64-bit) and [[Microsoft]] Windows 8 (32-bit and 64-bit) operating systems. It is required to have Microsoft Office Word installed on the PC because {{RadiMation}} uses Microsoft Office Word to generate and print the printout's.
The [[Microsoft]] [https://dotnet.microsoft.com/en-us/download/dotnet-framework/net48 .NET Framework 4.8] is also required, on Windows 10 1903 and higher this is installed by default. For Windows 10 October 2018 Update (Version 1809) the .NET framework 4.8 needs to be installed manually before installing {{RadiMation}}
==== Microsoft office ====
The components of the [[Microsoft]] office suite must be fully installed and working, so they can be used for [[Report Generator|report generation]].
It is required to have a supported, licensed and working version of Microsoft Office Word installed on the PC. This should be at least Microsoft Word 2016 or higher, as Microsoft Word 2013 and lower has no mainstream support anymore from Microsoft. {{RadiMation}} uses Microsoft Office Word to generate and print the printouts, besides creating the reports.
The {{RadiMation}} report generator can create reports in the following formats:
The {{RadiMation}} [[Report Generator|report generater]] can create reports in the following formats:
* Microsoft Word
* [[Microsoft Word]]
* Microsoft Excel
* [[Microsoft Excel]]
==== Installation ====
During the installation of {{RadiMation}} a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. DARE!! is not responsible for any additional costs or delays that may occurs if this support is not provided.
During the installation of {{RadiMation}} a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. {{CompanyName}} is not responsible for any additional costs or delays that may occurs if this support is not provided.
When {{RadiMation}} is installed for the first time, it needs to be installed with administrator rights. Upgrades have no need for full rights (although rights to perform installations are required). To update {{RadiMation}} the following rights needs to be set:
* Read rights to the 'C:\' drive
* Full rights to the installation directory (e.g. 'C:\Program Files\Dare Development' or 'C:\Program Files\Dare Instruments', please note that the path may be different depending on the used OS and language)
* Full rights to the installation directory (e.g. 'C:\Program Files\Raditeq', please note that the path may be different depending on the used OS and language)
== Conventions ==
== Software installation ==
=== First time software installation ===
To install the {{RadiMation}} software, please insert the {{RadiMation}} CD in the CD-ROM drive of your computer. The setup program of {{RadiMation}} should start automatically. If the program does not start automatically, run E:\setup.exe from the CD (where E:\ is the letter of your CD-ROM drive). The Installation Wizard will guide you through the installation process. After installation the computer must be restarted before running {{RadiMation}}. Please read [[#User rights and installation|User rights and installation]] about the user rights and information that is required for first time installation.
To install the {{RadiMation}} software, please download the {{RadiMation}} setup program from our download website at: https://www.raditeq.com/radimation-download/ . Once the download has been finished, the setup program can be started. The Installation Wizard will guide you through the installation process. After installation the computer must be restarted before running {{RadiMation}}. Please read [[#User rights and installation|User rights and installation]] about the user rights and information that is required for first time installation.
=== Software installation on a network ===
In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages compaired to a local installation. One of these benefits is that data can be shared between the connected computers. The {{RadiMation}} software is designed to optimally use this advantage. All the data that is used or generated by {{RadiMation}} can also be shared on the network. In most cases we recommend the use a of network configuration over a local configuration.
In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages compared to a local installation. One of these benefits is that data can be shared between the connected computers. The {{RadiMation}} software is designed to optimally use this advantage. All the data that is used or generated by {{RadiMation}} can also be shared on the network. In most cases we recommend the use a of network configuration over a local configuration.
{{Note|When {{RadiMation}} is using a network configuration the program itself will still be executed from the local computer. It is therefore neither cached nor running directly from the network.}}
Line 138: Line 136:
# The {{RadiMation}} program will use the <NetworkDir> as its main data directory from now on.
# Start the Configuration of the {{RadiMation}} software as described in the remaining part of this chapter. The complete configuration will be written to the <NetworkDir>.
When the {{RadiMation}} software is installed on another computer in the network, the following steps can be used:
Line 148: Line 145:
The newly installed {{RadiMation}} will now use the already configured <NetworkDir>.
=== Automated / Silent installation ===
{{RadiMation}} is not provided as a .MSI file that can be used by network administrators to perform an automated installation on a lot of PC's. However {{RadiMation}} is provided as an .EXE file that does accept additional command line arguments that can be used to perform an automated silent installation.
The following arguments can be specified to the {{RadiMation}} setup .EXE program:
{|class="wikitable"
!Argument || Purpose and explanation
|/SP- || Disables the 'This will install ... Do you wish to continue?' prompt at the beginning of the setup
|/SILENT || Instructs the setup program to be silent
|/VERYSILENT || Instructs the setup program to be very silent
|/SUPPRESSMSGBOXES || Instructs the setup to suppress message boxes
|/LOG || Causes setup to create a log file in the user's TEMP directory
|/LOG="filename" || Same as /LOG, except it allows to specify a fixed path/filename to use for the log file.
|/NOCANCEL || Prevents the user from cancelling during the installation process.
|/NORESTART || Prevents Setup from restarting the system following a successful installation, or after a Preparing to Install failure that requests a restart.
|/DIR="x:\dirname" || Overrides the default directory name
|/HELP || Shows a detailed explanation of all the command line parameters
The {{RadiMation}} SETUP.EXE also contains the [[DRIVERS.EXE]] that was actual at the moment that the SETUP.EXE was build. That embedded [[DRIVERS.EXE]] will automatically be installed during the installation of the {{RadiMation}} version. If in the same directory in which the SETUP.EXE is located, also an 'AUTOINSTALLDRIVERS.EXE' is present, then that 'AUTOINSTALLDRIVERS.EXE' will be installed instead of the embedded [[DRIVERS.EXE]]. The 'AUTOINSTALLDRIVERS.EXE' can be created by just renaming the desired [[DRIVERS.EXE]] to 'AUTOINSTALLDRIVERS.EXE'.
=== Installing software updates ===
Software updates can be distributed on CD or downloaded from our website at http://www.dare.eu/instruments/emc-test-software/download.
Software updates can be downloaded from the [https://www.raditeq.com/radimation-download/ {{RadiMation}} Download] page. These updates are supplied as {{RadiMation}} versions with a higher version number.
To install the downloaded {{RadiMation}} update, please execute the startup program.
Once the download of the new version has been finished, the setup program can be started. The setup program will install the new {{RadiMation}} version at the same location as the previous older, but in a separate installation folder. This will allow you to use the previously installed {{RadiMation}} version, as well as the new one.
To install the {{RadiMation}} update from a CD, please insert the CD in the CD-ROM drive of your computer. Run the installation program (i.e. E:\setup.exe, where E:\ is the CD-ROM drive letter). The setup program will detect that a previous version of {{RadiMation}} has already been installed on your PC. The setup program will install the new {{RadiMation}} version at the same location as the previously installed version. A back up of this older version will automatically be copied to the backup directory. All the settings of the older version will also be used for the newly installed {{RadiMation}}. If problems should occur with the newly installed {{RadiMation}}, the old version of {{RadiMation}} can be recovered.
==== Installing from a network directory ====
If the already installed {{RadiMation}} version was installed from a network directory, the new setup file can be copied to the <Setup> sub-directory of the <NetworkDir>.
{{Note|We strongly advice that a you make a complete backup of all your files and directories (e.g. TSF/Sequence/Correction/Calibration and EUT) before installing a {{RadiMation}} update.}}
The new {{RadiMation}} version can then be installed on each computer from that <Setup> sub-directory. This installation still has to be started manually, {{RadiMation}} will not automatically detect that a newer version is available on the network directory, and it will not automatically start the installation of the newer version.
{{Warning|'''Do ''NOT'' uninstall {{RadiMation}}!!''' This will cause removal of all user settings, TSF, sequence, calibration and correction files. This is unrecoverable.}}
==== Updating device drivers ====
After the new {{RadiMation}} version is installed, we strongly recommend that you install the latest [[device drivers]] as well. These device drivers can be installed by starting the ‘[[DRIVERS.EXE]]’ program that can be downloaded from our [https://wiki.dare.nl/wiki/index.php/Device_drivers RadiWiki website].
If the already installed {{RadiMation}} version was installed from a network directory, the new setup file can be copied to the <Setup> sub-directory of the <NetworkDir>. The new {{RadiMation}} version can then be installed on each computer from this directory.
The [[DRIVERS.EXE]] program will detect the different {{RadiMation}} versions installed on your PC. A dialog will start, asking you to select the {{RadiMation}} version that needs updated device drivers. Select the latest version of {{RadiMation}} that was just installed.
You have successfully obtained the latest released version of {{RadiMation}} when both the software and device drivers have been updated.
{{Warning| '''Do ''NOT'' uninstall {{RadiMation}}!!''' This will cause removal of all user settings, TSF, sequence, calibration and correction files. This is unrecoverable.}}
== User rights and installation ==
This chapter describes the rights a user must have to install or upgrade {{RadiMation}} software.
=== New installation ===
=== Upgrade ===
=== Update ===
To upgrade the {{RadiMation}} software, the user must have the rights to read and write in the {{RadiMation}} folder, which is normally located in the C:\Program Files (x86)\DARE Instruments\ directory.
To update the {{RadiMation}} software, the user must have the rights to read and write in the {{RadiMation}} folder, which is normally located in the C:\Program Files (x86)\Raditeq\ directory.
=== Daily usage ===
For daily usage of the {{RadiMation}} software, no specific rights are needed. Every standard end-user should be able to start and use the already installed {{RadiMation}} software.
== Software protection key ==
[[Image:Dongle-keyhanger.jpg|thumb|Software protection key]]
The {{RadiMation}} software is copy protected with an USB [[Software protection key]]. The USB [[Software protection key]] should be connected to the PC while a {{RadiMation}} test is running, or when the report generator or Excel Exporter is used. The USB Software protection key is not needed to open and review the test results.
A network license or a floating license of the USB [[Software protection key]] is not available, and each {{RadiMation}} test PC should thus have it's own USB [[Software protection key]] physically connected to the PC.
== Software configuration ==
Line 175: Line 217:
=== Password protection ===
{{RadiMation}} allows you to protect a number of configurations with a password.
{{RadiMation}} allows to protect a number of configurations by assigning applicable rights to an engineer, where also a login password can be configured for those engineers.
The password is standard turned on for the default used test engineer 'Administrator'.
The default password is '''Radimation'''.
* The password protection is by default turned on for the default engineer with the name: 'Administrator'.
* Configuring and using additional engineers with different passwords is optional.
* The configured engineers are stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured using the menu:
{{Menu|Configuration|Configuration|Directories}}
To change the password protection settings, go to the {{ScreenElement|Engineers}} window in the {{ScreenElement|Configuration}} tab.
To change the password protection settings, go to the ''' 'Engineers' window''' in the ''' 'Configuration' tab'''.
Select the {{ScreenElement|Administrator}} engineer and click {{ScreenElement|Edit}}.
Select the Administrator Engineer and click ''' 'Edit' '''.
[[Image:Engineers admin.png]]
In the (Administrator) ''' 'Test Engineer' window''' the password can be enabled, disabled or changed.
In the {{ScreenElement|Test Engineer}} window for the 'Administrator' user, the password can be enabled, disabled or modified.
{{ScreenElementDescription|Equipment|Prevents unauthorized changes to test equipment configuration.}}
Line 210: Line 251:
To enable, disable or change the password, click ''' '[[Change Password|Change Password]]' '''.
[[Image:Test engineer.png]]
To change the password, enter the old password and the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window.
To change the password, enter the old password and the new password (twice), then click ''' 'OK' '''. You will be returned to the ''' 'Test Engineer' window'''.
[[Image:Change password.png]]
To disable the password, enter the old password and leave the other two fields empty, then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window.
To disable the password, enter the old password and leave the other two fields empty, then click ''' 'OK' '''. You will be returned to the ''' 'Test Engineer' window'''.
(The password indication will change to: “Warning: Password is not set for this engineer.)
To enable the password, enter the new password (twice), then click ''' 'OK' '''. You will be returned to the ''' 'Test Engineer' window'''.
To enable the password, enter the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window.
(The password indication will change to: “Password is set for this engineer.)
[[Image:Enable password.png]]
To leave the {{ScreenElement|Engineers}} window, click {{ScreenElement|Ok}}.
To leave the ''' 'Engineers' window''', click ''' 'OK' '''.
=== Test engineers ===
Line 282: Line 320:
=== Default address information configuration ===
When a new [[EUT]] file is created, the name and address information of the manufacturer, test house and client can be entered separately.
* The default client
* The default manufacturer
* The default test site
Data entered under ''' 'Default Address Information' ''', will automatically be entered in the [[EUT]] window and therefore only needs to be entered once.
The default address information is stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured using the menu: {{Menu|Configuration|Configuration|Directories}}
Data entered under {{ScreenElement|Default Address Information}}, will automatically be copied and stored in the [[EUT]] file. Specifying this information is not required, however it has the benefit that address information that is always the same doesn't need to be entered over and over again in every new [[EUT]] file.
To add, modify or delete the default address information, go to the ''' 'Default Address Information' window''' in the ''' 'Configuration' tab'''.
To add, modify or delete the default address information, go to the {{ScreenElement|Default Address Information}} window in the {{ScreenElement|Configuration}} tab.
Line 305: Line 342:
To confirm the data input, click ''' 'OK' ''', to cancel the data input, click ''' 'Cancel' '''. The ''' 'Default Address Information' window''' will close.
To confirm the data input, click {{ScreenElement|OK}}, to cancel the data input, click {{ScreenElement|Cancel}}. The {{ScreenElement|Default Address Information}} window will close.
Line 350: Line 387:
=== Default directories configuration ===
Line 390: Line 428:
=== Device driver configuration ===
Line 441: Line 480:
To confirm the settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''.
More detailed device driver settings are described in [[#Device Driver settings|Chapter 14 "Device Driver settings"]].
More detailed device driver settings are described in the paragraph "Device Driver settings" in [[Chapter 14]].
[[Image:Device driver settings.png]]
Line 479: Line 518:
The '{{WindowTitle|Graph Customization Configuration}}' '''window''' contains three tabs; ''' 'Chart' ''', ''' 'Graph' ''' and ''' 'Markings' '''.
The ''' 'Chart' tab''' contains the settings related to the overall graph and can be used to determine the colors, layout and titles.
The ''' 'Chart' tab''' contains the settings related to the overall graph and can be used to determine the colours, layout and titles.
The ''' 'Graph' tab''' contains the settings related to the graph lines shown in the graph and can be used to determine the colors, scaling and layout of each and every shown graph line.
The ''' 'Graph' tab''' contains the settings related to the graph lines shown in the graph and can be used to determine the colours, scaling and layout of each and every shown graph line.
The ''' 'Markers' tab''' contains the settings related to the markers on the graph and can be used to determine the colors, style and scaling of each and every shown marker.
The ''' 'Markers' tab''' contains the settings related to the markers on the graph and can be used to determine the colours, style and scaling of each and every shown marker.
[[Image:GraphCustomization Chart.png]]
Line 497: Line 536:
=== Database Configuration ===
Line 531: Line 571:
{{Note|Not all languages are currently available. Please contact your local reseller for language availability.}}
=== Product Standard ===
==== Introduction ====
Line 558: Line 601:
==== Adding a Product Standard ====
To add a Product Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''.
Line 597: Line 641:
{{Note|When a product standard is selected or added in the ''' 'EUT' window''', the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the ''' 'Configuration' window''' of {{RadiMation}} is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested.}}
=== Basic Standard ===
To add a Basic Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''.
Line 624: Line 670:
=== Test equipment list configuration ===
The test equipment configuration contains a list of all the test equipment used during a test. The software uses this list to determine which device drivers to use to control the test equipment. {{RadiMation}} allows you to configure specific test equipment lists for each type of test as well as different test equipment lists for the same type of tests. (This can be useful if you have, for example, more than one signal generator to perform conducted immunity tests with).
The test equipment configuration contains a list of all the test equipment used during a test. The software uses this list to determine which device drivers should be used to control the test equipment. {{RadiMation}} allows you to configure specific test equipment lists for each type of test as well as different test equipment lists for the same type of tests. (This can be useful if you have, for example, more than one signal generator to perform conducted immunity tests with).
In the ''' 'Test Equipment' window''' a set of test equipment will be selected. When a calibration file is generated, {{RadiMation}} will record this test equipment configuration together with the calibration data. When a substitution test is started, the test equipment selected for the substitution test will be compared with the test equipment used (as recorded) during the calibration test. If the equipment does not match, an error message will be displayed.
Line 762: Line 809:
* Correction files for these devices
== EUT files ==
== EUT files ==
Line 776: Line 824:
=== Creating a new EUT file ===
Line 790: Line 837:
[[Image:Save eut as.png]]
=== Opening an existing EUT file ===
Line 826: Line 874:
More information about the content of these tabs can be found in [[Chapter 3|Chapter 3 "EUT Information"]].
== Limit line files ==
Line 850: Line 899:
{{ScreenElementDescription|Data|Enter frequency-value pairs for the limit line in the Data field (see ‘Enter, edit & remove data pairs’ below).}}
{{ScreenElementDescription|Options|Setup the Options field:
* Select a line style and color for the graph line.
* Select a line style and colour for the graph line.
* Select either logarithmic or linear interpolation. In between the specified frequency points, the correct value of the limit line will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard.
* Select the Master Unit (or Dimension) for the y-axis, (e.g. Electrical Field).
Line 860: Line 909:
[[Image:Limitline file.png]]
==== Enter, edit & remove data pairs====
To enter a frequency-value pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field.
* To enter a frequency-value pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field. Select the desired row and enter the frequency and value data, the limit line will be drawn directly in the graph.
To edit a data pair, select the desired value and/or unit of the data that you wish to edit and insert the new value and/or unit.
* To edit a data pair, select the desired value and/or unit of the data that you wish to edit and insert the new value and/or unit.
* To remove a frequency-value pair, select the desired data pair and click ''' 'Remove Row' '''.
Line 908: Line 954:
[[Image:Open Limit Line.png]]
== Correction files ==
Correction files can be used to correct the frequency response of cables, generators, clamps, antenna’s, couplers, etc. When the frequency response of a device is known, the operator can create a new correction file and assign it to the device.
=== Creating correction files ===
Line 920: Line 967:
The default ''' 'Correction File' window''' will appear.
This window contains three fields; Data, Options and the graph area.
[[Image:New correction file.png]]
Line 939: Line 986:
To select the desired columns (based on the use of the correction file), click ''' 'Columns/Units' '''. The ''' 'Correction Columns' selection window''' will appear where columns can be added and/or removed.
To select the desired columns (based on the use of the correction file), click ''' 'Columns/Units' '''. The ''' 'Correction Columns' selection window''' will appear where columns can be added and/or removed. Depending on the usage of the correction file, one or more columns can be selected. E.g. an amplifier limitation correction file will need two columns: frequency and power. A radiated emission test frequency list requires three columns: frequency, angle and height. A simple frequency list for an immunity test only contains a single column: frequency. Also the unit of each of the columns can be modified to represent the desired unit.
(E.g. an amplifier correction file will need two columns; frequency and power. A radiated emission test requires three columns; frequency, angle and height.)
Click ''' 'Close' ''' to return to the ''' 'Correction File' window'''.
Line 948: Line 993:
To enter a data pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field. Select the desired row and enter the information for the data pair, the correction value will be drawn directly in the graph.
* To enter a data pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field. Select the desired row and enter the information for the data pair, the correction value will be drawn directly in the graph.
To remove a data pair, select the desired data point row and click ''' 'Remove Row' '''.
* To remove a data pair, select the desired data point row and click ''' 'Remove Row' '''.
==== Defining the units for the column ====
As shown in the paragraph above you can use the button ''' 'Columns/Units' ''' to add columns for specific types of information.
[[Image:ColumnsCorrectionFile.png]]
Correction files are being used in {{RadiMation}} to correct the data in a lot of different situations. An example of this would be when you add a new current sensor device to your setup, which should be corrected with the correct resistance. The correction factor you need to use is usually supplied by the manufacturer, but different manufacturers use different units to express the correction factor. To accommodate for this you can select a specific unit (like Ohm or dBOhm) for the resistance column.
[[Image:ChoosingUnitsForCorrectionColumn.png]]
=== Saving correction files ===
Line 997: Line 1,054:
(See [[Chapter 11|Chapter 11 "Sequence Testing"]] for more information.)
== Viewing test data ==
Line 1,004: Line 1,062:
The ''' 'EUT' window''' will show a list of all the performed tests for the selected EUT (in the ''' 'Main' tab''').
The ''' 'EUT' window''' will show a list of all the performed tests for the selected EUT (in the 'Main' tab).
To view the extended data from a single test, select the desired test from the list and click the ''' 'Info' button ''' on the right side of the window.
Line 1,028: Line 1,086:
[[Image:EUT test data example.png]]
== Printing test data ==
Line 1,045: Line 1,105:
{{Warning|Printing from {{RadiMation}} will be done by your default Windows printer (with its default settings). To change the printer/settings used by {{RadiMation}}, please change the default printer settings in your Windows environment.}}
== Report generation ==
Line 1,050: Line 1,111:
This can be done with the [[Report Generator]] and [[Excel Exporter]]. Please read [[Chapter 13]] for detailed descriptions of these functions.
== Ambient information ==
Each test configuration menu allows you to enter a value for the temperature, pressure and relative humidity at the start of the test. However, a more sophisticated method can be used as a well. If, under the configuration, test site pull down menu a weather station is selected under the test site used, the temperature, pressure and relative humidity will be measured for each
Each test configuration menu allows you to enter a value for the temperature, pressure and relative humidity at the start of the test. This data can be entered, and viewed, in the ''' 'Environmental Data' window''', which can be opened from any ''' 'test data window' '''.
test point. The collected data will be stored together with the other test data and can be displayed in the three graphs as a function of the test frequency (see also “viewing test data”).
To open the ''' 'Environmental Data' window''', click the ''' 'Environment' button''' in the ''' 'test data window' '''.
[[Image:EUT new test environmental data.png]]
== The first test ==
Now that we successfully have installed RadiMation and know how to configure the general settings we can launch RadiMation with a (default or created) test engineer account.
Most importantly, we are able to create and configure the device drivers needed for our test setup and add them to our created test equipment.
Also, when we need Correction files and/or Limit Line Files we know how to handle them.
Creating or opening an EUT file and the general EUT overview has been explained.
To start performing tests it is necessary to have an EUT opened and to create your first TSF (Test Setup File).
Proceed to the next Chapters of the manual to read on how a test is configured. For some tests it is required to run a calibration first in order to run.
[[Category:Manual]]
[[Category:Manual]]
Latest revision as of 18:50, 6 October 2024
Getting started[edit]
Software overview[edit]
  • The Visual and Graph System
  • Sequence Testing
  • Calibration
  • Data Logging
  • Database Functionality
  • The Device Driver System
Conventions[edit]
   Menu.svg File
      Menu.svg Open
         Menu.svg EUT
Hardware requirements[edit]
  • Processor architecture: x64 (ARM64 is not supported)
  • Processor clock speed: 2 GHz minimum, > 3 GHz recommended
  • Free RAM memory: 2 GB minimal, > 4 GB recommended
Memory[edit]
GPIB card[edit]
Software requirements[edit]
Microsoft office[edit]
Installation[edit]
Software installation[edit]
    • CAL_File
    • CONFDVDR
    • COR_File
    • EUT_File
    • LLF_File
    • Printout
    • RGO_File
    • RGT_File
    • SEQ_File
    • TSF_File
Updating device drivers[edit]
New installation[edit]
Update[edit]
Daily usage[edit]
Software protection key[edit]
Software configuration[edit]
Password protection[edit]
  • The configured engineers are stored in the Device Driver Configuration Files directory, which can be configured using the menu:
   Menu.svg Configuration
         Menu.svg Directories
Test engineers[edit]
  • The default client
  • The default manufacturer
  • The default test site
Default Units[edit]
  • Radiated Immunity
  • Conducted Immunity
  • Radiated Emission
  • Conducted Emission
  • Pulsed Immunity
  • Antenna Diagram
  • RadiCal
Graphics configuration[edit]
Database Configuration[edit]
Language Selection[edit]
Product Standard[edit]
Basic Standard[edit]
  • Frequency range
  • Maximum input power
EUT files[edit]
Creating a new EUT file[edit]
EUT window overview[edit]
Limit line files[edit]
Creating limit lines[edit]
Saving limit line files[edit]
Correction files[edit]
In between the specified frequency points, the correct value of the correction value will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This depends on the requirements of the applicable standard.
Saving correction files[edit]
Viewing test data[edit]
  • Time and date of the test
  • Test engineer
  • Notes
  • Ambient information
  • Test graphs
  • Test configuration data
Printing test data[edit]
Report generation[edit]
Ambient information[edit]
The first test[edit]