| (132 intermediate revisions by 8 users not shown) | |||
| Line 4: | Line 4: | ||
For more (detailed) information, follow the links and references in the text, or go straight to the chapter related to your topic of interest. | For more (detailed) information, follow the links and references in the text, or go straight to the chapter related to your topic of interest. | ||
| Line 39: | Line 38: | ||
* Seamless data exchange between the different modules. | * Seamless data exchange between the different modules. | ||
* Calibration functionality enables an easy exchange of the calibration files. | * Calibration functionality enables an easy exchange of the calibration files. | ||
* Different types of tests can be performed, fully | * Different types of tests can be performed, fully automatically, after each other (Sequence testing). | ||
== Conventions == | |||
{{RadiMation}} uses the Windows structure of pull down menu’s. In this manual the pull down menu’s are visualized as followed: | |||
{{Menu|File|Open|EUT}} | |||
The example above shows that, to show the file open selection window, the ''' 'File' ''' item from the menu bar should be selected first, followed by the ''' 'Open' ''' item and then the ''' 'EUT' ''' item. | |||
A screenshot of these menu items is visible below: | |||
[[Image:MenuFileOpenEUT.png]] | |||
| Line 47: | Line 56: | ||
=== Hardware requirements === | === Hardware requirements === | ||
The {{RadiMation}} software makes extensive use of data calculations, bus interactions (to control the test equipment) and screen interactions (for the use of graphs). Because of this the processor performance is very important and a dedicated PC for {{RadiMation}} is required. Running other programs on the same computer can cause compatibility problems and is not supported. Programs which are supported are the programs used by {{RadiMation}}, such as Word, Excel and MySQL. | The {{RadiMation}} software makes extensive use of data calculations, bus interactions (to control the test equipment) and screen interactions (for the use of graphs). Because of this the processor performance is very important and a dedicated PC for {{RadiMation}} is required. Running other programs on the same computer can cause compatibility problems and is not supported. Programs which are supported are the programs used by {{RadiMation}}, such as Microsoft Word, Excel and MySQL. | ||
* Processor architecture: x64 (ARM64 is not supported) | |||
* Processor | * Processor clock speed: 2 GHz minimum, > 3 GHz recommended | ||
* Free RAM memory: | * Free RAM memory: 2 GB minimal, > 4 GB recommended | ||
* Monitor: | * Monitor resolution: Full HD (1920 x 1080) or larger | ||
{{ | {{RadiMation}} needs a considerable amount of memory to perform at its best. We recommend that you have at least 2 GB (GigaBytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM ''without'' having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps are to help you determine the amount of free memory in your PC: | ||
| Line 65: | Line 74: | ||
# Calculate the amount of free memory by subtracting the amount of memory of the commit charge from the amount of memory put in the computer. | # Calculate the amount of free memory by subtracting the amount of memory of the commit charge from the amount of memory put in the computer. | ||
If the amount of free memory is less then | If the amount of free memory is less then 2 GB, there are several options to obtain more free memory. Depending on the configuration and situation, the following options are available: | ||
* Insert additional memory modules into the machine. | * Insert additional memory modules into the machine. | ||
* Remove/uninstall programs with large memory usage that do not involve measuring, for example VNC viewers. | * Remove/uninstall programs with large memory usage that do not involve measuring, for example VNC viewers. | ||
| Line 71: | Line 80: | ||
If a GPIB connection to one of the measurement equipments is used, the GPIB hardware and software needs to be installed correctly before {{RadiMation}} is installed. {{CompanyName}} is not responsible for any additional costs or delays that may occur if this card is not installed prior to the installation of {{RadiMation}}. | |||
The GPIB | The used GPIB card has to be supported by the [[National Instruments]] NI-488.2 software library. At least version 15.0 or higher of the NI-488.2 software should be used. {{RadiMation}} is able to use any version of [[National Instruments]] NI-488.2 which is 15.0 or higher, so it is not needed to update the NI-488.2 library as long as the installed version is supported by National Instruments and {{RadiMation}}. | ||
=== Software requirements === | |||
The {{RadiMation}} software is supported on all the Microsoft Windows, Windows Enterprise and Windows Server 2019 (or higher) operating systems that are still supported by [[Microsoft]]. The IoT variant of these operating systems are not supported. If a specific version of Microsoft Windows is still supported, can be checked on [https://docs.microsoft.com/en-us/lifecycle/products/?products=windows Microsoft Lifecycle Policy pages]. Both the 32-bit and 64-bit versions of these operating systems are supported. In order to run RadiMation it is required that the used operating system is up to date. It can result in severe delays during an installation in the case that Windows Updates are not completely updated and installed. | |||
=== | ==== Microsoft .NET framework ==== | ||
The | The [[Microsoft]] [https://dotnet.microsoft.com/en-us/download/dotnet-framework/net48 .NET Framework 4.8] is also required. For all supported Windows operating systems, this is already installed by default. | ||
It is required to have a supported, licensed and working version of Microsoft Office Word installed on the PC. This should be at least Microsoft Word 2016 or higher, as Microsoft Word 2013 and lower has no mainstream support anymore from Microsoft. {{RadiMation}} uses Microsoft Office Word to generate and print the printouts, besides creating the reports. | |||
The {{RadiMation}} report generator can create reports in the following formats: | The {{RadiMation}} [[Report Generator|report generator]] can create reports in the following formats: | ||
* Microsoft Word | * [[Microsoft Word]] | ||
* Microsoft Excel | * [[Microsoft Excel]] | ||
During the installation of {{RadiMation}} a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. | During the installation of {{RadiMation}} a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. {{CompanyName}} is not responsible for any additional costs or delays that may occurs if this support is not provided. | ||
* Full rights to the installation directory (e.g. 'C:\Program Files\ | * Full rights to the installation directory (e.g. 'C:\Program Files\Raditeq', please note that the path may be different depending on the used OS and language) | ||
To install the {{RadiMation}} software, please | To install the {{RadiMation}} software, please download the {{RadiMation}} setup program from our download website at: https://www.raditeq.com/radimation-download/ . Once the download has been finished, the setup program can be started. The Installation Wizard will guide you through the installation process. After installation the computer must be restarted before running {{RadiMation}}. Please read [[#User rights and installation|User rights and installation]] about the user rights and information that is required for first time installation. | ||
In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages | In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages compared to a local installation. One of these benefits is that data can be shared between the connected computers. The {{RadiMation}} software is designed to optimally use this advantage. All the data that is used or generated by {{RadiMation}} can also be shared on the network. In most cases we recommend the use a of network configuration over a local configuration. | ||
| Line 138: | Line 136: | ||
| Line 148: | Line 145: | ||
=== Automated / Silent installation === | |||
{{RadiMation}} is not provided as a .MSI file that can be used by network administrators to perform an automated installation on a lot of PC's. However {{RadiMation}} is provided as an .EXE file that does accept additional command line arguments that can be used to perform an automated silent installation. | |||
The following arguments can be specified to the {{RadiMation}} setup .EXE program: | |||
{|class="wikitable" | |||
!Argument || Purpose and explanation | |||
|/SP- || Disables the 'This will install ... Do you wish to continue?' prompt at the beginning of the setup | |||
|/SILENT || Instructs the setup program to be silent | |||
|/VERYSILENT || Instructs the setup program to be very silent | |||
|/SUPPRESSMSGBOXES || Instructs the setup to suppress message boxes | |||
|/LOG || Causes setup to create a log file in the user's TEMP directory | |||
|/LOG="filename" || Same as /LOG, except it allows to specify a fixed path/filename to use for the log file. | |||
|/NOCANCEL || Prevents the user from cancelling during the installation process. | |||
|/NORESTART || Prevents Setup from restarting the system following a successful installation, or after a Preparing to Install failure that requests a restart. | |||
|/DIR="x:\dirname" || Overrides the default directory name | |||
|/HELP || Shows a detailed explanation of all the command line parameters | |||
The {{RadiMation}} SETUP.EXE also contains the [[DRIVERS.EXE]] that was actual at the moment that the SETUP.EXE was build. That embedded [[DRIVERS.EXE]] will automatically be installed during the installation of the {{RadiMation}} version. If in the same directory in which the SETUP.EXE is located, also an 'AUTOINSTALLDRIVERS.EXE' is present, then that 'AUTOINSTALLDRIVERS.EXE' will be installed instead of the embedded [[DRIVERS.EXE]]. The 'AUTOINSTALLDRIVERS.EXE' can be created by just renaming the desired [[DRIVERS.EXE]] to 'AUTOINSTALLDRIVERS.EXE'. | |||
Software updates can be | {{Note|We strongly advice to make a complete backup of all user files and directories (e.g. Device driver configuration files/TSF/Sequence/Correction/Calibration and EUT files) before installing a {{RadiMation}} update.}} | ||
Software updates can be downloaded from the [https://www.raditeq.com/radimation-download/ {{RadiMation}} Download] page. These updates are supplied as {{RadiMation}} versions with a higher version number. | |||
Once the download of the new version has been finished, the setup program can be started. The setup program will install the new {{RadiMation}} version at the same location as the previous older, but in a separate installation folder. This will allow you to use the previously installed {{RadiMation}} version, as well as the new one. | |||
==== Installing from a network directory ==== | |||
If the already installed {{RadiMation}} version was installed from a network directory, the new setup file can be copied to the <Setup> sub-directory of the <NetworkDir>. | |||
The new {{RadiMation}} version can then be installed on each computer from that <Setup> sub-directory. This installation still has to be started manually, {{RadiMation}} will not automatically detect that a newer version is available on the network directory, and it will not automatically start the installation of the newer version. | |||
{{ | ==== Updating device drivers ==== | ||
After the new {{RadiMation}} version is installed, we strongly recommend that you install the latest [[device drivers]] as well. These device drivers can be installed by starting the ‘[[DRIVERS.EXE]]’ program that can be downloaded from our [https://wiki.dare.nl/wiki/index.php/Device_drivers RadiWiki website]. | |||
The [[DRIVERS.EXE]] program will detect the different {{RadiMation}} versions installed on your PC. A dialog will start, asking you to select the {{RadiMation}} version that needs updated device drivers. Select the latest version of {{RadiMation}} that was just installed. | |||
You have successfully obtained the latest released version of {{RadiMation}} when both the software and device drivers have been updated. | |||
For a new installation or an update, different user rights can be setup. | |||
=== | === Update === | ||
To | To update the {{RadiMation}} software, the user must have the rights to read and write in the {{RadiMation}} folder, which is normally located in the 'C:\Program Files (x86)\Raditeq\' directory. | ||
=== Daily usage === | |||
For daily usage of the {{RadiMation}} software, no specific rights are needed. Every standard end-user should be able to start and use the already installed {{RadiMation}} software. | |||
== Software protection key == | |||
[[Image:Dongle-keyhanger.jpg|thumb|Software protection key]] | |||
The {{RadiMation}} software is copy protected with an USB [[Software protection key]]. The USB [[Software protection key]] should be connected to the PC while a {{RadiMation}} test is running, or when the report generator or Excel Exporter is used. The USB Software protection key is not needed to open and review the test results. | |||
A network license or a floating license of the USB [[Software protection key]] is not available, and each {{RadiMation}} test PC should thus have it's own USB [[Software protection key]] physically connected to the PC. | |||
| Line 175: | Line 215: | ||
The next paragraphs will explain the required configurations. | The next paragraphs will explain the required configurations. | ||
{{RadiMation}} allows | {{RadiMation}} allows to protect a number of configurations by assigning applicable rights to an engineer, where also a login password can be configured for those engineers. | ||
The password is | * The password protection is by default turned on for the default engineer with the name: 'Administrator'. | ||
* The default password is '''Radimation'''. | |||
* Configuring and using additional engineers with different passwords is optional. | |||
* The configured engineers are stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured using the menu: | |||
{{Menu|Configuration|Configuration|Directories}} | |||
To change the password protection settings, go to the {{ScreenElement|Engineers}} window in the {{ScreenElement|Configuration}} tab. | |||
To change the password protection settings, go to the | |||
Select the {{ScreenElement|Administrator}} engineer and click {{ScreenElement|Edit}}. | |||
Select the Administrator | |||
[[Image:Engineers admin.png]] | [[Image:Engineers admin.png]] | ||
In the | In the {{ScreenElement|Test Engineer}} window for the 'Administrator' user, the password can be enabled, disabled or modified. | ||
The current password setting for the selected engineer is displayed at the bottom of the window. This will indicate whether a password has been ‘set’ or ‘not set’. | The current password setting for the selected engineer is displayed at the bottom of the window. This will indicate whether a password has been ‘set’ or ‘not set’. | ||
In addition, the protection can be enabled or disabled for the following items | In addition, the protection can be enabled or disabled for the following items: | ||
| Line 210: | Line 249: | ||
To enable, disable or change the password, click | To enable, disable or change the password, click {{ScreenElement|[[Change Password]]}}. | ||
To change the password, enter the old password and the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | |||
To change the password, enter the old password and the new password (twice), then click | |||
To disable the password, enter the old password and leave the other two fields empty, then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | |||
(The password indication will change to: "Warning: Password is not set for this engineer.") | |||
(The password indication will change to: | |||
To enable the password, enter the new password (twice), then click | To enable the password, enter the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | ||
(The password indication will change to: | (The password indication will change to: "Password is set for this engineer.") | ||
To leave the {{ScreenElement|Engineers}} window, click {{ScreenElement|Ok}}. | |||
To leave the | |||
| Line 282: | Line 318: | ||
[[Image:Test engineer edit engineer.png]] | [[Image:Test engineer edit engineer.png]] | ||
When a new [[EUT]] file is created, the name and address information of the manufacturer, test house and client can be entered separately. | |||
When | |||
The default address information is stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured using the menu: {{Menu|Configuration|Configuration|Directories}} | |||
Data entered under {{ScreenElement|Default Address Information}}, will automatically be copied and stored in the [[EUT]] file. Specifying this information is not required, however it has the benefit that address information that is always the same doesn't need to be entered over and over again in every new [[EUT]] file. | |||
To add, modify or delete the default address information, go to the | To add, modify or delete the default address information, go to the {{ScreenElement|Default Address Information}} window in the {{ScreenElement|Configuration}} tab. | ||
[[Image:Config default address information.png]] | [[Image:Config default address information.png]] | ||
| Line 305: | Line 340: | ||
Enter the desired information as shown in the example below. | Enter the desired information as shown in the example below. | ||
To confirm the data input, click | To confirm the data input, click {{ScreenElement|OK}}, to cancel the data input, click {{ScreenElement|Cancel}}. The {{ScreenElement|Default Address Information}} window will close. | ||
| Line 350: | Line 385: | ||
Click ''' 'Close' ''' to leave the ''' 'Configuration' window'''. | Click ''' 'Close' ''' to leave the ''' 'Configuration' window'''. | ||
| Line 390: | Line 426: | ||
| Line 441: | Line 478: | ||
To confirm the settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | To confirm the settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | ||
More detailed device driver settings are described in | More detailed device driver settings are described in the paragraph "Device Driver settings" in [[Chapter 14]]. | ||
[[Image:Device driver settings.png]] | [[Image:Device driver settings.png]] | ||
| Line 479: | Line 516: | ||
The '{{WindowTitle|Graph Customization Configuration}}' '''window''' contains three tabs; ''' 'Chart' ''', ''' 'Graph' ''' and ''' 'Markings' '''. | The '{{WindowTitle|Graph Customization Configuration}}' '''window''' contains three tabs; ''' 'Chart' ''', ''' 'Graph' ''' and ''' 'Markings' '''. | ||
The ''' 'Chart' tab''' contains the settings related to the overall graph and can be used to determine the | The ''' 'Chart' tab''' contains the settings related to the overall graph and can be used to determine the colours, layout and titles. | ||
The ''' 'Graph' tab''' contains the settings related to the graph lines shown in the graph and can be used to determine the | The ''' 'Graph' tab''' contains the settings related to the graph lines shown in the graph and can be used to determine the colours, scaling and layout of each and every shown graph line. | ||
The ''' 'Markers' tab''' contains the settings related to the markers on the graph and can be used to determine the | The ''' 'Markers' tab''' contains the settings related to the markers on the graph and can be used to determine the colours, style and scaling of each and every shown marker. | ||
| Line 497: | Line 534: | ||
Click ''' 'Restore factory default' ''' to return to the factory settings. | Click ''' 'Restore factory default' ''' to return to the factory settings. | ||
| Line 504: | Line 542: | ||
{{RadiMation}} supports multi-language use. This means that the software package can be operated in several | {{RadiMation}} supports multi-language use. This means that the software package can be operated in several different languages. | ||
In addition, the multi- language support allows for easy translation from English to most other languages. | In addition, the multi-language support allows for easy translation from English to most other languages. | ||
Please contact your local reseller for the availability of your required language driver. | Please contact your local reseller for the availability of your required language driver. | ||
| Line 533: | Line 571: | ||
Product Standards describe the requirements (e.g. dimensional, technical, etc.) and the methods of testing (as well as grading, marking, etc.) of a product. | Product Standards describe the requirements (e.g. dimensional, technical, etc.) and the methods of testing (as well as grading, marking, etc.) of a product. | ||
| Line 558: | Line 598: | ||
To add a Product Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | To add a Product Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | ||
| Line 597: | Line 638: | ||
{{Note|When a product standard is selected or added in the ''' 'EUT' window''', the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the ''' 'Configuration' window''' of {{RadiMation}} is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested.}} | {{Note|When a product standard is selected or added in the ''' 'EUT' window''', the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the ''' 'Configuration' window''' of {{RadiMation}} is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested.}} | ||
To add a Basic Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | To add a Basic Standard to {{RadiMation}}, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | ||
| Line 624: | Line 667: | ||
The test equipment configuration contains a list of all the test equipment used during a test. The software uses this list to determine which device drivers | The test equipment configuration contains a list of all the test equipment used during a test. The software uses this list to determine which device drivers should be used to control the test equipment. {{RadiMation}} allows you to configure specific test equipment lists for each type of test as well as different test equipment lists for the same type of tests. (This can be useful if you have, for example, more than one signal generator to perform conducted immunity tests with). | ||
In the ''' 'Test Equipment' window''' a set of test equipment will be selected. When a calibration file is generated, {{RadiMation}} will record this test equipment configuration together with the calibration data. When a substitution test is started, the test equipment selected for the substitution test will be compared with the test equipment used (as recorded) during the calibration test. If the equipment does not match, an error message will be displayed. | In the ''' 'Test Equipment' window''' a set of test equipment will be selected. When a calibration file is generated, {{RadiMation}} will record this test equipment configuration together with the calibration data. When a substitution test is started, the test equipment selected for the substitution test will be compared with the test equipment used (as recorded) during the calibration test. If the equipment does not match, an error message will be displayed. | ||
| Line 651: | Line 695: | ||
To delete a test equipment from {{RadiMation}}, select the '''Test Equipment''' from the pull down menu and click the ''' 'Delete' button''' in the top of the window. A confirmation window will appear. | To delete a piece of test equipment from {{RadiMation}}, select the '''Test Equipment''' from the pull down menu and click the ''' 'Delete' button''' in the top of the window. A confirmation window will appear. | ||
Click ''' 'Yes' ''' to remove the device from the list, click ''' 'Cancel' ''' to cancel. You will be returned to the ''' 'Test Equipment' window'''. | Click ''' 'Yes' ''' to remove the device from the list, click ''' 'Cancel' ''' to cancel. You will be returned to the ''' 'Test Equipment' window'''. | ||
| Line 670: | Line 714: | ||
Another way to add a device is to click the desired '''Device Type''' in the ''' 'Test Equipment' overview'''. | Another way to add a device is to click the desired '''Device Type''' in the ''' 'Test Equipment' overview'''. | ||
A menu will open (at your mouse | A menu will open (at your mouse pointer) with the available devices for that Device Type. | ||
Select and click on the desired device. The device will then appear in the equipment overview. | Select and click on the desired device. The device will then appear in the equipment overview. | ||
| Line 739: | Line 783: | ||
Edit | Edit the information about your ‘user configurable device driver’ in the ''' 'Device Driver Settings' window'''. | ||
Click ''' 'Advanced' ''' and enter the required control commands for the device in the tabs of the dedicated window. | Click ''' 'Advanced' ''' and enter the required control commands for the device in the tabs of the dedicated window. | ||
| Line 776: | Line 820: | ||
Make sure that a network is used to store all the test data when performing tests at different locations. | Make sure that a network is used to store all the test data when performing tests at different locations. | ||
To perform test(s) on a new device, | To perform test(s) on a new device, start by creating a new EUT file. Select ''' 'File' , 'New' , 'EUT' ''' or press ''' 'Ctrl+E' '''. | ||
[[Image:File new eut.png]] | [[Image:File new eut.png]] | ||
{{RadiMation}} will ask you to specify and save a file name for this new EUT file. Name the EUT file and click | {{RadiMation}} will ask you to specify and save a file name for this new EUT file. Name the EUT file and click ''' 'Save' '''. | ||
(All the test data for this EUT will be stored under this EUT file name.) The ''' 'EUT' window''' will open. | |||
[[Image:Save eut as.png]] | [[Image:Save eut as.png]] | ||
| Line 792: | Line 837: | ||
To perform test(s) on a previously tested device, open the existing EUT file for that device. Select | To perform test(s) on a previously tested device, open the existing EUT file for that device. Select ''' 'File' , 'Open' , 'EUT' '''. | ||
[[Image:File open eut.png]] | [[Image:File open eut.png]] | ||
{{RadiMation}} will ask you to specify and open an existing EUT file. Select the desired EUT file and click | {{RadiMation}} will ask you to specify and open an existing EUT file. Select the desired EUT file and click ''' 'Open' '''. | ||
(The new test data will be added to the already existing EUT file.) | |||
[[Image:Open eut.png]] | [[Image:Open eut.png]] | ||
| Line 804: | Line 851: | ||
=== EUT window overview === | === EUT window overview === | ||
After performing a number of test the '''EUT''' | After performing a number of test the ''' 'EUT' window''' will look something like the example shown below. | ||
[[Image:Equipment under test example.png]] | [[Image:Equipment under test example.png]] | ||
The '''EUT''' | The ''' 'EUT' window''' contains five tabs; 'Main', 'EUT Information', 'Monitoring input channels', 'Standards' and 'Reports'. | ||
*The '''Main''' | *The ''' 'Main' tab''' consists of six areas where Client, Manufacturer, EUT, Order Number, Test Site and Test information can be found and specified. | ||
*The '''EUT information''' | *The ''' 'EUT information' tab''' consists of two areas; EUT and Cables. Here (more detailed) EUT information can be specified. This includes the cables that are connected to the EUT. | ||
*The '''Monitoring input channels''' | *The ''' 'Monitoring input channels' tab''' consists of two areas that can be used to monitor the input channels and setup short cut keys (that can be used during measurements). | ||
*The '''Standards''' | *The ''' 'Standards' tab''' consists of two areas; Product Standards and Basic Standards. Standards related to the EUT can be added in these fields. (Documentation for standards can be added to {{RadiMation}} in the 'Product standards' and 'Basic standards' tabs of the 'Configuration' window.) | ||
*The '''Reports''' | *The ''' 'Reports' tab''' can be used to generate reports with the configured report generator, export reports to a spread sheet program and print detailed EUT information. | ||
More information about the content of these tabs can be found in [[Chapter 3|Chapter 3 | More information about the content of these tabs can be found in [[Chapter 3|Chapter 3 "EUT Information"]]. | ||
== Limit line files == | == Limit line files == | ||
| Line 830: | Line 878: | ||
To create a Limit Line File, select | To create a Limit Line File, select ''' 'File' , 'New' , 'Limit Line' '''. The ''' 'Limit Line' window''' will open. | ||
[[Image:File New Limit Line.png]] | [[Image:File New Limit Line.png]] | ||
The default '''Limit Line''' | The default ''' 'Limit Line' window''' will appear. This window contains six fields; Description, Data, Options, Detectors, NB/BB and the graph area. | ||
[[Image:Limit Line Default.png]] | [[Image:Limit Line Default.png]] | ||
| Line 847: | Line 895: | ||
{{ScreenElementDescription|Data|Enter frequency-value pairs for the limit line in the Data field (see ‘Enter, edit & remove data pairs’ below).}} | {{ScreenElementDescription|Data|Enter frequency-value pairs for the limit line in the Data field (see ‘Enter, edit & remove data pairs’ below).}} | ||
{{ScreenElementDescription|Options|Setup the Options field: | {{ScreenElementDescription|Options|Setup the Options field: | ||
* Select a line style and | * Select a line style and colour for the graph line. | ||
* Select either logarithmic or linear interpolation. In between the specified frequency points, the correct value of the limit line will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. | * Select either logarithmic or linear interpolation. In between the specified frequency points, the correct value of the limit line will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This choice is depending on the requirements of the applicable standard. | ||
* Select the Master Unit (or Dimension) for the y-axis, (e.g. Electrical Field). | * Select the Master Unit (or Dimension) for the y-axis, (e.g. Electrical Field). | ||
| Line 857: | Line 905: | ||
[[Image:Limitline file.png]] | [[Image:Limitline file.png]] | ||
==== Enter, edit & remove data pairs==== | ==== Enter, edit & remove data pairs==== | ||
To enter a frequency-value pair, click the | * To enter a frequency-value pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field. Select the desired row and enter the frequency and value data, the limit line will be drawn directly in the graph. | ||
To edit a data pair, select the desired value and/or unit of the data that you wish to edit and insert the new value and/or unit. | * To edit a data pair, select the desired value and/or unit of the data that you wish to edit and insert the new value and/or unit. | ||
* To remove a frequency-value pair, select the desired data pair and click ''' 'Remove Row' '''. | |||
Before a limit line can be saved, make sure the required information has been entered into the | Before a limit line can be saved, make sure the required information has been entered into the '''Description''', '''Data''' and '''Detectors''' fields. | ||
Otherwise a ''' 'RadiMation Warning' window''' will appear when you try to save the limit line. | |||
To save a Limit Line File, select | To save a Limit Line File, select ''' 'File' , 'Save Limit line' ''' (or ''' 'Save Limit line As…' '''). | ||
The ''' 'Save Limit line As' window''' will open. (The 'Limit Line' window will remain open in the background.) | |||
[[Image:File Save Limit Line.png]] | [[Image:File Save Limit Line.png]] | ||
{{RadiMation}} will ask you to specify and save a file name for the (new) Limit Line File. This can be the same or a different filename as used in the limit line | {{RadiMation}} will ask you to specify and save a file name for the (new) Limit Line File. | ||
This can be the same or a different filename as used in the limit line '''Description''' field of the ''' 'Limit Line' window'''. | |||
Name the Limit Line File and click ''' 'Save' '''. (The 'Limit Line' window will remain open in the background.) | |||
[[Image:Save Limit line As.png]] | [[Image:Save Limit line As.png]] | ||
| Line 883: | Line 940: | ||
=== Opening limit line files === | === Opening limit line files === | ||
To open an existing limit line file, select | To open an existing limit line file, select ''' 'File' , 'Open' , 'Limit Line' '''. The ''' 'Limit Line' window''' will open. | ||
[[Image:File Open Limit Line.png]] | [[Image:File Open Limit Line.png]] | ||
{{RadiMation}} will ask you to specify and open an existing Limit Line File. Select the desired Limit Line File and click | {{RadiMation}} will ask you to specify and open an existing Limit Line File. | ||
Select the desired Limit Line File and click ''' 'Open' '''. The specified ''' 'Limit Line' window''' will appear. | |||
[[Image:Open Limit Line.png]] | [[Image:Open Limit Line.png]] | ||
Correction files can be used to correct the frequency response of cables, generators, clamps, antenna’s, couplers, etc. When the frequency response of a device is known, the operator can create a new correction file and assign it to the device. | Correction files can be used to for a lot of different purposes, for example to correct the frequency response of cables, generators, clamps, antenna’s, couplers, etc. When the frequency response of a device is known, the operator can create a new correction file and assign it to the device. From then on, every time that the device is used during a test, the values in selected correction are used to correct the measurement results. | ||
To create a Correction File, select | To create a Correction File, select ''' 'File' , 'New' , 'Correction' ''' or press ''' 'Ctrl+O' '''. | ||
[[Image:File new correction file.png]] | [[Image:File new correction file.png]] | ||
The default '''Correction File''' | The default ''' 'Correction File' window''' will appear. | ||
This window contains three areas: Data, Options and the graph area. | |||
| Line 911: | Line 973: | ||
{{ScreenElementDescription|Data|Enter data pairs for the correction value in the Data field (see ‘Enter, edit & remove data pairs’ below).}} | {{ScreenElementDescription|Data|Enter data pairs for the correction value in the Data field (see ‘Enter, edit & remove data pairs’ below).}} | ||
{{ScreenElementDescription|Options|Select either logarithmic or linear interpolation. In between the specified frequency points, the correct value of the correction value will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This | {{ScreenElementDescription|Options|Select either logarithmic or linear interpolation. | ||
In between the specified frequency points, the correct value of the correction value will be calculated by using either a logarithmic or linear interpolation over the frequency axis. This depends on the requirements of the applicable standard.}} | |||
==== Enter, edit & remove data pairs ==== | ==== Enter, edit & remove data pairs ==== | ||
The | The '''Data columns''' need to be defined before data pairs can be added. | ||
To select the desired columns (based on the use of the correction file), click | |||
To select the desired columns (based on the use of the correction file), click ''' 'Columns/Units' '''. The ''' 'Correction Columns' selection window''' will appear where columns can be added and/or removed. Depending on the usage of the correction file, one or more columns can be selected. E.g. an amplifier limitation correction file will need two columns: frequency and power. A radiated emission test frequency list requires three columns: frequency, angle and height. A simple frequency list for an immunity test only contains a single column: frequency. Also the unit of each of the columns can be modified to represent the desired unit. | |||
Click ''' 'Close' ''' to return to the ''' 'Correction File' window'''. | |||
[[Image:Correction file correction columns.png]] | [[Image:Correction file correction columns.png]] | ||
To enter a data pair, click the | * To enter a data pair, click the ''' 'Add Row' ''' or ''' 'Insert Row' ''' button, a new row will appear in the data field. Select the desired row and enter the information for the data pair, the correction value will be drawn directly in the graph. | ||
* To edit a data pair, select the desired value and/or unit of the data that you wish to edit and insert the new value and/or unit. | |||
* To remove a data pair, select the desired data point row and click ''' 'Remove Row' '''. | |||
==== Defining the units for the column ==== | |||
As shown in the paragraph above you can use the button ''' 'Columns/Units' ''' to add columns for specific types of information. | |||
[[Image:ColumnsCorrectionFile.png]] | |||
Correction files are being used in {{RadiMation}} to correct the data in a lot of different situations. An example of this would be when you add a new current sensor device to your setup, which should be corrected with the correct resistance. The correction factor you need to use is usually supplied by the manufacturer, but different manufacturers use different units to express the correction factor. To accommodate for this you can select a specific unit (like Ohm or dBOhm) for the resistance column. | |||
See the picture below on how to change the unit. | |||
[[Image:ChoosingUnitsForCorrectionColumn.png]] | |||
To save a Correction File, select | To save a Correction File, select ''' 'File' , 'Save Correction' (or 'Save Correction As…'). | ||
(The 'Save Correction As' window will open. The 'Correction File' window will remain open in the background.) | |||
[[Image:File save correction.png]] | [[Image:File save correction.png]] | ||
{{RadiMation}} will ask you to specify and save a file name for the (new) Correction File. Name the Correction File and click | {{RadiMation}} will ask you to specify and save a file name for the (new) Correction File. | ||
Name the Correction File and click ''' 'Save' '''. (The 'Correction File' window will remain open in the background.) | |||
[[Image:Save correction as.png]] | [[Image:Save correction as.png]] | ||
| Line 944: | Line 1,026: | ||
=== Opening correction files === | === Opening correction files === | ||
To open an existing limit line file, select | To open an existing limit line file, select ''' 'File' , 'Open', 'Correction' '''. (The 'Open Correction' window will open.) | ||
[[Image:File open correction.png]] | [[Image:File open correction.png]] | ||
{{RadiMation}} will ask you to specify and open an existing Correction File. Select the desired Correction File and click | {{RadiMation}} will ask you to specify and open an existing Correction File. | ||
Select the desired Correction File and click ''' 'Open' '''. (The specified 'Correction File' window will appear.) | |||
[[Image:Open correction.png]] | [[Image:Open correction.png]] | ||
| Line 962: | Line 1,046: | ||
A more advanced use of the TSF files takes place during sequence testing. In a sequence, different TSF files (even from different types of tests) can be lined up and started without the intervention of a test engineer (See [[Chapter 11|Chapter 11 "Sequence Testing"]] for more information.) | |||
A more advanced use of the TSF files takes place during sequence testing. In a sequence, different TSF files (even from different types of tests) can be lined up and started without the intervention of a test engineer. | |||
{{RadiMation}} stores all the test data in the corresponding EUT files. To view the test data, open the desired EUT file | {{RadiMation}} stores all the test data in the corresponding EUT files. To view the test data, open the desired EUT file. (Select 'File , 'Open' , 'EUT' to do so.) | ||
The '''EUT window''' will show a list of all the performed tests for the selected EUT (in the '' | The ''' 'EUT' window''' will show a list of all the performed tests for the selected EUT (in the 'Main' tab). | ||
To view the extended data from a single test, select the desired test from the list and click the ''' 'Info' button ''' on the right side of the window. | |||
[[Image:EUT main info.png]] | [[Image:EUT main info.png]] | ||
| Line 992: | Line 1,082: | ||
[[Image:EUT test data example.png]] | [[Image:EUT test data example.png]] | ||
Test data can be printed in a few different ways. | |||
Open the ''' 'EUT' window''' ('File' , 'Open' , 'EUT') that contains the desired test data and select the most appropriate print option for your situation: | |||
# To print part of the data from one test; | |||
#* Select the desired test (in the 'EUT' window) and click ''' 'Info' ''' at the right side of the list. | |||
#* Select the graph that you want to print (in the 'test data window'). The graph will be displayed on the screen. | |||
#* Click the ''' 'print graph' button ''' above the graph (containing a graph and printer icon). The selected graph will be printed. | |||
# To print all the data from one test; | |||
#* Select the desired test (in the 'EUT' window) and click ''' 'Print' ''' at the right side of the list. The data from the selected test will be printed. | |||
# To print the data of more than one test; | |||
#* Select the desired tests (in the 'EUT' window) while holding the ''' 'Shift' ''' button. | |||
#* Click ''' 'Print' ''' at the right side of the list, the selected tests will be printed. | |||
{{Warning|Printing from {{RadiMation}} will be done by your default Windows printer (with its default settings). To change the printer/settings used by {{RadiMation}}, please change the default printer settings in your Windows environment.}} | {{Warning|Printing from {{RadiMation}} will be done by your default Windows printer (with its default settings). To change the printer/settings used by {{RadiMation}}, please change the default printer settings in your Windows environment.}} | ||
{{RadiMation}} can use the information of the performed tests to automatically generate a report as well as export all the data to Microsoft Excel. | {{RadiMation}} can use the information of the performed tests to automatically generate a report as well as export all the data to Microsoft Excel. | ||
This can be done with the [[Report Generator]] and [[Excel Exporter]]. Please read [[Chapter 13]] for detailed descriptions of these functions. | |||
During EMC tests, ambient temperature, air pressure and relative humidity can influence test accuracy. Because of this most EMC standards define (at the very least) the temperature range in which EMC tests must be performed. | During EMC tests, ambient temperature, air pressure and relative humidity can influence test accuracy. Because of this most EMC standards define (at the very least) the temperature range in which EMC tests must be performed. | ||
Each test configuration menu allows you to enter a value for the temperature, pressure and relative humidity at the start of the test. | Each test configuration menu allows you to enter a value for the temperature, pressure and relative humidity at the start of the test. This data can be entered, and viewed, in the ''' 'Environmental Data' window''', which can be opened from any ''' 'test data window' '''. | ||
test | |||
To open the ''' 'Environmental Data' window''', click the ''' 'Environment' button''' in the ''' 'test data window' '''. | |||
[[Image:EUT new test environmental data.png]] | |||
== The first test == | |||
Now that we successfully have installed RadiMation and know how to configure the general settings we can launch RadiMation with a (default or created) test engineer account. | |||
Most importantly, we are able to create and configure the device drivers needed for our test setup and add them to our created test equipment. | |||
Also, when we need Correction files and/or Limit Line Files we know how to handle them. | |||
Creating or opening an EUT file and the general EUT overview has been explained. | |||
To start performing tests it is necessary to have an EUT opened and to create your first TSF (Test Setup File). | |||
Proceed to the next Chapters of the manual to read on how a test is configured. For some tests it is required to run a calibration first in order to run. | |||
Latest revision as of 10:27, 4 November 2025
Getting started[edit]
Software overview[edit]
- The Visual and Graph System
- Sequence Testing
- Calibration
- Data Logging
- Database Functionality
- The Device Driver System
- Report generator
- Excel Exporter
- Different types of tests can be performed, fully automatically, after each other (Sequence testing).
Conventions[edit]
File
Open
EUT
Hardware requirements[edit]
The RadiMation® software makes extensive use of data calculations, bus interactions (to control the test equipment) and screen interactions (for the use of graphs). Because of this the processor performance is very important and a dedicated PC for RadiMation® is required. Running other programs on the same computer can cause compatibility problems and is not supported. Programs which are supported are the programs used by RadiMation®, such as Microsoft Word, Excel and MySQL.
Memory[edit]
RadiMation® needs a considerable amount of memory to perform at its best. We recommend that you have at least 2 GB (GigaBytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM without having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps are to help you determine the amount of free memory in your PC:
GPIB card[edit]
Software requirements[edit]
The Microsoft .NET Framework 4.8 is also required. For all supported Windows operating systems, this is already installed by default.
Microsoft office[edit]
It is required to have a supported, licensed and working version of Microsoft Office Word installed on the PC. This should be at least Microsoft Word 2016 or higher, as Microsoft Word 2013 and lower has no mainstream support anymore from Microsoft. RadiMation® uses Microsoft Office Word to generate and print the printouts, besides creating the reports. The RadiMation® report generator can create reports in the following formats:
Installation[edit]
Software installation[edit]
- CAL_File
- CONFDVDR
- COR_File
- EUT_File
- LLF_File
- Printout
- RGO_File
- RGT_File
- SEQ_File
- TSF_File
| Argument | Purpose and explanation |
| /SP- | Disables the 'This will install ... Do you wish to continue?' prompt at the beginning of the setup |
| /SILENT | Instructs the setup program to be silent |
| /VERYSILENT | Instructs the setup program to be very silent |
| /SUPPRESSMSGBOXES | Instructs the setup to suppress message boxes |
| /LOG | Causes setup to create a log file in the user's TEMP directory |
| /LOG="filename" | Same as /LOG, except it allows to specify a fixed path/filename to use for the log file. |
| /NOCANCEL | Prevents the user from cancelling during the installation process. |
| /NORESTART | Prevents Setup from restarting the system following a successful installation, or after a Preparing to Install failure that requests a restart. |
| /DIR="x:\dirname" | Overrides the default directory name |
| /HELP | Shows a detailed explanation of all the command line parameters |
| Note: | We strongly advice to make a complete backup of all user files and directories (e.g. Device driver configuration files/TSF/Sequence/Correction/Calibration and EUT files) before installing a RadiMation® update. |
Updating device drivers[edit]
New installation[edit]
Update[edit]
To update the RadiMation® software, the user must have the rights to read and write in the RadiMation® folder, which is normally located in the 'C:\Program Files (x86)\Raditeq\' directory.
Daily usage[edit]
Software protection key[edit]
Software configuration[edit]
Password protection[edit]
Directories
Test engineers[edit]
- The default client
- The default manufacturer
- The default test site
Directories
Default Units[edit]
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Graphics configuration[edit]
Database Configuration[edit]
Language Selection[edit]
RadiMation® supports multi-language use. This means that the software package can be operated in several different languages.
| Note: | Not all languages are currently available. Please contact your local reseller for language availability. |
Product Standard[edit]
| Note: | When a product standard is selected or added in the 'EUT' window, the data of that standard is copied into the EUT Document. At that moment a copy of the data of the standard from the configuration is made. This means that, if the configuration of a standard in the 'Configuration' window of RadiMation® is changed, the data in the EUT file will still have the old data of the standard!!! This is as it should be, because this is the information of the standard that is used at the time that the EUT was tested. |
Basic Standard[edit]
To delete a piece of test equipment from RadiMation®, select the Test Equipment from the pull down menu and click the 'Delete' button in the top of the window. A confirmation window will appear.
Edit the information about your ‘user configurable device driver’ in the 'Device Driver Settings' window.
| Note: | Passive equipment, such as antennas, current probes, cables etc., also need a device driver. This is required because RadiMation® has to know a number of parameters for these devices. Among others the following information is relevant:
EUT files[edit]
Creating a new EUT file[edit]
EUT window overview[edit]
Limit line files[edit]
Creating limit lines[edit]
| Enter a file description for the limit line in the Description field (e.g. Limit Line Example). |
| Enter frequency-value pairs for the limit line in the Data field (see ‘Enter, edit & remove data pairs’ below). |
| Setup the Detectors field by selecting one or more detectors. During emission tests, RadiMation® will check if the measured emissions are below the limit line(s). To perform this check in the correct way, the software must know for which detector(s) the limit line is valid. |
| Setup the NB/BB field by selecting the option(s) Narrowband and/or Broadband. |
Saving limit line files[edit]
Correction files[edit]
| Enter data pairs for the correction value in the Data field (see ‘Enter, edit & remove data pairs’ below). |
Saving correction files[edit]
Viewing test data[edit]
- Time and date of the test
- Test engineer
- Notes
- Ambient information
- Test graphs
- Test configuration data
Printing test data[edit]
| Warning: | Printing from RadiMation® will be done by your default Windows printer (with its default settings). To change the printer/settings used by RadiMation®, please change the default printer settings in your Windows environment. |
Report generation[edit]
Ambient information[edit]
The first test[edit]