(→Software installation on a network) |
(→Default address information configuration) |
(307 intermediate revisions by 7 users not shown) | |||
__NOTOC__ | __NOTOC__ | ||
− | {{RadiMation}} consists of 7 | + | {{RadiMation}} consists of a software core and 7 complementing software modules. It can be used to perform 5 different types of EMC tests, calibrations, [[sequence]] testing and has an automatic report generation function. |
− | + | An overview of the {{RadiMation}} software is visible in the figure below: | |
[[Image:Modular overview.png]] | [[Image:Modular overview.png]] | ||
Line 33: | Line 21: | ||
* Database Functionality | * Database Functionality | ||
* The Device Driver System | * The Device Driver System | ||
+ | The different modules are integrated in one single software package. The separate modules are: | ||
+ | * [[Radiated immunity]] testing (RI) | ||
+ | * [[Conducted immunity]] testing (CI) | ||
+ | * [[Pulsed immunity]] (ESD, EFT, Surge) | ||
+ | * [[Radiated emission]] testing (RE) | ||
+ | * [[Conducted emission]] testing (CE) | ||
+ | * [[Report generator]] | ||
+ | * [[Excel Exporter]] | ||
+ | The implementation of different modules in one software package has a number of advantages compared to other software: | ||
+ | * Easy to use because all the modules have the same "look & feel". | ||
+ | * Seamless data exchange between the different modules. | ||
+ | * Calibration functionality enables an easy exchange of the calibration files. | ||
+ | * Different types of tests can be performed, fully automatic, after each other (Sequence testing). | ||
+ | * Fast automatic report generation, made possible by one data repository for all test data. | ||
+ | == Conventions == | ||
− | + | The example above shows that, to show the file open selection window, the ''' 'File' ''' item from the menu bar should be selected first, followed by the ''' 'Open' ''' item and then the ''' 'EUT' ''' item. | |
− | + | A screenshot of these menu items is visible below: | |
− | + | [[Image:MenuFileOpenEUT.png]] | |
− | + | There are certain requirements that the PC(-environment, used to perform the EMC tests) needs to comply with to run [[RadiMation|{{RadiMation}}]]. | |
=== Hardware requirements === | === Hardware requirements === | ||
− | The {{RadiMation}} software | + | The {{RadiMation}} software makes extensive use of data calculations, bus interactions (to control the test equipment) and screen interactions (for the use of graphs). Because of this the processor performance is very important and a dedicated PC for {{RadiMation}} is required. Running other programs on the same computer can cause compatibility problems and is not supported. Programs which are supported are the programs used by {{RadiMation}}, such as Word, Excel and MySQL. |
− | To use the software to its full extent | + | To use the software to its full extent the following hardware configuration is required: |
* Processor type: 2 GHz minimum, > 3 GHz recommended | * Processor type: 2 GHz minimum, > 3 GHz recommended | ||
− | * Free RAM memory: | + | * Free RAM memory: 1 GB minimal, > 2 GB recommended |
− | {{Radimation}} needs a considerable amount of memory to perform at its best. | + | {{Radimation}} needs a considerable amount of memory to perform at its best. We recommend that you have at least 512 MB (Mega Bytes) of free memory available, but more if possible. Free memory is defined as the amount of free RAM ''without'' having to use the swap file. Thus, not the specification of the total amount of memory that the machine has. The following steps are to help you determine the amount of free memory in your PC: |
− | # | + | # Cold boot your machine from power off, not from hibernate or restart. |
− | # Determine the total amount of memory put into the | + | # Determine the total amount of memory put into the computer. |
# Start the windows task manager and determine the commit charge. | # Start the windows task manager and determine the commit charge. | ||
− | # Calculate the amount of free memory by subtracting the amount of memory | + | # Calculate the amount of free memory by subtracting the amount of memory of the commit charge from the amount of memory put in the computer. |
− | If the amount of free memory is less then 512 MB there are several options to | + | If the amount of free memory is less then 512 MB, there are several options to obtain more free memory. Depending on the configuration and situation, the following options are available: |
− | * Insert | + | * Insert additional memory modules into the machine. |
− | * Remove programs with large memory usage | + | * Remove/uninstall programs with large memory usage that do not involve measuring, for example VNC viewers. |
− | * Prevent | + | * Prevent programs from starting automatically during the booting of the computer. |
− | The {{RadiMation}} software | + | The {{RadiMation}} software is supported on all the Microsoft Windows operating systems that are still supported by [[Microsoft]]. If a specific version of Microsoft Windows is still supported, can be checked on [https://docs.microsoft.com/en-us/lifecycle/products/?products=windows Microsoft Lifecycle Policy pages]. Both the 32-bit and 64-bit versions of these operating systems are supported. In order to run RadiMation it is required that the used operating system is up to date. It can result in severe delays during an installation in the case that Windows Updates are not completely updated and installed. |
+ | ==== Microsoft .NET framework ==== | ||
+ | The [[Microsoft]] [http://www.microsoft.com/en-US/download/details.aspx?id=49981 .NET Framework 4.6.1] is also required, but this will be automatically downloaded and installed by the RadiMation setup program, if it is not yet installed. A working internet connection is required in that case. | ||
− | The | + | The {{RadiMation}} [[Report Generator|report generater]] can create reports in the following formats: |
− | During the installation a network/PC administrator from the IT department must be available. This person must | + | During the installation of {{RadiMation}} a network/PC administrator from the IT department must be available. This person must be able to grant rights (for one day or more) and have sufficient knowledge of the (local) network. {{CompanyName}} is not responsible for any additional costs or delays that may occurs if this support is not provided. |
− | + | When {{RadiMation}} is installed for the first time, it needs to be installed with administrator rights. Upgrades have no need for full rights (although rights to perform installations are required). To update {{RadiMation}} the following rights needs to be set: | |
− | * Read rights | + | * Read rights to the 'C:\' drive |
− | * Full rights | + | * Full rights to the installation directory (e.g. 'C:\Program Files\Raditeq', please note that the path may be different depending on the used OS and language) |
− | == | + | == Software installation == |
− | To install the {{RadiMation}} software please | + | To install the {{RadiMation}} software, please download the {{RadiMation}} setup program from our download website at: https://www.raditeq.com/radimation-download/ . Once the download has been finished, the setup program can be started. The Installation Wizard will guide you through the installation process. After installation the computer must be restarted before running {{RadiMation}}. Please read [[#User rights and installation|User rights and installation]] about the user rights and information that is required for first time installation. |
− | In a network environment multiple computers are connected with each other | + | In a network environment multiple computers are connected with each other though a shared communication medium. This provides several advantages compared to a local installation. One of these benefits is that data can be shared between the connected computers. The {{RadiMation}} software is designed to optimally use this advantage. All the data that is used or generated by {{RadiMation}} can also be shared on the network. In most cases we recommend the use a of network configuration over a local configuration. |
− | {{Note|When {{RadiMation}} is | + | {{Note|When {{RadiMation}} is using a network configuration the program itself will still be executed from the local computer. It is therefore neither cached nor running directly from the network.}} |
− | To fully use | + | To fully use the functionality of data sharing, the configuration of {{RadiMation}} should be changed slightly. Make the necessary alterations by following the steps below: |
− | # Create a shared network directory with the appropriate rights on one of the Network Servers. Your Network Administrator can help you with this. This directory will be called <NetworkDir> from now on. | + | # Create a shared network directory (with the appropriate rights) on one of the Network Servers. Your Network Administrator can help you with this. This directory will be called <NetworkDir> from now on. |
+ | # Create a drive assignment for this specific <NetworkDir>. This will avoid the use of excessive path depths. Your Network Administrator can also help you with this step. | ||
# Create a sub-directory with the name <Setup> in the <NetworkDir>. | # Create a sub-directory with the name <Setup> in the <NetworkDir>. | ||
− | # Copy the {{RadiMation}} setup.exe file from the CD-ROM (or | + | # Copy the {{RadiMation}} setup.exe file from the CD-ROM (or, if you downloaded {{RadiMation}}, directly from local disk) to the <Setup> directory in the <NetworkDir>. |
# Install {{RadiMation}} on one of the network computers. This can be done by launching the setup.exe program. Use the installation procedure as described in the previous section. | # Install {{RadiMation}} on one of the network computers. This can be done by launching the setup.exe program. Use the installation procedure as described in the previous section. | ||
# In the <NetworkDir>, create the following sub-directories: | # In the <NetworkDir>, create the following sub-directories: | ||
Line 125: | Line 130: | ||
− | # Start the installed {{RadiMation}} program | + | # Start the installed {{RadiMation}} program and open the Configuration window. On the {{ScreenElement|Directories}} tab, change the location of the {{ScreenElement|Device Driver Configuration}} directory to the CONFDVDR directory in the <NetworkDir>. (Note: do not change the other directories yet.) |
# Restart the {{RadiMation}} program. | # Restart the {{RadiMation}} program. | ||
− | # Open the Configuration window again and on the {{ScreenElement|Directories}} tab, | + | # Open the Configuration window again and on the {{ScreenElement|Directories}} tab, change the location of all the other directories to the corresponding directories in the <NetworkDir>. |
# Restart the {{RadiMation}} program again. | # Restart the {{RadiMation}} program again. | ||
− | # | + | # The {{RadiMation}} program will use the <NetworkDir> as its main data directory from now on. |
− | When the {{RadiMation}} software | + | When the {{RadiMation}} software is installed on another computer in the network, the following steps can be used: |
# Install the {{RadiMation}} software from the <Setup> sub-directory in the <NetworkDir>. Use the installation procedure as described in the previous section. | # Install the {{RadiMation}} software from the <Setup> sub-directory in the <NetworkDir>. Use the installation procedure as described in the previous section. | ||
− | # Start the installed {{RadiMation}} program | + | # Start the installed {{RadiMation}} program and open the Configuration window. On the {{ScreenElement|Directories}} tab, change the location of the {{ScreenElement|Device Driver Configuration}} directory to the CONFDVDR directory in the <NetworkDir>. |
− | # Restart the {{RadiMation}} program | + | # Restart the {{RadiMation}} program. |
+ | {|class="wikitable" | ||
− | Software updates | + | Software updates can be downloaded from the [https://www.raditeq.com/radimation-download/ {{RadiMation}} Download] page. These updates are supplied as {{RadiMation}} versions with a higher version number. |
+ | The [[DRIVERS.EXE]] program will detect the different {{RadiMation}} versions installed on your PC. A dialog will start, asking you to select the {{RadiMation}} version that needs updated device drivers. Select the latest version of {{RadiMation}} that was just installed. | ||
− | + | {{Note|We strongly advice that a you make a complete backup of all your files and directories (e.g. TSF/Sequence/Correction/Calibration and EUT) before installing a {{RadiMation}} update.}} | |
+ | === Update === | ||
+ | === Daily usage === | ||
− | In the main menu, under the configuration pull down menu, a number of standard configurations can be made. The selected and / or entered configurations will be used as default. | + | In the main menu, under the configuration pull down menu, a number of standard configurations can be made. The selected and/or entered configurations will be used as the default settings for future use. |
− | The next paragraphs will explain | + | The next paragraphs will explain the required configurations. |
− | {{Menu|Configuration|Engineers|}} | + | * The password protection is by default turned on for the default engineer with the name: 'Administrator'. |
+ | To change the password protection settings, go to the {{ScreenElement|Engineers}} window in the {{ScreenElement|Configuration}} tab. | ||
− | + | Select the {{ScreenElement|Administrator}} engineer and click {{ScreenElement|Edit}}. | |
− | [[Image: | + | [[Image:Engineers admin.png]] |
+ | The current password setting for the selected engineer is displayed at the bottom of the window. This will indicate whether a password has been ‘set’ or ‘not set’. | ||
+ | In addition, the protection can be enabled or disabled for the following items: | ||
− | {{ScreenElementDescription|Equipment| | + | {{ScreenElementDescription|Equipment|Prevents unauthorized changes to test equipment configuration.}} |
− | {{ScreenElementDescription|Engineer| | + | {{ScreenElementDescription|Engineer|Prevents unauthorized changes to test engineers configuration.}} |
− | {{ScreenElementDescription|TSF Files| | + | {{ScreenElementDescription|TSF Files|Prevents unauthorized changes to [[Test Setup File]]s (TSF) files.}} |
− | {{ScreenElementDescription|Limit Line| | + | {{ScreenElementDescription|Limit Line|Prevents unauthorized changes to [[Chapter 18|Limit Line File]]s (LLF) files.}} |
− | {{ScreenElementDescription|Correction File| | + | {{ScreenElementDescription|Correction File|Prevents unauthorized changes to correction files.}} |
− | {{ScreenElementDescription|General Configuration| | + | {{ScreenElementDescription|General Configuration|Prevents unauthorized changes to the general configuration.}} |
− | + | To enable, disable or change the password, click {{ScreenElement|[[Change Password]]}}. | |
+ | [[Image:Test engineer.png]] | ||
+ | To change the password, enter the old password and the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | ||
+ | (The password indication will change to: "Warning: Password is not set for this engineer.") | ||
+ | [[Image:Disable password.png]] | ||
+ | To enable the password, enter the new password (twice), then click {{ScreenElement|OK}}. You will be returned to the {{ScreenElement|Test Engineer}} window. | ||
+ | [[Image:Enable password.png]] | ||
Line 188: | Line 278: | ||
− | + | To add, delete or edit (the names of) test engineers, go to the ''' 'Engineers' window''' in the ''' 'Configuration' tab'''. | |
+ | The Engineers window will show the current configured list of engineers. To leave the ''' 'Engineers' window''', click ''' 'OK' '''. | ||
+ | To add a new test engineer, click ''' 'Add' '''. | ||
+ | Enter the name and set the rights of the new engineer, then click ''' 'OK' '''. You will be returned to the ''' 'Engineers' window'''. | ||
+ | The new engineer will not have a protection password (yet), this can be enabled by clicking ''' '[[Change Password|Change Password]]' '''. Please read [[#Password protection|Password protection]] for more information. | ||
+ | To delete a test engineer, select the engineer and click ''' 'Delete' '''. | ||
+ | To confirm that you want to delete the engineer from the configuration list, click ''' 'Yes' '''. | ||
− | + | To edit the name, rights and/or password of an engineer, select the engineer and click ''' 'Edit' '''. | |
− | + | [[Image:Engineer edit engineer.png]] | |
− | + | Make the desired changes in the ''' 'Test Engineer' window''' and click ''' 'OK' '''. | |
+ | [[Image:Test engineer edit engineer.png]] | ||
+ | In many situations one or more of this information will be the same for different [[EUT]] files (i.e.: when a test house owns the software, the test house data will always be the same). | ||
− | + | To prevent users from having to enter repetitive data input, {{RadiMation}} provides the option to set default address information for: | |
− | Data entered under {{ScreenElement| | + | The default address information is stored in the {{ScreenElement|Device Driver Configuration Files}} directory, which can be configured using the menu: {{Menu|Configuration|Configuration|Directories}} |
+ | [[Image:Config default address information.png]] | ||
− | + | Enter the desired information as shown in the example below. | |
− | {{ | + | To confirm the data input, click {{ScreenElement|OK}}, to cancel the data input, click {{ScreenElement|Cancel}}. The {{ScreenElement|Default Address Information}} window will close. |
Line 227: | Line 348: | ||
− | {{RadiMation}} allows the user to define the default units. These units are used throughout the package | + | {{RadiMation}} allows the user to define the default units. These units are used throughout the software package and can be changed at any time during measurements and viewing. |
− | + | To configure the unit settings, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | |
− | [[Image: | + | [[Image:Config configuration.png]] |
+ | The default settings for the different units can be set in the ''' 'Units' tab'''. | ||
− | It is possible to define different units | + | It is possible to define different units for different tests. The following can be selected: |
Line 245: | Line 368: | ||
* RadiCal | * RadiCal | ||
− | + | To change the settings of a unit, choose the desired test (e.g. Radiated Immunity) from the drop down menu in the ''' 'Units' tab'''. | |
− | + | [[Image:Configuration units test selection.png]] | |
+ | Select the measurement (e.g. Power) and click ''' 'Change'. | ||
− | + | [[Image:Configuration Units.png]] | |
Line 271: | Line 403: | ||
+ | To modify a path name, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. | ||
− | + | Go to the ''' 'Directories' tab''', select the directory that you wish to change and click ''' 'Change' '''. | |
− | In {{RadiMation}}, the device drivers | + | Device drivers let your software know which equipment has to be used and controlled during EMC tests. |
− | + | To configure a device driver, go to the ''' 'Device Drivers' tab''' in the ''' 'Configuration' window'''. | |
− | [[ | + | Select the desired ''' '[[Device Types|Device Driver Type]]' ''' from the drop-down menu (e.g. Signal Generator). The available drivers for that type will be shown. |
− | + | [[Image:Device driver type.png]] | |
+ | The ''' 'Device Driver Settings' window''' will appear with information on the selected device driver. In this window information can be added, edited and removed. | ||
+ | To edit an existing device driver, select the desired (device driver) type and device driver, then click ''' 'Edit' '''. (The ''' 'Device Driver Settings' window''' will appear.) | ||
− | [[Image: | + | [[Image:Device driver remove.png]] |
− | + | A confirmation window will appear. To remove the selected device driver, click ''' 'Yes' ''', to cancel, click ''' 'No' '''. You will be returned to the ''' 'Configuration' window'''. | |
− | + | Click ''' 'Close' ''' to leave the ''' 'Configuration' window'''. | |
− | To configure the | + | To configure the style of the shown graphs, go to the ''' 'Configuration' window''' in the ''' 'Configuration' tab'''. |
{{:Customer database}} | {{:Customer database}} | ||
− | + | {{RadiMation}} supports multi-language use. This means that the software package can be operated in several language. | |
− | + | In addition, the multi- language support allows for easy translation from English to most other languages. | |
+ | The default language of {{RadiMation}} is English. | ||
+ | Go to the ''' 'Language' tab'''. The current language setting is marked with the text '''(current)'''. | ||
+ | For example: | ||
− | + | In addition, one single Technical Setup File (TSF) can be generated for each Basic Standard, configured to perform the test as described in the standard. | |
− | The test equipment configuration contains a list of all test equipment | + | The test equipment configuration contains a list of all the test equipment used during a test. The software uses this list to determine which device drivers should be used to control the test equipment. {{RadiMation}} allows you to configure specific test equipment lists for each type of test as well as different test equipment lists for the same type of tests. (This can be useful if you have, for example, more than one signal generator to perform conducted immunity tests with). |
− | In the | + | |
− | + | In the ''' 'Test Equipment' window''' a set of test equipment will be selected. When a calibration file is generated, {{RadiMation}} will record this test equipment configuration together with the calibration data. When a substitution test is started, the test equipment selected for the substitution test will be compared with the test equipment used (as recorded) during the calibration test. If the equipment does not match, an error message will be displayed. | |
+ | To close the ''' 'Test Equipment' window''', click ''' 'Close' '''. | ||
− | To | + | To save the changes, click ''' 'Yes' ''', to not save the changes, click ''' 'No' '''. The ''' 'Test Equipment' window''' will close. |
− | [[Image:Test | + | [[Image:Test equipment close.png]] |
− | + | [[Image:Device driver configuration2.png]] | |
− | + | Edit he information about your ‘user configurable device driver’ in the ''' 'Device Driver Settings' window'''. | |
− | After | + | Click ''' 'Advanced' ''' and enter the required control commands for the device in the tabs of the dedicated window. |
− | + | [[Image:Configurable signal generator edit.png]] | |
− | {{Note|Passive equipment such as | + | |
+ | To confirm the advanced settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Device Driver Settings' window'''. | ||
+ | To confirm your (final) settings, click ''' 'OK' ''', to cancel your changes, click ''' 'Cancel' '''. You will be returned to the ''' 'Configuration' window'''. | ||
+ | Your new ‘user configurable device driver’ is now visible on the '''Available Device Drivers''' list in the ''' 'Configuration' window''' and can be used as any other device driver (by selecting it in the Equipment list). | ||
+ | {{Note|Passive equipment, such as antennas, current probes, cables etc., also need a device driver. This is required because {{RadiMation}} has to know a number of parameters for these devices. Among others the following information is relevant: | ||
Line 466: | Line 809: | ||
* Correction files for these devices | * Correction files for these devices | ||
+ | When, at a later time, tests are to be performed on the same EUT, the old EUT file can be opened and the new test results can be stored under this (old) EUT window. | ||
− | + | [[Image:File new eut.png]] | |
+ | [[Image:Equipment under test example.png]] | ||
− | + | *The ''' 'Main' tab''' consists of six areas where Client, Manufacturer, EUT, Order Number, Test Site and Test information can be found and specified. | |
− | The | + | *The ''' 'Monitoring input channels' tab''' consists of two areas that can be used to monitor the input channels and setup short cut keys (that can be used during measurements). |
− | [[ | + | More information about the content of these tabs can be found in [[Chapter 3|Chapter 3 "EUT Information"]]. |
== Limit line files == | == Limit line files == | ||
Line 521: | Line 882: | ||
− | + | [[Image:File New Limit Line.png]] | |
− | The | + | The default ''' 'Limit Line' window''' will appear. This window contains six fields; Description, Data, Options, Detectors, NB/BB and the graph area. |
− | To create a | + | To create a Correction File, select ''' 'File' , 'New' , 'Correction' ''' or press ''' 'Ctrl+O' '''. |
− | + | Name the Correction File and click ''' 'Save' '''. (The 'Correction File' window will remain open in the background.) | |
− | + | [[Image:Save correction as.png]] | |
+ | == Using Test Setup Files (TSF) == | ||
+ | {{RadiMation}} is a modular integrated test software package, which, in its full configuration, can perform full automatic EMC tests. The software package includes [[Radiated immunity]] tests [R.I.], [[Conducted immunity]] tests [C.I.], [[Radiated emission]] tests [R.E.] and [[Conducted emission]] tests [C.E.]. (For more information about the modular test package, consult the [[#Software overview|software overview]].) | ||
+ | The test configuration data for each type of test (such as start/stop frequency, test level, dwell time, etc.) can be saved in a Test Setup File (*.TSF). This file also stores the calibration file and test site information. Save the TSF file with a clear name/description for future reference. A maximum of 40 characters can be used for the description of the TSF file. When the same type of test must be performed, one can simply load the matching TSF file and start the test. | ||
+ | {{RadiMation}} enables the user to create a directory tree to organize and store the different TSF files in a structured way. | ||
− | A more | + | |
+ | [[Image:EUT main info.png]] | ||
+ | * Test engineer | ||
+ | * Notes | ||
− | + | * Test graphs | |
− | + | * Test configuration data | |
− | + | * Calibration file used during the test (if applicable) | |
− | * Test | + | |
− | * Test configuration data | + | |
− | * | + | [[Image:EUT test data example.png]] |
+ | == The first test == | ||
Latest revision as of 08:38, 29 August 2023
Getting started
Software overview
- The Visual and Graph System
- Sequence Testing
- Calibration
- Data Logging
- Database Functionality
- The Device Driver System
- Report generator
- Excel Exporter
Conventions
Hardware requirements
- Free RAM memory: 1 GB minimal, > 2 GB recommended
Memory
GPIB card
Software requirements
The RadiMation® software is supported on all the Microsoft Windows operating systems that are still supported by Microsoft. If a specific version of Microsoft Windows is still supported, can be checked on Microsoft Lifecycle Policy pages. Both the 32-bit and 64-bit versions of these operating systems are supported. In order to run RadiMation it is required that the used operating system is up to date. It can result in severe delays during an installation in the case that Windows Updates are not completely updated and installed.
Microsoft .NET framework
The Microsoft .NET Framework 4.6.1 is also required, but this will be automatically downloaded and installed by the RadiMation setup program, if it is not yet installed. A working internet connection is required in that case.
Microsoft office
Installation
Software installation
- CAL_File
- CONFDVDR
- COR_File
- EUT_File
- LLF_File
- Printout
- RGO_File
- RGT_File
- SEQ_File
- TSF_File
Installing software updates
Updating device drivers
New installation
Update
Daily usage
Software protection key
Software configuration
Password protection
Test engineers
- The default client
- The default manufacturer
- The default test site
Default Units
- Radiated Immunity
- Conducted Immunity
- Radiated Emission
- Conducted Emission
- Pulsed Immunity
- Antenna Diagram
- RadiCal
Graphics configuration
Database Configuration
Introduction
Language Selection
Product Standard
Introduction
Adding a Product Standard
Add(ing) Basic Standard(s)
Basic Standard
EUT files
EUT window overview
Limit line files
Correction files
Viewing test data
- Time and date of the test
- Test engineer
- Notes
- Ambient information
- Test graphs
- Test configuration data
Printing test data
Report generation
Ambient information
The first test